Re: [volt-nuts] Fluke 5100B Calibrator? Repair Help

2014-03-13 Thread David Smith
Thank you all for your help on repairing my 5100B. I am totally blown away by your willingness to help and your knowledge! I am really green with this instrument and appreciate your input. I now have replacement part numbers as well as a way to determine what the values should be. Thanks to all!

Re: [volt-nuts] Fluke 5100B Calibrator‏ Repair Help

2014-03-13 Thread Tom Miller
They are MOVs. They protect the Guard circuit to chassis and outguard DC supply. Tom - Original Message - From: "David McQuate" To: Sent: Thursday, March 13, 2014 2:41 PM Subject: Re: [volt-nuts] Fluke 5100B Calibrator‏ Repair Help Question: are the vaporized parts varistors (u

Re: [volt-nuts] Fluke 5100B Calibratorþ Repair Help

2014-03-13 Thread Tom Miller
Hi Dave, It looks like these are just 68 volt GE MOVs. MFR Code is 09214 - GE Fluke Part 460394 GE part V68MA3A It limits the voltage from the guard circuit to chassis ground. The unit limit is 20 V pp from the front panel. Should not be too critical. Regards, Tom - Original Messag

Re: [volt-nuts] Fluke 5100B Calibrator‏ Repair Help

2014-03-13 Thread David McQuate
Question: are the vaporized parts varistors (usually put in series for inrush current limiting) or MOVs (placed in shunt to limit transient peak voltages)? My comments concerned the former... Dave On 3/13/2014 8:34 AM, David Smith wrote: I need some help with repairing a Fluke 5100B Calibrat

Re: [volt-nuts] Fluke 5100B Calibratorþ Repair Hel p

2014-03-13 Thread Charles Steinmetz
Dave wrote: I need some help with repairing a Fluke 5100B Calibrator. * * * According to the manual's parts list these varistors are in-house parts by Fluke. How hard did you look? According to the 5100B manual on Fluke's web site (PN 926634, October 1992, Rev.1 1/94), they are GE

Re: [volt-nuts] Fluke 5100B Calibrator‏ Repair Help

2014-03-13 Thread David McQuate
Estimate the steady-state maximum current (eg from the unit's normal mains current) then search Digikey for "inrush current limiters". Select values near or a little greater than your estimated mains current in the 6th list-box (select multiple values), titled "Current - Steady State Max", an

Re: [volt-nuts] Fluke 5100B Calibrator‏ Repair Help

2014-03-13 Thread Dallas Smith
Are they the ones on page 7-12 in the service manual? http://www.ko4bb.com/manuals/index.php?dir=Fluke Dallas On 3/13/2014 12:34 PM, David Smith wrote: I need some help with repairing a Fluke 5100B Calibrator. The capacitor support bracket fell into the power supply module and vaporized both

Re: [volt-nuts] Fluke 5100B Calibrator Repair Help

2014-03-13 Thread John Phillips
Try calling 510 651 5112. They were Fluke repair for 5100B, (EECS Fremont ca) On Thu, Mar 13, 2014 at 9:34 AM, David Smith wrote: > I need some help with repairing a Fluke 5100B Calibrator. > The capacitor support bracket fell into the power supply module and > vaporized both varistors, RV1 and

Re: [volt-nuts] Fluke 5100B Calibrator‏ Repair Help

2014-03-13 Thread Tom Knox
Someone will most likely have an exact value, but assuming that RV1 and RV2 are being used as excessive transient voltage protection another way to determine value is to use a variac to power up the 5100B and measure the normal voltages at RV1 and RV2. Then replace with a properly rated MOV. T

[volt-nuts] Fluke 5100B Calibrator‏ Repair Help

2014-03-13 Thread David Smith
I need some help with repairing a Fluke 5100B Calibrator. The capacitor support bracket fell into the power supply module and vaporized both varistors, RV1 and RV2. A call to Fluke tells me they no longer can supply these MOV's. According to the manual's parts list these varistors are in-house

Re: [volt-nuts] fluke 332d

2014-03-13 Thread pa4tim
I have a 332 that was unused on a shelf for 25-30 years. Together with a Guildline 4 cell standard, a 720, 845 and a ESI KV divider. (and more) I use the guildline and 720 to set the 332.,Then set a 731 as a control to check during calibration if one is changed. My 332 seems to be not bad but