Dear Crispin and all,

We have Zero for one calibration, we have range, linear response and detection limit. Working in a lab one can get overly critical of reported values that does not go through all the steps at the time of testing. I realize not as important for much of the stuff we do but if we are to place values on CO in the air, set limits and the such, we all need to be using calibrated equipment.

Does Emerson provide the supply gas? or use electrical signal for the calibration? I suggest it can only be done using real pure gas.

Regards
Frank







Crispin Pemberton-Pigott wrote:

Dear Frank

I think the calibration is needed now and then. Real calibration against a calibrated source. Daily, we look for how it treats air. I like Philip Lloyd's practise of removing the probe from time to time to see if it has CO go to zero. It is rare for cells other than CO to have problems. CO cells can get saturated and report high for quite a long time (an hour). In UB we use air as a start and finish calibration. The instruments are so accurate it is amazing. The CO2 will happily report 17% and after 5 hours of zipping up and down, settle on air being 0.039% within a few seconds.
However, the issue of absolute calibration is important enough that Emerson 
offers an attachment that will recalibrate each time the instrument changes 
ranges!

Regards
Crispin



--
Frank Shields
Soil Control Lab
42 Hangar way
Watsonville, CA  95076
(831) 724-5422 tel
(831) 724-3188 fax
[email protected]
www.compostlab.com



_______________________________________________
The Stoves list has moved to [email protected] - please
update your email contacts to reflect the change.

Please visit BioEnergy Discussion Lists
http://info.bioenergylists.org/
Thank you, Stoves
Administrator
http://stoves.bioenergylists.org
http://info.bioenergylists.org

UNSUBSCRIBE HERE;
http://listserv.repp.org/mailman/listinfo/stoves_listserv.repp.org

Reply via email to