On 7/27/2015 10:26 PM, Mike S wrote: > On 7/27/2015 7:54 PM, Bill Gold wrote: >> You will observe that the HP/Agilent/Keysight manual for the 3458A >> does not give any "zero" stability specs, at least that I can find. > > On DC 100 mV range, the standard model is spec'd for 14+3 ppm of range @ > 2 years from calibration, not considering temperature. That's 1.7 uV, > from my reckoning. The OP's reading of 0.7 uV is well within that - it > even beats the 90 day spec.
Doh. That 14+3 actually equates to ~=3 ppm ~= 0.3 uV (since reading is essentially zero), but there's also temperature, which contributes > 0.115 / C. _______________________________________________ volt-nuts mailing list -- [email protected] To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.