Hi Adrian,

Although I have not used the Pomona spade lugs or a nano-voltmeter, my experience is consistent with yours otherwise.

Anyone who has calibrated (CAL 0 anyway) a 3458A has enough information to deduce that lowest input short voltage is going to be a copper wire since the meter is set to zero volts during calibration using a heavy gauge (14 to 16) copper wire to short the top four input terminals. This is very convenient since it is so easy to duplicate in the field and makes simple inexpensive test leads best for high precision measurements. In order for the 3458A to make full accuracy measurements (8 digit) NPLC must be set to 1000 (according to the User's Guide). If you use a lower NPLC value there is a table in the User's Guide that can be used to determine how accurate your measurements are going to be.

Just for fun I ran several shorts for NPLC 1000 on my 3458A. It has been about a year since I clid my last CAL 0 so it was going to be interesting at least for me. The first shunt was my test "U' shaped shunt that I used for my last CAL 0. Note: STP = Shielded Twisted Pair.

Calibration shunt -0.00021mv +/- 10 pV Equilibration time 5 minutes. 14 gauge per Calibration Manual. "U" heavy wire -0.00021mv +/- 10pV Equil. time 5 minutes. Used the through holes in the Input banana posts only. Copper wire -0.00019mv +/- 10pV NAPA PVC covered automobile wire at same contact points as CAL shunt Standard Ground Plate -0.00040mv +/- 10pV Equil. time 5 minutes. Gold plated ground plate from my Datron 4910 Copper wire -0.00019mv +/- 10Pv Equil. time 2 seconds. Used the banana through holes. STP 2 meter test lead -0.00021mv +/- 10pV Equil. time 2 seconds. M27500 24 gauge STP Tefzel insulation. Banana plugs -0.00021mv +/- 30pV Equil. time 20 minutes. My best "Perfect" gold plated plugs with copper wire.

Charlie


On 8/25/2014 3:36 AM, acb...@gmx.de wrote:
I have used the pomona spades, mainly to interface the low emf pomona banana 
cables to binding posts. I have stopped this, reasons being, they are large and 
worse, that the pomona spring loaded insulation tube that covers the banana 
plug conductor uses such a strong spring that slowly the plug works its way out 
of the spade. this btw also happend to me when I used the pomona low emf 
binding posts together with the pomona low emf banana cables. overall I m not 
happy with these.
so, due to lack of options, I changed to self-made twisted shielded pair of 
high grade teflon/kapton silver plated copper cable with gold plated copper 
spades (crimped). I use them not only with the 3458a but also with nanovolt 
meters. these have higher resolution and accuracy in low level measurements 
than the 3458a. emf voltages were never an issue with these cables if properly 
used. I have posted some results doing 34420a stabilty measurements on the pmel 
forum, and the results are convincing (purpose was actually not to test the 
cables but the stability of the 34420a, but the emf issue is a part of this of 
course. we use the 34420a to do low voltage precision measurements on thermal 
converters where the full scale signal sometimes is 1mV).
that btw also relates to don's statements below, I do not concurr with his 
comments about copper telurium as cable and spade material and so on. this 
material, as stated here many times, is used because it is machinable, for 
copper spades one would not use it. the 34420a factory cable uses copper cable 
and copper spades, not telurium-copper. if there was a problem, it would be 
worse with the 34420a than with the 3458a because of its low level ranges. and 
again, I have not seen any problems in a chain of (output to input):
1.copper-tellurium post from e.g. 8 digit calibrator
2.crimped copper spade, gold plated
3.silver plated tsp copper cable
4a.crimped copper spade to copper-tellurium post or
4b.soldered copper connector(34420)
my consistent results over more than a year using them.



Gesendet: Montag, 25. August 2014 um 06:33 Uhr
Von: "Orin Eman" <orin.e...@gmail.com>
An: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
Betreff: Re: [volt-nuts] 732A and Prologix received

On Sun, Aug 24, 2014 at 1:46 PM, Don@True-Cal <truecalservi...@gmail.com>
wrote:

Randy & all,

You have correctly concluded that some (maybe not all) of your measurement
problem is thermal EMF being added or subtracted in series within your
measurement interconnect. This thermal EMF is generated at the junction of
dissimilar metals when accompanied with thermal gradients between the test
lead and device terminals. You have to eliminate both the dissimilarity of
the metal junctions as well as minimize the thermal differences. The
terminals of the 3458A as well as the 732A are Beryllium Copper so you want
to use the same test lead terminals. Forget the typical Tin plated lugs or
even Gold plated as both are not Beryllium Copper and constitute dissimilar
metals. The best solution (as usually the most expensive) is to use a set
of
Fluke 5440A-7005 (48") cables. I also have just as good results using the
much more flexible Pomona 11174A (lugs end always stay connected to the
732A) or 11058A with more convenient shielded banana plugs. The Fluke cable
has the added Guard built in but be sure to also use a Guard lead with the
Pomona cabled. The Guard lead does not need to be low thermal EMF. DIY
cables is usually not a good idea because the lead wire to terminal also
constitutes just as critical of junction. The above cables use Tellurium
Copper wire which is usually hard to find and hard to crimp properly and
NEVER solder.


11058A and 11174A are discontinued at Keysight.  However, Pomona 5295 spade
to banana cables are available (5295-36 at Mouser et al) and claim that
they are designed to minimize thermal EMFs.  Datasheet is here:
http://www.mouser.com/ds/2/159/d5295_1_01-51722.pdf  Any comments on these
as an alternative?

Orin.
_______________________________________________
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
_______________________________________________
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.





_______________________________________________
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Reply via email to