Hi

Shouldn´t 5x10^-11 over 50C be 1x10^-12 / C? So with 2-4°C variation it is 
2-4x10^-12.

What about pressure variations for the HP5065 and other Rb´s? My LPRO have 
about 7x10^-14/mBar (hPa) so with 15-20mbar change, that can happen quite 
quick, it is also in the ^-12 range. The tempco for my LPRO is 7x10^-13/°C and 
drift in the high ^-14 per day so my GPSDO controller mostly fights the 
temperature and pressure variations I think. I like having a GPS disciplined Rb 
as I haven´t had to adjust it during the last years. Of course a OCXO based 
GPSDO will also stay on frequency. For me the Rb have been good when I have 
tested GPS modules and GPSDO´s just out of curiosity. In hold mode it have been 
useful to get the ADEV out to say 10000 secs (low ^-13).

Lars

Från: Bob Camp<mailto:kb...@n1k.org>
Skickat: den 14 mars 2016 02:01
Till: Discussion of precise time and frequency 
measurement<mailto:time-nuts@febo.com>
Ämne: Re: [time-nuts] HP 5065A repair

Hi

Some math:

5x10^-11 over 50C

You have 1x10^-13 / C

If you have pretty good HVAC you get 2C cycles. On a typical home system, you 
get 2X that or more.

Net is a bump at 2x10^-13 (or more).

That assumes no hysteresis. (Hint: there always is hysteresis).

That assumes you have no rate dependent effects. (… they almost always are 
present ..).

If you are at 10X the data sheet level, the bump is more like 2x10^-12 (or 
more). Either one will likely show up on a good test plot.

Can you take care of all this? Of course you can. Does modeling and correcting 
all this fall into the “quick and easy fix” category? Nope, not at all. The 
thread is about a request for a simple approach to an Rb setup. That sort of 
thing does not include fancy models and all sorts of corrections.

Bob



> On Mar 13, 2016, at 5:10 PM, Poul-Henning Kamp <p...@phk.freebsd.dk> wrote:
>
> --------
> In message <aabmqmgvvauqc...@smtpout04.dca.untd.com>, cdel...@juno.com writes:
>
>> As far a tempco goes, unless your lab swings tens of degrees will you
>> really see it?
>
> Well, I do...
>
> My air-con is far from optimal, but it clearly makes a very obvious
> bump in my AVAR plots.
>
>
> --
> Poul-Henning Kamp       | UNIX since Zilog Zeus 3.20
> p...@freebsd.org         | TCP/IP since RFC 956
> FreeBSD committer       | BSD since 4.3-tahoe
> Never attribute to malice what can adequately be explained by incompetence.
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.

_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to