I skimmed through it, one thing that struck me was that they hit 1372°C for
10 minutes.  I have serious doubts that their stainless steel vessel could
have survived such a temperature (barely bellow melting) - which makes me
suspicious of an error somewhere, this is above where k-type thermocouples
would typically be expected to be accurate or reliable.  Also melting point
depression would have melted nickel powder at such temps destroying (what I
thought was important) nickel surface morphology.

They seem to have good resources so hoping they do better calorimetry in
future.

On 31 May 2015 at 04:19, Bob Higgins <rj.bob.higg...@gmail.com> wrote:

> Type B thermocouples are expensive; even for fine wire, short, uninsulated
> couples, because they are made from platinum.  They may be 10x more
> expensive than type-K and extension wires are just as expensive.
> Additionally the signal level is smaller with type-B which means more noise
> in the measurement.
>
>
> On Sat, May 30, 2015 at 11:41 AM, Jed Rothwell <jedrothw...@gmail.com>
> wrote:
>
>> a.ashfield <a.ashfi...@verizon.net> wrote:
>>
>>  Beats me why they don't use type S or type B thermocouples that are
>>> common in the glass industry.
>>>
>>
>> That probably would be better. You should suggest it to Jiang. (His
>> e-mail in the slides. He is a good guy.)
>>
>> The K-type thermocouple maxed out.
>>
>> They have to replace the inner thermocouple (T3) in any case.
>>
>> - Jed
>>
>>
>