> Since the ion pump current was dropping slowly.(80uA to 76uA in 1 hour), is 
> it possible to be back to normal?

Hi Li,

Yes, is it possible. Never give up on ion current. Give it hours, give it days 
if necessary.

Monitor the LCD or syst:print? periodically and plot the trend. The trend may 
be linear (you may only need to wait for hours). The trend my be logarithmic 
(you may need to wait for days). If the trend looks like it will eventually 
drop below 1 uA that is good news. You just have to wait. If the 5071A gives a 
fatal error and turns off the ion pump for safety, then power cycle to try 
again.

What you are seeing is not normal behavior. But given the chance that you might 
get a tube that works, even for a little while, that is worth something. Do you 
know where these 5071A were installed, or what they were used for?

If it looks like the ion pump current will never drop to 1 uA, even after 
several days, then you may have a vacuum leak or failed ion pump or too much 
outgassing or something. Try swapping tubes to see if the new frame shows a 
similar ion pump reading and trend. If so, then both frames are ok and the tube 
is bad. I don't know of a fix for this. Cesium tubes fail for many reasons, not 
just running out of cesium.

/tvb