Hello Andrew,

To the best of my knowledge, IEC 61010-1:2001 is not intended to be the base
standard for previously existing 61010 part 2's or 3's. The second edition
of IEC 61010-1 stands alone until the various parts 2 and 3 are updated and
reference the second edition. In my opinion, you are better off in
continuing to use the first edition of IEC 61010-1 until the appropriate
part 2's and 3's you use are updated to the second edition and the first
amendment to the second edition is published. 

Regards,
+=================================================================+
|Ronald R. Wellman                |Voice : 408-345-8229           |
|Agilent Technologies             |FAX   : 408-553-2412           |
|5301 Stevens Creek Blvd.,        |E-Mail: ron_well...@agilent.com|
|Mailstop 54L-BB                  |WWW   : http://www.agilent.com |
|Santa Clara, California 95052 USA|                               |
+=================================================================+
| "Common sense is the collection of prejudices acquired by age   |
|  eighteen." - Albert Einstein                                   |
+=================================================================+



-----Original Message-----
From: Andrew Wood [mailto:andrew.w...@landinst.com]
Sent: Thursday, March 28, 2002 2:20 AM
To: 'emc-pstc'
Subject: EN 61010-1 2001 and IEC61010-3



All
I have a question regarding interpretation of section 6.10.2 Fitting of
non-detachable MAINS supply cords
in EN 61010-1:2001.

Cord anchorages shall meet the following requirements
a)...d)

e) A compression bushing shall not be used as a cord anchorage unless it has
provision for
   clamping all types and sizes of MAINS supply cords which meet the
requirements of 6.10.1
   and are suitable for connection to the TERMINALS provided, or the bushing
has been 
   designed to terminate a screened MAINS supply cord.

f)...

I would interpet this as follows:
If the bush has been designed to accept a screened cable then it is ok for
the bush to accept only a specific cable type 
which should be specified.
Otherwise the bush must be capable of accepting any mains cable which has
the neccessary voltage/current rating and will
be suitable for the terminals provided.

However in IEC61010-3-1 first edition 1997-07 conformity verification report
for IEC 61010-1
there is a quite different interpretation.

This document is basically a sample test report sheet for IEC61010-1. It is
based on earlier versions of 61010-1
with ammendments 1 and 2 added. 

Here we have:
6.10.2.2 cord anchorages
compression bushing:
a) clamps all types and sizes of mains cords

and

b) is suitable:
                     1)  for connection to TERMINALS provided
              or
                     2) it is designed for screened mains cord.


I know that this document refers to an earlier version of 61010. However
1997 edition is still the only one listed on IEC website.
I do not currently have access to an earlier version of 61010-1 to look at,
but the interpretation above seems so anti- common sense
 that I am doubtful that it was faithful to the earlier version of the
standard.

So bottom line is: is my understanding of the current requirements of
EN61010-1 6.10.2 correct?

Secondly and out of curiosity does anyone know why IEC61010-1-3 1997 seems
to contradict this?

Thank in advance
Andy.

Andrew Wood
Land Instruments International
England.



            






This e-mail and its contents may be confidential, privileged and protected 
by law. Access is only authorised by the intended recipient. The contents 
of this e-mail may not be disclosed to, or used by, anyone other than the 
intended recipient, or stored or copied in any medium. If you are not the 
intended recipient, please advise the sender immediately.



-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
     majord...@ieee.org
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Ron Pickard:              emc-p...@hypercom.com
     Dave Heald:               davehe...@attbi.com

For policy questions, send mail to:
     Richard Nute:           ri...@ieee.org
     Jim Bacher:             j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
    http://ieeepstc.mindcruiser.com/
    Click on "browse" and then "emc-pstc mailing list"

-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
     majord...@ieee.org
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Ron Pickard:              emc-p...@hypercom.com
     Dave Heald:               davehe...@attbi.com

For policy questions, send mail to:
     Richard Nute:           ri...@ieee.org
     Jim Bacher:             j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
    http://ieeepstc.mindcruiser.com/
    Click on "browse" and then "emc-pstc mailing list"

Reply via email to