Thanks to Tom, Jean-Louis and Chris for all of the helpful advice.

Problem is now fixed.

First problem was obvious - the 74HC390 I use as a divider clocks on 
negative-going transition. Added inverter, and the problem appeared to still be 
present (!)

Next problem was not so obvious - my very cheap benchtop chinese frequency 
counter (a Zhaoxin HC-2700L, bought on ebay, and retrofitted with a 
divided-by-two Pletronics 26 MHz VCOCXO as timebase, which were for sale 
cheaply on ebay a few years ago) turns out to have a bug/defect in the 
"totalizer" mode; when count transitions from 99 999 999 to 00 000 000, the 
counter ignores its input for a few milliseconds while it processes something, 
so my original plan to count for 10 seconds (expecting to see 30 000 000, which 
is 130 million with the hundred-million digit wrapped off) didn't pan out. 
Small change to the counter circuit and I now count a series of 1-second 
intervals, each of 13 MHz reference oscillator in the counter.

I adjusted the counter's timebase VCOCXO bias input so that counting for seven 
1-second intervals shows 91 000 000 (13 000 000 * 7). There seems to be a +/- 1 
count uncertainty, likely due to absence of synchronization between the 
counter's timebase and the 1 PPS signal, but this accomplishes what I set out 
to do, which was a cheap-and-dirty (but certainly not quick!) calibration of 
the counter's timebase.

Thanks for everyone's advice and support!

Dave

>To: jl.on...@free.fr, Discussion of precise time and frequency
>       measurement     <time-nuts@febo.com>
>Subject: Re: [time-nuts] Beginner question - unexpected possible
>       jitter in 1 PPS output of Motorola ONCORE UT+ module
>Message-ID:
>       <CABbxVHvpyUYZfP7=zztTKRXGr+F=k7DBYCzNLwAG04Tr=3_...@mail.gmail.com>
>Content-Type: text/plain; charset=UTF-8
>
>I posted here some days ago an admission that I did exactly this by
>accident once.  It is a very hard problem to debug because everything works
>just fine, except for the very high jitter.   So I go looking for noise and
>what not.    The problem was a wrong number of inverters.
>
>
>On Sat, May 31, 2014 at 7:58 PM, Jean-Louis Oneto <jl.on...@free.fr> wrote:
>
>> ..... So if your divider is clocked on the falling edge of the 1PPS, you
>> can expect a 1 ms jitter...
>
>-- 
>
>Chris Albertson
>Redondo Beach, California