Yes, that means the product works on the lower voltages and to use it on the 
higher voltages you have to move a switch. There is an IEC standard:

IEC 61293  
Marking of electrical equipment with ratings
related to electrical supply – Safety
requirements

With best wishes DESIGN IT IN! OOO – Own Opinions Only
www.jmwa.demon.co.uk J M Woodgate and Associates Rayleigh England

Sylvae in aeternum manent.


-----Original Message-----
From: Kunde, Brian [mailto:brian_ku...@lecotc.com] 
Sent: Monday, April 10, 2017 4:25 PM
To: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: Re: [PSES] Voltage Rating vs Voltage Specification

Is something like this allowed?

100-115-120/208-220-230-240

Will a ±10% tolerance always be assumed? If your tolerance was something 
different, such as -15%/+10%, does this information have to be on the device or 
is the manual good enough?

The Other Brian

-----Original Message-----
From: John Woodgate [mailto:jmw1...@btinternet.com]
Sent: Friday, April 07, 2017 5:01 PM
To: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: Re: [PSES] Voltage Rating vs Voltage Specification

These runes differ in meaning:

I have seen products rated 85-264V and others rated 100/120/208/230/240.

85-264 means any voltage within that range. You could put in 165 V and expect 
no problem.

100/120/208/230/240 means only those voltages, with whatever the relevant 
standard says about tolerances. 165 V would not work for this product.

With best wishes DESIGN IT IN! OOO – Own Opinions Only www.jmwa.demon.co.uk J M 
Woodgate and Associates Rayleigh England

Sylvae in aeternum manent.


-----Original Message-----
From: Ralph McDiarmid [mailto:ralph.mcdiar...@schneider-electric.com]
Sent: Friday, April 7, 2017 8:29 PM
To: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: Re: [PSES] Voltage Rating vs Voltage Specification

I don't know, but I suppose it could be addressed by an "abnormal test" to see 
if UUT fails in a way which then renders it potentially unsafe by way of 
non-compliance with a criterion in the standard.

I have seen products rated 85-264V and others rated 100/120/208/230/240.   I 
would expect the first one to pass thermal test criteria at 85V, at rated 
power, at highest rated ambient .  One corner of the "performance envelope" if 
you will.  And then, do I test the latter at 100V -10%   ?

And, I don't think that compliance with a standard proves a product safe; only 
that it complies with a specific set of criteria.  Product safety is hard to 
define, much harder to accurately assess, even with use of the AFMEA and FTA 
tools, which are subjective so it seems to me.

Ralph McDiarmid
Product Compliance
Engineering
Solar Business
Schneider Electric



-----Original Message-----
From: Richard Nute [mailto:ri...@ieee.org]
Sent: Friday, April 07, 2017 10:51 AM
To: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: Re: [PSES] Voltage Rating vs Voltage Specification

Supposedly, since the ratings are specified in the standard, they must involve 
safety if not done according to the standard.  So, what is the safety issue if 
the ratings are not in accordance with the standard?  What is the injury?

What is the safety issue if the applied voltage is less than or more than the 
marked ratings but still within the nominal from the electric power utility?  
Again, what is the injury?

Rich


> -----Original Message-----
> From: John Woodgate
> [mailto:jmw1...@btinternet.com]
> Sent: Friday, April 07, 2017 9:11 AM
> To: EMC-PSTC@LISTSERV.IEEE.ORG
> Subject: Re: [PSES] Voltage Rating vs Voltage Specification
>
> There are in fact two IEC resources, Electropedia, which has all the
> formal definitions produced by TC1 and Glossary, which has a selection
> of terms, culled from many standards, that have not been adopted by
> TC1.
>
> http://www.electropedia.org/?ref=extfooter
>
> http://std.iec.ch/glossary?ref=extfooter
>
> Neither can be comprehensive at one instant, because new terms are
> being added all the time.
>
> With best wishes DESIGN IT IN! OOO – Own Opinions Only
> www.jmwa.demon.co.uk J M Woodgate and Associates Rayleigh England
>
> Sylvae in aeternum manent.
>
> -----Original Message-----
> From: Ralph McDiarmid
> [mailto:Ralph.McDiarmid@SCHNEIDER-
> ELECTRIC.COM]
> Sent: Friday, April 7, 2017 4:42 PM
> To: EMC-PSTC@LISTSERV.IEEE.ORG
> Subject: Re: [PSES] Voltage Rating vs Voltage Specification
>
> Pete, I wonder if the onus to define the terminology like “rated
> voltage” should really be on the technical committees, not academia.
> I know that is some standards, terms like "disconnect" and "trip" are
> loosely defined.  I wonder if there should be one IEC document, which
> could serve as a reference to all others for terminology.  I think
> there is one, but it is likely not comprehensive.
>
> Ralph McDiarmid
> Product Compliance
> Engineering
> Solar Business
> Schneider Electric
>
>
>
> From: Pete Perkins [mailto:00000061f3f32d0c-dmarc- requ...@ieee.org]
> Sent: Thursday, April 06, 2017 10:20 PM
> To: EMC-PSTC@LISTSERV.IEEE.ORG
> Subject: Re: [PSES] Voltage Rating vs Voltage Specification
>
> All,
>
>                Yes, the consultant or safety engineers
> dream/nightmare.  We have to realize that the glass is half empty for
> most of the world and we have an ongoing opportunity to strike them
> across the knuckles with a ruler (as the nuns did in primary school)
> and begin the teaching mode.
>
>                As PT Barnum (the American circus
> entrepreneur) once said (and quoted often) ‘There is a fool born every
> minute’.
>
>                If the technical schools provided all of this detailed
> training we wouldn’t have anything to do.
>
>                So fill your peddler’s sack with all of these important
> stories and smile, but not laugh out loud, when you run into the same
> situation again (and again, and again).
>
>                Every project and every design team is an opportunity
> to straighten out the world.
>
> :>)     br,      Pete
>
> Peter E Perkins, PE
> Principal Product Safety & Regulatory Affairs Consultant PO Box 23427
> Tigard, ORe  97281-3427
>
> 503/452-1201
>
> mailto:p.perk...@ieee.org
>
> From: john Allen
> [mailto:john_e_al...@blueyonder.co.uk]
> Sent: Thursday, April 6, 2017 12:57 AM
> To: mailto:EMC-PSTC@LISTSERV.IEEE.ORG
> Subject: Re: [PSES] Voltage Rating vs Voltage Specification
>
> As Dave said, this is “age old problem” that we also faced at HP
> Bristol in the 1980s when we built peripherals that had to operate
> across the World – and that meant from 100V 50 & 60Hz in Japan and up
> to 240V 50Hz for the UK and a few other places, AND +/- to cope with
> the relevant required local tolerances (and so effectively meant +/-
> 10% across the board). This meant careful selection and testing of
> PSUs and of the ratings to be marked on the end-use products, but
> fortunately most of our products did not have directly mains-powered
> motors.
>
> In practice, if you have a product that does have such motors then it
> may well mean that you need to produce separate models with different
> motors for the geographical areas that operate at the extremes of the
> voltage/frequency ranges – especially  those at the lower end thereof
> – or else change the designs to use DC motors supplied from full
> voltage/frequency range- capable PSUs (or, possibly, use AC motors
> rated for the lowest “worst case” voltage/ frequency / tolerance
> combination, but with solid state control systems which ensure that
> those motors are operated within that regime regardless of the actual
> supplied mains voltages/ frequencies/ tolerances?).
>
> OTOH, the latter approaches are probably impracticable in most cases
> for cost /space /weight / technology reasons , and so that  means you
> need a “horses for courses” approach.
>
> John E Allen
> W. London, UK
>
> -
> ----------------------------------------------------------------
> This message is from the IEEE Product Safety Engineering Society
> emc-pstc discussion list. To post a message to the list, send your
> e-mail to <emc- p...@ieee.org>
>
> All emc-pstc postings are archived and searchable on the web at:
> http://www.ieee-pses.org/emc-pstc.html
>
> Attachments are not permitted but the IEEE PSES Online Communities
> site at http://product- compliance.oc.ieee.org/ can be used for
> graphics (in well-used formats), large files, etc.
>
> Website:  http://www.ieee-pses.org/
> Instructions:  http://www.ieee-pses.org/list.html
> (including how to unsubscribe)
> List rules: http://www.ieee-pses.org/listrules.html
>
> For help, send mail to the list administrators:
> Scott Douglas <sdoug...@ieee.org>
> Mike Cantwell <mcantw...@ieee.org>
>
> For policy questions, send mail to:
> Jim Bacher:  <j.bac...@ieee.org>
> David Heald: <dhe...@gmail.com>
>
> -
> ----------------------------------------------------------------
> This message is from the IEEE Product Safety Engineering Society
> emc-pstc discussion list. To post a message to the list, send your
> e-mail to <emc- p...@ieee.org>
>
> All emc-pstc postings are archived and searchable on the web at:
> http://www.ieee-pses.org/emc-pstc.html
>
> Attachments are not permitted but the IEEE PSES Online Communities
> site at http://product- compliance.oc.ieee.org/ can be used for
> graphics (in well-used formats), large files, etc.
>
> Website:  http://www.ieee-pses.org/
> Instructions:  http://www.ieee-pses.org/list.html
> (including how to unsubscribe)
> List rules: http://www.ieee-pses.org/listrules.html
>
> For help, send mail to the list administrators:
> Scott Douglas <sdoug...@ieee.org>
> Mike Cantwell <mcantw...@ieee.org>
>
> For policy questions, send mail to:
> Jim Bacher:  <j.bac...@ieee.org>
> David Heald: <dhe...@gmail.com>

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<emc-p...@ieee.org>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://www.ieee-pses.org/list.html (including how to 
unsubscribe) List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <sdoug...@ieee.org>
Mike Cantwell <mcantw...@ieee.org>

For policy questions, send mail to:
Jim Bacher:  <j.bac...@ieee.org>
David Heald: <dhe...@gmail.com>

______________________________________________________________________
This email has been scanned by the Symantec Email Security.cloud service.
______________________________________________________________________

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<emc-p...@ieee.org>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://www.ieee-pses.org/list.html (including how to 
unsubscribe) List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <sdoug...@ieee.org>
Mike Cantwell <mcantw...@ieee.org>

For policy questions, send mail to:
Jim Bacher:  <j.bac...@ieee.org>
David Heald: <dhe...@gmail.com>

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<emc-p...@ieee.org>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://www.ieee-pses.org/list.html (including how to 
unsubscribe) List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <sdoug...@ieee.org>
Mike Cantwell <mcantw...@ieee.org>

For policy questions, send mail to:
Jim Bacher:  <j.bac...@ieee.org>
David Heald: <dhe...@gmail.com>
________________________________

LECO Corporation Notice: This communication may contain confidential 
information intended for the named recipient(s) only. If you received this by 
mistake, please destroy it and notify us of the error. Thank you.

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<emc-p...@ieee.org>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://www.ieee-pses.org/list.html (including how to unsubscribe)
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <sdoug...@ieee.org>
Mike Cantwell <mcantw...@ieee.org>

For policy questions, send mail to:
Jim Bacher:  <j.bac...@ieee.org>
David Heald: <dhe...@gmail.com>

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<emc-p...@ieee.org>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://www.ieee-pses.org/list.html (including how to unsubscribe)
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <sdoug...@ieee.org>
Mike Cantwell <mcantw...@ieee.org>

For policy questions, send mail to:
Jim Bacher:  <j.bac...@ieee.org>
David Heald: <dhe...@gmail.com>

Reply via email to