[ccp4bb] Help with reducing crystal mosaicity

2007-07-09 Thread Mary Fitzgerald

Help please!

I'm looking for some new ideas.  I have crystals that come out of a
sitting drop with a mixture of sodium cacodylate at pH 6.5, magnesium
acetate and MPD for the well solution.  The MPD concentration is
sufficient to act as a cryoprotectant.  Currently, I directly freeze
these crystals in liquid nitrogen.  When I collect data, I typically
have high anisotropic mosaicity; it ranges from 0.8 to 1.2.  This is
further complicated with a weakly diffracting crystal (4-5 A) that has
a long unit cell axis of ~500 and often twinning.

It has been suggested to me that the cryoprotectent is a problem.  I
haven't checked the diffraction at room temperature, yet.  Please no
suggestions of finding a different crystal form as that's not a
consideration at the moment.  I have my reasons.  I did find one
crystal that has lower mosaicity (0.5 to 0.8) but had weaker
diffraction then the typical crystal.  Attempts at flash cryoannealing
have not helped.

So, what's a good way to change the cryoprotectant if the
cryoprotectant is the precipitant?  I've considered trying dehydration
but wasn't certain if that would help with the mosaicity.

Thanks for any ideas,

Mary X. Fitzgerald
Postdoctoral Associate


Re: [ccp4bb] Help with reducing crystal mosaicity

2007-07-09 Thread Jim Pflugrath
What is the mosaicity of the unfrozen crystal?

Jim

-Original Message-
From: Mary Fitzgerald <[EMAIL PROTECTED]>
Date: Mon, 9 Jul 2007 18:05:10 
To:CCP4BB@JISCMAIL.AC.UK
Subject: [ccp4bb] Help with reducing crystal mosaicity

Help please!

I'm looking for some new ideas.  I have crystals that come out of a
sitting drop with a mixture of sodium cacodylate at pH 6.5, magnesium
acetate and MPD for the well solution.  The MPD concentration is
sufficient to act as a cryoprotectant.  Currently, I directly freeze
these crystals in liquid nitrogen.  When I collect data, I typically
have high anisotropic mosaicity; it ranges from 0.8 to 1.2.  This is
further complicated with a weakly diffracting crystal (4-5 A) that has
a long unit cell axis of ~500 and often twinning.

It has been suggested to me that the cryoprotectent is a problem.  I
haven't checked the diffraction at room temperature, yet.  Please no
suggestions of finding a different crystal form as that's not a
consideration at the moment.  I have my reasons.  I did find one
crystal that has lower mosaicity (0.5 to 0.8) but had weaker
diffraction then the typical crystal.  Attempts at flash cryoannealing
have not helped.

So, what's a good way to change the cryoprotectant if the
cryoprotectant is the precipitant?  I've considered trying dehydration
but wasn't certain if that would help with the mosaicity.

Thanks for any ideas,

Mary X. Fitzgerald
Postdoctoral Associate


Re: [ccp4bb] Help with reducing crystal mosaicity

2007-07-09 Thread David J. Schuller
Here's a method that has had some success in reducing mosaicity, either
with or without cryoprotectant:

Chae Un Kim, Raphael Kapfer and Sol M. Gruner (2005), High Pressure
Cooling of Protein Crystals without Cryoprotectants, ActaCryst. D61,
881-890

This method may be available at a synchrotron near you.

Cheers,

-- 
===
With the single exception of Cornell, there is not a college in the
United States where truth has ever been a welcome guest - R.G. Ingersoll
===
  David J. Schuller
  modern man in a post-modern world
  MacCHESS, Cornell University
  [EMAIL PROTECTED]


On Mon, 2007-07-09 at 18:05 -0400, Mary Fitzgerald wrote:
> Help please!
> 
> I'm looking for some new ideas.  I have crystals that come out of a
> sitting drop with a mixture of sodium cacodylate at pH 6.5, magnesium
> acetate and MPD for the well solution.  The MPD concentration is
> sufficient to act as a cryoprotectant.  Currently, I directly freeze
> these crystals in liquid nitrogen.  When I collect data, I typically
> have high anisotropic mosaicity; it ranges from 0.8 to 1.2.  This is
> further complicated with a weakly diffracting crystal (4-5 A) that has
> a long unit cell axis of ~500 and often twinning.
> 
> It has been suggested to me that the cryoprotectent is a problem.  I
> haven't checked the diffraction at room temperature, yet.  Please no
> suggestions of finding a different crystal form as that's not a
> consideration at the moment.  I have my reasons.  I did find one
> crystal that has lower mosaicity (0.5 to 0.8) but had weaker
> diffraction then the typical crystal.  Attempts at flash cryoannealing
> have not helped.
> 
> So, what's a good way to change the cryoprotectant if the
> cryoprotectant is the precipitant?  I've considered trying dehydration
> but wasn't certain if that would help with the mosaicity.
> 
> Thanks for any ideas,
> 
> Mary X. Fitzgerald
> Postdoctoral Associate


Re: [ccp4bb] Help with reducing crystal mosaicity

2007-07-09 Thread Raji Edayathumangalam
Hi Mary,

What final percent MPD do you have prior to flashcooling? Karolin Luger (and 
perhaps others) found 
that the final percent MPD had a significant effect on crystal mosaicity and 
final diffraction 
limit for nucleosome crystals. 

For example, if the crystals are left in mother liquor containing 15% MPD, that 
is suboptimal. On 
the other hand, if the MPD concentration was increased beyond 24%, the the 
crystal mosacity 
increased and the diffraction worsened. We stuck to 24% MPD. Did you try 
playing with the MPD 
concentrations? Also, try supplementing MPD with a small percent of sugars like 
sucrose, trehalose.

Another note: Nucleosome crystals are prone to temperature-dependent lattice 
changes (always along 
c-axis) with a concomitant worsening of diffraction. Even if it sounds 
anectodal, mythological or 
voodoo-ish, we found that it greatly helped diffraction anisotropy and 
mosaicity if we dunked the 
crystals in liquid propane first. After the initial dunking, it was just fine 
to freeze the 
crystals in liq. N2 for longer storage and/or transport. It was the first dunk 
that was critical.

Good luck!
Raji



-Included Message--
>Help please!
>
>I'm looking for some new ideas.  I have crystals that come out of a
>sitting drop with a mixture of sodium cacodylate at pH 6.5, magnesium
>acetate and MPD for the well solution.  The MPD concentration is
>sufficient to act as a cryoprotectant.  Currently, I directly freeze
>these crystals in liquid nitrogen.  When I collect data, I typically
>have high anisotropic mosaicity; it ranges from 0.8 to 1.2.  This is
>further complicated with a weakly diffracting crystal (4-5 A) that has
>a long unit cell axis of ~500 and often twinning.
>
>It has been suggested to me that the cryoprotectent is a problem.  I
>haven't checked the diffraction at room temperature, yet.  Please no
>suggestions of finding a different crystal form as that's not a
>consideration at the moment.  I have my reasons.  I did find one
>crystal that has lower mosaicity (0.5 to 0.8) but had weaker
>diffraction then the typical crystal.  Attempts at flash cryoannealing
>have not helped.
>
>So, what's a good way to change the cryoprotectant if the
>cryoprotectant is the precipitant?  I've considered trying dehydration
>but wasn't certain if that would help with the mosaicity.
>
>Thanks for any ideas,
>
>Mary X. Fitzgerald
>Postdoctoral Associate
>
>
-End of Included Message--


Re: [ccp4bb] Help with reducing crystal mosaicity

2007-07-09 Thread Mary Fitzgerald

Wow, thanks.   I'm going to try to answer most of the questions I've
received in one message as I'm overwhelmed by the quick multitude of
responses.

As I stated earlier, I haven't collected any room temperature data,
yet.  So, I don't know the unfrozen mosaicity.  It is very possible
that the crystals are simply mosaic at RT, too.  I have tried
flash-cooling in liquid nitrogen and "gas cooling" in the cryo-stream,
I didn't notice a difference in the diffraction pattern.

My solvent content is on the high side about 67%.  Additional crystal
details, I set the trays up at 20 degrees.  The crystals take about 2
weeks to appear and another week to grow to full size.  The crystals
are bipyrimidal and often 150x150x300 although sometimes smaller.
Typically, there's about 34-39% MPD in the well solutions.  The well
solution freezes clear and I don't get any ice rings in my diffraction
patterns.  I've tried looping the crystals directly out of the drop or
adding a few microletters of well solution and then looping, I didn't
notice a difference.  I've been hesitant to start adding extras
fearing crystal cracking.   I did however switch to freezing directly
in the automounter puck which was my one low mosacity crystal but I
had plenty of other crystals in the puck that weren't better then
before.

I was wondering if the MPD concentration could cause problems and
sounds like it might be.  So, with the crystals I have I'll try higher
MPD to dehydrate and try freezing in lower MPD concentration or add
some other cryoprotectants to my current well solution.  Although, I
might have to go find some propane.

If that doesn't work maybe, I'll try seeding at lower MPD
concentrations or pressure freezing.

Thanks again,
Mary


Re: [ccp4bb] Help with reducing crystal mosaicity

2007-07-09 Thread Tim Grune
On Tuesday 10 July 2007 09:16, Mary Fitzgerald wrote:
> If that doesn't work maybe, I'll try seeding at lower MPD
> concentrations or pressure freezing.
>
> Thanks again,
> Mary

With such a high MPD condition I would probably first try micro-seeding at 
25-35% MPD before doing anything else. It sounds like a very good candidate.
-- 
Tim Grune
Australian Synchrotron
800 Blackburn Road
Clayton, VIC 3168
Australia


pgpEJuDKsMqme.pgp
Description: PGP signature


Re: [ccp4bb] Help with reducing crystal mosaicity

2007-07-09 Thread Juergen Bosch

Hi Mary,

not sure if you received this suggestion already.
Despite the low resolution diffraction 4-5 A as you report, try mounting 
smaller crystals and see how well they perform. Very often our crystals 
had a better freezing experience when they were small, even if 
sufficient cryo protectant was present.


Another thing to try is once you have mounted a crystal, how does it 
behave after flash annealing / wet annealing / multiple rounds of 
annealing ?


Juergen

--
Jürgen Bosch
University of Washington
Dept. of Biochemistry, K-426
1705 NE Pacific Street
Seattle, WA 98195
Box 357742
Phone:   +1-206-616-4510
FAX: +1-206-685-7002


Re: [ccp4bb] Help with reducing crystal mosaicity

2007-07-10 Thread mesters

Mary,

freezing habitually increases mosaicity. In your case, the high water 
content adds to the problem.
Try not to freeze the crystal but collect at sub-zero temperature (in 
short glass capillaries or use oil plugs instead).

You have to optimize the "close to freezing" data-collection temperature.

I collected complete synchrotron datasets (of GCPII in buffer with 
PEG1500 and PEG400) at 260-263 Kelvin which resulted in mosaicity values 
of as small as 0.07 degrees! At 277 K, the crystals only last for a few 
images and freezing did not work (for the buffer mentioned before).


- J. -


-Original Message-
From: Mary Fitzgerald <[EMAIL PROTECTED]>
Date: Mon, 9 Jul 2007 18:05:10 
To:CCP4BB@JISCMAIL.AC.UK

Subject: [ccp4bb] Help with reducing crystal mosaicity

Help please!

I'm looking for some new ideas.  I have crystals that come out of a
sitting drop with a mixture of sodium cacodylate at pH 6.5, magnesium
acetate and MPD for the well solution.  The MPD concentration is
sufficient to act as a cryoprotectant.  Currently, I directly freeze
these crystals in liquid nitrogen.  When I collect data, I typically
have high anisotropic mosaicity; it ranges from 0.8 to 1.2.  This is
further complicated with a weakly diffracting crystal (4-5 A) that has
a long unit cell axis of ~500 and often twinning.

It has been suggested to me that the cryoprotectent is a problem.  I
haven't checked the diffraction at room temperature, yet.  Please no
suggestions of finding a different crystal form as that's not a
consideration at the moment.  I have my reasons.  I did find one
crystal that has lower mosaicity (0.5 to 0.8) but had weaker
diffraction then the typical crystal.  Attempts at flash cryoannealing
have not helped.

So, what's a good way to change the cryoprotectant if the
cryoprotectant is the precipitant?  I've considered trying dehydration
but wasn't certain if that would help with the mosaicity.

Thanks for any ideas,

Mary X. Fitzgerald
Postdoctoral Associate
  



--
Jeroen Raymundus Mesters, Ph.D.
Institut fuer Biochemie, Universitaet zu Luebeck
Zentrum fuer Medizinische Struktur und Zellbiologie
Ratzeburger Allee 160, D-23538 Luebeck
Tel: +49-451-5004070, Fax: +49-451-5004068
E-mail: [EMAIL PROTECTED]
Http://www.biochem.uni-luebeck.de
Http://www.iobcr.org
Http://www.opticryst.org
--
If you can look into the seeds of time and say
which grain will grow and which will not - speak then to me  (Macbeth)
--


Re: [ccp4bb] Help with reducing crystal mosaicity

2007-07-10 Thread Flip Hoedemaeker
 You might also want to loook into using parathone for freezing... Or
collection at 260K indeed!

Flip

-Original Message-
From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of
mesters
Sent: Tuesday, July 10, 2007 11:01
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] Help with reducing crystal mosaicity

Mary,

freezing habitually increases mosaicity. In your case, the high water 
content adds to the problem.
Try not to freeze the crystal but collect at sub-zero temperature (in 
short glass capillaries or use oil plugs instead).
You have to optimize the "close to freezing" data-collection temperature.

I collected complete synchrotron datasets (of GCPII in buffer with 
PEG1500 and PEG400) at 260-263 Kelvin which resulted in mosaicity values 
of as small as 0.07 degrees! At 277 K, the crystals only last for a few 
images and freezing did not work (for the buffer mentioned before).

- J. -

> -Original Message-
> From: Mary Fitzgerald <[EMAIL PROTECTED]>
> Date: Mon, 9 Jul 2007 18:05:10 
> To:CCP4BB@JISCMAIL.AC.UK
> Subject: [ccp4bb] Help with reducing crystal mosaicity
>
> Help please!
>
> I'm looking for some new ideas.  I have crystals that come out of a
> sitting drop with a mixture of sodium cacodylate at pH 6.5, magnesium
> acetate and MPD for the well solution.  The MPD concentration is
> sufficient to act as a cryoprotectant.  Currently, I directly freeze
> these crystals in liquid nitrogen.  When I collect data, I typically
> have high anisotropic mosaicity; it ranges from 0.8 to 1.2.  This is
> further complicated with a weakly diffracting crystal (4-5 A) that has
> a long unit cell axis of ~500 and often twinning.
>
> It has been suggested to me that the cryoprotectent is a problem.  I
> haven't checked the diffraction at room temperature, yet.  Please no
> suggestions of finding a different crystal form as that's not a
> consideration at the moment.  I have my reasons.  I did find one
> crystal that has lower mosaicity (0.5 to 0.8) but had weaker
> diffraction then the typical crystal.  Attempts at flash cryoannealing
> have not helped.
>
> So, what's a good way to change the cryoprotectant if the
> cryoprotectant is the precipitant?  I've considered trying dehydration
> but wasn't certain if that would help with the mosaicity.
>
> Thanks for any ideas,
>
> Mary X. Fitzgerald
> Postdoctoral Associate
>   


-- 
Jeroen Raymundus Mesters, Ph.D.
Institut fuer Biochemie, Universitaet zu Luebeck
Zentrum fuer Medizinische Struktur und Zellbiologie
Ratzeburger Allee 160, D-23538 Luebeck
Tel: +49-451-5004070, Fax: +49-451-5004068
E-mail: [EMAIL PROTECTED]
Http://www.biochem.uni-luebeck.de
Http://www.iobcr.org
Http://www.opticryst.org
--
If you can look into the seeds of time and say
which grain will grow and which will not - speak then to me  (Macbeth)
--


Re: [ccp4bb] Help with reducing crystal mosaicity

2007-07-10 Thread Edward Snell
Hi Mary,

There are many excellent suggestions already but you may want to try
tweaking your cryo a bit. This is described very nicely by Ed Mitchell
and Elspeth Garman in a paper on mosaic spread as a function of
cryoprotectants concentration - J. Appl. Cryst. (1994) 27, 1070-1074.
You already have a cryoprotectant, the MPD, and could up the
concentration a little by successive transfer or adding a little to the
drop. I've also found that (2R,3R)-(-)-2,3-butanediol works as a 'magic'
elixir in very small concentrations when added directly to the drop. 

I'd say try room temperature but at that resolution you are going to
need all the X-rays you can get onto it.

Cheers,

Eddie.

Edward Snell Ph.D.
Assistant Prof. Department of Structural Biology, SUNY Buffalo,
Hauptman-Woodward Medical Research Institute
700 Ellicott Street, Buffalo, NY 14203-1102
Phone: (716) 898 8631 Fax: (716) 898 8660
Email: [EMAIL PROTECTED]  Telepathy: 42.2 GHz
 
Heisenberg was probably here!
 


Mary,

freezing habitually increases mosaicity. In your case, the high water 
content adds to the problem.
Try not to freeze the crystal but collect at sub-zero temperature (in 
short glass capillaries or use oil plugs instead).
You have to optimize the "close to freezing" data-collection
temperature.

I collected complete synchrotron datasets (of GCPII in buffer with 
PEG1500 and PEG400) at 260-263 Kelvin which resulted in mosaicity values

of as small as 0.07 degrees! At 277 K, the crystals only last for a few 
images and freezing did not work (for the buffer mentioned before).

- J. -

> -Original Message-
> From: Mary Fitzgerald <[EMAIL PROTECTED]>
> Date: Mon, 9 Jul 2007 18:05:10 
> To:CCP4BB@JISCMAIL.AC.UK
> Subject: [ccp4bb] Help with reducing crystal mosaicity
>
> Help please!
>
> I'm looking for some new ideas.  I have crystals that come out of a
> sitting drop with a mixture of sodium cacodylate at pH 6.5, magnesium
> acetate and MPD for the well solution.  The MPD concentration is
> sufficient to act as a cryoprotectant.  Currently, I directly freeze
> these crystals in liquid nitrogen.  When I collect data, I typically
> have high anisotropic mosaicity; it ranges from 0.8 to 1.2.  This is
> further complicated with a weakly diffracting crystal (4-5 A) that has
> a long unit cell axis of ~500 and often twinning.
>
> It has been suggested to me that the cryoprotectent is a problem.  I
> haven't checked the diffraction at room temperature, yet.  Please no
> suggestions of finding a different crystal form as that's not a
> consideration at the moment.  I have my reasons.  I did find one
> crystal that has lower mosaicity (0.5 to 0.8) but had weaker
> diffraction then the typical crystal.  Attempts at flash cryoannealing
> have not helped.
>
> So, what's a good way to change the cryoprotectant if the
> cryoprotectant is the precipitant?  I've considered trying dehydration
> but wasn't certain if that would help with the mosaicity.
>
> Thanks for any ideas,
>
> Mary X. Fitzgerald
> Postdoctoral Associate
>   


-- 
Jeroen Raymundus Mesters, Ph.D.
Institut fuer Biochemie, Universitaet zu Luebeck
Zentrum fuer Medizinische Struktur und Zellbiologie
Ratzeburger Allee 160, D-23538 Luebeck
Tel: +49-451-5004070, Fax: +49-451-5004068
E-mail: [EMAIL PROTECTED]
Http://www.biochem.uni-luebeck.de
Http://www.iobcr.org
Http://www.opticryst.org
--
If you can look into the seeds of time and say
which grain will grow and which will not - speak then to me  (Macbeth)
--


Re: [ccp4bb] Help with reducing crystal mosaicity

2007-07-10 Thread Patrick Loll
Jeroen brings up a good point.  Back in the old days, around 5 B. C.  
(Before Cryo), we would use a chilled air generator to blow a stream  
of cold air along the capillary axis to keep the crystals just above  
their freezing point--it made a huge difference in crystal lifetime.  
I recall a colleague devising an apparatus from a 50 ml conical tube.  
The bottom was cut off and cold air was blown in from the other end.  
Windows were cut in either side to allow the beam to pass & covered  
in mylar.  This way the entire capillary was contained within the  
cold tube, so no temperature gradients formed along the length of the  
capillary (temp gradient => distillation => dead crystal).  Later, we  
purchased a very clever goniometer head from Nonius that had a  
plastic cylinder attached to goniometer head, with a swivel, so the  
hose supplying cold air didn't get tangled during data collection...


I've often thought duplicating this apparatus when we encounter cryo  
problems, but I'm always stymied when trying to find a cheap and  
simple source of cold air.  Any bright ideas?



On Jul 10, 2007, at 5:00 AM, mesters wrote:


Mary,

freezing habitually increases mosaicity. In your case, the high  
water content adds to the problem.
Try not to freeze the crystal but collect at sub-zero temperature  
(in short glass capillaries or use oil plugs instead).
You have to optimize the "close to freezing" data-collection  
temperature.


I collected complete synchrotron datasets (of GCPII in buffer with  
PEG1500 and PEG400) at 260-263 Kelvin which resulted in mosaicity  
values of as small as 0.07 degrees! At 277 K, the crystals only  
last for a few images and freezing did not work (for the buffer  
mentioned before).


- J. -


 
---

Patrick J. Loll, Ph. D. (215) 762-7706
Associate Professor FAX: (215) 762-4452
Department of Biochemistry & Molecular Biology
Director, Biochemistry Graduate Program
Drexel University College of Medicine
Room 10-102 New College Building
245 N. 15th St., Mailstop 497
Philadelphia, PA  19102-1192  USA

[EMAIL PROTECTED]



Re: [ccp4bb] Help with reducing crystal mosaicity

2007-07-10 Thread Matthew . Franklin
CCP4 bulletin board  wrote on 07/10/2007 02:07:37
PM:

> Jeroen brings up a good point.  Back in the old days, around 5 B. C.
> (Before Cryo), we would use a chilled air generator to blow a stream
> of cold air along the capillary axis to keep the crystals just above
> their freezing point--it made a huge difference in crystal lifetime.
> I recall a colleague devising an apparatus from a 50 ml conical
> tube. The bottom was cut off and cold air was blown in from the
> other end. Windows were cut in either side to allow the beam to pass
> & covered in mylar.  This way the entire capillary was contained
> within the cold tube, so no temperature gradients formed along the
> length of the capillary (temp gradient => distillation => dead
> crystal).  Later, we purchased a very clever goniometer head from
> Nonius that had a plastic cylinder attached to goniometer head, with
> a swivel, so the hose supplying cold air didn't get tangled during
> data collection...
>
> I've often thought duplicating this apparatus when we encounter cryo
> problems, but I'm always stymied when trying to find a cheap and
> simple source of cold air.  Any bright ideas?
>

Hi Patrick -

Many cryosystems (definitely the Cryojet, and I believe the Cryostream) can
be set to run at any temperature between room temp and liquid nitrogen
temp.  I'm not sure of the temperature stability at temps > 0 C, and you
might burn out the heaters prematurely if you do this all the time, but it
should work.  Then you just need to move the nozzle so it's coaxial with
your goniostat's rotation axis, and aim the capillary down the nozzle.  You
could probably even move the capillary *inside* the cryo nozzle a bit, so
only the bit with the crystal is in the free stream.

It's not a cheap solution, but you've almost certainly got a cryosystem
already...


- Matt

--
Matthew Franklin , Ph.D.
Senior Scientist, ImClone Systems
180 Varick Street, 6th floor
New York, NY 10014
phone:(917)606-4116   fax:(212)645-2054


Confidentiality Note:  This e-mail, and any attachment to it, contains
privileged and confidential information intended only for the use of the
individual(s) or entity named on the e-mail.  If the reader of this e-mail
is not the intended recipient, or the employee or agent responsible for
delivering it to the intended recipient, you are hereby notified that
reading it is strictly prohibited.  If you have received this e-mail in
error, please immediately return it to the sender and delete it from your
system.  Thank you.

Re: [ccp4bb] Help with reducing crystal mosaicity

2007-07-11 Thread mesters

Matthew,

I have not had any problem with our Cryojet XL (90 - 300 Kelvin), it is 
very stable at sub-zero temperatures, I checked it over a longer period 
using a highly sensitive electronic device. Also, you can buy the 
Cryojet HT that can be operated from 90 – 490 Kelvin!

The heaters will be okay for both devices.

But more important, coaxial mounting does not help unless the capillary 
is really short! If the capillary is to long, water will again collect 
at the point closest to the nozzle, especially if the data collection 
takes more than one or two hours!


I am not aware of any cheaper devices that deliver a stable temperature 
over a prolonged time-span.


- J. -



[EMAIL PROTECTED] wrote:

Hi Patrick -

Many cryosystems (definitely the Cryojet, and I believe the Cryostream) can
be set to run at any temperature between room temp and liquid nitrogen
temp.  I'm not sure of the temperature stability at temps > 0 C, and you
might burn out the heaters prematurely if you do this all the time, but it
should work.  Then you just need to move the nozzle so it's coaxial with
your goniostat's rotation axis, and aim the capillary down the nozzle.  You
could probably even move the capillary *inside* the cryo nozzle a bit, so
only the bit with the crystal is in the free stream.

It's not a cheap solution, but you've almost certainly got a cryosystem
already...


- Matt

--
Matthew Franklin , Ph.D.
Senior Scientist, ImClone Systems
180 Varick Street, 6th floor
New York, NY 10014
phone:(917)606-4116   fax:(212)645-2054


Confidentiality Note:  This e-mail, and any attachment to it, contains
privileged and confidential information intended only for the use of the
individual(s) or entity named on the e-mail.  If the reader of this e-mail
is not the intended recipient, or the employee or agent responsible for
delivering it to the intended recipient, you are hereby notified that
reading it is strictly prohibited.  If you have received this e-mail in
error, please immediately return it to the sender and delete it from your
system.  Thank you.



CCP4 bulletin board  wrote on 07/10/2007 02:07:37
PM:

  

Jeroen brings up a good point.  Back in the old days, around 5 B. C.
(Before Cryo), we would use a chilled air generator to blow a stream
of cold air along the capillary axis to keep the crystals just above
their freezing point--it made a huge difference in crystal lifetime.
I recall a colleague devising an apparatus from a 50 ml conical
tube. The bottom was cut off and cold air was blown in from the
other end. Windows were cut in either side to allow the beam to pass
& covered in mylar.  This way the entire capillary was contained
within the cold tube, so no temperature gradients formed along the
length of the capillary (temp gradient => distillation => dead
crystal).  Later, we purchased a very clever goniometer head from
Nonius that had a plastic cylinder attached to goniometer head, with
a swivel, so the hose supplying cold air didn't get tangled during
data collection...

I've often thought duplicating this apparatus when we encounter cryo
problems, but I'm always stymied when trying to find a cheap and
simple source of cold air.  Any bright ideas?




  



--
Jeroen Raymundus Mesters, Ph.D.
Institut fuer Biochemie, Universitaet zu Luebeck
Zentrum fuer Medizinische Struktur und Zellbiologie
Ratzeburger Allee 160, D-23538 Luebeck
Tel: +49-451-5004070, Fax: +49-451-5004068
E-mail: [EMAIL PROTECTED]
Http://www.biochem.uni-luebeck.de
Http://www.iobcr.org
Http://www.opticryst.org
--
If you can look into the seeds of time and say
which grain will grow and which will not - speak then to me  (Macbeth)
--


Re: [ccp4bb] Help with reducing crystal mosaicity

2007-07-11 Thread Marius Schmidt
We used to use FTS-cooler for stable 
temperatures in Laue-experiments on
protein crystals in capillaries. 
Check out BioCARS at the APS for this device.
http://cars9.uchicago.edu/biocars/pages/timeresolved.html
It has a big opening for constant air flow
and temperature along the capillary and
it can go to sub 0 deg C.

Marius

> Matthew,
> 
> I have not had any problem with our Cryojet XL (90 - 300 Kelvin), it
> is 
> very stable at sub-zero temperatures, I checked it over a longer
> period 
> using a highly sensitive electronic device. Also, you can buy the 
> Cryojet HT that can be operated from 90 – 490 Kelvin!
> The heaters will be okay for both devices.
> 
> But more important, coaxial mounting does not help unless the
> capillary 
> is really short! If the capillary is to long, water will again
> collect 
> at the point closest to the nozzle, especially if the data collection
> takes more than one or two hours!
> 
> I am not aware of any cheaper devices that deliver a stable
> temperature 
> over a prolonged time-span.
> 
> - J. -
> 
> 
> [EMAIL PROTECTED] wrote:
> 
> Hi Patrick -
> 
> Many cryosystems (definitely the Cryojet, and I believe the
> Cryostream) can
> be set to run at any temperature between room temp and liquid
> nitrogen
> temp.  I'm not sure of the temperature stability at temps > 0 C, and
> you
> might burn out the heaters prematurely if you do this all the time,
> but it
> should work.  Then you just need to move the nozzle so it's coaxial
> with
> your goniostat's rotation axis, and aim the capillary down the
> nozzle.  You
> could probably even move the capillary *inside* the cryo nozzle a
> bit, so
> only the bit with the crystal is in the free stream.
> 
> It's not a cheap solution, but you've almost certainly got a
> cryosystem
> already...
> 
> - Matt
> 
> --
> Matthew Franklin , Ph.D.
> Senior Scientist, ImClone Systems
> 180 Varick Street, 6th floor
> New York, NY 10014
> phone:(917)606-4116   fax:(212)645-2054
> 
> Confidentiality Note:  This e-mail, and any attachment to it,
> contains
> privileged and confidential information intended only for the use of
> the
> individual(s) or entity named on the e-mail.  If the reader of this
> e-mail
> is not the intended recipient, or the employee or agent responsible
> for
> delivering it to the intended recipient, you are hereby notified that
> reading it is strictly prohibited.  If you have received this e-mail
> in
> error, please immediately return it to the sender and delete it from
> your
> system.  Thank you.
> 
>> CCP4 bulletin board  wrote on 07/10/2007 02:07:37
>> PM:
>>
>>   
>>> Jeroen brings up a good point.  Back in the old days, around 5 B. C.
>>> (Before Cryo), we would use a chilled air generator to blow a stream
>>> of cold air along the capillary axis to keep the crystals just above
>>> their freezing point--it made a huge difference in crystal lifetime.
>>> I recall a colleague devising an apparatus from a 50 ml conical
>>> tube. The bottom was cut off and cold air was blown in from the
>>> other end. Windows were cut in either side to allow the beam to pass
>>> & covered in mylar.  This way the entire capillary was contained
>>> within the cold tube, so no temperature gradients formed along the
>>> length of the capillary (temp gradient => distillation => dead
>>> crystal).  Later, we purchased a very clever goniometer head from
>>> Nonius that had a plastic cylinder attached to goniometer head, with
>>> a swivel, so the hose supplying cold air didn't get tangled during
>>> data collection...
>>>
>>> I've often thought duplicating this apparatus when we encounter cryo
>>> problems, but I'm always stymied when trying to find a cheap and
>>> simple source of cold air.  Any bright ideas?
>>>
>>> 
>>
>>   
> 
> -- 
> Jeroen Raymundus Mesters, Ph.D.
> Institut fuer Biochemie, Universitaet zu Luebeck
> Zentrum fuer Medizinische Struktur und Zellbiologie
> Ratzeburger Allee 160, D-23538 Luebeck
> Tel: +49-451-5004070, Fax: +49-451-5004068
> E-mail: [EMAIL PROTECTED]
> Http://www.biochem.uni-luebeck.de
> Http://www.iobcr.org
> Http://www.opticryst.org
> --
> If you can look into the seeds of time and say
> which grain will grow and which will not - speak then to me  (Macbeth)
> --

PD Dr. habil. Marius Schmidt
Physikdepartment E17
Technische Universitaet Muenchen
James Franck Strasse
85747 Garching/Germany
email: [EMAIL PROTECTED]
http://users.physik.tu-muenchen.de/marius/
phone: +49-(0)89-2891-2550
fax:   +49-(0)89-2891-2548


Re: [ccp4bb] Help with reducing crystal mosaicity

2007-07-11 Thread Moody, Dr P.C.E.
a couple of  points from long BC about approx. 0 C cooling...for what its worth!
 
i) don't forget that high concentrations of salts can have a significant effect 
on the freezing point ( remember freezing point depression?). We used to 
collect data at -15C from crystals grown in 2.4M Ammonium Sulphate
 
ii) cool slowly and gently over a couple of hours
 
iii) put the mother liquor at only the colder end of the capillary to prevent 
condensation
 
iv) we found a transition at about -5 for above crystals, so be careful.
 
 
 
 
Peter Moody
Henry Wellcome Laboratories of Structural Biology
Department of Biochemistry
Henry Wellcome Building
University of Leicester
LE1 9HN
0116 229 7097



From: CCP4 bulletin board on behalf of mesters
Sent: Wed 11/07/2007 08:46
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] Help with reducing crystal mosaicity



snipped


Re: [ccp4bb] Help with reducing crystal mosaicity

2007-07-11 Thread Patrick Shaw Stewart
Just a thought, Mary - going back to your original question about MPD.  I 
extracted the crystallization conditions from REMARK 280 of 3939 PDB entries a 
couple of years ago.  The average concentration of the MPD used was high - 
38.6%, while PEGs tended to be used at lower concs, e.g. PEG400 25.7%.  You can 
see the data at www.douglas.co.uk/top14.htm 

I thought this information could be useful if you want to replace some of the 
MPD with another precipitant (or cryoprotectant).

Best wishes

Patrick Shaw Stewart


--  
[EMAIL PROTECTED] Douglas Instruments Ltd.  
DouglasHouse, EastGarston, Hungerford, Berkshire, RG177HD, UK 
Directors: Peter Baldock, Patrick Shaw Stewart, James Smith 
http://douglas.co.uk or http://www.douglasinstruments.com 
Tel: 44 (0) 148-864-9090    US toll-free 1-877-225-2034 
Regd. England 2177994, VAT Reg. GB 480 7371 36


> -Original Message-
> From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of Mary
> Fitzgerald
> Sent: 09 July 2007 23:05
> To: CCP4BB@JISCMAIL.AC.UK
> Subject: [ccp4bb] Help with reducing crystal mosaicity
> 
> Help please!
> 
> I'm looking for some new ideas.  I have crystals that come out of a
> sitting drop with a mixture of sodium cacodylate at pH 6.5, magnesium
> acetate and MPD for the well solution.  The MPD concentration is
> sufficient to act as a cryoprotectant.  Currently, I directly freeze
> these crystals in liquid nitrogen.  When I collect data, I typically
> have high anisotropic mosaicity; it ranges from 0.8 to 1.2.  This is
> further complicated with a weakly diffracting crystal (4-5 A) that has
> a long unit cell axis of ~500 and often twinning.
> 
> It has been suggested to me that the cryoprotectent is a problem.  I
> haven't checked the diffraction at room temperature, yet.  Please no
> suggestions of finding a different crystal form as that's not a
> consideration at the moment.  I have my reasons.  I did find one
> crystal that has lower mosaicity (0.5 to 0.8) but had weaker
> diffraction then the typical crystal.  Attempts at flash cryoannealing
> have not helped.
> 
> So, what's a good way to change the cryoprotectant if the
> cryoprotectant is the precipitant?  I've considered trying dehydration
> but wasn't certain if that would help with the mosaicity.
> 
> Thanks for any ideas,
> 
> Mary X. Fitzgerald
> Postdoctoral Associate


Re: [ccp4bb] Help with reducing crystal mosaicity

2007-07-11 Thread Benini, Stefano
Dear Mary,

the problem encountered with cryoprotectans is the change in the solution 
surrounding your crystal as they may not be present in your crystallisation 
conditions or you need to increase their concentration to make them act as 
cryoprotectant.
So I don't thik is the cryoprotectant itself as it is you mother liquor..

test it in a capillary so you could rule out that freezing is the problem

of course you don't want to find new crystallisation conditions when you tried 
everything and I know what it means!!

I would try to improve crystal quality using additives or ligand etc,

your long axis is very worrying as well!!! may be some magic additive could 
change the packing and givew better diffraction too

I also had crystals growing from MPD, unfortunately there was a non-cleavable 
his-tag 
so I could not hope that metals would help. and other additives did not 
make any difference
then my contract finished and so did the project!

I hope you still have a lot of time available to try different things but I 
would not waste time trying to change your cryo

ciao

Stefano

Stefano Benini PhD
AstraZeneca UK
Structural Biology 
Mereside 50S38
Alderley Park








-Original Message-
From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] Behalf Of
Patrick Shaw Stewart
Sent: 11 July 2007 12:10
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] Help with reducing crystal mosaicity


Just a thought, Mary - going back to your original question about MPD.  I 
extracted the crystallization conditions from REMARK 280 of 3939 PDB entries a 
couple of years ago.  The average concentration of the MPD used was high - 
38.6%, while PEGs tended to be used at lower concs, e.g. PEG400 25.7%.  You can 
see the data at www.douglas.co.uk/top14.htm 

I thought this information could be useful if you want to replace some of the 
MPD with another precipitant (or cryoprotectant).

Best wishes

Patrick Shaw Stewart


--  
[EMAIL PROTECTED] Douglas Instruments Ltd.  
DouglasHouse, EastGarston, Hungerford, Berkshire, RG177HD, UK 
Directors: Peter Baldock, Patrick Shaw Stewart, James Smith 
http://douglas.co.uk or http://www.douglasinstruments.com 
Tel: 44 (0) 148-864-9090    US toll-free 1-877-225-2034 
Regd. England 2177994, VAT Reg. GB 480 7371 36


> -Original Message-
> From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of Mary
> Fitzgerald
> Sent: 09 July 2007 23:05
> To: CCP4BB@JISCMAIL.AC.UK
> Subject: [ccp4bb] Help with reducing crystal mosaicity
> 
> Help please!
> 
> I'm looking for some new ideas.  I have crystals that come out of a
> sitting drop with a mixture of sodium cacodylate at pH 6.5, magnesium
> acetate and MPD for the well solution.  The MPD concentration is
> sufficient to act as a cryoprotectant.  Currently, I directly freeze
> these crystals in liquid nitrogen.  When I collect data, I typically
> have high anisotropic mosaicity; it ranges from 0.8 to 1.2.  This is
> further complicated with a weakly diffracting crystal (4-5 A) that has
> a long unit cell axis of ~500 and often twinning.
> 
> It has been suggested to me that the cryoprotectent is a problem.  I
> haven't checked the diffraction at room temperature, yet.  Please no
> suggestions of finding a different crystal form as that's not a
> consideration at the moment.  I have my reasons.  I did find one
> crystal that has lower mosaicity (0.5 to 0.8) but had weaker
> diffraction then the typical crystal.  Attempts at flash cryoannealing
> have not helped.
> 
> So, what's a good way to change the cryoprotectant if the
> cryoprotectant is the precipitant?  I've considered trying dehydration
> but wasn't certain if that would help with the mosaicity.
> 
> Thanks for any ideas,
> 
> Mary X. Fitzgerald
> Postdoctoral Associate


Re: [ccp4bb] Help with reducing crystal mosaicity

2007-07-11 Thread Kris Tesh
At the risk of sounding too commercial, here are some suggestions:

1.  See if you can either place your crystallization tray in a cold room or
crystallize it there.  Mount your crystal in a capillary here so that at
whatever lower temperature you use, the distillation of solvent will be
reduced.

2.  If capillaries are too hard to work with, even a quick diffraction image
out of a loop, while the crystal is drying out, may give you some
information.  Alternately, I have saturated my stock solution with sugar
from the coffee room and swished my crystal through this for getting a quick
diffraction image.  In a rare case, the crystal actually diffracted for
several hours at room temperature.

3.  Someone mentioned using Paratone oil.  This is good, but a little thick
and sticky.  So, I use PerFluoroPolyEther (PFPE) as it is much less viscous
and is colorless.  I buy FOMBLIN Y VAC 14/6 from SigmaAldrich (Cat #
317934-100G).  PFPE is mildly hydroscopic, so it is best to open the bottle
infrequently (I withdraw ten 0.5 ml aliquots whenever I open the bottle).

4.  The Free Mounting System (FMS) is ideal for RT screening.  So long as
you do not have a highly volatile component in solution, then you can use
the standard FMS at temperatures in the range of 15 to 25 C.  If you do have
a volatile component, then Proteros bistructures GmbH (www.proteros.com) has
the only FMS with the capability of adding solvents to the gas stream...and
with some compensation, they will gladly help you with this.

Kris


Kris F. Tesh, Ph D
Director, Macromolecular Products
Rigaku Americas Corporation
9009 New Trails Drive
The Woodlands, TX  77381  USA
001 281 362 2300 x 144
-Original Message-
From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of
Benini, Stefano
Sent: Wednesday, July 11, 2007 7:17 AM
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] Help with reducing crystal mosaicity

Dear Mary,

the problem encountered with cryoprotectans is the change in the solution
surrounding your crystal as they may not be present in your crystallisation
conditions or you need to increase their concentration to make them act as
cryoprotectant.
So I don't thik is the cryoprotectant itself as it is you mother liquor..

test it in a capillary so you could rule out that freezing is the problem

of course you don't want to find new crystallisation conditions when you
tried everything and I know what it means!!

I would try to improve crystal quality using additives or ligand etc,

your long axis is very worrying as well!!! may be some magic additive could
change the packing and givew better diffraction too

I also had crystals growing from MPD, unfortunately there was a
non-cleavable his-tag 
so I could not hope that metals would help. and other additives did not
make any difference
then my contract finished and so did the project!

I hope you still have a lot of time available to try different things but I
would not waste time trying to change your cryo

ciao

Stefano

Stefano Benini PhD
AstraZeneca UK
Structural Biology 
Mereside 50S38
Alderley Park








-Original Message-
From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] Behalf Of
Patrick Shaw Stewart
Sent: 11 July 2007 12:10
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] Help with reducing crystal mosaicity


Just a thought, Mary - going back to your original question about MPD.  I
extracted the crystallization conditions from REMARK 280 of 3939 PDB entries
a couple of years ago.  The average concentration of the MPD used was high -
38.6%, while PEGs tended to be used at lower concs, e.g. PEG400 25.7%.  You
can see the data at www.douglas.co.uk/top14.htm 

I thought this information could be useful if you want to replace some of
the MPD with another precipitant (or cryoprotectant).

Best wishes

Patrick Shaw Stewart


--  
[EMAIL PROTECTED] Douglas Instruments Ltd.  
DouglasHouse, EastGarston, Hungerford, Berkshire, RG177HD, UK 
Directors: Peter Baldock, Patrick Shaw Stewart, James Smith 
http://douglas.co.uk or http://www.douglasinstruments.com 
Tel: 44 (0) 148-864-9090    US toll-free 1-877-225-2034 
Regd. England 2177994, VAT Reg. GB 480 7371 36


> -Original Message-
> From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of Mary
> Fitzgerald
> Sent: 09 July 2007 23:05
> To: CCP4BB@JISCMAIL.AC.UK
> Subject: [ccp4bb] Help with reducing crystal mosaicity
> 
> Help please!
> 
> I'm looking for some new ideas.  I have crystals that come out of a
> sitting drop with a mixture of sodium cacodylate at pH 6.5, magnesium
> acetate and MPD for the well solution.  The MPD concentration is
> sufficient to act as a cryoprotectant.  Currently, I directly freeze
> these crystals in liquid nitrogen.  When I collect data, I typically
> have high anisotropic mosaicity; it ranges from 0.8 to 1.2.  This is
> further complicated with a weakly diffracting crystal (4-5 A) that 

Re: [ccp4bb] Help with reducing crystal mosaicity

2007-07-11 Thread Das, Debanu
Hi Mary,
There is another expt. you might look into after trying out the different 
cryoprotectants, small vs. large crystals, capillary mounts, etc. as suggested.
 
Dr. Robert Thorne's group (from Mitegen & Cornell U.)) has been doing some 
studies on flash freezing procedures and effect on mosaicity/diffraction 
quality. Some of the conclusions and recommendations from these experiments 
deals with the effect of cooling rate during flash freezing (by dunking in 
liquid nitrogen) and the effect on mosaicity. 
 
   If I remember correctly,  they found that crystals they tested had much 
lower mosaicity during diffraction data collection if instead of dunking the 
looped-out crystals directly in the container with liquid nitrogen, they dunk 
it in the container which has dry liquid nitrogen blowing laterally along the 
surface. This changes the environment at the liquid/air interface near the 
container allowing for much faster cooling rates than otherwise, resulting in 
an improved diffraction quality and reduced mosaicity.
 
   I haven't tried it myself but the results seemed convincing. I know that 
sector 8 beamlines at the ALS have a modified apparatus to do this (look for 
Corie Ralston). You can get more information from them or contact Robert Thorne 
directly. 
 
BTW, ALS beamline 12.3.1 has the FMS system if you want to try that out. 
 
Regards,
Debanu.
 
---
Debanu Das,
Stanford Synchrotron Research Laboratory,
SLAC, Menlo Park, California.
 

 


From: CCP4 bulletin board on behalf of Kris Tesh
Sent: Wed 7/11/2007 8:35 AM
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] Help with reducing crystal mosaicity



At the risk of sounding too commercial, here are some suggestions:

1.  See if you can either place your crystallization tray in a cold room or
crystallize it there.  Mount your crystal in a capillary here so that at
whatever lower temperature you use, the distillation of solvent will be
reduced.

2.  If capillaries are too hard to work with, even a quick diffraction image
out of a loop, while the crystal is drying out, may give you some
information.  Alternately, I have saturated my stock solution with sugar
from the coffee room and swished my crystal through this for getting a quick
diffraction image.  In a rare case, the crystal actually diffracted for
several hours at room temperature.

3.  Someone mentioned using Paratone oil.  This is good, but a little thick
and sticky.  So, I use PerFluoroPolyEther (PFPE) as it is much less viscous
and is colorless.  I buy FOMBLIN Y VAC 14/6 from SigmaAldrich (Cat #
317934-100G).  PFPE is mildly hydroscopic, so it is best to open the bottle
infrequently (I withdraw ten 0.5 ml aliquots whenever I open the bottle).

4.  The Free Mounting System (FMS) is ideal for RT screening.  So long as
you do not have a highly volatile component in solution, then you can use
the standard FMS at temperatures in the range of 15 to 25 C.  If you do have
a volatile component, then Proteros bistructures GmbH (www.proteros.com) has
the only FMS with the capability of adding solvents to the gas stream...and
with some compensation, they will gladly help you with this.

Kris


Kris F. Tesh, Ph D
Director, Macromolecular Products
Rigaku Americas Corporation
9009 New Trails Drive
The Woodlands, TX  77381  USA
001 281 362 2300 x 144
-Original Message-
From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of
Benini, Stefano
Sent: Wednesday, July 11, 2007 7:17 AM
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] Help with reducing crystal mosaicity

Dear Mary,

the problem encountered with cryoprotectans is the change in the solution
surrounding your crystal as they may not be present in your crystallisation
conditions or you need to increase their concentration to make them act as
cryoprotectant.
So I don't thik is the cryoprotectant itself as it is you mother liquor..

test it in a capillary so you could rule out that freezing is the problem

of course you don't want to find new crystallisation conditions when you
tried everything and I know what it means!!

I would try to improve crystal quality using additives or ligand etc,

your long axis is very worrying as well!!! may be some magic additive could
change the packing and givew better diffraction too

I also had crystals growing from MPD, unfortunately there was a
non-cleavable his-tag
so I could not hope that metals would help. and other additives did not
make any difference
then my contract finished and so did the project!

I hope you still have a lot of time available to try different things but I
would not waste time trying to change your cryo

ciao

Stefano

Stefano Benini PhD
AstraZeneca UK
Structural Biology
Mereside 50S38
Alderley Park








-Original Message-
From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] Behalf Of
Patrick Shaw Stewart
Sent: 11 July 2007 12:10
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [cc

Re: [ccp4bb] Help with reducing crystal mosaicity

2007-07-11 Thread Raji Edayathumangalam
You can also try to use the micromount loops (look like ink-pen nibs) sold by 
Mitegen. They are a 
much easier alternative to capillary mounts; they are easy to handle and work 
well for room 
temperature data collection.

Good luck.
Raji


-Included Message--
>At the risk of sounding too commercial, here are some suggestions:
>
>1.  See if you can either place your crystallization tray in a cold room or
>crystallize it there.  Mount your crystal in a capillary here so that at
>whatever lower temperature you use, the distillation of solvent will be
>reduced.
>
>2.  If capillaries are too hard to work with, even a quick diffraction image
>out of a loop, while the crystal is drying out, may give you some
>information.  Alternately, I have saturated my stock solution with sugar
>from the coffee room and swished my crystal through this for getting a quick
>diffraction image.  In a rare case, the crystal actually diffracted for
>several hours at room temperature.
>
>3.  Someone mentioned using Paratone oil.  This is good, but a little thick
>and sticky.  So, I use PerFluoroPolyEther (PFPE) as it is much less viscous
>and is colorless.  I buy FOMBLIN Y VAC 14/6 from SigmaAldrich (Cat #
>317934-100G).  PFPE is mildly hydroscopic, so it is best to open the bottle
>infrequently (I withdraw ten 0.5 ml aliquots whenever I open the bottle).
>
>4.  The Free Mounting System (FMS) is ideal for RT screening.  So long as
>you do not have a highly volatile component in solution, then you can use
>the standard FMS at temperatures in the range of 15 to 25 C.  If you do have
>a volatile component, then Proteros bistructures GmbH (www.proteros.com) has
>the only FMS with the capability of adding solvents to the gas stream...and
>with some compensation, they will gladly help you with this.
>
>Kris
>
>
>Kris F. Tesh, Ph D
>Director, Macromolecular Products
>Rigaku Americas Corporation
>9009 New Trails Drive
>The Woodlands, TX  77381  USA
>001 281 362 2300 x 144
>-Original Message-
>From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of
>Benini, Stefano
>Sent: Wednesday, July 11, 2007 7:17 AM
>To: CCP4BB@JISCMAIL.AC.UK
>Subject: Re: [ccp4bb] Help with reducing crystal mosaicity
>
>Dear Mary,
>
>the problem encountered with cryoprotectans is the change in the solution
>surrounding your crystal as they may not be present in your crystallisation
>conditions or you need to increase their concentration to make them act as
>cryoprotectant.
>So I don't thik is the cryoprotectant itself as it is you mother liquor..
>
>test it in a capillary so you could rule out that freezing is the problem
>
>of course you don't want to find new crystallisation conditions when you
>tried everything and I know what it means!!
>
>I would try to improve crystal quality using additives or ligand etc,
>
>your long axis is very worrying as well!!! may be some magic additive could
>change the packing and givew better diffraction too
>
>I also had crystals growing from MPD, unfortunately there was a
>non-cleavable his-tag 
>so I could not hope that metals would help. and other additives did not
>make any difference
>then my contract finished and so did the project!
>
>I hope you still have a lot of time available to try different things but I
>would not waste time trying to change your cryo
>
>ciao
>
>Stefano
>
>Stefano Benini PhD
>AstraZeneca UK
>Structural Biology 
>Mereside 50S38
>Alderley Park
>
>
>
>
>
>
>
>
>-Original Message-
>From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] Behalf Of
>Patrick Shaw Stewart
>Sent: 11 July 2007 12:10
>To: CCP4BB@JISCMAIL.AC.UK
>Subject: Re: [ccp4bb] Help with reducing crystal mosaicity
>
>
>Just a thought, Mary - going back to your original question about MPD.  I
>extracted the crystallization conditions from REMARK 280 of 3939 PDB entries
>a couple of years ago.  The average concentration of the MPD used was high -
>38.6%, while PEGs tended to be used at lower concs, e.g. PEG400 25.7%.  You
>can see the data at www.douglas.co.uk/top14.htm 
>
>I thought this information could be useful if you want to replace some of
>the MPD with another precipitant (or cryoprotectant).
>
>Best wishes
>
>Patrick Shaw Stewart
>
>
>--  
>[EMAIL PROTECTED]Douglas Instruments Ltd.  
>DouglasHouse, EastGarston, Hungerford, Berkshire, RG177HD, UK 
>Directors: Peter Baldock, Patrick Shaw Stewart, James Smith 
>http://douglas.co.uk or http://www.douglasinstruments.com 
>Tel: 44 (0) 148-864-9090US toll-free 1-877-225-2034 
>Regd. England 2177994, VAT Reg. GB 480 7371 36
>
>
>> -Original Message-