Hi Dave and Net,

Transmitting at 160m (1.8MHz) in an operating position with about 600V/m in
my amateur radio stations (HB9CVQ/DK2VQ) would certainly make me search for
a reduction method to lower the fields to less than 20 V/m. I tested this
with 1 kW output and  a calibrated field strength monitor (Narda Safety Test
Systems) in CW-mode.
Setup: Antenna G5RV , short dipole 2x15.4m, normally  9.85m 450 Ohm open
wire feeder, at 160m extension with extra 17,5m 450 Ohm line, common mode
choke (35dB) , 3.5m RG213, asymmetrical tuner, TL922, IC-761. All equipment
is shielded and properly grounded (max. I CM < 10mA anywhere at
1kW-reference ground plane below equipment desk installed-open zoning
concept). The antenna feed point is about 7m away with the feed line being
perfectly in the symmetry plane of the ant. The end of the feed line is
about 2m away from the operating position.
With just the 9.85m feeder line and 15  uH inductance , about 30 cm
diameter, open-no shield, real low loss inductors at the end, to tune the
line on each side, I got 1.5 KV/m in the operating position-about 1,3m away
only with 100W output on 1.85MHz! That made me worry to change very quickly.
Efficiency of this new antenna system (additional extended feeder by 17.5m-
coiled up on 40cm diameter) is naturally reduced over a full half wave
dipole, but reasonable for the restricted space. In publicly accessible
areas near the antenna (feed point 12m up and ends down to 9m height above
ground) the E-field drops to about 1V/m at full legal 1kW level. By the way
measuring very near a metallic surface can sometimes produce strange
readings with E-field probes, better check the surface currents with a
suitable probe.
I hope these comments are useful

Diethard Hansen

Best regards

Dr. Diethard Hansen

EES President, Principal Consultant, Trainer
US Senior NARTE certif. EMC Eng. 001937NE
EMC+Automotive+Telecom QM+techn. EA-Auditor

Euro EMC SERVICE (EES) Dr.-Ing. D. Hansen
POB 64, Bahnhofstr.39, CH-8965 Berikon 2,
Switzerland, Swiss VAT (Mwst.)No.:323214

tel./fax: +41 566 33 73 81
German mobile phone: +49(0)1736015909
www.euro-emc-service.de <http://www.euro-emc-service.de>
euro.emc.serv...@swissonline.ch <mailto:euro.emc.serv...@swissonline.ch>
( mailto:diethard.han...@ieee.org <mailto:diethard.han...@ieee.org> )

Consulting+Seminars+R&D-Projects
Marketing+CE-Testing+ww.Audits


From: owner-emc-p...@listserv.ieee.org
[mailto:owner-emc-p...@listserv.ieee.org]On Behalf Of drcuthb...@micron.com
Sent: Freitag, 19. November 2004 15:47
To: israe...@oms-it.co.il; emc-p...@ieee.org
Subject: RE: Heart pacemakers - Vulnerability to EM


I don't know what the standards call for but it is quite possible for a
person to be subjected to an E-field exceeding 500 V/m very near to an
amateur radio station. This is the field at my operating position when I am
transmitting at 1.8 MHz. A cell phone held close to the body will cause a
field, at the body, of over 30 V/m.

   Dave Cuthbert

From: owner-emc-p...@listserv.ieee.org
[mailto:owner-emc-p...@listserv.ieee.org] On Behalf Of Israel
Sent: Thursday, November 18, 2004 4:55 PM
To: emc-p...@ieee.org
Subject: Heart pacemakers - Vulnerability to EM



Hi all



Heart pacemakers may be vulnerable to Electro Magnetic energy.  Has anyone
idea about safe immunity levels ? about relevant standards regarding this
matter ?



Thanks !

 This
message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org


Instructions: http://listserv.ieee.org/listserv/request/user-guide.html


List rules: http://www.ieee-pses.org/listrules.html


For help, send mail to the list administrators:


Ron Pickard: emc-p...@hypercom.com Scott Douglas emcp...@ptcnh.net


For policy questions, send mail to:


Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org


All emc-pstc postings are archived and searchable on the web at:


http://www.ieeecommunities.org/emc-pstc

 This
message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org


Instructions: http://listserv.ieee.org/listserv/request/user-guide.html


List rules: http://www.ieee-pses.org/listrules.html


For help, send mail to the list administrators:


Ron Pickard: emc-p...@hypercom.com Scott Douglas emcp...@ptcnh.net


For policy questions, send mail to:


Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org


All emc-pstc postings are archived and searchable on the web at:


http://www.ieeecommunities.org/emc-pstc



This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message

Re: Heart pacemakers - Vulnerability to EM

2004-11-20 Thread owner-emc-p...@listserv.ieee.org
It may be worth noting that in the medium wave broadcast band the FCC limit
is over 600 volts per meter. We don't incur inherently harmful induced body
currents with less at these frequencies -- unless we're hanging onto an
antenna.

But considering the low potentials to be measured, and the energy delivered
by a pacer or defibrillator, it seems difficult to adequately filter device
leads from all fields they might see, so we have the fall-back constant
pacing mode mentioned before. In my opinion, the immunity levels required
by the Medical Directive are not unreasonable, even if they may be
encountered rarely; one report extant on the Web is of a patient who
subjected to search by an airport security wand, which caused his
implanted, computerized drug delivery system to deliver a non-survivable
level of drugs.  Inductive coupling directly to devices and leads is quite
probably a harsher environment than being near a broadcast antenna. But all
too common!


Cortland Richmond


This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/listserv/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Ron Pickard:  emc-p...@hypercom.com
 Scott Douglas emcp...@ptcnh.net

For policy questions, send mail to:

 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc



Re: Heart pacemakers - Vulnerability to EM

2004-11-20 Thread owner-emc-p...@listserv.ieee.org
But there is a very big difference between the two fields Mr. Cuthbert cites,
aside from frequency.  The cell phone field likely has a field impedance close
to that of free space, because the antenna is a good match to the transmitter.
 Unless your tower is around 100' and operating as a base fed quarter-wave
stub, you are likely using a tuning device to step up the output potential of
your transmitter to more effectively drive a shorter stub radiator.  The net
effect is that in the very near field the ratio of electric to magnetic field
is very much higher than 377 Ohms.  The susceptibility effect of such a
quasi-static electric field is very different than for a plane wave,
especially if the EUT is surrounded by a shielding material, such as the human
body, or any metallic shielding material one might choose to employ.  A
conclusion I would draw from this is that if it were deemed necessary to
simulate the near field of an amateur radio site or an AM broadcast station, I
would use an open-circuited parallel plate, as opposed to one loaded by 50
Ohms or other matching impedance. 



From: drcuthb...@micron.com
List-Post: emc-pstc@listserv.ieee.org
Date: Fri, 19 Nov 2004 07:46:59 -0700
To: , 
Subject: RE: Heart pacemakers - Vulnerability to EM




I don't know what the standards call for but it is quite possible for a person
to be subjected to an E-field exceeding 500 V/m very near to an amateur radio
station. This is the field at my operating position when I am transmitting at
1.8 MHz. A cell phone held close to the body will cause a field, at the body,
of over 30 V/m.

  Dave Cuthbert 

From: owner-emc-p...@listserv.ieee.org 
mailto:owner-emc-p...@listserv.ieee.org] On Behalf Of Israel
Sent: Thursday, November 18, 2004 4:55 PM
To: emc-p...@ieee.org
Subject: Heart pacemakers - Vulnerability to EM

Hi all 



Heart pacemakers may be vulnerable to Electro Magnetic energy.  Has anyone
idea about safe immunity levels ? about relevant standards regarding this
matter ? 



Thanks !
 This message
is from the IEEE Product Safety Engineering Society emc-pstc discussion list.
Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to emc-p...@ieee.org 

Instructions: http://listserv.ieee.org/listserv/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators: 

Ron Pickard: emc-p...@hypercom.com Scott Douglas emcp...@ptcnh.net 

For policy questions, send mail to: 

Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org 

All emc-pstc postings are archived and searchable on the web at: 

http://www.ieeecommunities.org/emc-pstc
 This message
is from the IEEE Product Safety Engineering Society emc-pstc discussion list. 
  Website:  http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to emc-p...@ieee.org 

Instructions:  http://listserv.ieee.org/listserv/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators: 

Ron Pickard:  emc-p...@hypercom.com Scott Douglas
emcp...@ptcnh.net 

For policy questions, send mail to: 

Richard Nute:   ri...@ieee.org Jim Bacher:
j.bac...@ieee.org 

All emc-pstc postings are archived and searchable on the web at: 

http://www.ieeecommunities.org/emc-pstc 



 This message
is from the IEEE Product Safety Engineering Society emc-pstc discussion list.
Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to emc-p...@ieee.org 


Instructions: http://listserv.ieee.org/listserv/request/user-guide.html 


List rules: http://www.ieee-pses.org/listrules.html 


For help, send mail to the list administrators: 


Ron Pickard: emc-p...@hypercom.com Scott Douglas emcp...@ptcnh.net 


For policy questions, send mail to: 


Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org 


All emc-pstc postings are archived and searchable on the web at: 


http://www.ieeecommunities.org/emc-pstc 




RE: Heart pacemakers - Vulnerability to EM

2004-11-19 Thread owner-emc-p...@listserv.ieee.org
I have just a bit of personal experience in testing implantable pacemakers and
volumetric infusion pumps. I had been doing RF immunity studies for a
manufacturer, circa 1990, using an FDA standard (MDS-401-xxx?) for medical
electronic devices.

  
I exposed devices to fields as strong as I could create (10 kHz through 18
GHz, always at least 200 V/M, sometimes as high as 600 V/M), using both 1 kHz,
50% duty cycle and a 200 Hz, 10% duty cycle, gated at a 1 Hz rate. The devices
were tested first in open air, and then immersed in a rectangular tank of
saline solution (a crude attempt to simulate body tissue). I also exposed them
to 20 Oersted DC & 60 Hz magnetic fields (the infusion pumps had a magnetic
communications link for programming and monitoring).

  
The nice part is that, after some design changes, I was unable to elicit any
improper responses from the devices. Since these devices were so nicely
immune, I imagine that a person would feel the effects of tissue heating well
before the electronics are jeopardized.

  
I hope they still make them as good now.  
  
Ed 

Ed Price 
ed.pr...@cubic.com WB6WSN 
NARTE Certified EMC Engineer & Technician 
Electromagnetic Compatibility Lab 
Cubic Defense Applications 
San Diego, CA USA 
858-505-2780 (Voice) 
858-505-1583 (Fax) 
Military & Avionics EMC Is Our Specialty 

  




 

From: Barker, Neil [ mailto:neil.bar...@e2v.com] 
Sent: Friday, November 19, 2004 8:20 AM 
To: 'drcuthb...@micron.com'; israe...@oms-it.co.il; emc-p...@ieee.org 
    Subject: RE: Heart pacemakers - Vulnerability to EM 


I am no expert in this subject, and some of what I am about to say is
hearsay, but here it is for what it's worth.

 
Safety levels are generally based not on field strength, but on power
density. However, these levels are set for the power absorption of the body
(in W/kg) based on the heating effect. For comparison purposes consider the
acceptance levels applied to SAR testing. On this basis, I suspect that Dave's
amateur rig probably doesn't meet SAR acceptance levels, and should probably
have an exclusion zone around the antenna!

 
The original message was concerned with the effect on a pacemaker
rather than directly on the body itself. Clearly, if you disrupt a pacemaker
then it is the pacemaker that has an effect on the body.

I believe that the susceptibility depends greatly on whether they are
external pacemakers or implanted pacemakers. The implanted devices are much
less susceptible as they are completely surrounded by soft body tissue, which
is largely water in any case. External devices are much more susceptible, due
largely to the relatively long electrode wire. being long, this is generally
susceptible at fairly low frequencies, and I have heard that they can be
affected by the magnetic loops used in store doorways for security
(merchandise protection) purposes. I have no knowledge of their susceptibility
to E fields.

 
Best regards, 

Neil R. Barker C.Eng. MIEE FSEE MIEEE 
Manager 
Compliance Engineering 
e2v technologies (uk) ltd 
106 Waterhouse Lane 
Chelmsford 
Essex 
CM1 2QU 
UK 

Tel: +44 (0)1245 453616 
Fax: +44 (0)1245 453410 
e-mail: neil.bar...@e2v.com 
Web: http://www.e2v.com < http://www.e2v.com/>  

-Original Message- 
From: drcuthb...@micron.com [ mailto:drcuthb...@micron.com] 
Sent: 19 November 2004 14:47 
To: israe...@oms-it.co.il; emc-p...@ieee.org 
        Subject: RE: Heart pacemakers - Vulnerability to EM 


I don't know what the standards call for but it is quite
possible for a person to be subjected to an E-field exceeding 500 V/m very
near to an amateur radio station. This is the field at my operating position
when I am transmitting at 1.8 MHz. A cell phone held close to the body will
cause a field, at the body, of over 30 V/m.

 
   Dave Cuthbert 
-Original Message- 
From: owner-emc-p...@listserv.ieee.org [
mailto:owner-emc-p...@listserv.ieee.org] On Behalf Of Israel 
Sent: Thursday, November 18, 2004 4:55 PM 
To: emc-p...@ieee.org 
Subject: Heart pacemakers - Vulnerability to EM 



Hi all 

 

Heart pacemakers may be vulnerable to Electro Magnetic energy.
 Has anyone idea about safe immunity levels ? about relevant standards
regarding this matter ?

 

Thanks !

 This message
is from the IEEE Product Saf

RE: Heart pacemakers - Vulnerability to EM

2004-11-19 Thread owner-emc-p...@listserv.ieee.org
The standards EN 50061 /EN 45502-2-1 are dealing with the immunity of pace
makers. 

Immunity is tested in a torso simulator and a specific antenna to transmit
close to water at the right frequency.

The problem is not the pacemaker itself, it is the fact that the sensors
provide a good antenna.

Beside this modern pace makers additionally have a mode of operation they go
into if they got interference with a certain level above the normal
interference they have. 

They don’t sense anymore and got to a fixed  rhythm.

 

Even with that relatively high level of immunity there are devices people with
pace makers have to stay away like high power transformers and all devices
creating a high magnetic field (MRT).  Electric shavers, hairdryers and tolls
like electric drills also have a potential to disturb the function if used
close to the pace maker.

 

 

 

Lothar Schmidt

 

Technical Manager EMC, SAR, Antenna testing and BQB

 

CETECOM Inc. 

411 Dixon Landing Road

Milpitas, CA 95035

 

Phone +1 408 586 6214

Fax +1 408 586 6299

 

This e-mail may contain proprietary, privileged and confidential information
for the sole use of the named intended recipient. Any review or distribution
of this e-mail by any party other than the intended recipient or that person's
agent is strictly prohibited. If you are not the intended recipient, please
delete all copies and immediately Contact the sender. You must not, directly
or indirectly, use, disclose, distribute, print, or copy any part of this
message if you are not the intended recipient. 

Thank you.

 

  _  

From: owner-emc-p...@listserv.ieee.org 
mailto:owner-emc-p...@listserv.ieee.org] On Behalf Of Barker, Neil
Sent: Friday, November 19, 2004 8:20 AM
To: 'drcuthb...@micron.com'; israe...@oms-it.co.il; emc-p...@ieee.org
Subject: RE: Heart pacemakers - Vulnerability to EM

 

I am no expert in this subject, and some of what I am about to say is hearsay,
but here it is for what it's worth.

 

Safety levels are generally based not on field strength, but on power density.
However, these levels are set for the power absorption of the body (in W/kg)
based on the heating effect. For comparison purposes consider the acceptance
levels applied to SAR testing. On this basis, I suspect that Dave's amateur
rig probably doesn't meet SAR acceptance levels, and should probably have an
exclusion zone around the antenna!

 

The original message was concerned with the effect on a pacemaker rather than
directly on the body itself. Clearly, if you disrupt a pacemaker then it is
the pacemaker that has an effect on the body.

I believe that the susceptibility depends greatly on whether they are external
pacemakers or implanted pacemakers. The implanted devices are much less
susceptible as they are completely surrounded by soft body tissue, which is
largely water in any case. External devices are much more susceptible, due
largely to the relatively long electrode wire. being long, this is generally
susceptible at fairly low frequencies, and I have heard that they can be
affected by the magnetic loops used in store doorways for security
(merchandise protection) purposes. I have no knowledge of their susceptibility
to E fields.

 

Best regards, 

Neil R. Barker C.Eng. MIEE FSEE MIEEE 
Manager 
Compliance Engineering 
e2v technologies (uk) ltd 
106 Waterhouse Lane 
Chelmsford 
Essex 
CM1 2QU 
UK 

Tel: +44 (0)1245 453616 
Fax: +44 (0)1245 453410 
e-mail: neil.bar...@e2v.com 
Web: http://www.e2v.com <http://www.e2v.com/>  


From: drcuthb...@micron.com [mailto:drcuthb...@micron.com]
Sent: 19 November 2004 14:47
To: israe...@oms-it.co.il; emc-p...@ieee.org
Subject: RE: Heart pacemakers - Vulnerability to EM

I don't know what the standards call for but it is quite possible for a person
to be subjected to an E-field exceeding 500 V/m very near to an amateur radio
station. This is the field at my operating position when I am transmitting at
1.8 MHz. A cell phone held close to the body will cause a field, at the body,
of over 30 V/m.

 

   Dave Cuthbert 


From: owner-emc-p...@listserv.ieee.org 
mailto:owner-emc-p...@listserv.ieee.org] On Behalf Of Israel
Sent: Thursday, November 18, 2004 4:55 PM
To: emc-p...@ieee.org
Subject: Heart pacemakers - Vulnerability to EM

Hi all

 

Heart pacemakers may be vulnerable to Electro Magnetic energy.  Has anyone
idea about safe immunity levels ? about relevant standards regarding this
matter ?

 

Thanks !

 This message
is from the IEEE Product Safety Engineering Society emc-pstc discussion list.
Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to emc-p...@ieee.org 


Instructions: http://listserv.ieee.org/listserv/request/user-guide.html 


List rules: http://www.ieee-pses.org/listrules.html 


For help, send mail to the list administrators: 


Ron Pickard: emc-p...@hypercom.com Scott Douglas 

RE: Heart pacemakers - Vulnerability to EM

2004-11-19 Thread owner-emc-p...@listserv.ieee.org
I am no expert in this subject, and some of what I am about to say is hearsay,
but here it is for what it's worth.
 
Safety levels are generally based not on field strength, but on power density.
However, these levels are set for the power absorption of the body (in W/kg)
based on the heating effect. For comparison purposes consider the acceptance
levels applied to SAR testing. On this basis, I suspect that Dave's amateur
rig probably doesn't meet SAR acceptance levels, and should probably have an
exclusion zone around the antenna!
 
The original message was concerned with the effect on a pacemaker rather than
directly on the body itself. Clearly, if you disrupt a pacemaker then it is
the pacemaker that has an effect on the body.
I believe that the susceptibility depends greatly on whether they are external
pacemakers or implanted pacemakers. The implanted devices are much less
susceptible as they are completely surrounded by soft body tissue, which is
largely water in any case. External devices are much more susceptible, due
largely to the relatively long electrode wire. being long, this is generally
susceptible at fairly low frequencies, and I have heard that they can be
affected by the magnetic loops used in store doorways for security
(merchandise protection) purposes. I have no knowledge of their susceptibility
to E fields.
 
Best regards, 

Neil R. Barker C.Eng. MIEE FSEE MIEEE 
Manager 
Compliance Engineering 
e2v technologies (uk) ltd 
106 Waterhouse Lane 
Chelmsford 
Essex 
CM1 2QU 
UK 

Tel: +44 (0)1245 453616 
Fax: +44 (0)1245 453410 
e-mail: neil.bar...@e2v.com 
Web: http://www.e2v.com <http://www.e2v.com/>  


From: drcuthb...@micron.com [mailto:drcuthb...@micron.com]
Sent: 19 November 2004 14:47
To: israe...@oms-it.co.il; emc-p...@ieee.org
Subject: RE: Heart pacemakers - Vulnerability to EM


I don't know what the standards call for but it is quite possible for a person
to be subjected to an E-field exceeding 500 V/m very near to an amateur radio
station. This is the field at my operating position when I am transmitting at
1.8 MHz. A cell phone held close to the body will cause a field, at the body,
of over 30 V/m.
 
   Dave Cuthbert 

From: owner-emc-p...@listserv.ieee.org 
mailto:owner-emc-p...@listserv.ieee.org] On Behalf Of Israel
Sent: Thursday, November 18, 2004 4:55 PM
To: emc-p...@ieee.org
Subject: Heart pacemakers - Vulnerability to EM



Hi all

 

Heart pacemakers may be vulnerable to Electro Magnetic energy.  Has anyone
idea about safe immunity levels ? about relevant standards regarding this
matter ?

 

Thanks !

 This message
is from the IEEE Product Safety Engineering Society emc-pstc discussion list.
Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to emc-p...@ieee.org 


Instructions: http://listserv.ieee.org/listserv/request/user-guide.html 


List rules: http://www.ieee-pses.org/listrules.html 


For help, send mail to the list administrators: 


Ron Pickard: emc-p...@hypercom.com Scott Douglas emcp...@ptcnh.net 


For policy questions, send mail to: 


Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org 


All emc-pstc postings are archived and searchable on the web at: 


http://www.ieeecommunities.org/emc-pstc 




RE: Heart pacemakers - Vulnerability to EM

2004-11-19 Thread owner-emc-p...@listserv.ieee.org
I don't know what the standards call for but it is quite possible for a person
to be subjected to an E-field exceeding 500 V/m very near to an amateur radio
station. This is the field at my operating position when I am transmitting at
1.8 MHz. A cell phone held close to the body will cause a field, at the body,
of over 30 V/m.
 
   Dave Cuthbert 

From: owner-emc-p...@listserv.ieee.org 
mailto:owner-emc-p...@listserv.ieee.org] On Behalf Of Israel
Sent: Thursday, November 18, 2004 4:55 PM
To: emc-p...@ieee.org
Subject: Heart pacemakers - Vulnerability to EM



Hi all

 

Heart pacemakers may be vulnerable to Electro Magnetic energy.  Has anyone
idea about safe immunity levels ? about relevant standards regarding this
matter ?

 

Thanks !

 This message
is from the IEEE Product Safety Engineering Society emc-pstc discussion list.
Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to emc-p...@ieee.org 


Instructions: http://listserv.ieee.org/listserv/request/user-guide.html 


List rules: http://www.ieee-pses.org/listrules.html 


For help, send mail to the list administrators: 


Ron Pickard: emc-p...@hypercom.com Scott Douglas emcp...@ptcnh.net 


For policy questions, send mail to: 


Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org 


All emc-pstc postings are archived and searchable on the web at: 


http://www.ieeecommunities.org/emc-pstc 

 This message
is from the IEEE Product Safety Engineering Society emc-pstc discussion list.
Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to emc-p...@ieee.org 


Instructions: http://listserv.ieee.org/listserv/request/user-guide.html 


List rules: http://www.ieee-pses.org/listrules.html 


For help, send mail to the list administrators: 


Ron Pickard: emc-p...@hypercom.com Scott Douglas emcp...@ptcnh.net 


For policy questions, send mail to: 


Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org 


All emc-pstc postings are archived and searchable on the web at: 


http://www.ieeecommunities.org/emc-pstc