   1.   What equipment do you use for Alpha smear and air sample counting?

   Canberra / Tennelec 5XLB  and Ludlum 2000 with 43/10 detector, we have
      iSOLO but not  yet in service

   2.   What equipment do you use for Beta smear and air sample counting?

   Canberra / Tennelec 5XLB

   3.   What statistical tests (LAC) do you perform on your scaler counting
      equipment, e.g., Chi-Squared Testing, Confidence Interval Testing,
      etc.?

   Chi-square test when establishing QC chart data following calibration,
      daily checks of source count with +/- 3 sigma

   4.   Where / when do you perform long term performance checks, e.g., are
      after instrument calibration, in the Cal Lab facility or at the use
      location?  Is this a RP Tech function or is a separate qualification
      required?

   daily checks when in use, RP tech function

   5.   For daily performance checks, what source confidence levels do you
      use, e.g., within 2 sigma standard deviation?

    source counts within +/- 3 sigma, in process of implementing option to
      use +/ 3% of mean (from MARLAP manual).  background values use fixed
      (upper only) levels for pass / fail

Benjy Bertossi
Fermi2 - RP Supervisor