Greetings All,

 

The TLD should work no problem unless you have specifically told your vendor to 
use a particular algorithm path to evaluate your doses (unlikely).  Your other 
hand held type instruments will require some correction factors if you are 
using them.  Particularly anything using a NaI detector (uR meter for example).


Yours,

Rob

 

Robert J. Gunter, MSc, CHP

CHP Consultants/CHP Dosimetry

www.chpconsultants.com

www.chpdosimetry.com

Toll Free: (888) 766-4833

Fax:  (866) 491-9913 

Cel:  (865) 387-0028

[email protected]

 

From: [email protected] [mailto:[email protected]] On Behalf Of 
June Scott
Sent: Thursday, July 18, 2013 2:19 AM
To: [email protected]
Subject: Re: Powernet: X-Ray Radiography Experience?

 

I would discuss with your TLD vendor to determine if it is best to use a TLD 
just for this job that can be read for the identified energy spectrum as 
opposed to your normal plant calibration spectrum. Also applying specific fade 
factors for the job may be appropriate. 

Sent from my iPhone


On Jul 17, 2013, at 9:02 AM, "Holmes, Stephen J" <[email protected]> 
wrote:

We have a proposal for a vendor to come on site and perform radiography using 
x-ray equipment. This equipment will emit x-rays of considerably lower energy 
(150 kVp max) than our typical gamma source term, and we are concerned about 
whether our instrumentation/dosimetry (calibrated to Cs-137) will adequately 
measure the dose from this device. We have previously performed pulsed x-ray 
radiography at Ginna, and had proven that our instrumentation performed 
satisfactorily, but this was at a significantly higher x-ray energy (300-400 
kVp max). We plan on using the Fluke-Biomedical 451B ion chamber, Landauer BA 
series OSLs, and SRDs (pocket ion chambers) paired with the DMC2000s EDs to 
monitor the work.

 

Do you have experience with x-ray radiography at your facility?

 

If so, did you do any testing, or anything else (such as a white paper), to 
justify your ability to properly monitor the exposure from the x-ray source?

 

If so, would you be willing to share the results or can you provide contact 
information?

 

Thanks,

 

-Steve

 

Stephen J Holmes, CHP, PE

Sr. Plant Health Physicist, CENG-RE Ginna NPP

[email protected] 

585-771-3577

 

This e-mail and any attachments are confidential, may contain legal, 
professional or other privileged information, and are intended solely for the 
addressee. If you are not the intended recipient, do not use the information 
in this e-mail in any way, delete this e-mail and notify the sender. -EXCIP

Reply via email to