On 28/12/12 20:33, Fabio Eboli wrote:
Bruce, Bob, now I have more questions
than before :)

First and most important:
Where can I find references about the
statistical method you mentioned?
I tried to search online but didnt't find
any info.

The bucket statistic method is very simple.
You divide a scale into a suitable set of ranges, equally wide. The you count how many "hits" you get within each of those ranges or "buckets".

As your oscillators drift in phase relative to each other, you could divide that into say 1024 "buckets". In this case, you would choose the scale to stretch over the range of ADC values. Another way would be to stretch them over the saw-tooth ramp.

What do you think about the simple setup
I mentioned before (the 2 Rb osc)?
If I set accurately the frequency difference
(easy with a scope and a timer) I will know
the cycle-by-cycle pulse increment
(or reduction).

Yes. If you have a 1 s rate of measurement, dial in 1E-11 in frequency error, the advancement will be 10 ps (1E-11 s). The increment will be y*T where y is the relative frequency error and T is the interval.

Cheers,
Magnus

_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to