Randy: Look at page #23 of the 3458A Calibration Manual. Make yourself a "4 Terminal Short" as shown and put it in as shown.
Before I did an ACAL I had around -000.00025 mVDC. After ACAL I now read +000.00002 mVDC. I did the "CAL 0" myself so I would expect that the meter should return to a low value, and it does. I do get a variation of +/- 30 nVDC using 100 PLC and just observing the variations. As I remember I have never seen a spec on the ZERO stability over temperature. If I turn on the MATH function and then do 40 measurements with 100 PLC the statistics show: Low reading -70 nVDC Mean reading -28 nVDC High reading +3.5 nVDC Total Variation 73 nVDC So that correlates with my visual observation of 60 nVDC. After an hour the room had gone up around 1 degree C. Then I observed -000.00023 mVDC. After another ACAL the reading was again +000.00002 mVDC. This particular meter has a negative tempco as the room temp goes up. Obviously do an ACAL before any precision measurements requiring low nanovolts. Go to the Keysight website and go to "Technical Support" and choose "Parts". Then enter in the "Part Number" "03458-66517" which is the replacement "03458-66507" assembly and you will see the replacement part number on the right hand side. Click on that and you will get the information about the exchange program and so on. Looks like you can just order this part online and pay for it with a credit card, but you have to create or use an existing login account. I needed a new display a few years ago. At that time you could order just the display for around $80. Being extremely good at removing and then inserting and soldering I ordered the part. The problem was that the spacing from top to bottom of the pins had changed. It went from around 1.3 inches to around 1.5 inches. So I had to bend the pins to fit my display board and then get something like 72 pins into the holes on the PC Board. It took hours. This change is probably why HP/Agilent/Keysight doesn't let you just get the display anymore but wants you to get the whole PC Board assembly. It did work just fine once installed. Hope this helps your decision to keep or not. Bill ----- Original Message ----- From: "Randy Evans" <randyevans2...@gmail.com> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com> Sent: Saturday, August 16, 2014 10:36 PM Subject: Re: [volt-nuts] RE "new" 3458A > Interesting note. After the room cooled down from about 79F to 73F, and > another ACAL, the meter now reads +000.00035 mVDC, a more reasonable value, > although it does bounce around a couple of tenths of a uV. > > Maybe that is OK? If so, then the only issue would seem to be the display > has some faint pixels, which a new display should fix. > > Randy > > > On Sat, Aug 16, 2014 at 9:59 PM, Randy Evans <randyevans2...@gmail.com> > wrote: > > > > > The unit seems to be working so far except for one issue. After doing an > > ACAL, and making sure the Auto Zero is ON, I short the input leads with a > > copper wire shunt across the inputs and the reading is approximately > > -000.0023 mVDC. That seems rather high. I would expect the unit to short > > the input leads internally and force a zero reading during the ACAL. > > Anyone have any comments on this reading? > > > > Thanks, > > > > Randy > > > > > > On Sat, Aug 16, 2014 at 3:37 PM, Richard Moore <richiem5...@gmail.com> > > wrote: > > > >> Hi Randy -- sounds like your unit is in cal, based on your measurements > >> of DCV and precision 10k resistor. > >> > >> Using autocal all is recommended before doing precision measurements, and > >> I do that if it's been more than a day or two since last use. The autocal > >> uses the internal Vref and an internal 10K resistor to do cal on everything > >> else, so that tells you what the basic cal procedure is. I just got my 3458 > >> back from Loveland, and that's what they did for me -- warmed it up, then > >> ran autocal, then measured everything against a Fluke 5700, aided by an HP > >> 3325, and another 3458. > >> > >> It has been 5 years since I replaced the display board (no "exchange" > >> deal was available then AFAIK, so I don't know what's changed) and also the > >> NVRAM board, which was dead, with one with the Snap-cap RAM chips. I did > >> those replacements, then sent it home for cal, which was complete, since > >> all the RAM was new. Now after 5 years, the unit passed all incoming > >> performance tests and was sent back to me without a cal process of any > >> kind. This tells me that an old, well-aged Vref module is a good thing. The > >> 10VDC test had changed by a bit under 5ppm, or roughly 1ppm/year. > >> > >> They have a cal deal -- use code 1.090 -- press them for it -- and that > >> saved me 30% off the normal price. I think this deal lasts until > >> mid-September, so my recent "cal" ended up at just under $400 including > >> shipping. I'm not sure the deal is available on new or first-time cals; my > >> unit was in their data bank. > >> > >> But this is a long way of saying I don't think you need to send it for > >> cal -- just push Auto Cal and Enter and wait about 10 minutes and you > >> should be good to go. > >> _______________________________________________ > >> volt-nuts mailing list -- volt-nuts@febo.com > >> To unsubscribe, go to https://www.febo.com/cgi-bin/ > >> mailman/listinfo/volt-nuts > >> and follow the instructions there. > >> > > > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.