TNO made a suggestive title by referring to RFID devices, where most problems
are induced by only 2 of the available systems. Probably to draw more
attention for their article, and they succeeded.

TNO is a Netherlands institute for applied science and have been forced to
change their way of being financed from public to private.

 

Apparently, most of the problems occur with high power 868 MHz backscatter
RFID systems. These are conditionally allowed to transmit up to 3.8 watts EIRP
(Friss tells us that that might be up to  50 V/m ) For frequency power
allocations in NL :

 

http://www.agentschaptelecom.nl/nfr/main_nfr_uk.html

 

This is close to  the  GSM-standard and power range, and we all know that
these devices are to be used with care (forbidden) in medical environments.

 

These RFID’s  are definitely not the items we usually use as RFID such as 
Legic, Mifare or other low power RFID systems.

The article should therefore be read as a warning against high power RFID
backscatter technologies.

While we can discuss about if  Medical equipment should or should not be
immune to this field strength, we may discuss the suitability of these 868 MHz
backscatter radar type RFID systems,  in medical and  residential 
environments where fields of 3 or maximum 10V/m should be expected.

I wonder btw how this allowance of  2W ERP (3.8 W EIRP) got into the European
frequency tables. Maybe because the power and frequency are close to GSM. Not
taking into account the much closer operating distance that goes with RFID.

 

For the time being we will have to live with RFID technologies  that interfere
with medical (and other) electronics, until the  standards committees again 
noticed their lag on technology. and raise the immunity levels to  50 V/m or
more.

 

(EN 55024  for example still uses 3V/m and limits the test range to 1 GHz only
, where 1.8 GHz GSM are used all over Europe, and even higher frequency UMTS
systems are being rolled out: new problems are to be expected!)

 

Regarding the discussion: both interfering RFID systems should not be
‘simulated’ with the EN 6100-4-6 method, as the 125 kHz is below it’s 
f-limits (and would go unnoticed), and the other is way above it’s upper
f-limits (80  or 230 MHz)

 

 

Gert Gremmen

ce-test, qualified testing bv

 

Van: emc-p...@ieee.org [mailto:emc-p...@ieee.org] Namens Ken Javor
Verzonden: Friday, June 27, 2008 12:57 AM
Aan: Untitled
Onderwerp: Re: EMC in the news: RFID  Medical

 

None of the below.  61000-4-3 specified rf source impedance is 150 Ohms,
calibrated into 150 Ohms load impedance. The 1, 3, and 10 Volt levels are
open-circuit quantities, thus the potential developed in the matched load is
one-half the limit.  The current flowing on a cable under test will depend on
the cable impedance, which is not controlled, except on the AE side, because
the injection device is designed to insert a high common mode impedance on the
AE side of the cable (high relative to 150 Ohms).  As far as I know, you do
not measure injected current doing the test this way.  There is an alternative
current injection test when a suitable coupling device is not available, but
then you inject and measure current, not rf potential.  The current limit is
the short circuit current available from the open circuit potential limit
divided by the 150 Ohm source impedance, even though you will not have a 150
Ohm source impedance when injecting current.
 
Ken Javor

Phone: (256) 650-5261





From: Daniel Roman dan.ro...@dialogic.com
List-Post: emc-pstc@listserv.ieee.org
List-Post: emc-pstc@listserv.ieee.org
List-Post: emc-pstc@listserv.ieee.org
Date: Thu, 26 Jun 2008 17:53:59 -0400
To: Ken Javor ken.ja...@emccompliance.com, Untitled emc-p...@ieee.org
Conversation: EMC in the news: RFID  Medical
Subject: RE: EMC in the news: RFID  Medical

While slightly off-topic, this discussion got me thinking about the
measurement of the voltage on the cable.  If done with a RF cable current
clamp, do you have to do a conversion of some sort if the cable is assumed to
be 150 ohms?
 
For a 50 ohm system, a 3V level measured with a clamp would be 96 dBuA minus
the transfer impedance in dB ohms.  If the cable is 150 ohms though does that
mean you are measuring the equivalent of 1V from the clamp instead of 3V
because the clamp is measuring current?  Stated another way, assuming the
transfer impedance of the current clamp is zero to simply things, for an
assumed 150 ohm cable should you be reading 96 dBuV on a spectrum analyzer
hooked up to the probe or something less that than equivalent to 1V?
 
Never worked with this stuff so I’m curious.
 
Dan
 

From: emc-p...@ieee.org [mailto:emc-p...@ieee.org] On Behalf Of Ken Javor
Sent: Thursday, June 26, 2008 11:14 AM
To: Untitled
Subject: Re: EMC in the news: RFID  Medical

Comment (e) is unrealistic.  Assuming 61000-4-6 was applied, the amount of
power required to directly inject 1, 3 or 10 Volts oc into 150 Ohms is 

Re: EMC in the news: RFID Medical

2008-06-27 Thread John Woodgate

In message 48640ca5.3060...@orionscientific.com, dated Thu, 26 Jun 
2008, E. Robert Bonsen robert.bon...@orionscientific.com writes:


Assessment took place^ according to an international test protocol.

Really? Don't they cite the reference? I don't know of one that covers 
125 kHz and 868 MHz.
-- 
OOO - Own Opinions Only. Try www.jmwa.demon.co.uk and www.isce.org.uk
Either we are causing global warming, in which case we may be able to stop it,
or natural variation is causing it, and we probably can't stop it. You choose!
John Woodgate, J M Woodgate and Associates, Rayleigh, Essex UK

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:emc-p...@daveheald.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc






Re: Power Spectral Density Measurement

2008-06-27 Thread Grace Lin
Dear Members,
 
Thank you very much to those whe have commented.
 
FYI.  We found out the data discrepancy was caused by a test fixture/board
(support test commands to set up channels and power level) and the length of
power lead (from EUT to DC power supply).  

Best regards,
Grace
 
On 6/18/08, Grace Lin graceli...@gmail.com wrote: 

Dear Members,
 
I need your help to find out what may be wrong/inaccurate for the power
spectral density measurement per 47 CFR 15.247.
 
I was told the data I took was about 6 dB higher.  I follow the 
measurement
procedure (PSD Option 1) for the digital transmission system, published by the
FCC ( 

https://fjallfoss.fcc.gov/oetcf/kdb/fo
ms/FTSSearchResultPage.cfm?id=21124switch=P).  The antenna port of the
transmitter was connected to the RS ESU40 receiver (in spectrum analyzer
mode).  Data was taken with and without 30dB attenuators to avoid overload
issue.  RBW=3kHz, VBW=10kHz, Span=1.5MHz, Sweep=500s.  I couldn't figure out
where might be wrong.

 

Thank you and look forard to your instructions.

 

Best regards,

Grace


-  This
message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to emc-p...@ieee.org 

Instructions: http://listserv.ieee.org/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators: 

Scott Douglas emcp...@ptcnh.net Mike Cantwell mcantw...@ieee.org 

For policy questions, send mail to: 

Jim Bacher: j.bac...@ieee.org David Heald: emc-p...@daveheald.com 

All emc-pstc postings are archived and searchable on the web at: 

http://www.ieeecommunities.org/emc-pstc 



RE: EMC in the news: RFID Medical

2008-06-27 Thread Dean Gerard (Medical Physics)
  

Whilst I've seen the original article (Journal of the American Medical
Association), unfortunately I can't circulate it without breaking our
subscription license agreement.

The investigation was apparently commissioned by the Dutch Ministry of Health. 

Standard they refer to is - 

American National Standard Recommended Practice for On-site Ad Hoc Test Method
for Estimating Radiated Electromagnetic Immunity of Medical Devices to
Specific Radio-frequency Transmitters (Standard C63.18). (AAMI:/ IEEE; 1997).

(not a rigorous EMC product test standard - but rather a suggested basic
methodology for equipment users to investigate possible EMI in a consistent
way)


The project was apparently driven by the search for a system to monitor blood
products, which influenced their choice of RFID systems to test, as follows -

The selection of 2 RFID systems tested in this study was based on 3
characteristics: (1) the systems needed to comply with RFID standards set by
the European Telecommunications Standards Institute; (2) radio frequencies
needed to fall within the most common internationally used RFID frequency
bands and (3) performance needed to fulfill the operational requirements of
the project including availability of temperature sensitive RFID tags,
low-cost tags suitable for disposable materials, contemporary integration with
the local communications network, and location accuracy within a health care
facility.

Tests were performed in a 1-patient room, free of reflective objects, based
upon methodology in the standard above. 


Ged Dean 
Nottingham University Hospitals NHS Trust 


___
_ 

-Original Message- 
From: emc-p...@ieee.org [mailto:emc-p...@ieee.org mailto:emc-p...@ieee.org ]
On Behalf Of John Woodgate 
Sent: 27 June 2008 08:29 
To: emc-p...@ieee.org 
Subject: Re: EMC in the news: RFID  Medical 

In message 48640ca5.3060...@orionscientific.com, dated Thu, 26 Jun 2008, E.
Robert Bonsen robert.bon...@orionscientific.com writes:


Assessment took place^ according to an international test protocol. 

Really? Don't they cite the reference? I don't know of one that covers 
125 kHz and 868 MHz. 
-- 
OOO - Own Opinions Only. Try www.jmwa.demon.co.uk
file://www.jmwa.demon.co.uk  and www.isce.org.uk file://www.isce.org.uk  
Either we are causing global warming, in which case we may be able to stop it, 
or natural variation is causing it, and we probably can't stop it. You choose! 
John Woodgate, J M Woodgate and Associates, Rayleigh, Essex UK 

- 
 



The information contained in this email may be the subject of public
disclosure under the
Freedom of Information Act 2000 - unless legally exempt from disclosure, the
confidentiality of this email and your reply cannot be guaranteed.

Copyright of this email and any attachments created by us remains vested in the
Nottingham University Hospitals NHS Trust.

This email and any files transmitted with it are confidential and intended
solely for 
the use of the individual or entity to whom they are addressed. 

This footnote also confirms that this email message has been swept by 
security software for the presence of known computer viruses.

 

Nottingham University Hospitals NHS Trust
City Hospital campus | Hucknall Road | Nottingham | NG5 1PB | Tel: 0115 969
1169
QMC campus | Derby Road | Nottingham | NG7 2UH | Tel: 0115 924 9924

If you have received this email in error please notify the sender. 


ID:QMCICT1 - B486510b70002.0001.0001.mml - 17:09:27  06/27/08 

-  This
message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to emc-p...@ieee.org 

Instructions: http://listserv.ieee.org/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators: 

Scott Douglas emcp...@ptcnh.net Mike Cantwell mcantw...@ieee.org 

For policy questions, send mail to: 

Jim Bacher: j.bac...@ieee.org David Heald: emc-p...@daveheald.com 

All emc-pstc postings are archived and searchable on the web at: 

http://www.ieeecommunities.org/emc-pstc 



Re: EMC in the news: RFID Medical

2008-06-27 Thread John Woodgate

In message c1bb4f6866f4d144bb37968587d03a64e47...@cexc01.nuh.LOCAL, 
dated Fri, 27 Jun 2008, Dean Gerard (Medical Physics) 
gerard.d...@nuh.nhs.uk writes:


Whilst I've seen the original article (Journal of the American Medical 
Association), unfortunately I can't circulate it without breaking our 
subscription license agreement.

Thanks for the information you have given, anyway. The elephant in the 
room is the field strength information. Is there anything about that in 
the paper?
-- 
OOO - Own Opinions Only. Try www.jmwa.demon.co.uk and www.isce.org.uk
Either we are causing global warming, in which case we may be able to stop it,
or natural variation is causing it, and we probably can't stop it. You choose!
John Woodgate, J M Woodgate and Associates, Rayleigh, Essex UK

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:emc-p...@daveheald.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc






FW: EMC in the news: RFID Medical

2008-06-27 Thread Dean Gerard (Medical Physics)


Just nominal power outputs and separation distances between interfering and
susceptible equipment. 

Output info given is - 
 The passive RFID system selected for this study (OBID, Feig Electronic,
Weilburg, Germany) had an 868-MHz reader (2-4 W). The active RFID system

(Eureka RFID, Avonwood, England) had a 125-kHz reader (68_10E-3 µT at 1m)
that forces tags to transmit in its proximity. The active RFID tag had an
operational frequency of 868 MHz at 2 µW

Interference effects were provoked at separation distances ranging from 5 -
600cm, depending upon equipment affected. 



Ged Dean 
Nottingham University Hospitals NHS Trust 



 

-Original Message- 
From: emc-p...@ieee.org [mailto:emc-p...@ieee.org mailto:emc-p...@ieee.org ]
On Behalf Of John Woodgate 
Sent: 27 June 2008 17:28 
To: emc-p...@ieee.org 
Subject: Re: EMC in the news: RFID  Medical 

In message c1bb4f6866f4d144bb37968587d03a64e47...@cexc01.nuh.LOCAL, 
dated Fri, 27 Jun 2008, Dean Gerard (Medical Physics) 
gerard.d...@nuh.nhs.uk writes: 


Whilst I've seen the original article (Journal of the American Medical 
Association), unfortunately I can't circulate it without breaking our 
subscription license agreement. 

Thanks for the information you have given, anyway. The elephant in the 
room is the field strength information. Is there anything about that in 
the paper? 
-- 
OOO - Own Opinions Only. Try www.jmwa.demon.co.uk
file://www.jmwa.demon.co.uk  and www.isce.org.uk file://www.isce.org.uk  
Either we are causing global warming, in which case we may be able to stop it, 
or natural variation is causing it, and we probably can't stop it. You choose! 
John Woodgate, J M Woodgate and Associates, Rayleigh, Essex UK 

- 
 
This message is from the IEEE Product Safety Engineering Society 
emc-pstc discussion list.Website:  http://www.ieee-pses.org/
http://www.ieee-pses.org/  

To post a message to the list, send your e-mail to emc-p...@ieee.org 

Instructions:  http://listserv.ieee.org/request/user-guide.html
http://listserv.ieee.org/request/user-guide.html  

List rules: http://www.ieee-pses.org/listrules.html
http://www.ieee-pses.org/listrules.html  

For help, send mail to the list administrators: 

 Scott Douglas   emcp...@ptcnh.net 
 Mike Cantwell   mcantw...@ieee.org 

For policy questions, send mail to: 

 Jim Bacher: j.bac...@ieee.org 
 David Heald:emc-p...@daveheald.com 

All emc-pstc postings are archived and searchable on the web at: 

http://www.ieeecommunities.org/emc-pstc
http://www.ieeecommunities.org/emc-pstc  



The information contained in this email may be the subject of public
disclosure under the
Freedom of Information Act 2000 - unless legally exempt from disclosure, the
confidentiality of this email and your reply cannot be guaranteed.

Copyright of this email and any attachments created by us remains vested in the
Nottingham University Hospitals NHS Trust.

This email and any files transmitted with it are confidential and intended
solely for 
the use of the individual or entity to whom they are addressed. 

This footnote also confirms that this email message has been swept by 
security software for the presence of known computer viruses.

 

Nottingham University Hospitals NHS Trust
City Hospital campus | Hucknall Road | Nottingham | NG5 1PB | Tel: 0115 969
1169
QMC campus | Derby Road | Nottingham | NG7 2UH | Tel: 0115 924 9924

If you have received this email in error please notify the sender. 


ID:QMCICT1 - B48651855.0001.0001.mml - 17:41:57  06/27/08 

-  This
message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to emc-p...@ieee.org 

Instructions: http://listserv.ieee.org/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators: 

Scott Douglas emcp...@ptcnh.net Mike Cantwell mcantw...@ieee.org 

For policy questions, send mail to: 

Jim Bacher: j.bac...@ieee.org David Heald: emc-p...@daveheald.com 

All emc-pstc postings are archived and searchable on the web at: 

http://www.ieeecommunities.org/emc-pstc 



RE: EMC in the news: RFID Medical

2008-06-27 Thread Conway, Patrick R (Houston)
 
This is great information.  
Thank you for providing some details.  
 
The comment: had an 868-MHz reader (2-4 W).   is informative.  It
indicates that the RFID was not the only transmitter in the proximity of the
medical EUT.  This RFID reader adds another parameter that requires control
and investigation during the test.  
 
continuing the list of possibilities:
 
f)  is it possible that the RFID tag has much less effect than that of the
RFID reader?
Would be informative to find out if the RFID reader, in the absence of the
tags themselves, can duplicate any of the EUT problems.

 

Best Regards, 

Patrick. 
p.con...@hp.com 

 




From: emc-p...@ieee.org [mailto:emc-p...@ieee.org] On Behalf Of Dean Gerard
(Medical Physics)
Sent: Friday, June 27, 2008 11:48 AM
To: emc-p...@ieee.org
Subject: FW: EMC in the news: RFID  Medical






Just nominal power outputs and separation distances between interfering and
susceptible equipment. 

Output info given is - 
 The passive RFID system selected for this study (OBID, Feig Electronic,
Weilburg, Germany) had an 868-MHz reader (2-4 W). The active RFID system

(Eureka RFID, Avonwood, England) had a 125-kHz reader (68_10E-3 µT at 1m)
that forces tags to transmit in its proximity. The active RFID tag had an
operational frequency of 868 MHz at 2 µW

Interference effects were provoked at separation distances ranging from 5 -
600cm, depending upon equipment affected. 



Ged Dean 
Nottingham University Hospitals NHS Trust 

 

-  This
message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to emc-p...@ieee.org 

Instructions: http://listserv.ieee.org/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators: 

Scott Douglas emcp...@ptcnh.net Mike Cantwell mcantw...@ieee.org 

For policy questions, send mail to: 

Jim Bacher: j.bac...@ieee.org David Heald: emc-p...@daveheald.com 

All emc-pstc postings are archived and searchable on the web at: 

http://www.ieeecommunities.org/emc-pstc 



Re: EMC in the news: RFID Medical

2008-06-27 Thread pat.lawler

Hi John: 

No, the article doesn't mention field strength.  However, the test procedure
is one I'm not familiar with: 
   ANSI C63.18-1997, 'American national standard recommended practice for an
on-site, ad hoc test method for 
   estimating radiated electromagnetic immunity of medical devices to specific
radio-frequency transmitters' 
In short, they vary the separation between the RFID device and the medical
equipment from 600 cm (20 feet) down to 0 cm (touching the equipment case),
and record the distance where the equipment failed.  Random RFID devices,
random power levels, with a test distance down to zero? 

I appreciate RFID interference is something to be considered.  However, it's
not fair to compare the immunity standard used in the report, to the pass/fail
testing with the fixed amplitude/fixed distance conditions in present IEC
immunity tests (mentioned elsewhere in the email thread.) 

Pat Lawler
EMC Engineer
SL Power Electronics Corp.

John Woodgate j...@jmwa.demon.co.uk wrote on 06/27/2008 09:28:20 AM:
 In message c1bb4f6866f4d144bb37968587d03a64e47...@cexc01.nuh.LOCAL,
 dated Fri, 27 Jun 2008, Dean Gerard (Medical Physics)
 gerard.d...@nuh.nhs.uk writes:
 
 Whilst I've seen the original article (Journal of the American Medical
 Association), unfortunately I can't circulate it without breaking our
 subscription license agreement.

 Thanks for the information you have given, anyway. The elephant in the
 room is the field strength information. Is there anything about that in
 the paper?
 --
 OOO - Own Opinions Only. Try www.jmwa.demon.co.uk and www.isce.org.uk
 Either we are causing global warming, in which case we may be able to stop
it,
 or natural variation is causing it, and we probably can't stop it. You
choose!
 John Woodgate, J M Woodgate and Associates, Rayleigh, Essex UK -
 This message
is from the IEEE Product Safety Engineering Society emc-pstc discussion list.
Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to emc-p...@ieee.org 

Instructions: http://listserv.ieee.org/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators: 

Scott Douglas emcp...@ptcnh.net Mike Cantwell mcantw...@ieee.org 

For policy questions, send mail to: 

Jim Bacher: j.bac...@ieee.org David Heald: emc-p...@daveheald.com 

All emc-pstc postings are archived and searchable on the web at: 

http://www.ieeecommunities.org/emc-pstc