I just did a quickie experiment to get higher resolution. I rigged up an external meter movement about the same size as the internal one, hooked to the monitor out, which is 3V FS. The external meter is set up for about ten times the sensitivity, so 300 mV FS. I wanted to see how the noise looks, with similar meter ballistics, and expand the zone around zero to assess the bias current. The net result is the external meter reads about 10 fA FS, and the noise and drift are still low enough to be quite usable. The latest bias current estimate is about 50-100 aA, slowly dithering around less than plus one percent of FS, with occasional jumps to about 500 aA or less. I can't tell yet if these jumps are part of the 1/f noise, or line noise and transients getting through the power supplies.

Anyway, presuming I didn't make any mistakes in my measuring and figuring, this is quite impressive. I hope I just didn't get lucky with one particular part, but I think the others will be similar. I will eventually be checking them all, and other part types, in test setups.

For reference, I confirmed the scaling. With the 417 set up for maximum sensitivity, it's 100 fA FS. Putting the suppression supply at 100 mV through the top 1E12 resistor, it reads near full scale. Setting at 10 mV, the internal meter reads ten percent of FS, and the external meter reads near its full scale. BTW I can set the suppression directly with mV resolution. Long ago, I changed the last digit pot to a ten-turn precision type with a kilodial indicator - it's nice.

This is now over a hundred times better than the original spec for the 417's grid current. When it's all said and done, I'll likely add a 10X meter switch arrangement in the 417, to get a 10 fA FS range capability too. It may be good even at much higher sensitivity, but 10X is pretty good, and I likely won't need any more, practically speaking. Besides, putting a DVM (and some filtering) on the output can expand it even more, if ever necessary.

One problem I did find, is there's some line ripple appearing in the output signal, which I need to investigate. It's pretty small - about one percent FS - and the meters don't notice it, but it still bugs me, since it's way bigger than the apparent random noise.

Ed