[Histonet] Ventana ultra

2011-04-13 Thread Barbara.Crill
Does anyone use the Benchmark Ultra - care to share your comments?

We are considering switching from the Benchmark XT to the Ultra but would like 
to hear from users about this instrument.
Is there an increase in reagent cost?

Can you really add more slides without adding time to the run?


ANTOINETTE CRILL, MBA,CT(ASCP)
TEAM LEADER ANATOMIC PATHOLOGY
DANVILLE REGIONAL MEDICAL CENTER
(O) 434.799.4470
(F) 434.773.6806
E-mail:  barbara.cr...@lpnt.netmailto:barbara.cr...@lpnt.net

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


RE: [Histonet] Ventana ultra

2011-04-13 Thread Sebree Linda A
Please Reply All as I'd like to hear feedback also.

Thanks. 


Linda A. Sebree
University of Wisconsin Hospital  Clinics
IHC/ISH Laboratory
DB1-223 VAH
600 Highland Ave.
Madison, WI 53792
(608)265-6596


-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of
barbara.cr...@lpnt.net
Sent: Wednesday, April 13, 2011 8:39 AM
To: histonet@lists.utsouthwestern.edu
Subject: [Histonet] Ventana ultra

Does anyone use the Benchmark Ultra - care to share your comments?

We are considering switching from the Benchmark XT to the Ultra but
would like to hear from users about this instrument.
Is there an increase in reagent cost?

Can you really add more slides without adding time to the run?


ANTOINETTE CRILL, MBA,CT(ASCP)
TEAM LEADER ANATOMIC PATHOLOGY
DANVILLE REGIONAL MEDICAL CENTER
(O) 434.799.4470
(F) 434.773.6806
E-mail:  barbara.cr...@lpnt.netmailto:barbara.cr...@lpnt.net

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Renal Catecholamines

2011-04-13 Thread Lynne Cates
Does anyone know any places that perform catecholamines on porcine renal
tissue and contact numbers?

 

Lynne 

 

 

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


RE: [Histonet] Ventana ultra

2011-04-13 Thread Kim Tournear
I've heard that adding slides and/or reagents can add as much as 8 minutes to 
the run time. Again, that's just what I've heard, don't know if it's fact or 
fiction

~Kim~   
OU ROCKS
~Don't be afraid your life will end, 
be afraid it will never begin~

--- On Wed, 4/13/11, Sebree Linda A lseb...@uwhealth.org wrote:


From: Sebree Linda A lseb...@uwhealth.org
Subject: RE: [Histonet] Ventana ultra
To: barbara.cr...@lpnt.net, histonet@lists.utsouthwestern.edu
Date: Wednesday, April 13, 2011, 2:09 PM


Please Reply All as I'd like to hear feedback also.

Thanks. 


Linda A. Sebree
University of Wisconsin Hospital  Clinics
IHC/ISH Laboratory
DB1-223 VAH
600 Highland Ave.
Madison, WI 53792
(608)265-6596


-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of
barbara.cr...@lpnt.net
Sent: Wednesday, April 13, 2011 8:39 AM
To: histonet@lists.utsouthwestern.edu
Subject: [Histonet] Ventana ultra

Does anyone use the Benchmark Ultra - care to share your comments?

We are considering switching from the Benchmark XT to the Ultra but
would like to hear from users about this instrument.
Is there an increase in reagent cost?

Can you really add more slides without adding time to the run?


ANTOINETTE CRILL, MBA,CT(ASCP)
TEAM LEADER ANATOMIC PATHOLOGY
DANVILLE REGIONAL MEDICAL CENTER
(O) 434.799.4470
(F) 434.773.6806
E-mail:  barbara.cr...@lpnt.netmailto:barbara.cr...@lpnt.net

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet




___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


Re: [Histonet] Ventana ultra

2011-04-13 Thread Angela Bitting
each of the 30 drawers essentially is it's own run. You don't have to wait 
for all 30 drawers to finish their runs before adding more slides.
There are 2 reagents that are exclusive to the Ultra and they are a little more 
expensive. Ultra CC1 and Ultra LCS,(and ultra CC2, if you use CC2). The EZ 
Prep, Reaction buffer and SSC are the same reagents that you use on your XT.
The software can be a little quirky sometimes. But, overall, we really like 
ours because of the continuous feed ability.
 
 
Angela Bitting, HT(ASCP), QIHC
Technical Specialist, Histology
Geisinger Medical Center 
100 N Academy Ave. MC 23-00
Danville, PA 17822
phone  570-214-9634
fax  570-271-5916 
 
No trees were hurt in the sending of this email
However many electrons were severly inconvienienced!

 barbara.cr...@lpnt.net 4/13/2011 9:39 AM 
Does anyone use the Benchmark Ultra - care to share your comments?

We are considering switching from the Benchmark XT to the Ultra but would like 
to hear from users about this instrument.
Is there an increase in reagent cost?

Can you really add more slides without adding time to the run?


ANTOINETTE CRILL, MBA,CT(ASCP)
TEAM LEADER ANATOMIC PATHOLOGY
DANVILLE REGIONAL MEDICAL CENTER
(O) 434.799.4470
(F) 434.773.6806
E-mail:  barbara.cr...@lpnt.netmailto:barbara.cr...@lpnt.net

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu 
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


IMPORTANT WARNING: The information in this message (and the documents attached 
to it, if any) is confidential and may be legally privileged. It is intended 
solely for the addressee. Access to this message by anyone else is 
unauthorized. If you are not the intended recipient, any disclosure, copying, 
distribution or any action taken, or omitted to be taken, in reliance on it is 
prohibited and may be unlawful. If you have received this message in error, 
please delete all electronic copies of this message (and the documents attached 
to it, if any), destroy any hard copies you may have created and notify me 
immediately by replying to this email. Thank you.BEGIN:VCARD
VERSION:2.1
X-GWTYPE:USER
FN:Bitting, Angela
TEL;WORK:570-271-6844
ORG:;Histology
EMAIL;WORK;PREF;NGW:akbitt...@geisinger.edu
N:Bitting;Angela
END:VCARD

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


AW: [Histonet] Benchmark Ultra question - follow up

2011-04-13 Thread Gudrun Lang
Thank you all for your replies. After all I have to admit, that the cause
was human failure. We found, that the protocols have been changed and noone
did remember. That was after a stopped run, where HIER war clicked off for
the repetition. And afterward only short HIER was clicked on again instead
of mild HIER. – stupid me!

 

Bye Gudrun

 

  _  

Von: Lynn Lee [mailto:lynnlee2...@live.com] 
Gesendet: Mittwoch, 13. April 2011 04:14
An: gu.l...@gmx.at
Betreff: FW: [Histonet] Benchmark Ultra question - follow up

 


One last thought- do you filter the new antibody solution you made  EVERY
time when refilling the prep kit? If not, there could be some particulates
in the diluent causing a blockage also.  Hope these suggestions help or
maybe you have already checked this out.
 
Lynn Lee



 

  _  

From: lynnlee2...@live.com
To: gu.l...@gmx.at
Subject: RE: [Histonet] Benchmark Ultra question - follow up
Date: Tue, 12 Apr 2011 20:10:43 -0600

It's possible the flip lid on the prep kit dispenser top is not completed
closed and seated properly. When the dispense hammer goes down, the pressure
may not be the same, every time if this is the case.  I always check the tip
of the dispenser for air bubbles before putting on the machine for EVERY one
whether it is a prefilled Ventana dispenser or my prep kit dispenser.
Finally something could be clogged in a line somewhere. 

I worked in R  D for Ventana for almost 4 years and have since used the XT
and Ultra in other labs and never had a problem with incorrect amount of
reagent being dispensed. I guess there's a first time for anything though. I
would call Ventana Customer Support. They are great!
 
Lynn Lee B.S., HT(ASCP)
Tucson, AZ

 
 From: gu.l...@gmx.at
 To: histonet@lists.utsouthwestern.edu
 Date: Tue, 12 Apr 2011 18:23:27 +0200
 Subject: [Histonet] Benchmark Ultra question - follow up
 
 Today I made a further experiment. I did two titration runs. On one slide
I
 added the working solution by pipette on the other slide I added it by
 pushing the filled PrepKit.
 Both stainings came out beautiful. - There has to be something wrong with
 the dispension-amount. Ventana is informed, I hope they find the culprit.
 
 Bye
 Gudrun
 
 -Ursprüngliche Nachricht-
 Von: histonet-boun...@lists.utsouthwestern.edu
 [mailto:histonet-boun...@lists.utsouthwestern.edu] Im Auftrag von Gudrun
 Lang
 Gesendet: Montag, 11. April 2011 18:48
 An: 'Angela Bitting'
 Cc: histonet@lists.utsouthwestern.edu
 Betreff: AW: [Histonet] Benchmark Ultra question
 
 Yes, that could be true. Some of the PrepKits go harder than others.
 Perhaps the pressure of the hammer is not big enough. Sometimes the
nozzles
 are not filled equally. (We press always reagens in the nozzle when we
 prepare the run.)
 
 With the Ventana-System there is an additional dilution, because the
 working-solution is added to a reaction-buffer film under the LCS. Perhaps
 it makes a difference if you dispense the solution with the pipett-tip
under
 the LCS or if the drop falls on the surface of the LCS. That could be a
 matter of the LCS-quality, or buffer-quantity , or . I think I'll become
 mad!
 
 Our pathologists have an uncertain unhappiness with the overall staining
 results, but most of the antibodies work well. Perhaps the majority isn't
 very sensitiv to small changes in the system and the quality-difference
 isn't big enough, but some are easier to kill.
 
 Regards, Gudrun
 
 
 
 
 
 _ 
 
 Von: Angela Bitting [mailto:akbitt...@geisinger.edu] 
 Gesendet: Montag, 11. April 2011 18:06
 An: gu.l...@gmx.at
 Betreff: Re: [Histonet] Benchmark Ultra question
 
 
 
 When I was trained to do titration runs on the BenchmarkXT and Ultras, I
was
 told to titrate 100ul of antibody. I have been suspecting for some time
now
 that the dispensers DON'T dispense a full 100ul. I haven't taken the time
to
 prove it, but your question may have motivated me.
 
 That could explain the weak staining. 
 
  Gudrun Lang gu.l...@gmx.at 4/11/2011 11:48 AM 
 Hi!
 
 I have some issues with our new Ultra. Since a few days antibodies, that
 work usually well, show up very faint. For example the TTF1 (Novocastra).
 
 I ordered a new bottle, made a titration and found, that the old titer
 1:50 was again well enough. I filled the old PrepKit - and the result was
 very weak.
 
 Then I thought, the PrepKit itself is the culprit. I changed the PrepKit -
 and again the result was very weak. At that time the working-solution was
 only three days old.
 
 
 
 Has anyone an explanation, why the titration works well and the automated
 dispension doesn't?
 
 
 
 Regards
 
 Gudrun Lang
 
 ___
 Histonet mailing list
 Histonet@lists.utsouthwestern.edu
 http://lists.utsouthwestern.edu/mailman/listinfo/histonet
 
 
 _ 
 
 IMPORTANT WARNING: The information in this message (and the documents
 attached to it, if any) is confidential and may be legally privileged. It
is
 intended solely for the 

RE: [Histonet] Ventana ultra

2011-04-13 Thread Kuhnla, Melissa
I currently have one XT and 3 Ultras.  I love the Ultras. There is no
difference in reagent costs.  The only thing that I can say remotely
related is that if you keep both platforms, the XT and the ultras have
separate CC1, CC2, and LCS. I am fairly certain the contents of the
bottle are the same, but the label and part numbers are different.
I love the continuous access. I have it set up where the three ultras
are 'top loaded' with set antibodies every day. This eliminates bouncing
things around all the time. Contact me if you would like anymore info. I
would be happy to help. I notice no delay in run times.
melissa

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of
barbara.cr...@lpnt.net
Sent: Wednesday, April 13, 2011 9:39 AM
To: histonet@lists.utsouthwestern.edu
Subject: [Histonet] Ventana ultra

Does anyone use the Benchmark Ultra - care to share your comments?

We are considering switching from the Benchmark XT to the Ultra but
would like to hear from users about this instrument.
Is there an increase in reagent cost?

Can you really add more slides without adding time to the run?


ANTOINETTE CRILL, MBA,CT(ASCP)
TEAM LEADER ANATOMIC PATHOLOGY
DANVILLE REGIONAL MEDICAL CENTER
(O) 434.799.4470
(F) 434.773.6806
E-mail:  barbara.cr...@lpnt.netmailto:barbara.cr...@lpnt.net

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


The information in this e-mail, and any attachments therein, is confidential 
and for use by the intended addressee only. If this message is received by you 
in error please do not disseminate or read further. Please reply to the sender 
that you have received the message in error, then delete the message. Although 
Catholic Health Services of Long Island attempts to sweep e-mail  and 
attachments for viruses, it does not guarantee that either are virus-free and 
accepts no liability for any damage sustained as a result of viruses. Thank you.



___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


Re: [Histonet] FW: 1 HE slide vs. 2

2011-04-13 Thread Rene J Buesa
There is not such a study. Cutting 2 vs. 1 HE/block is merely a preference 
issue amongst pathologists. Now, if you refer to TWO different levels, that is 
a different issue.
Two consecutive-serial sections from each block should not have a diagnostic 
impact BUT two different levels, separated by 20-25 sections, may.
That is a decision to take by the pathologist.
René J.

--- On Tue, 4/12/11, Eugenia Thomas eugenia90...@hotmail.com wrote:


From: Eugenia Thomas eugenia90...@hotmail.com
Subject: [Histonet] FW: 1 HE slide vs. 2
To: histonet@lists.utsouthwestern.edu
Date: Tuesday, April 12, 2011, 11:14 PM







From: eugenia90...@hotmail.com
To: histonet@lists.utsouthwestern.edu
Subject: 1 HE slide vs. 2
Date: Mon, 11 Apr 2011 13:06:59 -0400




Good afternoon everyone,

Does anyone know of an article or statistics discussing the impact (medical 
diagnosing) of cutting 2 HE slides per block verses 1 for all routine work?


Genia
              ___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet




___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


RE: [Histonet] Ventana ultra

2011-04-13 Thread Amber McKenzie
I recently got an Ultra about a month ago and doing test runs, I have
really enjoyed the continuous feed as well but I have a question...if I
deactivate a product from the Ultra's inventory list I notice it's still
in my XT's inventory list.  So, I have to deactivate everything off of
each instrument? 

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Kuhnla,
Melissa
Sent: Wednesday, April 13, 2011 9:45 AM
To: barbara.cr...@lpnt.net; histonet@lists.utsouthwestern.edu
Subject: RE: [Histonet] Ventana ultra

I currently have one XT and 3 Ultras.  I love the Ultras. There is no
difference in reagent costs.  The only thing that I can say remotely
related is that if you keep both platforms, the XT and the ultras have
separate CC1, CC2, and LCS. I am fairly certain the contents of the
bottle are the same, but the label and part numbers are different.
I love the continuous access. I have it set up where the three ultras
are 'top loaded' with set antibodies every day. This eliminates bouncing
things around all the time. Contact me if you would like anymore info. I
would be happy to help. I notice no delay in run times.
melissa

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of
barbara.cr...@lpnt.net
Sent: Wednesday, April 13, 2011 9:39 AM
To: histonet@lists.utsouthwestern.edu
Subject: [Histonet] Ventana ultra

Does anyone use the Benchmark Ultra - care to share your comments?

We are considering switching from the Benchmark XT to the Ultra but
would like to hear from users about this instrument.
Is there an increase in reagent cost?

Can you really add more slides without adding time to the run?


ANTOINETTE CRILL, MBA,CT(ASCP)
TEAM LEADER ANATOMIC PATHOLOGY
DANVILLE REGIONAL MEDICAL CENTER
(O) 434.799.4470
(F) 434.773.6806
E-mail:  barbara.cr...@lpnt.netmailto:barbara.cr...@lpnt.net

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


The information in this e-mail, and any attachments therein, is
confidential and for use by the intended addressee only. If this message
is received by you in error please do not disseminate or read further.
Please reply to the sender that you have received the message in error,
then delete the message. Although Catholic Health Services of Long
Island attempts to sweep e-mail  and attachments for viruses, it does
not guarantee that either are virus-free and accepts no liability for
any damage sustained as a result of viruses. Thank you.



___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


RE: [Histonet] Ventana ultra

2011-04-13 Thread Setlak, Lisa
We like ours as well. It's not as continual as we originally thought to 
maximize its potential I would recommend placing frequently used antibodies on 
the wheel with the first run. As long as you have the antibody you want to 
stain for on the wheel you can add the slide anytime you want and it will run 
it. If the antibody is not on the wheel you have to set a landing zone in 
order to add it. This usually works out fine, however once the run that is 
already going reaches a certain point, you aren't able to set a landing zone 
and may have to wait awhile to load new slides. It is pretty cool that you can 
run FITC, FISH, and IHC at the same time (however we tend to put our EBER on 
overnight because it's a fairly long run).
Lisa

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Angela Bitting
Sent: Wednesday, April 13, 2011 9:42 AM
To: histonet@lists.utsouthwestern.edu; barbara.cr...@lpnt.net
Subject: Re: [Histonet] Ventana ultra

each of the 30 drawers essentially is it's own run. You don't have to wait 
for all 30 drawers to finish their runs before adding more slides.
There are 2 reagents that are exclusive to the Ultra and they are a little more 
expensive. Ultra CC1 and Ultra LCS,(and ultra CC2, if you use CC2). The EZ 
Prep, Reaction buffer and SSC are the same reagents that you use on your XT.
The software can be a little quirky sometimes. But, overall, we really like 
ours because of the continuous feed ability.
 
 
Angela Bitting, HT(ASCP), QIHC
Technical Specialist, Histology
Geisinger Medical Center 
100 N Academy Ave. MC 23-00
Danville, PA 17822
phone  570-214-9634
fax  570-271-5916 
 
No trees were hurt in the sending of this email
However many electrons were severly inconvienienced!

 barbara.cr...@lpnt.net 4/13/2011 9:39 AM 
Does anyone use the Benchmark Ultra - care to share your comments?

We are considering switching from the Benchmark XT to the Ultra but would like 
to hear from users about this instrument.
Is there an increase in reagent cost?

Can you really add more slides without adding time to the run?


ANTOINETTE CRILL, MBA,CT(ASCP)
TEAM LEADER ANATOMIC PATHOLOGY
DANVILLE REGIONAL MEDICAL CENTER
(O) 434.799.4470
(F) 434.773.6806
E-mail:  barbara.cr...@lpnt.netmailto:barbara.cr...@lpnt.net

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu 
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


IMPORTANT WARNING: The information in this message (and the documents attached 
to it, if any) is confidential and may be legally privileged. It is intended 
solely for the addressee. Access to this message by anyone else is 
unauthorized. If you are not the intended recipient, any disclosure, copying, 
distribution or any action taken, or omitted to be taken, in reliance on it is 
prohibited and may be unlawful. If you have received this message in error, 
please delete all electronic copies of this message (and the documents attached 
to it, if any), destroy any hard copies you may have created and notify me 
immediately by replying to this email. Thank you.

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


AW: [Histonet] Ventana ultra

2011-04-13 Thread Gudrun Lang
What we like about the Ultra is, that there's no need for cleaning runs
between the runs and that you can take out the bulk and waste containers
during the run.
We don't use the continious loading, because it doesn't fit to our workplan
and because we use various antibodies from run to run.
The runs are faster, but more than 2 runs per 8 hour shift and 1 run over
night are not possible for us. That's the same workload we did with the XT.

There are windows during the run, when you can add new slides and reagens.
If the needed reagens in on the instrument, you can add them any time, if
there's enough place. But the run will last longer and you have to wait for
a window of the longer run to get the reagens off the instrument. Ready
stained slides can be taken out independently. But the free pads cannot be
filled and runs started, until you have the possibility to add the right
reagens. So we decided to stay with patch-workload and real go-away.

Bye
Gudrun

-Ursprüngliche Nachricht-
Von: histonet-boun...@lists.utsouthwestern.edu
[mailto:histonet-boun...@lists.utsouthwestern.edu] Im Auftrag von
barbara.cr...@lpnt.net
Gesendet: Mittwoch, 13. April 2011 15:39
An: histonet@lists.utsouthwestern.edu
Betreff: [Histonet] Ventana ultra

Does anyone use the Benchmark Ultra - care to share your comments?

We are considering switching from the Benchmark XT to the Ultra but would
like to hear from users about this instrument.
Is there an increase in reagent cost?

Can you really add more slides without adding time to the run?


ANTOINETTE CRILL, MBA,CT(ASCP)
TEAM LEADER ANATOMIC PATHOLOGY
DANVILLE REGIONAL MEDICAL CENTER
(O) 434.799.4470
(F) 434.773.6806
E-mail:  barbara.cr...@lpnt.netmailto:barbara.cr...@lpnt.net

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] RE: Histonet Digest, Von Kossa Stain

2011-04-13 Thread Mayer,Toysha N
We used the lamp at our cutting stations, coplin jar in our waterbath or on top 
of the embedder, wrapped in tin foil for 1 hr.

Toysha N. Mayer, MBA, HT (ASCP)
Education Coordinator
Program in Histotechnology
School of Health Professions
MD Anderson Cancer Center
(713) 563-3481
tnma...@mdanderson.org




What is everyone using for their light when developing the silver in the 
VonKossa stain when you have no sunlight to use?  We used to use a 60 watt 
lamp, but haven't done one for years and am bringing this stain back to our 
repetiore due to pathologist request.  Thanks much!

Dorothy Webb, HT (ASCP)
Regions Histology Technical Supervisor
651-254-2962




 







--




___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] IHC platforms/Sales Representatives

2011-04-13 Thread Paula Lucas
Hello,

 

We are inquiring about IHC staining platforms and we've already contacted
the Leica sales rep for the Bond.  I would also like to contact Ventana to
get a quote on their system, and I've tried to get in contact with a sales
rep in our area but I haven't had anyone returning my calls.  If there is a
Vantana rep reading this, would you please contact the sales representative
for my area and let him/her know we are interested in a quote?  We are
located in Orange County (Fountain Valley), California.  

 

For those who have experience with these two platforms, would you mind
taking the time to share with me your views and experiences about the two
systems?

 

Thanks so much,

Paula Lucas

Lab Manager

Bio-Path Medical Group

Fountain Valley, CA

714.433.1330

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


Re: AW: [Histonet] Benchmark Ultra question - follow up

2011-04-13 Thread Angela Bitting
Augh! I hate when that happens!

 Gudrun Lang gu.l...@gmx.at 4/13/2011 10:44 AM 
Thank you all for your replies. After all I have to admit, that the
cause
was human failure. We found, that the protocols have been changed and
noone
did remember. That was after a stopped run, where HIER war clicked off
for
the repetition. And afterward only short HIER was clicked on again
instead
of mild HIER. – stupid me!



Bye Gudrun



  _  

Von: Lynn Lee [mailto:lynnlee2...@live.com] 
Gesendet: Mittwoch, 13. April 2011 04:14
An: gu.l...@gmx.at 
Betreff: FW: [Histonet] Benchmark Ultra question - follow up




One last thought- do you filter the new antibody solution you made 
EVERY
time when refilling the prep kit? If not, there could be some
particulates
in the diluent causing a blockage also.  Hope these suggestions help
or
maybe you have already checked this out.

Lynn Lee





  _  

From: lynnlee2...@live.com 
To: gu.l...@gmx.at 
Subject: RE: [Histonet] Benchmark Ultra question - follow up
Date: Tue, 12 Apr 2011 20:10:43 -0600

It's possible the flip lid on the prep kit dispenser top is not
completed
closed and seated properly. When the dispense hammer goes down, the
pressure
may not be the same, every time if this is the case.  I always check
the tip
of the dispenser for air bubbles before putting on the machine for
EVERY one
whether it is a prefilled Ventana dispenser or my prep kit dispenser.
Finally something could be clogged in a line somewhere. 

I worked in R  D for Ventana for almost 4 years and have since used
the XT
and Ultra in other labs and never had a problem with incorrect amount
of
reagent being dispensed. I guess there's a first time for anything
though. I
would call Ventana Customer Support. They are great!

Lynn Lee B.S., HT(ASCP)
Tucson, AZ


 From: gu.l...@gmx.at 
 To: histonet@lists.utsouthwestern.edu 
 Date: Tue, 12 Apr 2011 18:23:27 +0200
 Subject: [Histonet] Benchmark Ultra question - follow up
 
 Today I made a further experiment. I did two titration runs. On one
slide
I
 added the working solution by pipette on the other slide I added it
by
 pushing the filled PrepKit.
 Both stainings came out beautiful. - There has to be something wrong
with
 the dispension-amount. Ventana is informed, I hope they find the
culprit.
 
 Bye
 Gudrun
 
 -Ursprüngliche Nachricht-
 Von: histonet-boun...@lists.utsouthwestern.edu 
 [mailto:histonet-boun...@lists.utsouthwestern.edu] Im Auftrag von
Gudrun
 Lang
 Gesendet: Montag, 11. April 2011 18:48
 An: 'Angela Bitting'
 Cc: histonet@lists.utsouthwestern.edu 
 Betreff: AW: [Histonet] Benchmark Ultra question
 
 Yes, that could be true. Some of the PrepKits go harder than
others.
 Perhaps the pressure of the hammer is not big enough. Sometimes the
nozzles
 are not filled equally. (We press always reagens in the nozzle when
we
 prepare the run.)
 
 With the Ventana-System there is an additional dilution, because the
 working-solution is added to a reaction-buffer film under the LCS.
Perhaps
 it makes a difference if you dispense the solution with the
pipett-tip
under
 the LCS or if the drop falls on the surface of the LCS. That could be
a
 matter of the LCS-quality, or buffer-quantity , or . I think I'll
become
 mad!
 
 Our pathologists have an uncertain unhappiness with the overall
staining
 results, but most of the antibodies work well. Perhaps the majority
isn't
 very sensitiv to small changes in the system and the
quality-difference
 isn't big enough, but some are easier to kill.
 
 Regards, Gudrun
 
 
 
 
 
 _ 
 
 Von: Angela Bitting [mailto:akbitt...@geisinger.edu] 
 Gesendet: Montag, 11. April 2011 18:06
 An: gu.l...@gmx.at 
 Betreff: Re: [Histonet] Benchmark Ultra question
 
 
 
 When I was trained to do titration runs on the BenchmarkXT and
Ultras, I
was
 told to titrate 100ul of antibody. I have been suspecting for some
time
now
 that the dispensers DON'T dispense a full 100ul. I haven't taken the
time
to
 prove it, but your question may have motivated me.
 
 That could explain the weak staining. 
 
  Gudrun Lang gu.l...@gmx.at 4/11/2011 11:48 AM 
 Hi!
 
 I have some issues with our new Ultra. Since a few days antibodies,
that
 work usually well, show up very faint. For example the TTF1
(Novocastra).
 
 I ordered a new bottle, made a titration and found, that the old
titer
 1:50 was again well enough. I filled the old PrepKit - and the result
was
 very weak.
 
 Then I thought, the PrepKit itself is the culprit. I changed the
PrepKit -
 and again the result was very weak. At that time the working-solution
was
 only three days old.
 
 
 
 Has anyone an explanation, why the titration works well and the
automated
 dispension doesn't?
 
 
 
 Regards
 
 Gudrun Lang
 
 ___
 Histonet mailing list
 Histonet@lists.utsouthwestern.edu 
 http://lists.utsouthwestern.edu/mailman/listinfo/histonet 
 
 
 _ 
 
 IMPORTANT WARNING: The information in this message (and the
documents
 attached to it, if any) 

RE: [Histonet] Ventana ultra

2011-04-13 Thread Angela Bitting
We load our Ultras with the most commonly used antibodies each morning. So that 
minimizes the number of landing zones we need to use.
Also, (if you run both XTs and Ultras)  run your antibodies with the longest 
protocols on Ultra and that will shorten your XT runs.
 
 Setlak, Lisa lset...@childrensmemorial.org 4/13/2011 10:54 AM 
We like ours as well. It's not as continual as we originally thought to 
maximize its potential I would recommend placing frequently used antibodies on 
the wheel with the first run. As long as you have the antibody you want to 
stain for on the wheel you can add the slide anytime you want and it will run 
it. If the antibody is not on the wheel you have to set a landing zone in 
order to add it. This usually works out fine, however once the run that is 
already going reaches a certain point, you aren't able to set a landing zone 
and may have to wait awhile to load new slides. It is pretty cool that you can 
run FITC, FISH, and IHC at the same time (however we tend to put our EBER on 
overnight because it's a fairly long run).
Lisa

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Angela Bitting
Sent: Wednesday, April 13, 2011 9:42 AM
To: histonet@lists.utsouthwestern.edu; barbara.cr...@lpnt.net 
Subject: Re: [Histonet] Ventana ultra

each of the 30 drawers essentially is it's own run. You don't have to wait 
for all 30 drawers to finish their runs before adding more slides.
There are 2 reagents that are exclusive to the Ultra and they are a little more 
expensive. Ultra CC1 and Ultra LCS,(and ultra CC2, if you use CC2). The EZ 
Prep, Reaction buffer and SSC are the same reagents that you use on your XT.
The software can be a little quirky sometimes. But, overall, we really like 
ours because of the continuous feed ability.


Angela Bitting, HT(ASCP), QIHC
Technical Specialist, Histology
Geisinger Medical Center 
100 N Academy Ave. MC 23-00
Danville, PA 17822
phone  570-214-9634
fax  570-271-5916 

No trees were hurt in the sending of this email
However many electrons were severly inconvienienced!

 barbara.cr...@lpnt.net 4/13/2011 9:39 AM 
Does anyone use the Benchmark Ultra - care to share your comments?

We are considering switching from the Benchmark XT to the Ultra but would like 
to hear from users about this instrument.
Is there an increase in reagent cost?

Can you really add more slides without adding time to the run?


ANTOINETTE CRILL, MBA,CT(ASCP)
TEAM LEADER ANATOMIC PATHOLOGY
DANVILLE REGIONAL MEDICAL CENTER
(O) 434.799.4470
(F) 434.773.6806
E-mail:  barbara.cr...@lpnt.netmailto:barbara.cr...@lpnt.net

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu 
http://lists.utsouthwestern.edu/mailman/listinfo/histonet 


IMPORTANT WARNING: The information in this message (and the documents attached 
to it, if any) is confidential and may be legally privileged. It is intended 
solely for the addressee. Access to this message by anyone else is 
unauthorized. If you are not the intended recipient, any disclosure, copying, 
distribution or any action taken, or omitted to be taken, in reliance on it is 
prohibited and may be unlawful. If you have received this message in error, 
please delete all electronic copies of this message (and the documents attached 
to it, if any), destroy any hard copies you may have created and notify me 
immediately by replying to this email. Thank you.




-
This message was secured by ZixCorp(R).
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] slide brite and coverslip drying time

2011-04-13 Thread Dawn Herron
Hello all. We have just made the transition from xylene in our linear
stainer to Slide Brite. We are also coverslipping from Slide Brite with
Permount. So far the slides look good (though we have to be careful about
water contamination in the dehydrating alcohols) but we are having a problem
with extremely long coverslipping drying time. Some of the slides have been
on the warmer for over 18 hours and you can still move the coverslips!
(Normally the slides would dry within 4 hours on the warmer with xylene and
acrymount) Any suggestions on how to speed drying time?

Thanks,
Dawn
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Tissue transfer

2011-04-13 Thread Bartlett, Jeanine (CDC/OID/NCEZID)
Hello all,

We have some unstained slides we received from outside the U.S. on 
non-charged/un-coated slides. We would like to transfer these sections to 
charged slides before proceeding with testing.  Any suggestions?

Thanks!

Jeanine Bartlett, BS, HT(ASCP)QIHC
Centers for Disease Control and Prevention
Infectious Diseases Pathology Branch
1600 Clifton Road, MS/G-32
18/SB-114
Atlanta, GA  30333
(404) 639-3590
jeanine.bartl...@cdc.hhs.gov



___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] General Data Cassette Labeler

2011-04-13 Thread Robin Negron
 

Can anyone that uses the General Data cassette labeler
give me information regarding your overall opinion of this instrument. 

 

 

1.   I would like to have any information regarding the speed of
machine. When we did a demo of this equipment it seems fast.

2.I was also questioning the cassettes, we have found that the
only company to purchase these from is general data (at a high cost). 

3.   Has anyone had the equipment break down and how is the
dependability and  the customer support service.

4.Our company is very interested in the on demand version. How
has anyone found this to be an advantage.  

5.   Does anyone purchase the cassettes from another dealer besides
General Data

6.   Has anyone had problems with the black scatching off.

7.We currently use TBS and they are very slow and constantly
break down this does not meet our demand as we currently print approx.
400-500 per day.

 

 

I would like to hear your opinions.

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] RE: How long should autopsy/necropsy tissue be in fixation for a suspected TB case?

2011-04-13 Thread Thurby, Christina
Hi,
Do anyone have references for how long should autopsy/necropsy tissue be in 
fixation for a suspected TB case?  Is 72-96 hours fixation sufficient or should 
the time be extended to 5-7 days.
Thanks,
Christina Thurby
Bristol-Myers Squibb Company
Research Scientist I
812-307-2093 (tie 625)




This message (including any attachments) may contain confidential, proprietary, 
privileged and/or private information.  The information is intended to be for 
the use of the individual or entity designated above.  If you are not the 
intended recipient of this message, please notify the sender immediately, and 
delete the message and any attachments.  Any disclosure, reproduction, 
distribution or other use of this message or any attachments by an individual 
or entity other than the intended recipient is prohibited.

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


Re: [Histonet] Tissue transfer

2011-04-13 Thread Diane . Craft
I use Harleco Krystalon liquid coverslip medium. Cover the tissue, laying 
horizontally in oven for a few hours until it hardens, than soak in water 
bath, peel, and transfer to charged slide with same side facing down, dry 
in oven again, dissolve in xylene when ready for staining. 

Diane Craft
Pathology Department
Animal Medical Center
510 East 62nd St
New York NY 10065-8314
212-329-8675 (phone)
212-759-5878 (fax)



Bartlett, Jeanine (CDC/OID/NCEZID) j...@cdc.gov 
Sent by: histonet-boun...@lists.utsouthwestern.edu
04/13/2011 01:06 PM

To
histonet@lists.utsouthwestern.edu histonet@lists.utsouthwestern.edu
cc

Subject
[Histonet] Tissue transfer





Hello all,

We have some unstained slides we received from outside the U.S. on 
non-charged/un-coated slides. We would like to transfer these sections to 
charged slides before proceeding with testing.  Any suggestions?

Thanks!

Jeanine Bartlett, BS, HT(ASCP)QIHC
Centers for Disease Control and Prevention
Infectious Diseases Pathology Branch
1600 Clifton Road, MS/G-32
18/SB-114
Atlanta, GA  30333
(404) 639-3590
jeanine.bartl...@cdc.hhs.gov



___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


Re: [Histonet] RE: How long should autopsy/necropsy tissue be in fixation for a suspected TB case?

2011-04-13 Thread Jay Lundgren
 As long as possible.  I remember reading somewhere about TB in lung
tissue that was still viable after *years* of 10% NBF.  Make sure the lungs
are thoroughly perfused, or they will still be unfixed in the middle.  There
is a technique to inflate lungs at autopsy using a large syringe full of
NBF and a clamp.  Make sure that the specimen is not contacting the walls or
bottom of the specimen container.  You can suspend the whole organ with
string inside the container, if needed. One can also use gauze or paper
towels to keep the organ under the surface of the fixative.  Lung tissue is
notoriously difficult to fix properly, due both to the high lipid content of
the pleural parenchyma, and the fact that fresh lungs float.

Sincerely,

  Jay A.
Lundgren, M.S., HTL (ASCP)
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


Re: [Histonet] slide brite and coverslip drying time

2011-04-13 Thread Rene J Buesa
That is exactly one of the problems with Slide Brite, and is the trade-off 
you have to pay.
The alternative (although it may seem odd) is to get away with alcohols and 
xylene altogether.
After staining the slides, wash them in distilled water, shake them well and 
place them in an over at 60ºC for 5 minutes. After drying apply the mounting 
medium more diluted (consistency of light mineral oil), and coverslip.
You will get away with alcohols, and xylene (or its substitutes).
Do not frown, just try it!
René J.

--- On Wed, 4/13/11, Dawn Herron dwenze...@gmail.com wrote:


From: Dawn Herron dwenze...@gmail.com
Subject: [Histonet] slide brite and coverslip drying time
To: Histonet@lists.utsouthwestern.edu
Date: Wednesday, April 13, 2011, 12:16 PM


Hello all. We have just made the transition from xylene in our linear
stainer to Slide Brite. We are also coverslipping from Slide Brite with
Permount. So far the slides look good (though we have to be careful about
water contamination in the dehydrating alcohols) but we are having a problem
with extremely long coverslipping drying time. Some of the slides have been
on the warmer for over 18 hours and you can still move the coverslips!
(Normally the slides would dry within 4 hours on the warmer with xylene and
acrymount) Any suggestions on how to speed drying time?

Thanks,
Dawn
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet




___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


Re: [Histonet] RE: How long should autopsy/necropsy tissue be in fixation for a suspected TB case?

2011-04-13 Thread Rene J Buesa
5 days (96 h) is more than enough, but always handle the tissue and 
blocks using safety precautions.
René J.

--- On Wed, 4/13/11, Thurby, Christina christina.thu...@bms.com wrote:


From: Thurby, Christina christina.thu...@bms.com
Subject: [Histonet] RE: How long should autopsy/necropsy tissue be in fixation 
for a suspected TB case?
To: histonet@lists.utsouthwestern.edu histonet@lists.utsouthwestern.edu
Date: Wednesday, April 13, 2011, 2:04 PM


Hi,
Do anyone have references for how long should autopsy/necropsy tissue be in 
fixation for a suspected TB case?  Is 72-96 hours fixation sufficient or should 
the time be extended to 5-7 days.
Thanks,
Christina Thurby
Bristol-Myers Squibb Company
Research Scientist I
812-307-2093 (tie 625)




This message (including any attachments) may contain confidential, proprietary, 
privileged and/or private information.  The information is intended to be for 
the use of the individual or entity designated above.  If you are not the 
intended recipient of this message, please notify the sender immediately, and 
delete the message and any attachments.  Any disclosure, reproduction, 
distribution or other use of this message or any attachments by an individual 
or entity other than the intended recipient is prohibited.

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet




___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] DAB staining for Liver frozen sections

2011-04-13 Thread Nasreen Bashir
Hi,
   I need some information on how to block endogenous peroxidase effectively
in liver frozen sections, as I need to do DAB staining.
   What blocking buffer will work best for blocking in liver frozen
sections.
   I appreciate any information on this issue.

  Thanks,
  Reena
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


RE: [Histonet] RE: How long should autopsy/necropsy tissue be in fixation for a suspected TB case?

2011-04-13 Thread Kumar, Devender
Hi, 

We fix 'Category-A' agent infected murine lungs in 10% buffered formalin for 7 
days in ABSL-3. We also confirm tissue sterility by plating the tissue 
homogenates on suitable media before taking the samples out of ABSL-3 facility 
for sectioning. In your case 7 days should be enough, but it is always better 
to confirm the sterility of tissue.

Thanks,
Devender


Devender Kumar, D.V.M., Ph.D.
Postdoctorate fellow,
Center for Immunology  Microbial Disease,
Albany Medical College,
47 New Scotland Avenue, MC151
Albany, NY 12208
518.262.6220(LAB)
--


-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Rene J Buesa
Sent: Wednesday, April 13, 2011 3:57 PM
To: histonet@lists.utsouthwestern.edu; ChristinaThurby
Subject: Re: [Histonet] RE: How long should autopsy/necropsy tissue be in 
fixation for a suspected TB case?

5 days (96 h) is more than enough, but always handle the tissue and 
blocks using safety precautions.
René J.

--- On Wed, 4/13/11, Thurby, Christina christina.thu...@bms.com wrote:


From: Thurby, Christina christina.thu...@bms.com
Subject: [Histonet] RE: How long should autopsy/necropsy tissue be in fixation 
for a suspected TB case?
To: histonet@lists.utsouthwestern.edu histonet@lists.utsouthwestern.edu
Date: Wednesday, April 13, 2011, 2:04 PM


Hi,
Do anyone have references for how long should autopsy/necropsy tissue be in 
fixation for a suspected TB case?  Is 72-96 hours fixation sufficient or should 
the time be extended to 5-7 days.
Thanks,
Christina Thurby
Bristol-Myers Squibb Company
Research Scientist I
812-307-2093 (tie 625)




This message (including any attachments) may contain confidential, proprietary, 
privileged and/or private information.  The information is intended to be for 
the use of the individual or entity designated above.  If you are not the 
intended recipient of this message, please notify the sender immediately, and 
delete the message and any attachments.  Any disclosure, reproduction, 
distribution or other use of this message or any attachments by an individual 
or entity other than the intended recipient is prohibited.

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet



  
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

-
CONFIDENTIALITY NOTICE: This email and any attachments may contain
confidential information that is protected by law and is for the
sole use of the individuals or entities to which it is addressed.
If you are not the intended recipient, please notify the sender by
replying to this email and destroying all copies of the
communication and attachments. Further use, disclosure, copying,
distribution of, or reliance upon the contents of this email and
attachments is strictly prohibited. To contact Albany Medical
Center, or for a copy of our privacy practices, please visit us on
the Internet at www.amc.edu.

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


Re: [Histonet] Tissue transfer

2011-04-13 Thread Rene J Buesa
Mission impossible (successfully, you can always ruin them)! 
It is better to do the staining after a prolonged oven heating and do it 
delicately.
René J.

--- On Wed, 4/13/11, Bartlett, Jeanine (CDC/OID/NCEZID) j...@cdc.gov wrote:


From: Bartlett, Jeanine (CDC/OID/NCEZID) j...@cdc.gov
Subject: [Histonet] Tissue transfer
To: histonet@lists.utsouthwestern.edu histonet@lists.utsouthwestern.edu
Date: Wednesday, April 13, 2011, 1:06 PM


Hello all,

We have some unstained slides we received from outside the U.S. on 
non-charged/un-coated slides. We would like to transfer these sections to 
charged slides before proceeding with testing.  Any suggestions?

Thanks!

Jeanine Bartlett, BS, HT(ASCP)QIHC
Centers for Disease Control and Prevention
Infectious Diseases Pathology Branch
1600 Clifton Road, MS/G-32
18/SB-114
Atlanta, GA  30333
(404) 639-3590
jeanine.bartl...@cdc.hhs.gov



___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet




___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Part-time Histotech Opening

2011-04-13 Thread Norm Burnham
Dear Histonetters, 

 

We have the following opening:

 

 

HISTOTECHNOLOGIST

 

ProPath, a high volume, pathology practice, located in Dallas, Texas, has an
immediate opening for a part-time Histotechnologist.  Responsibilities
include embedding tissue specimens, microtomy of paraffin-embedded tissue,
operation of automated stainer and coverslipper, equipment maintenance and
record retention.  

 

The ideal candidate will have a high school diploma or equivalent.  We
prefer, HT, HTL (ASCP) registered or eligible.

 

The hours are 2:00 p.m. to 6:00 p.m. Monday - Friday.

 

For consideration send resume to:

ProPath, Human Resources

1355 River Bend Drive

Dallas, TX 75247 

FAX: 214/237-1825

Job-Line: 214/237-1775

Email address: j...@propath.com 

Website: www.propath.com http://www.propath.com/ 

 

Don't Follow the Leader!  Join the Leader!

 

EOE

 

 

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


Re: [Histonet] Update regarding question on Plastic Embedding

2011-04-13 Thread Joseph Saby
Mahesh-

If you have access to a sonicator, you can etch the slides by first soaking 
them 
in 1% formic acid in the sonicator for 5-10 minutes, then after a brief water 
rinse soak them in 50% ethanol for another 5-10 minutes in the sonicator.

You may have to work with the staining times, but you will find that many of 
your paraffin embedding stains will work.

Good luck!

Joe Saby, BA HT
NAMSA




From: Mahesh Polavarapu polavarapu.mah...@gmail.com
To: histonet histonet@lists.utsouthwestern.edu
Sent: Tue, April 12, 2011 8:25:30 PM
Subject: [Histonet] Update regarding question on Plastic Embedding

Looking for a protocol to visualize vascularization and collagen deposition
at the bone-tendon interface of a rabbit rotator cuff embedded in a plastic
system. Sections will be rather thick (~50um) b/c they are being made
through a titanium anchor. Using MMA with a cold-curing resin, Technovit
9100. Thanks in advance!

- Mahesh
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet



  
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Seeking techs all over the place!

2011-04-13 Thread Cheryl

Help!  
 
We're getting busy again (the economy rebounds??) and need both permanent and 
temporary histotechs. 
 
Are you looking for...
...a new place to live?
...a bigger city?
...a smaller city?
...a promotion?
...a raise?
...a place to learn new things?
 
We will actively consider...
...new grads (and almost grads)
...experienced bench techs
...bench techs looking for a step up
...experienced management
...retirees (we know you've still 'got it')
 
We have a number of permanent openings and seeking to successfully fill 9 
thirteen-week temp positions.  We are working histotechs so the conversation is 
easy and usually kinda fun.
Send a resume or ask for an application - 
tkngfl...@yahoo.com
Fax/Phone: 800.756.3309
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Histology Supervisor ( Hospital Scientist)-F/T

2011-04-13 Thread Fawaz Zouabi
A Histology Supervisor ( Hospital Scientist) -F/T  position is currently 
available at The Department of Forensic Medicine (Sydney local health network / 
Sydney south West), GLEBE 2037 NSW AUSTRALIA.
To enquire please check the NSW health web site on 
http://www.health.nsw.gov.au/jobs/index.asp
The successful applicant will be responsible of the supervision of the DOFM 
histology laboratory which involves the collection and processing of forensic 
histological specimens. 
 
mailto:fawaz.zou...@sswahs.nsw.gov.au  
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet