Update:  After changing the resistor I added in series with R1008 (A decade 
position 1.0) from 204.8 to 202.4 Ohms (I wonder if this means that R302 is 
slowly increasing in value back towards nominal) and re-calibrating again, the 
A decade is now within 1ppm of linear (using the 10V range and nominally 10V 
input).

I note that when I'd done this, with the A decade set to 0.9 (reading 9.000000V 
on the 3458A), the actual input voltage needed was 10.000040 or so.  Is that to 
be expected?   

Two of the positions on the B decade wouldn't quite adjust for a null when 
calibrating it.   In this situation I think I have two options:

        1) Add series resistors to compensate for slightly low value 9.898k 
resistors in the relevant two positions of the decade and recalibrate.

        2) Adjust R203 to change the bridge balance slightly to the 10K ohm 
decades and recalibrate.

Your thoughts on which approach to take is much appreciated.

Thanks
Dave

_______________________________________________
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Reply via email to