Jesus,

The formic acid (5%) will decalcify the hydroxyapatite which can be very
hard and crunchy to cut but since you ground the section I guess you don't
care about that.  I would be concerned though that the formic acid might
also cause you to lose staining of some cell components that depend on the
calcium.  The methyl methacrylate can be removed from the section by xylene
like deparaffinizing paraffin sections.  It works for methyl methacrylate
but nothing removes glycol methacrylate.

Regards,

Patsy 

Patsy Ruegg, HT(ASCP)QIHC
IHCtech
12635 Montview Blvd. Ste.215
Aurora, CO 80045
720-859-4060
fax 720-859-4110
www.ihctech.net 
www.ihcrg.org


-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Jesus
Hernandez
Sent: Friday, November 04, 2011 3:41 PM
To: histonet@lists.utsouthwestern.edu
Subject: [Histonet] Formic Acid Recipe

Dear all,

I was having trouble staining Human embryonic palatal mesenchymal cells with
multi-stain solution. My PI told me to use formic acid on the samples so
that it could increase the permeability of the methyl methacrylate. I am not
sure if this is enough information, but the cells were loaded on
hydroxyapatite scaffolds. I have also tried aniline blue and villaneuva
stain. Any information on how I can make a solution of formic acid is
appreciated along with maybe other stains I could try using. Each sample was
grinded to about 50 microns. Thank you.

Best Regards,

 Jesus W. Hernandez
Graduate Student
Department of Biomedical Engineering
University of Texas at San Antonio
jesus.w@gmail.com



___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


RE: [Histonet] low profile vs. hi profile blades

2011-11-06 Thread Patsy Ruegg
It depends on the microtome being used and how the blade holder is set up to
take high profile or low, many of them (at least the microms I have can take
either) on my microm to use low profile I leave the bar on the knife holder
so the holder is set to hold shorter low profile blades, I can take that bar
out with a couple of little screws on each side which deepens the knife
holder space so it can take high profile knives.

One of my cryostats knife holder only takes high profile blades so I like to
use Sturkey gold high profile there, the other cryostat I have only takes
low profile blades.



Patsy Ruegg, HT(ASCP)QIHC
IHCtech
12635 Montview Blvd. Ste.215
Aurora, CO 80045
720-859-4060
fax 720-859-4110
www.ihctech.net 
www.ihcrg.org


-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Nancy
Schmitt
Sent: Friday, November 04, 2011 7:50 AM
To: histonet@lists.utsouthwestern.edu
Subject: [Histonet] low profile vs. hi profile blades

Hi-
Please educate me on this.  There seems to be a lot of personal opinion
involved in regards to which type of blade to use.
Thanks for your help
Nancy



NOTICE: This email may contain legally privileged information. The
information
is for the use of only the intended recipient(s) even if addressed
incorrectly. If you are not the intended recipient, please notify the sender
that you have received it in error and then delete it along with any
attachments. Thank you.



___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


RE: [Histonet] low profile vs. hi profile blades

2011-11-06 Thread Patsy Ruegg
In my experience even with the harder high profile blades for denser tissues
such as bone I still prefer to use permanent tungsten carbide blades I have
to send out to get sharpened.  The high profile disposable blades seem to
vibrate too much for me even the heavier tungsten ones for bone.

Patsy Ruegg, HT(ASCP)QIHC
IHCtech
12635 Montview Blvd. Ste.215
Aurora, CO 80045
720-859-4060
fax 720-859-4110
www.ihctech.net 
www.ihcrg.org


-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Monfils,
Paul
Sent: Friday, November 04, 2011 8:17 AM
To: Nancy Schmitt; histonet@lists.utsouthwestern.edu
Subject: RE: [Histonet] low profile vs. hi profile blades

In my opinion it doesn't make a lot of difference which type you use,
since all you actually use is the 2mm or so that extends beyond the edge
of the knife holder.  What is crucial is that you use the type of blade
your knife holder is designed for.  If you try to use high profile
blades in a low profile knife holder, or vice versa, you will have major
problems.  Also, high profile blades are available in a heavier, less
flexible form for cutting denser tissues.  I don't think the low profile
blades offer that option, though I am not certain since I use only the
high profile format.


___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


RE: [Histonet] Training and Competency Assessment for HE Slide Review

2011-11-06 Thread Patsy Ruegg
This was what I was going to suggest, I bet they let you participate in
HistoQip even if you are from Canada.

Patsy Ruegg, HT(ASCP)QIHC
IHCtech
12635 Montview Blvd. Ste.215
Aurora, CO 80045
720-859-4060
fax 720-859-4110
www.ihctech.net 
www.ihcrg.org


-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Janice
Mahoney
Sent: Friday, November 04, 2011 10:10 AM
To: joellewea...@hotmail.com; diana.har...@viha.ca; histo net
Subject: RE: [Histonet] Training and Competency Assessment for HE Slide
Review


A great way to do this is by using the results of the Histo QIP from CAP and
the NSH.  If you participate you get wonderful study and competency
materials to use for routine HE's special stains and IHC.
Jan Mhaoney
Omaha
 

 From: joellewea...@hotmail.com
 To: diana.har...@viha.ca; histonet@lists.utsouthwestern.edu
 Date: Thu, 3 Nov 2011 17:42:53 +
 Subject: Re: [Histonet] Training and Competency Assessment for HE Slide
Review
 CC: 
 
 Any certified histologist has gone through this, but the NSH has good
resources for this.
 Sent from my Verizon Wireless BlackBerry
 
 -Original Message-
 From: Harris Diana diana.har...@viha.ca
 Date: Thu, 3 Nov 2011 17:34:03 
 To: histonet@lists.utsouthwestern.edu
 Subject: [Histonet] Training and Competency Assessment for HE Slide
Review
 
 Any suggestions for creating a Training and Competency (TC) Assessment
for HE slide review?  We currently QC all HE slides macroscopically and
15% microscopically.  I would like to have all Histotechs trained and
competent to QC HE slides.  Has anyone gone thru this process?
 
 Thanks
 Diana Harris
 QC  Method Development Technologist
 Royal Jubilee Hospital
 Victoria, BC  Canada
 ___
 Histonet mailing list
 Histonet@lists.utsouthwestern.edu
 http://lists.utsouthwestern.edu/mailman/listinfo/histonet
 
 ___
 Histonet mailing list
 Histonet@lists.utsouthwestern.edu
 http://lists.utsouthwestern.edu/mailman/listinfo/histonet
 
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Re: Celestin blue B (was Help)

2011-11-06 Thread Bob Richmond
Corrie Vernick writes:

I am currently a histology student at Keiser University. I am doing a project 
for my routine staining class about Celestine Blue. I've been able to find 
information on why it was created, the chemical make up, and some of it's 
uses including the trichrome stain. I am having trouble finding images of 
slides stained with Celestine Blue. Any additional information about the uses 
would be helpful as well! Thank you, Corrinne Vernick, Keiser University FL 
U.S.A.

I don't have access to my library this week, but you can get a good
bit information by Googling celestin blue B. This dye was often used
as a sort of backup or substitute for hematoxylin, particularly in the
old outmoded Pearse stain for pituitary cells. R.D. Lillie as I
remember didn't think much of the dye, and I don't think this dye is a
very good topic for a study such as the one you describe.

Bob Richmond
Samurai Pathologist
Knoxville TN

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] RE: processing program for biopsies

2011-11-06 Thread Tony Henwood
Only 40 minutes in formalin is a problem.

Increase the formalin time to several hours and this should suffice for small 
2-3mm punch biopsies of skin.

For larger skin sections the alcohol, xylene and wax times will need to be 
tripled at least (as well as fixing the skins for longer (3-4 hours on the 
machine would be great)).

Regards 
Tony Henwood JP, MSc, BAppSc, GradDipSysAnalys, CT(ASC), FFSc(RCPA) 
Laboratory Manager  Senior Scientist 
Tel: 612 9845 3306 
Fax: 612 9845 3318 
the children's hospital at westmead
Cnr Hawkesbury Road and Hainsworth Street, Westmead
Locked Bag 4001, Westmead NSW 2145, AUSTRALIA 

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Carol Bryant
Sent: Saturday, 5 November 2011 3:51 AM
To: histonet@lists.utsouthwestern.edu
Subject: [Histonet] processing program for biopsies
Importance: High

Would you experienced histotechs mind to look at this program and see if it 
would cause skin biopsies to be poorly processed? I think too much time in 
alcohols.
  The pathologists say they look bad.
Thanks in advance for your help!

   Station  SolutionTime   Temp   P/VMix
   ---  --  -  -  -  --
   1Formalin 0:20   OFF   P/VSlow
   2Formalin 0:20   OFF   P/VSlow
   370% Alcohol  0:15   OFF   P/VSlow
   480% Alcohol  0:15   OFF   P/VSlow
   595% Alcohol  0:15   OFF   P/VSlow
   695% Alcohol  0:25   OFF   P/VSlow
   7100% Alcohol 0:25   OFF   P/VSlow
   8100% Alcohol 0:25   OFF   P/VSlow
   9Xylene   0:35   OFF   P/VSlow
  10Xylene   0:35   OFF   P/VSlow
  11Paraffin 0:15   63C.  P/VSlow
  12Paraffin 0:15   63C.  P/VSlow
  13Paraffin 0:15   63C.  P/VSlow
  14Paraffin 0:15   63C.  P/VSlow


Carol Bryant, CT (ASCP)
Cytology/Histology Manager
Lexington Clinic
Phone (859) 258-4082
Fax (859) 258-4081
cb...@lexclin.com



NOTICE OF CONFIDENTIALITY

This message, including any attachments, is intended only for the sole use of 
the addressee and may contain confidential or privileged information that is 
protected by the State of Kentucky and/or Federal regulations.  If you are not 
the intended recipient, do not read, copy, retain or disseminate this message 
or any attachment. If you have received this message in error, please call the 
sender immediately at (859)258-4000 and delete all copies of this message and 
any attachment. Any unauthorized review, use, disclosure, copying or 
distribution is strictly prohibited.  Neither the transmission of this message 
or any attachment, nor any error in transmission or misdelivery shall 
constitute waiver of any applicable legal privilege. 
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

*
This email and any files transmitted with it are confidential and intended 
solely for the use of the individual or entity to whom they are addressed. If 
you are not the intended recipient, please delete it and notify the sender.

Views expressed in this message and any attachments are those of the individual 
sender, and are not necessarily the views of The Children's Hospital at Westmead

This note also confirms that this email message has been virus scanned and 
although no computer viruses were detected, The Childrens Hospital at Westmead 
accepts no liability for any consequential damage resulting from email 
containing computer viruses.
*

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


Re: [Histonet] Re: Celestin blue B (was Help)

2011-11-06 Thread Lee Peggy Wenk
With the hematoxylin shortage of a couple of years ago (real, not imagined 
in about 2007-2008), several companies tried to come up with a synthetic dye 
substitute.


A little background: Celestine blue (CI 51050, also known as Mordant blue 
14) is a substitute touted many years ago (late 1960s/early 1970s if I 
remember, when there was another shortage for different reasons). I've used 
it years ago, in a stain called MSB for fibrin (look in a Bancroft book). 
The MSB stain used a double nuclear stain of aluminum hematoxylin (e.g. 
Mayer or Harris) and celestine blue. We didn't like the celestine blue, 
because when we mixed the iron mordant salt with the celestine blue dye, it 
worked right away, but, the next time someone asked for a MSB stain (several 
months later), the celestine blue was overoxidized, and wouldn't work. So we 
would have to stop and make the dye up immediately. It was a nuisance. But 
the double nuclear stain to to try to keep the nuclei a blue color, after 
going through the next 3 dyes (MSB is sort of a quatrachrome). Eventually, 
we just started using the Weigert hematoxylin from the trichrome stain in 
place of the double nuclear stain in the MSB. Our pathologists like the MSB 
stain better with the Weigert hematoxylin.


Current use: Since 2008, three companies that I know of came up with 
hematoxylin substitutes, in response to that shortage.
- Anatech Ltd, which used Mordant blue 3, CI 43820, and called it Tango 
Blue. So this is NOT celestine blue.

http://www.anatechltdusa.com/MSDS_pdf/TangoStainMSDS.pdf
- Newcomer Supply, which calls their substitute Newly Blue, but says it is a 
Celestine blue in one solution, ferric ammonium sulfate in the other 
solution. I don't know if you mix the two solutions before use, or if you 
dip the slides in first one solution and then the other. I didn't get around 
to their booth this year at NSH (sorry Marcia), and have never seen a 
procedure sheet on this stain.

http://www.newcomersupply.com/products/standard-special-stains?page=N#181
- ThermoFisher/Richard Allan, which calls their substitute Phoenix blue.The 
problem is, they are being secretive (proprietary) about what dye/reagents 
are in their solutions. I looked at one time, and couldn't find MSDS on 
their websites And their flyer doesn't mention the dyes name or CI #. 
However, the photo on the flyer, at least to me, looks like celestine blue. 
That's not proof that it IS celestine blue. Could be a close look-alike 
cousin.

http://www.thermo.com/eThermo/CMA/PDFs/Product/productPDF_6906.pdf

So look to these companies for more information. Anatech has a newsletter, 
available on their website, about the hematoxylin shortage (and their Tango 
blue product, of course).  (Not that this will help you with your Celestine 
blue project, but it will fill you in on the hematoxylin shortage history.)

http://www.anatechltdusa.com/Innovators/Inn08Summer.pdf

Hope that helps some.

Peggy A. Wenk, HTL(ASCP)SLS
Schools of Histotechnology
Beaumont Hospital
Royal Oak, MI 48073

(Opinions expressed are my own, and not Beaumont Hospitals'.)

-Original Message- 
From: Bob Richmond

Sent: Sunday, November 06, 2011 3:42 PM
To: histonet@lists.utsouthwestern.edu
Subject: [Histonet] Re: Celestin blue B (was Help)

Corrie Vernick writes:

I am currently a histology student at Keiser University. I am doing a 
project for my routine staining class about Celestine Blue. I've been able 
to find information on why it was created, the chemical make up, and some 
of it's uses including the trichrome stain. I am having trouble finding 
images of slides stained with Celestine Blue. Any additional information 
about the uses would be helpful as well! Thank you, Corrinne Vernick, 
Keiser University FL U.S.A.


I don't have access to my library this week, but you can get a good
bit information by Googling celestin blue B. This dye was often used
as a sort of backup or substitute for hematoxylin, particularly in the
old outmoded Pearse stain for pituitary cells. R.D. Lillie as I
remember didn't think much of the dye, and I don't think this dye is a
very good topic for a study such as the one you describe.

Bob Richmond
Samurai Pathologist
Knoxville TN

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet 



___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Microwave processing

2011-11-06 Thread Jamie Gomez
Hello, I am a histology technology student and I need
information on the melting point of the paraffin used in Microwave Processing…
Freida Carson stated in her book something regarding the paraffin temperature
set at 84 degrees. The class had a discussion regarding the same and concluded
that the tissue morphology would be damage at such high temperature. Can someone
please provide me with some input on this issue. 
 
Thank you,
Jamie G.
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] microwave processing

2011-11-06 Thread Zoe rosa

Hi there,
 
On an Instrumentation course at keiser, we had a discussion about the right 
temperature fro the paraffin when the proceesing is done using microwaves. Can 
some tell me what should be the ideal temperature with out damaging the 
processing. The text book mentions 84 degrees, but isn't this to high; perhaps 
if the time is less, it wont affect it?
 
Thanks,
 
Ismael Lopez  
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


RE: [Histonet] microwave processing

2011-11-06 Thread Tony Henwood (SCHN)
I believe that the higher temperature is needed to evaporate off the 
isopropanol.
Just make sure that your tissue is well fixed.

Regards 
Tony Henwood JP, MSc, BAppSc, GradDipSysAnalys, CT(ASC), FFSc(RCPA) 
Laboratory Manager  Senior Scientist 
Tel: 612 9845 3306 
Fax: 612 9845 3318 
the children's hospital at westmead
Cnr Hawkesbury Road and Hainsworth Street, Westmead
Locked Bag 4001, Westmead NSW 2145, AUSTRALIA 


-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Zoe rosa
Sent: Monday, 7 November 2011 1:41 PM
To: histonet@lists.utsouthwestern.edu
Subject: [Histonet] microwave processing


Hi there,
 
On an Instrumentation course at keiser, we had a discussion about the right 
temperature fro the paraffin when the proceesing is done using microwaves. Can 
some tell me what should be the ideal temperature with out damaging the 
processing. The text book mentions 84 degrees, but isn't this to high; perhaps 
if the time is less, it wont affect it?
 
Thanks,
 
Ismael Lopez  
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

*
This email and any files transmitted with it are confidential and intended 
solely for the use of the individual or entity to whom they are addressed. If 
you are not the intended recipient, please delete it and notify the sender.

Views expressed in this message and any attachments are those of the individual 
sender, and are not necessarily the views of The Children's Hospital at Westmead

This note also confirms that this email message has been virus scanned and 
although no computer viruses were detected, The Childrens Hospital at Westmead 
accepts no liability for any consequential damage resulting from email 
containing computer viruses.
*

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


Re: [Histonet] Re: Celestin blue B (was Help)

2011-11-06 Thread Tony Reilly
Hello
 
I have used Celestin Blue-Haemalum (Mayers) considerably in the past as a 
substitute for Weigert's and as you say the Celestin Blue stain does not keep 
and is best made fresh for each use.  Our method to simplify this was to make 
separate double strength solutions of the Celestin blue and the mordant which 
were mixed 50/50 just prior to use to make a solution of the correct strength.  
Both of these solutions store well and the preparation time is minimal.
 
regards
Tony
 
 
 

Tony Reilly  B.App.Sc. , M.Sc.
Chief Scientist, Anatomical Pathology
Pathology Queensland-PA Laboratory
_
Clinical and Statewide Services Division| QueenslandHealth
 
Level 1, Building 15,Princess Alexandra Hospital
Ipswich Road,WOOLLOONGABBA  Qld4102
Ph: 07 3176 2412
Mob: 0402 139411
Fax: 07 3176 2930
Email: tony_rei...@health.qld.gov.au
Web:  www.health.qld.gov.au/qhcss/
 
 


 Lee  Peggy Wenk lpw...@sbcglobal.net 11/7/2011 8:36 am 
With the hematoxylin shortage of a couple of years ago (real, not imagined 
in about 2007-2008), several companies tried to come up with a synthetic dye 
substitute.

A little background: Celestine blue (CI 51050, also known as Mordant blue 
14) is a substitute touted many years ago (late 1960s/early 1970s if I 
remember, when there was another shortage for different reasons). I've used 
it years ago, in a stain called MSB for fibrin (look in a Bancroft book). 
The MSB stain used a double nuclear stain of aluminum hematoxylin (e.g. 
Mayer or Harris) and celestine blue. We didn't like the celestine blue, 
because when we mixed the iron mordant salt with the celestine blue dye, it 
worked right away, but, the next time someone asked for a MSB stain (several 
months later), the celestine blue was overoxidized, and wouldn't work. So we 
would have to stop and make the dye up immediately. It was a nuisance. But 
the double nuclear stain to to try to keep the nuclei a blue color, after 
going through the next 3 dyes (MSB is sort of a quatrachrome). Eventually, 
we just started using the Weigert hematoxylin from the trichrome stain in 
place of the double nuclear stain in the MSB. Our pathologists like the MSB 
stain better with the Weigert hematoxylin.

Current use: Since 2008, three companies that I know of came up with 
hematoxylin substitutes, in response to that shortage.
- Anatech Ltd, which used Mordant blue 3, CI 43820, and called it Tango 
Blue. So this is NOT celestine blue.
http://www.anatechltdusa.com/MSDS_pdf/TangoStainMSDS.pdf 
- Newcomer Supply, which calls their substitute Newly Blue, but says it is a 
Celestine blue in one solution, ferric ammonium sulfate in the other 
solution. I don't know if you mix the two solutions before use, or if you 
dip the slides in first one solution and then the other. I didn't get around 
to their booth this year at NSH (sorry Marcia), and have never seen a 
procedure sheet on this stain.
http://www.newcomersupply.com/products/standard-special-stains?page=N#181 
- ThermoFisher/Richard Allan, which calls their substitute Phoenix blue.The 
problem is, they are being secretive (proprietary) about what dye/reagents 
are in their solutions. I looked at one time, and couldn't find MSDS on 
their websites And their flyer doesn't mention the dyes name or CI #. 
However, the photo on the flyer, at least to me, looks like celestine blue. 
That's not proof that it IS celestine blue. Could be a close look-alike 
cousin.
http://www.thermo.com/eThermo/CMA/PDFs/Product/productPDF_6906.pdf 

So look to these companies for more information. Anatech has a newsletter, 
available on their website, about the hematoxylin shortage (and their Tango 
blue product, of course).  (Not that this will help you with your Celestine 
blue project, but it will fill you in on the hematoxylin shortage history.)
http://www.anatechltdusa.com/Innovators/Inn08Summer.pdf 

Hope that helps some.

Peggy A. Wenk, HTL(ASCP)SLS
Schools of Histotechnology
Beaumont Hospital
Royal Oak, MI 48073

(Opinions expressed are my own, and not Beaumont Hospitals'.)

-Original Message- 
From: Bob Richmond
Sent: Sunday, November 06, 2011 3:42 PM
To: histonet@lists.utsouthwestern.edu 
Subject: [Histonet] Re: Celestin blue B (was Help)

Corrie Vernick writes:

I am currently a histology student at Keiser University. I am doing a 
project for my routine staining class about Celestine Blue. I've been able 
to find information on why it was created, the chemical make up, and some 
of it's uses including the trichrome stain. I am having trouble finding 
images of slides stained with Celestine Blue. Any additional information 
about the uses would be helpful as well! Thank you, Corrinne Vernick, 
Keiser University FL U.S.A.

I don't have access to my library this week, but you can get a good
bit information by Googling celestin blue B. This dye was often used
as a sort of backup or substitute for hematoxylin, particularly in the