As I thought about it last night, I think I would be a hard drive changer as 
well.

The SAM-11 has less than 10 paramaters to change that could cause confusion 
whereas the SAM-12 has ~137.  There is greater complexity.  We  know from 
SAM-11 operation, that it is highly unlikely that the detectors, pm tube, and 
counting circuitry are really the issue.

If I had to come to your site, I'd probably have ~5hrs on the ground before I 
had to start driving back to the airport.  I can verify bkg and source count 
rates in the first two hrs.  If I don't have a solution there, then the problem 
is most likely on the hard drive.  It could be a customer error in one of the 
many settings, bad software settings affecting HV or discriminators, a random 
file corruption, or a systematic software issue.  With 2^137 binary tests to 
perform, I'll never get done in hrs 3-5.  While I don't know exactly what 
happened, I am >97% sure replacing the drive will fix the problem as everything 
is software driven.

If I replace the hard drive, I can take it back and trouble shoot it, have 
others remote in and check it out.  I can run file comparisons against known 
good files to look for differences.  If I get another customer drive with 
similar symptoms, I'm more likely to figure out the true cause than two diff 
reps in the field at two diff locations.  I can do this high quality 
troubleshooting on my time, not your time.

With the simple variables of the SAM-11, there was nothing you could not solve 
in the field in a short period of time.  With the new SAM-12 and all of the 
features and complexity, there is a good chance that the solution may not be 
found in the first couple hrs of the 5hr visit and put you into replacement and 
cal in hrs 3-5.  I'd probably swap in a clean set up file first, but the hard 
drive would be next.

It is likely the SAM-11 will still be the apex of reliability for the near 
future.  You can tell they added a lot of features to make customers all around 
the world happy, but that increased the complexity of the system increased the 
overall likelihood of having a problem.  I think drive changes are a key part 
of troubleshooting for near-term learning and long-term ability to resolve 
issues quickly.

If I had  to come to your site during a refueling outage, you'd need the issue 
fixed quickly, not extended troubleshooting...

Glen


From: [email protected] [mailto:[email protected]] On Behalf Of 
Vickers, Glen:(GenCo-Nuc)
Sent: Thursday, December 26, 2013 8:19 PM
To: [email protected]
Cc: [email protected]
Subject: RE: Powernet: RE: Quick benchmarking question: Tool Monitors

Thermo sent out a notification several months back that errors appearing to be 
hard drive failures were due to a software issue where the Bkg file was not 
purging properly and filling up the hard drive and choking the monitor.  The 
message appears to be hard drive related because you can't write to the hard 
drive any more.  We also had multiple hard drive changes.  There have been 
other issues where they weren't sure of the cause and replaced the hard drive 
anyway.  In this example, the Thermo tech got a message they didn't understand 
and replaced the hard drive as a conservative measure.  The counting circuitry 
is fine, so just swap out Frankenstein's brain with another one until you get 
what you like.

We recently had a PM-12 where a detector replacement caused some form of reset 
and the message wasn't understood, so the hard drive was replaced.  They seem 
to be pretty good guys, but you can tell they're still catching up to the new 
architecture.

I really like the user interface with the PM-12 and SAM-12.  Historically you 
had to diagnose monitor issues from behavioral symptoms and now these monitors 
will give you data files to support the diagnosis.  Currently working with them 
to incorporate any of their learnings into our procedures as necessary for 
troubleshooting or enhanced settings.  Work in progress...

Glen Vickers
Exelon Corp RP Technical Lead, CHP
815-216-2723 (work/cell)


From: [email protected]<mailto:[email protected]> 
[mailto:[email protected]] On Behalf Of 
[email protected]<mailto:[email protected]>
Sent: Monday, December 23, 2013 12:13 PM
To: [email protected]<mailto:[email protected]>
Cc: [email protected]<mailto:[email protected]>
Subject: Re: Powernet: RE: Quick benchmarking question: Tool Monitors

DC Cook employs two, SAM-12 monitors at each of our two, Restricted Area Exits. 
We also use SAM-11s for equipment releases in outlying areas such as our Rad 
Waste Handling Building (outside the Protected Area).
We have had multiple hard drive failures on our SAM-12s.  A recent upgrade to 
solid state hard drives have proven only slightly more reliable.
We have also had about one door handle and/or door latching mechanism failure 
per year, over the years as well.
We do constantly coach workers to close SAM doors with care.

Dave Raye
RP Operations General Supervisor
Cook Nuclear Plant
Bridgman, MI 49106
269-466-2803

Be careful how you practice, as you shall surely perfect it.



From:        "Barber, Jerry" 
<[email protected]<mailto:[email protected]>>
To:        "[email protected]<mailto:[email protected]>" 
<[email protected]<mailto:[email protected]>>,
Date:        12/11/2013 08:21 AM
Subject:        Powernet: RE: Quick benchmarking question:  Tool Monitors
Sent by:        [email protected]<mailto:[email protected]>
________________________________



This is an EXTERNAL email. STOP. THINK before you CLICK links or OPEN 
attachments.
________________________________

Robinson uses Canberra Cronos monitors.  They work great and we have had no 
issues with them.

From: [email protected]<mailto:[email protected]> 
[mailto:[email protected]] On Behalf Of Johns, Alex M.
Sent: Wednesday, December 11, 2013 6:58 AM
To: [email protected]<mailto:[email protected]>
Subject: Powernet: RE: Quick benchmarking question: Tool Monitors

*** This is an EXTERNAL email. Exercise caution. DO NOT open attachments or 
click links from unknown senders or unexpected email. ***
Salem still using SAM-9(s) and SAM-11(s).  Replaced hinges on the 9(s).  No 
other significant issues.

Alex Johns
RPS Salem Station
856-339-3909
[email protected]<mailto:[email protected]>

From: [email protected]<mailto:[email protected]> 
[mailto:[email protected]] On Behalf Of Holmes, Stephen J:(GINNA)
Sent: Wednesday, December 11, 2013 6:48 AM
To: [email protected]<mailto:[email protected]>
Subject: Powernet: RE: Quick benchmarking question: Tool Monitors

Ginna uses SAM-12; we've had them for a couple years now and haven't had any 
significant problems with them

Stephen J Holmes, CHP, PE
Sr. Plant Health Physicist, RE Ginna NPP
[email protected]<mailto:[email protected]>
585-771-3577

From: [email protected]<mailto:[email protected]> 
[mailto:[email protected]] On Behalf Of Kinmark, Michael L.
Sent: Tuesday, December 10, 2013 2:01 PM
To: [email protected]<mailto:[email protected]>
Subject: Powernet: Quick benchmarking question: Tool Monitors

Question from Columbia:

What tool monitor do you currently use?  SAM-12, Canberra Cronos, or something 
other?   Any issues?

Thank you in advance.


Michael Kinmark
HP Staff Advisor III
Radiological Support
Columbia Generating Station
509-377-2091
[email protected]<mailto:[email protected]>



This e-mail and any attachments are confidential, may contain legal,
professional or other privileged information, and are intended solely for the
addressee. If you are not the intended recipient, do not use the information
in this e-mail in any way, delete this e-mail and notify the sender. -EXCIP

________________________________

The information contained in this e-mail, including any attachment(s), is 
intended solely for use by the named addressee(s). If you are not the intended 
recipient, or a person designated as responsible for delivering such messages 
to the intended recipient, you are not authorized to disclose, copy, distribute 
or retain this message, in whole or in part, without written authorization from 
PSEG. This e-mail may contain proprietary, confidential or privileged 
information. If you have received this message in error, please notify the 
sender immediately. This notice is included in all e-mail messages leaving 
PSEG. Thank you for your cooperation.

Reply via email to