[volt-nuts] Fluke 5200A (and other) extender boards

2017-09-26 Thread Mark Sims
I got several inquiries when I mentioned I did a 5200A extender cable. I didn't like my solution since it uses ribbon cables and there is +/- (300V?) on some of the slots which is pushing what a ribbon cable can do, although it did pass a test with a 1000V insulation tester. There is a guy on

Re: [volt-nuts] Fluke 5200A instability... A resolution...

2015-08-27 Thread John Phillips
The HP 3455A had a much better AC section but it was a pain to cal. HP work hard to make cal easy and settled for a weak AC converter so they could do a closed case cal. The 3455A, 3457A and 3458A all integrate at line frequency and all have a better AC converter. Integration is tied to line fre

Re: [volt-nuts] Fluke 5200A instability... A resolution...

2015-08-27 Thread Chuck Harris
Hi David, I don't mind so much the journey, as I do being disappointed in the performance of the HP3456A's AC section. DC and Ohms, it is a wonderful DVM. AC, not so much. Since it times its integrators to the power line frequency, it is dooming itself to work stably only at those exact freque

Re: [volt-nuts] Fluke 5200A instability... A resolution...

2015-08-27 Thread David C. Partridge
ded to turn on the filter for <400Hz signals. Cheers Dave -Original Message- From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Chuck Harris Sent: 27 August 2015 17:41 To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 5200A instability... A resol

Re: [volt-nuts] Fluke 5200A instability... A resolution...

2015-08-27 Thread Chuck Harris
Back in July, I posted this note about a stability problem I was having with my Fluke 5200A. I didn't get much response, so I dove in with both feet. The symptom was viewed on a trio of HP3456A 6-1/2 digit DVM's, when measuring the 10V 60Hz output of a Fluke 5200A. The 3456A's showed a 20mv off

Re: [volt-nuts] Fluke 5200A instability...

2015-07-29 Thread Chuck Harris
Hi Tom, That's an idea, but they didn't use any. They used a lot of tantalums... a mix of the orange drop style, and the hermetically sealed axial leaded variety. Some are operated disgustingly close to the voltage ratings... and some are used in that suspicious back-to-back way of making an NP

Re: [volt-nuts] Fluke 5200A instability...

2015-07-29 Thread Tom Miller
Message - From: "Chuck Harris" To: "Discussion of precise voltage measurement" Sent: Tuesday, July 28, 2015 11:35 PM Subject: [volt-nuts] Fluke 5200A instability... Before I spend a lot more time on it, does anyone have any experience with the Fluke 5200A's feedb

[volt-nuts] Fluke 5200A instability...

2015-07-28 Thread Chuck Harris
Before I spend a lot more time on it, does anyone have any experience with the Fluke 5200A's feedback loop oscillating when in the 100Hz band position? Mine started to oscillate at about a 1/2Hz rate a while back. It is two to three orders of magnitude better when in the 1KHz and higher band posi

Re: [volt-nuts] fluke 5200a extenders

2014-11-18 Thread Ken Goodhew
;volt-nuts@febo.com" Subject: [volt-nuts] Fluke 5200a extender Message-ID: Content-Type: text/plain; charset="iso-8859-1" I recently bought a small insulation tester DMM off of Ebay ($20, search for "bm500a insulation tester"). It can test at 250/500/1000V. I built up

[volt-nuts] Fluke 5200a extender

2014-11-17 Thread Mark Sims
I recently bought a small insulation tester DMM off of Ebay ($20, search for "bm500a insulation tester"). It can test at 250/500/1000V. I built up three of the 5200A extender cables and tested them. There were no problems at 1000V (even though the ribbon cables are only rated for 300V and the

[volt-nuts] fluke 5200a

2014-10-27 Thread Ken Goodhew
Hi There, Any chance one of the members might have or know where I might be able to buy an A7 power amplifier board for the Fluke 5200a calibrator. Ken. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.

[volt-nuts] fluke 5200a extender

2014-09-24 Thread Kgoodhew
Had a look at the pin outs on both the power amplifier and the power supply regulator boards and the maximum potential difference between any two adjacent pins is 190 volts. EG. +300 v to +190 v to 190v return to -190v to -300v on consecutive pins on the power amplifier board. A similar situatio

[volt-nuts] Fluke 5200a extender

2014-09-23 Thread Mark Sims
That would be me... I have the boards in and they probably work with all the cards, but they scare me... Three of the 5200 boards have +/-300V and +/- 190V on some of the pins and I don't totally trust them to handle that properly. The ribbon cable is rated at 300V and the box headers are (

[volt-nuts] Fluke 5200a extender

2014-09-23 Thread Kgoodhew
One of the members was getting some pcb's made to fabricate some extender boards, has there been any developments yet? --- This email is free from viruses and malware because avast! Antivirus protection is active. http://www.avast.com ___ volt-nuts ma

Re: [volt-nuts] fluke 5200a extender boards

2014-08-29 Thread Kgoodhew
Hi all, 1. Charles I looked at the item you mentioned on digikey but unfortunately it will not do the job as the card extender has to sit vertically above the instrument and plug into the main board socket which also sits vertically, also the original card extender is almost 12" long by 3" deep,

Re: [volt-nuts] Fluke 5200A extender board.

2014-08-29 Thread Tom Miller
Would that be at the maximum temperature? Service would be at room temp. Sounds like a good excuse, right? - Original Message - From: "Mark Sims" To: Sent: Friday, August 29, 2014 2:28 PM Subject: [volt-nuts] Fluke 5200A extender board. Those are not holes... they ar

Re: [volt-nuts] Fluke 5200A extender board.

2014-08-29 Thread Tom Miller
Learn something new every day. Thanks. - Original Message - From: "Mike S" To: Sent: Friday, August 29, 2014 2:27 PM Subject: Re: [volt-nuts] Fluke 5200A extender board. On 8/29/2014 2:16 PM, Tom Miller wrote: Did you remove the two holes? :) You mean the two pin 1

[volt-nuts] Fluke 5200A extender board.

2014-08-29 Thread Mark Sims
Those are not holes... they are polarity markers on the silkscreen for the connectors. I do have one concern about doing the extender with ribbon cable. It is rated for 300V and three of the card slots do have 300V on them. _

Re: [volt-nuts] Fluke 5200A extender board.

2014-08-29 Thread Mike S
On 8/29/2014 2:16 PM, Tom Miller wrote: Did you remove the two holes? :) You mean the two pin 1 markers on the silkscreen? ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts

Re: [volt-nuts] Fluke 5200A extender board.

2014-08-29 Thread Tom Miller
Did you remove the two holes? :) - Original Message - From: "Mark Sims" To: Sent: Friday, August 29, 2014 1:53 PM Subject: [volt-nuts] Fluke 5200A extender board. 5200A extender cables are a go! Connectors and boards are on order (from a different board fab in Hong

[volt-nuts] Fluke 5200A extender board.

2014-08-29 Thread Mark Sims
5200A extender cables are a go! Connectors and boards are on order (from a different board fab in Hong Kong this time). One board goes on each end of the extender. Boards are connected by a 50 pin and 40 pin ribbon cable. __

Re: [volt-nuts] Fluke 5200A extender board.

2014-08-29 Thread David C. Partridge
o.com Subject: [volt-nuts] Fluke 5200A extender board. Ok, I did a 5200A extender cable board layout. It has headers for a 50 pin and 40 pin ribbon cable. One copy of the board would be on each end of the cable, with the edge connector soldered to one of the boards. So, how many people are inter

[volt-nuts] Fluke 5200a extender board.

2014-08-28 Thread Kgoodhew
Hi all, Thankyou to all who replied with suggestions and information about this extender board, I have looked up the surplus store as suggested and found they do have stock of the 43/86 card edge connectors. Something I have missed in my searches. Another contributor has given me

[volt-nuts] Fluke 5200A extender board.

2014-08-28 Thread Mark Sims
Ok, I did a 5200A extender cable board layout. It has headers for a 50 pin and 40 pin ribbon cable. One copy of the board would be on each end of the cable, with the edge connector soldered to one of the boards. So, how many people are interested in a 5200A extender cable? Depending upon q

[volt-nuts] Fluke 5200A extender board.

2014-08-28 Thread Mark Sims
There is also one listed on Ebay for $9... http://www.ebay.com/itm/TRW-Cinch-50-86C-30-43-Pos-Double-Wire-Wrap-PC-Edge-Connector-New-Old-Stock-/171213916405?pt=LH_DefaultDomain_0&hash=item27dd2508f5 If you have access to a PCB layout program you can easily make the circuit board/edge tabs. OSHPAR

Re: [volt-nuts] Fluke 5200A extender board.

2014-08-28 Thread Bill Gold
" together which I don't want to do. I will do a PM to you with a pdf attached of the backside of the extender. Bill - Original Message - From: "Kgoodhew" To: Sent: Wednesday, August 27, 2014 11:44 PM Subject: [volt-nuts] Fluke 5200A extender board. > Hi,

Re: [volt-nuts] Fluke 5200A extender board.

2014-08-28 Thread Charles Black
Hi Ken, Nigel is setting you on the right track with Vectorbord. I have used them for the edge connectors over the years with good results. I don't know if you are able to use Digikey of not but they list a 3690-26 ($80) that will be easy to modify to work for your project. I probably should

Re: [volt-nuts] Fluke 5200A extender board.

2014-08-28 Thread GandalfG8
Hi Ken, I don't know if it helps but there's a 43/86 edge connector offered here for 14USD, (12USD for 5+).. https://www.surplussales.com/ComputerAccess/con_edge.html Single and double sided PCBs with 0.1 inch spaced strips were quite common at one time as part of the Veroboard range

Re: [volt-nuts] fluke 5200a extender board.

2014-08-28 Thread David C. Partridge
If it should help, most connector mfrs will make up a small run to order - small typically meaning 30 or so. If you are mega lucky (as I once was) you sometimes find they have one or two hanging around (worth a telephone call) ... Regards, David Partridge

[volt-nuts] Fluke 5200A extender board.

2014-08-27 Thread Kgoodhew
Hi, The pin spacing on the card edge connectors is 0.1" (2.54 mm) and is a 43/86 dual row socket. I know what the extender looks like as I found one on ebay several months ago that had already sold for $50 (damm it!!) and I have been looking ever since, but it had a picture of it. It is

[volt-nuts] fluke 5200a extender board.

2014-08-27 Thread Mark Sims
What is the pin spacing on the edge 5200a connector? BTW, I got in my circuit boards for the TM500/5000 and HP5370 extenders from China. They look great. I am now waiting for the rest of the edge connectors to arrive. They have been shipped... from the US. Should be here in a day or two.

[volt-nuts] fluke 5200a extender board.

2014-08-27 Thread Ken Goodhew1
Hi all, Over the last 6 months I have been steadily repairing a faulty 5200A that I brought, it had numerous faults in the power supply regulator and power amplifier boards that I have managed to repair with a lot of difficulty as I do not have an extender board to operate the boar

[volt-nuts] Fluke 5200A repair

2014-07-14 Thread Kgoodhew
Hi Charles, Thanks for the comments. I have already lifted one end of R3 to isolate the input to the power amplifier and the fault is definitely in the power amplifier board. I have done quite a bit of thinking about the problem with the circuit in hand to try and make sense of the voltage

Re: [volt-nuts] Fluke 5200A repair

2014-07-14 Thread Charles Steinmetz
Ken wrote: Yes R6 changes the output of the op amp by about 15 mv.. During further investigation I found the input to Q1 was sitting at -400 mv when it should be at 0 +/- 100uv. Suggest lifting one end of R3 to isolate U1/Q1 from the oscillator. Using the transistor tester I first tested al

[volt-nuts] Fluke 5200A repair

2014-07-13 Thread Kgoodhew
, as well as making it difficult to do measurements. So when I have the board disassembled I will run some cables to allow me to operate the board out of the chassis. Ken. Date: Sun, 13 Jul 2014 21:42:23 +1000 From: Gerd To: Discussion of precise voltage measurement Subject: [volt-nuts] Fluke

[volt-nuts] Fluke 5200A repair.

2014-07-13 Thread Gerd
vary it by R6. Will do some further checking today. Ken. Message: 5 Date: Fri, 11 Jul 2014 12:38:05 +1000 From: Gerd To: Discussion of precise voltage measurement Subject: [volt-nuts] fluke 5200a repair Message-ID: <53bf4e0d.9030...@controlelectronics.com.au> Content-Type: text/plain;

[volt-nuts] Fluke 5200A repair.

2014-07-10 Thread Kgoodhew
some further checking today. Ken. Message: 5 Date: Fri, 11 Jul 2014 12:38:05 +1000 From: Gerd To: Discussion of precise voltage measurement Subject: [volt-nuts] fluke 5200a repair Message-ID: <53bf4e0d.9030...@controlelectronics.com.au> Content-Type: text/plain; charset=ISO-8859-1;

[volt-nuts] fluke 5200a repair

2014-07-10 Thread Gerd
to see if they may be leaking at a higher voltage. Thanks, Ken - Message: 2 Date: Wed, 09 Jul 2014 14:21:13 -0400 From: Charles Steinmetz To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 5200A repair Message-ID: <20140709222123.lmbgv...@smtp2o.mail.yandex.net&g

[volt-nuts] fluke 5200a repair

2014-07-10 Thread Kgoodhew
2014 14:21:13 -0400 From: Charles Steinmetz To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 5200A repair Message-ID: <20140709222123.lmbgv...@smtp2o.mail.yandex.net> Content-Type: text/plain; charset="us-ascii"; format=flowed Ken wrote: >R6 slide

[volt-nuts] Fluke 5200A repair

2014-07-10 Thread Gerd
maybe I am wrong. Thanks again for your input. Ken. Message: 6 Date: Wed, 09 Jul 2014 22:19:46 +1000 From: Gerd To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 5200A repair. Message-ID: <53bd3362.5030...@controlelectronics.com.au> Content-Type: text/plain;

Re: [volt-nuts] Fluke 5200A repair

2014-07-09 Thread Charles Steinmetz
Oops! the divider ratio of R8/R9 is 2100, not 210. Best regards, Charles Ken wrote: R6 slider to U1 varies from +15v to -15v but the output of U1 only varies by mv's, I put this down to the loading of R9 to earth on the input to pin3 of U1 (221 ohm),only allowing a small variation to the

Re: [volt-nuts] Fluke 5200A repair

2014-07-09 Thread Charles Steinmetz
Ken wrote: R6 slider to U1 varies from +15v to -15v but the output of U1 only varies by mv's, I put this down to the loading of R9 to earth on the input to pin3 of U1 (221 ohm),only allowing a small variation to the input on pin 3 of U1, maybe this needs further investigation! Look at the circ

[volt-nuts] Fluke 5200A repair

2014-07-09 Thread Kgoodhew
ge: 6 Date: Wed, 09 Jul 2014 22:19:46 +1000 From: Gerd To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 5200A repair. Message-ID: <53bd3362.5030...@controlelectronics.com.au> Content-Type: text/plain; charset=ISO-8859-1; format=flowed Hello Ken, Glad to hear the

Re: [volt-nuts] Fluke 5200A repair.

2014-07-09 Thread Gerd
m Subject: Re: [volt-nuts] Fluke 5200a Message-ID: <53bb3ba7.2070...@controlelectronics.com.au> Content-Type: text/plain; charset=ISO-8859-1; format=flowed Hello Ken, From what I remember of this circuit, R26 is for output stage bias. R6 is for adjusting offset. By the way, which voltages a

[volt-nuts] Fluke 5200A repair.

2014-07-08 Thread Kgoodhew
5200a (John Phillips) -- Message: 1 Date: Tue, 08 Jul 2014 10:30:31 +1000 From: Gerd To: volt-nuts@febo.com Subject: Re: [volt-nuts] Fluke 5200a Message-ID: <53bb3ba7.2070...@controlelectronics.com.au> Content-Type: text/plain;

Re: [volt-nuts] Fluke 5200a

2014-07-07 Thread John Phillips
The problem is dc offset. On Mon, Jul 7, 2014 at 5:30 PM, Gerd wrote: > Hello Ken, > > From what I remember of this circuit, R26 is for output stage bias. R6 is > for adjusting offset. By the way, which voltages are high and how much? > > The 5200 is an old beast, expect a lot of marginal elect

Re: [volt-nuts] Fluke 5200a

2014-07-07 Thread Gerd
Hello Ken, From what I remember of this circuit, R26 is for output stage bias. R6 is for adjusting offset. By the way, which voltages are high and how much? The 5200 is an old beast, expect a lot of marginal electrolytics, carbon resistors that have drifted well out of tolerance and even tran

[volt-nuts] Fluke 5200a

2014-07-06 Thread Ken Goodhew1
Hi all, I am currently repairing a Fluke 5200a ac voltage calibrator. After pulling cards in and out and soldering multiple wires on to test I now have the unit working about 98%. Anybody might have or know of where I might be able to obtain an extender card for this unit as it w