TNO made a suggestive title by referring to RFID devices, where most problems
are induced by only 2 of the available systems. Probably to draw more
attention for their article, and they succeeded.

TNO is a Netherlands institute for applied science and have been forced to
change their way of being financed from public to private.

 

Apparently, most of the problems occur with high power 868 MHz backscatter
RFID systems. These are conditionally allowed to transmit up to 3.8 watts EIRP
(Friss tells us that that might be up to  50 V/m ) For frequency power
allocations in NL :

 

http://www.agentschaptelecom.nl/nfr/main_nfr_uk.html

 

This is close to  the  GSM-standard and power range, and we all know that
these devices are to be used with care (forbidden) in medical environments.

 

These RFID’s  are definitely not the items we usually use as RFID such as 
Legic, Mifare or other low power RFID systems.

The article should therefore be read as a warning against high power RFID
backscatter technologies.

While we can discuss about if  Medical equipment should or should not be
immune to this field strength, we may discuss the suitability of these 868 MHz
backscatter radar type RFID systems,  in medical and  residential 
environments where fields of 3 or maximum 10V/m should be expected.

I wonder btw how this allowance of  2W ERP (3.8 W EIRP) got into the European
frequency tables. Maybe because the power and frequency are close to GSM. Not
taking into account the much closer operating distance that goes with RFID.

 

For the time being we will have to live with RFID technologies  that interfere
with medical (and other) electronics, until the  standards committees again 
noticed their lag on technology. and raise the immunity levels to  50 V/m or
more.

 

(EN 55024  for example still uses 3V/m and limits the test range to 1 GHz only
, where 1.8 GHz GSM are used all over Europe, and even higher frequency UMTS
systems are being rolled out: new problems are to be expected!)

 

Regarding the discussion: both interfering RFID systems should not be
‘simulated’ with the EN 6100-4-6 method, as the 125 kHz is below it’s 
f-limits (and would go unnoticed), and the other is way above it’s upper
f-limits (80  or 230 MHz)

 

 

Gert Gremmen

ce-test, qualified testing bv

 

Van: emc-p...@ieee.org [mailto:emc-p...@ieee.org] Namens Ken Javor
Verzonden: Friday, June 27, 2008 12:57 AM
Aan: Untitled
Onderwerp: Re: EMC in the news: RFID & Medical

 

None of the below.  61000-4-3 specified rf source impedance is 150 Ohms,
calibrated into 150 Ohms load impedance. The 1, 3, and 10 Volt levels are
open-circuit quantities, thus the potential developed in the matched load is
one-half the limit.  The current flowing on a cable under test will depend on
the cable impedance, which is not controlled, except on the AE side, because
the injection device is designed to insert a high common mode impedance on the
AE side of the cable (high relative to 150 Ohms).  As far as I know, you do
not measure injected current doing the test this way.  There is an alternative
current injection test when a suitable coupling device is not available, but
then you inject and measure current, not rf potential.  The current limit is
the short circuit current available from the open circuit potential limit
divided by the 150 Ohm source impedance, even though you will not have a 150
Ohm source impedance when injecting current.
 
Ken Javor

Phone: (256) 650-5261



________________________________

From: Daniel Roman <dan.ro...@dialogic.com>
List-Post: emc-pstc@listserv.ieee.org
List-Post: emc-pstc@listserv.ieee.org
List-Post: emc-pstc@listserv.ieee.org
Date: Thu, 26 Jun 2008 17:53:59 -0400
To: Ken Javor <ken.ja...@emccompliance.com>, Untitled <emc-p...@ieee.org>
Conversation: EMC in the news: RFID & Medical
Subject: RE: EMC in the news: RFID & Medical

While slightly off-topic, this discussion got me thinking about the
measurement of the voltage on the cable.  If done with a RF cable current
clamp, do you have to do a conversion of some sort if the cable is assumed to
be 150 ohms?
 
For a 50 ohm system, a 3V level measured with a clamp would be 96 dBuA minus
the transfer impedance in dB ohms.  If the cable is 150 ohms though does that
mean you are measuring the equivalent of 1V from the clamp instead of 3V
because the clamp is measuring current?  Stated another way, assuming the
transfer impedance of the current clamp is zero to simply things, for an
assumed 150 ohm cable should you be reading 96 dBuV on a spectrum analyzer
hooked up to the probe or something less that than equivalent to 1V?
 
Never worked with this stuff so I’m curious.
 
Dan
 

From: emc-p...@ieee.org [mailto:emc-p...@ieee.org] On Behalf Of Ken Javor
Sent: Thursday, June 26, 2008 11:14 AM
To: Untitled
Subject: Re: EMC in the news: RFID & Medical

Comment (e) is unrealistic.  Assuming 61000-4-6 was applied, the amount of
power required to directly inject 1, 3 or 10 Volts oc into 150 Ohms is well
above the radiated power from an rf id device. 
Ken Javor

Phone: (256) 650-5261

________________________________

From: "Conway, Patrick R (Houston)" <p.con...@hp.com>
List-Post: emc-pstc@listserv.ieee.org
List-Post: emc-pstc@listserv.ieee.org
List-Post: emc-pstc@listserv.ieee.org
Date: Thu, 26 Jun 2008 14:17:00 +0000
To: Gert Gremmen <administra...@ce-test.info>, "Rudd, Adam"
<ar185...@ncr.com>, "emc-p...@ieee.org" <emc-p...@ieee.org>
Conversation: EMC in the news: RFID & Medical
Subject: RE: EMC in the news: RFID & Medical

Gert-
    You mention one possible reason for the test results is that the EUT's
have immunity deficits.
      I agree, that is one possibility.  

  here are a few other possibilities:   
        (BTW- not affiliated in any way with medical devices nor with, all
comments OOO )

 
b)    Could it be a systemic testing error?
            Was a shield room used?   
            Were the ambients controlled and eliminated?
            Were the devices connected to a patient simulator?
            Did the test engineer have his personal GSM phone "OFF"?  etc. 
(...crazier things have happened.)

 
c)    Maybe the EUT are old.
            Through outdated design specs perhaps RFID proximity was not a
consideration during their design.
          And yet, in today's hospitals, the two types of devices may be in
close proximity.
          If this is the case then the study has done a great service to the
community by uncovering a problem that was unknown.

 
d)    Maybe the EUT are old (not a repeat) 
          Through many years of use perhaps once immune equipment has lost
some of their designed immunity?
          Again- if this is the case this study may have uncovered a
previously unknown problem.

    
e)    Is it possible for an RF ID device to overwhelm the immunity levels of
the EUT?
        If a medical device is tested at 10 V/m and an RFID device TXout is in
the mW range- is it possible for a RFID mW transmitter to generate 10 V/m?
        Perhaps.
        For instance- since an RFID device operating at 125 kHz in not
transmitting in the classic sense, then there may be near-field resonant
effects that are not previously understood in the medical device immunity
requirements?
       

 
  It seems that we, as professionals in this field, have the unique ability to
analyze these reports like no other community can.  I wonder if we will find
the answers to the large number of questions raised by the article.  

 
 
All comments OOO.
Best Regards, 

Patrick. 
p.con...@hp.com 

  

________________________________


From: Gert Gremmen [mailto:administra...@ce-test.info] 
Sent: Thursday, June 26, 2008 2:57 AM
To: Conway, Patrick R (Houston); Rudd, Adam; emc-p...@ieee.org
Subject: RE: EMC in the news: RFID & Medical

The report was produced by TNO, a Dutch private organization
(http://www.tno.nl/content.cfm?&context
markten&content=markt_persbericht&laag1=189&item_id=200806250026&Taal=2
<http://www.tno.nl/content.cfm?&amp;con
ext=markten&amp;content=markt_persberic
t&amp;laag1=189&amp;item_id=200806250026&amp;Taal=2>
<http://www.tno.nl/content.cfm?&context
markten&content=markt_persbericht&laag1=189&item_id=200806250026&Taal=2> 
<http://www.tno.nl/content.cfm?&amp;context=markten&amp;content=markt_persbericht&amp
laag1=189&amp;item_id=200806250026&amp;Taal=2>
<http://www.tno.nl/content.cfm?&context
markten&content=markt_persbericht&laag1=189&item_id=200806250026&Taal=2>  )
 
and some results are available here:
 
http://www.amc.nl/?pid=5266
 
Manufacturers name and equipment type included.
 
Please note that the energy levels of RFID are in the milliwatt range,
so all problems are to be categorized as immunity deficits.
 
 
It is astonishing that the security of healthy persons (like car drivers) 
is taken much more seriously (by car manufacturers for example )
as the security  of people with bad health like in hospitals.
Most medical equipment is tested  at 10 V/meter or less.
where critical car parts must  be tested up to 200V/m.
 
Cars are to be sold at low prices (relatively) , medical equipment
at sky-high costs.  It seems that emc quality is the inverse of the
costs of equipment. Where much attention is given to
reliability and electrical safety of medical equipment,
emc is still  neglected. 
Note that this investigation was made in a Dutch
University Hospital;
On  European soil , where immunity requirements have been 
virtually law since 1996 !!!!
 
The lack of EMC care might be related to the fact that
medical accidents are easy to cover up, (more easy then car accidents)
and liability of medical staff is difficult to prove, let alone
the liability of a medical equipment manufacturer.
See the discussion on the Therac-25.
 
Gert Gremmen
 
 
 
 
 
 
 

Van: emc-p...@ieee.org [mailto:emc-p...@ieee.org] Namens Conway, Patrick R
(Houston)
Verzonden: Wednesday, June 25, 2008 4:08 PM
Aan: Rudd, Adam; emc-p...@ieee.org
Onderwerp: RE: EMC in the news: RFID & Medical

...and if you put four of them in a circle, you can pop corn.  :)

 

 

Best Regards, 
Patrick. 
p.con...@hp.com 

 
   

________________________________



From: emc-p...@ieee.org [mailto:emc-p...@ieee.org] On Behalf Of Rudd, Adam
Sent: Wednesday, June 25, 2008 7:19 AM
To: emc-p...@ieee.org
Subject: EMC in the news: RFID & Medical
“The latest research, conducted at Vrije University in Amsterdam, tested the
effect of holding both "passive" and powered RFIDs close to 41 medical
devices, including ventilators, syringe pumps, dialysis machines and
pacemakers.

A total of 123 tests, three on each machine, were carried out, and 34 produced
an "incident" in which the RFID appeared to have an effect - 24 of which were
deemed either "significant" or "hazardous".

In some tests, RFIDs either switched off or changed the settings on mechanical
ventilators, completely stopped the working of syringe pumps, caused external
pacemakers to malfunction, and halted dialysis machines.

The device did not have to be held right up to the machine to make this happen
- some "hazardous" incidents happened when the RFID was more than 10 inches
away.”

--http://news.bbc.co.uk/2/hi/health/7471008.stm
<http://news.bbc.co.uk/2/hi/health/7471008.stm>
<http://news.bbc.co.uk/2/hi/health/7471008.stm>  

Best Regards,

Adam Rudd

Electrical Engineer (EMC)

NCR Corporation, RHSS

Duluth, GA

(770) 495-2825

- ---------------------------------------------------------------- This
message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/ 
To post a message to the list, send your e-mail to emc-p...@ieee.org 

Instructions: http://listserv.ieee.org/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators: 

Scott Douglas emcp...@ptcnh.net Mike Cantwell mcantw...@ieee.org 

For policy questions, send mail to: 

Jim Bacher: j.bac...@ieee.org David Heald: emc-p...@daveheald.com 

All emc-pstc postings are archived and searchable on the web at: 

http://www.ieeecommunities.org/emc-pstc

- ---------------------------------------------------------------- This
message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/ 
To post a message to the list, send your e-mail to emc-p...@ieee.org 

Instructions: http://listserv.ieee.org/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators: 

Scott Douglas emcp...@ptcnh.net Mike Cantwell mcantw...@ieee.org 

For policy questions, send mail to: 

Jim Bacher: j.bac...@ieee.org David Heald: emc-p...@daveheald.com 

All emc-pstc postings are archived and searchable on the web at: 

http://www.ieeecommunities.org/emc-pstc 
- ---------------------------------------------------------------- This
message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list.    Website:  http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to emc-p...@ieee.org 

Instructions:  http://listserv.ieee.org/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators: 

 Scott Douglas           emcp...@ptcnh.net Mike Cantwell          
mcantw...@ieee.org 

For policy questions, send mail to: 

 Jim Bacher:             j.bac...@ieee.org David Heald:           
emc-p...@daveheald.com 

All emc-pstc postings are archived and searchable on the web at: 

 http://www.ieeecommunities.org/emc-pstc

- ---------------------------------------------------------------- This
message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/ 
To post a message to the list, send your e-mail to emc-p...@ieee.org 

Instructions: http://listserv.ieee.org/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators: 

Scott Douglas emcp...@ptcnh.net Mike Cantwell mcantw...@ieee.org 

For policy questions, send mail to: 

Jim Bacher: j.bac...@ieee.org David Heald: emc-p...@daveheald.com 

All emc-pstc postings are archived and searchable on the web at: 

http://www.ieeecommunities.org/emc-pstc 
- ---------------------------------------------------------------- This
message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list.    Website:  http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to emc-p...@ieee.org 

Instructions:  http://listserv.ieee.org/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators: 

 Scott Douglas           emcp...@ptcnh.net Mike Cantwell          
mcantw...@ieee.org 

For policy questions, send mail to: 

 Jim Bacher:             j.bac...@ieee.org David Heald:           
emc-p...@daveheald.com 

All emc-pstc postings are archived and searchable on the web at: 

 http://www.ieeecommunities.org/emc-pstc

- ---------------------------------------------------------------- This
message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to emc-p...@ieee.org 

Instructions: http://listserv.ieee.org/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators: 

Scott Douglas emcp...@ptcnh.net Mike Cantwell mcantw...@ieee.org 

For policy questions, send mail to: 

Jim Bacher: j.bac...@ieee.org David Heald: emc-p...@daveheald.com 

All emc-pstc postings are archived and searchable on the web at: 

http://www.ieeecommunities.org/emc-pstc 

- ---------------------------------------------------------------- This
message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to emc-p...@ieee.org 

Instructions: http://listserv.ieee.org/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators: 

Scott Douglas emcp...@ptcnh.net Mike Cantwell mcantw...@ieee.org 

For policy questions, send mail to: 

Jim Bacher: j.bac...@ieee.org David Heald: emc-p...@daveheald.com 

All emc-pstc postings are archived and searchable on the web at: 

http://www.ieeecommunities.org/emc-pstc 

Reply via email to