Further information:

Shorting pins 5 and 6 of PL2 causes the same symptoms, as does pressing and 
holding the DIG FILT key on another meter.

The odd thing is though that an ohms test between 5 and 6 of keyboard is OC - 
which does not make sense to me?

Dave
-----Original Message-----
From: [email protected] [mailto:[email protected]] On Behalf 
Of David C. Partridge
Sent: 30 January 2013 17:15
To: 'Discussion of precise voltage measurement'
Subject: Re: [volt-nuts] Solartron 7071 beeping about twice a second

Joe,

Looks like you win the prize for remote diagnosis! Disconnecting PL2 which goes 
to the keyboard, and it all comes up nicely. 

What I can do about that though I am not at all sure ... :(

Dave
-----Original Message-----
From: [email protected] [mailto:[email protected]] On Behalf 
Of J. L. Trantham
Sent: 29 January 2013 03:14
To: 'Discussion of precise voltage measurement'
Subject: Re: [volt-nuts] Solartron 7071 beeping about twice a second

I wonder if it might mean multiple 'stuck' buttons on the front panel.  Any way 
to check that?  Perhaps disconnecting the front panel and reconnecting?
Perhaps making measurements on the front panel itself?

Good luck.

Joe

_______________________________________________
volt-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Reply via email to