Hi Stephan,

If there were overflows on the detector, which cause lines due to the spill over of the wells of the CCD, the lines would be visible in the readout direction of the CCD detector, which generally is from bottom to top or top to bottom. You can also see this with your own (pocket) camera if you point it towards a bright light source, before you take the actual picture. The lines cannot be in a skew direction and will always be connected to very strong reflections. As you can see in the images Margriet sent, the lines are parallel to one of the reciprocal planes and there are plenty of lines in the image that are not connected to reflections.

With very (and I mean very) strong reflections you will sometimes see an speckled arc around the reflection. This is diffraction of the beryllium window of the detector, where the reflected beam acts as the primary beam (see attached picture.)

Bram


Stephan Ginell wrote:
Hi.... Such line can occur on a CCD detector if reflection are saturating a pixel and is generally in the direction of the detector readout. 1) what detector are you using, 2) are reflections in the black center saturated i.e. Greater than the dynamic range of the detector. 3) what is your exposure time, 4) do you see such streaks on short / attenuated exposures? 5 do you see such streaks on dark images of (no x-rays) but same time. Steve
----------------------------------------------------------------------------