Hi Bob,
Normally I see somewhere between 2E-11 and 4E-11 at 1S tau on my 5370A, as in 
the blue trace on the attached plot.  Am I misunderstanding your meaning?  
Granted, I am clocking the 5370A with a GPSDO, but I believe I see about the 
same thing with the HP10811.  This test was 1PPS vs 1PPS on two different units.
The plot also has a test run by Tom, in orange, using his H Maser and a Timepod 
to show how poor the 5370 is compared to the Timepod below about 60S tau.  
These are essentially apples vs apples tests.

Bob




      From: Bob Camp <kb...@n1k.org>
 To: Discussion of precise time and frequency measurement <time-nuts@febo.com> 
 Sent: Thursday, January 12, 2017 8:27 AM
 Subject: Re: [time-nuts] General questions about making measurements with time 
interval counter.
   
Hi

There are a number of ways to improve the resolution (and accuracy) of your 
data without spending 
big piles of cash. They have been discussed here on the list many times over 
the last few years. 
What I’m suggesting is that you dig into that ahead of taking data. You will 
dive into it eventually as you 
look more and more at devices that are locked to some sort of stable reference 
internally. 

Ideally you would like a device with a floor 5X to 10X better than what you are 
measuring. For ADEV style
data, the 5370 is a 1x10^-10 sort of device single shot (so 1x10^-9 is the 
limit at 10:1). With a lot of averaging 
(which is not something you do with ADEV) you can get about 5X better than that 
as a floor. In either case, it is getting in the way of any
readings that are much below 1x10^-9 at one second. A low cost XO can hit that 
level of performance.  