Re: [Histonet] Cell block processing

2019-10-28 Thread John Garratt via Histonet
There's been excellent discussion on fixation, IHC and cells blocks, with Joe 
Walker's email summing up the situation nicely. For those interested in side by 
side comparison of IHC staining of cytology cell blocks using different 
fixatives go to http://cpqa.ca/main/wp-content/uploads/2014/06/2014-Thomson.pdf

Regards

John Garratt


www.ciqc.ca

‐‐‐ Original Message ‐‐‐
On Monday, October 28, 2019 7:38 AM, Joe W. Walker, Jr. via Histonet 
 wrote:

> Hi Terri,
>
> At one time we did the same thing but have changed our approach in light of 
> the FDA's and CAP's view point on ASRs. The potential problem is that IHCs 
> are all validated/tested by the manufacturer on FFPE tissue. By introducing 
> methanol/ethanol as the first step in fixation, you potentially have altered 
> the initial fixation steps. I've attended several meetings on this topic and 
> have been advised to stop performing IHC on methanol/ethanol fixed specimens 
> unless we validated that this fixation step doesn't alter the expression of 
> the target antigen in the tissue. Formalin fixation after an alcohol fixation 
> doesn't change/reverse any alterations to the antigen in the tissue.
>
> We utilize an IBF tissue fixative but have also validated this fixative with 
> our antibody panels that we offer. The IBF does contain a small amount of 
> alcohol and the fixative is slightly different than 10% buffered formalin.
>
> I agree that CytoLyt is excellent at lysing red blood cells but would just 
> caution you on using the specimen for IHC without a disclaimer within your 
> report or validating your IHCs on these specimens to ensure they work as 
> expected. Keep in mind that most control tissue is FFPE and using it to 
> compare if the IHC worked in a first fixed alcohol specimen is not an apples 
> to apples comparison.
>
> Cheers,
>
> Joe W. Walker, Jr. MS, SCT(ASCP)
> Anatomical Pathology Manager
> joewal...@rrmc.org, www.rrmc.org
>
> -Original Message-
> From: Terri Braud via Histonet histonet@lists.utsouthwestern.edu
> Sent: Monday, October 28, 2019 10:14 AM
> To: 'histonet@lists.utsouthwestern.edu' histonet@lists.utsouthwestern.edu
> Subject: Re: [Histonet] Cell block processing
>
> [External Email] This email originated from outside of the organization. 
> Think before you click: Don’t click on links, open attachments or respond to 
> requests for sensitive information if the email looks suspicious or you don’t 
> recognize the sender.
>
> We collect our FNAs in CytoLyt. We also use it to wash all our non-gyn 
> fluids, but then we fix the cell block "pellet" in formalin.
> We have had no problems with immunos, and are able to lyse the RBCs to 
> provide a nice, clear specimen.
> Hope this helps.
>
> Terri L. Braud, HT(ASCP)
> Anatomic Pathology Supervisor
> Laboratory
> Holy Redeemer Hospital
> 1648 Huntingdon Pike
> Meadowbrook, PA 19046
> ph: 215-938-3689
> fax: 215-938-3874
> Care, Comfort, and Heal
>
> Histonet mailing list
> Histonet@lists.utsouthwestern.edu
> https://nam02.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.com%2Fv3%2F__http%3A%2F%2Flists.utsouthwestern.edu%2Fmailman%2Flistinfo%2Fhistonet__%3B!5JlSGkcjda6Eqs5J!rHP2_dFONxwRmR87ddavejT_o6RIcXBah8jw7nMUO0tWdC5KpksUnk2bttyIMkU%24=02|01|jwwalker%40rrmc.org|eeac9b3afe444de27cd208d75bb1634b|0e55647d438e4a448437e959c3cf2240|0|0|637078689748526729=1%2BpvxH9TcvKxAo6jgjr7fobCAUUz35sHbsLfoBvHmGE%3D=0
> [https://www.rrmc.org/app/files/public/2633/2019_hyht_sig-_jan2019_final.jpg]
>
> Histonet mailing list
> Histonet@lists.utsouthwestern.edu
> http://lists.utsouthwestern.edu/mailman/listinfo/histonet



___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


Re: [Histonet] Cell block processing

2019-10-28 Thread Joe W. Walker, Jr. via Histonet
Hi Terri,

At one time we did the same thing but have changed our approach in light of the 
FDA's and CAP's view point on ASRs.  The potential problem is that IHCs are all 
validated/tested by the manufacturer on FFPE tissue.  By introducing 
methanol/ethanol as the first step in fixation, you potentially have altered 
the initial fixation steps.  I've attended several meetings on this topic and 
have been advised to stop performing IHC on methanol/ethanol fixed specimens 
unless we validated that this fixation step doesn't alter the expression of the 
target antigen in the tissue.  Formalin fixation after an alcohol fixation 
doesn't change/reverse any alterations to the antigen in the tissue.

We utilize an IBF tissue fixative but have also validated this fixative with 
our antibody panels that we offer.  The IBF does contain a small amount of 
alcohol and the fixative is slightly different than 10% buffered formalin.

I agree that CytoLyt is excellent at lysing red blood cells but would just 
caution you on using the specimen for IHC without a disclaimer within your 
report or validating your IHCs on these specimens to ensure they work as 
expected.  Keep in mind that most control tissue is FFPE and using it to 
compare if the IHC worked in a first fixed alcohol specimen is not an apples to 
apples comparison.

Cheers,

Joe W. Walker, Jr. MS, SCT(ASCP)
Anatomical Pathology Manager
joewal...@rrmc.org, www.rrmc.org

-Original Message-
From: Terri Braud via Histonet 
Sent: Monday, October 28, 2019 10:14 AM
To: 'histonet@lists.utsouthwestern.edu' 
Subject: Re: [Histonet] Cell block processing

[External Email] This email originated from outside of the organization. Think 
before you click: Don’t click on links, open attachments or respond to requests 
for sensitive information if the email looks suspicious or you don’t recognize 
the sender.


We collect our FNAs in CytoLyt.  We also use it to wash all our non-gyn fluids, 
but then we fix the cell block "pellet" in formalin.
We have had no problems with immunos, and are able to lyse the RBCs to provide 
a nice, clear specimen.
Hope this helps.

Terri L. Braud, HT(ASCP)
Anatomic Pathology Supervisor
Laboratory
Holy Redeemer Hospital
1648 Huntingdon Pike
Meadowbrook, PA 19046
ph: 215-938-3689
fax: 215-938-3874
Care, Comfort, and Heal



___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
https://nam02.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.com%2Fv3%2F__http%3A%2F%2Flists.utsouthwestern.edu%2Fmailman%2Flistinfo%2Fhistonet__%3B!5JlSGkcjda6Eqs5J!rHP2_dFONxwRmR87ddavejT_o6RIcXBah8jw7nMUO0tWdC5KpksUnk2bttyIMkU%24data=02%7C01%7Cjwwalker%40rrmc.org%7Ceeac9b3afe444de27cd208d75bb1634b%7C0e55647d438e4a448437e959c3cf2240%7C0%7C0%7C637078689748526729sdata=1%2BpvxH9TcvKxAo6jgjr7fobCAUUz35sHbsLfoBvHmGE%3Dreserved=0
[https://www.rrmc.org/app/files/public/2633/2019_hyht_sig-_jan2019_final.jpg]
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


Re: [Histonet] Cell block processing

2019-10-28 Thread Terri Braud via Histonet
We collect our FNAs in CytoLyt.  We also use it to wash all our non-gyn fluids, 
but then we fix the cell block "pellet" in formalin.
We have had no problems with immunos, and are able to lyse the RBCs to provide 
a nice, clear specimen.
Hope this helps.

Terri L. Braud, HT(ASCP)
Anatomic Pathology Supervisor
Laboratory
Holy Redeemer Hospital
1648 Huntingdon Pike
Meadowbrook, PA 19046
ph: 215-938-3689
fax: 215-938-3874
Care, Comfort, and Heal



___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


Re: [Histonet] Cell block processing

2019-10-25 Thread Tony Henwood (SCHN) via Histonet
Hi Charles,

I have had excellent success with lysing the red blood cells (using  Isotonic 
Ammonium Chloride) prior to cell block preparation with thromboplastin-plasma. 
The lysing solution contains EDTA so you will need to add a few drops of 1% 
calcium chloride. Method as follows:

Lysis solution
Ammonium Chloride4.5g
Potassium carbonate  0.5g
EDTA 0.0186g
Distilled water  500mls

Method:
1.  Centrifuge bloody fluid.
2.  Remove supernatant and add equal volume of lysis solution.
3.  Resuspend and incubate for 5 minutes at 4oC.
4.  Centrifuge, if blood still remains, then repeat from step 2.
5. Rinse in Hanks or RPMI, centrifuge.
6.  Mix pellet in a few drops of plasma.
7. Add thromboplastin and a few drops  of 1% Calcium Chloride, mix gently 
and allow clot to form.
8. Add 10% buffered formalin and fix and process as usual.

Reference:  
Kuenen-Boumiester etal (1996) Acta Cytolog 40:475-479

The lysis solution can also be purchased commercially from several companies 
(eg Biolegend). It is commonly used for sample preparation for flow cytometry. 
Check the SDS to make sure it does not contain formaldehyde.


Regards
Tony Henwood JP, MSc, BAppSc, GradDipSysAnalys, CT(ASC), FFSc(RCPA)
Principal Scientist, the Children’s Hospital at Westmead
Adjunct Fellow, School of Medicine, University of Western Sydney
Tel: 612 9845 3306
Fax: 612 9845 3318
Pathology Department
the children's hospital at westmead
Cnr Hawkesbury Road and Hainsworth Street, Westmead
Locked Bag 4001, Westmead NSW 2145, AUSTRALIA


From: Charles Riley via Histonet 
Sent: Friday, 25 October 2019 23:12
To: Histo List
Subject: [Histonet] Cell block processing

Does anyone have any tips or suggestions on how to better process extremely
bloody FNA  specimens?Is there anyway to clear out some or all of the
blood without destroying the other tissues?

--

Charles Riley BS  HT, HTL(ASCP)CM

Histopathology Coordinator/ Mohs
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

This message is intended for the addressee named and may contain confidential 
information. If you are not the intended recipient, please delete it and notify 
the sender.

Views expressed in this message are those of the individual sender, and are not 
necessarily the views of NSW Health or any of its entities.


___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


Re: [Histonet] Cell block processing

2019-10-25 Thread Cartun, Richard via Histonet
I agree with Joe.  We used to use ETOH for cell blocks, but stopped using it 
when we started doing IHC biomarker testing on these specimens.  Alcohol is 
good for some proteomic targets, but can be a disaster for others.  We also fix 
all of our cell block specimens that are collected in saline or RPMI in 
formalin.

Richard

Richard W. Cartun, MS, PhD
Director, Histology & The Martin M. Berman, MD Immunopathology/Morphologic 
Proteomics Laboratory
Director, Biospecimen Collection Programs
Assistant Director, Anatomic Pathology
Hartford Hospital
80 Seymour Street
Hartford, CT  06102
(860) 972-1596 (Office)
(860) 545-2204 (Fax)
richard.car...@hhchealth.org

-Original Message-
From: Joe W. Walker, Jr. via Histonet [mailto:histonet@lists.utsouthwestern.edu]
Sent: Friday, October 25, 2019 4:36 PM
To: Charles Riley
Cc: histonet@lists.utsouthwestern.edu
Subject: Re: [Histonet] Cell block processing

EXTERNAL email is from outside HHC. DO NOT open attachments or click links from 
unknown senders.

As a cytotech, that wouldn’t be my first choice for collections and FNA 
specimens.  The main reason is that once fixed in 95% ETOH you are limited if 
you need to perform IHC stains on the cell block unless you have validated your 
IHCs on ETOH fixed specimens.  How do you process the FNA rinses that in in 
ETOH: Only Cell blocks or do you have another cytology liquid prep?



Without knowing your prep process, I’d suggest collecting the FNA needle rinses 
in Hank’s Balanced Salt solution.  After making the cell block, you could then 
formalin fix them.  I can send you a procedure that we utilize for this 
process.  The cell blocks cut great, look great, and you can perform IHC an 
molecular testing if needed.



Joe Walker



From: Charles Riley 

Sent: Friday, October 25, 2019 12:57 PM

To: Joe W. Walker, Jr. 

Cc: histonet@lists.utsouthwestern.edu

Subject: Re: [Histonet] Cell block processing



[External Email] This email originated from outside of the organization. Think 
before you click: Don’t click on links, open attachments or respond to requests 
for sensitive information if the email looks suspicious or you don’t recognize 
the sender.



Our tech said they use 95% alcohol to collect the specimen.



On Fri, Oct 25, 2019 at 12:23 PM Joe W. Walker, Jr. 
mailto:jwwal...@rrmc.org>> wrote:

Hi Charles,



What are you collecting the FNA into?  Cytorich? Cytolyt? Other?



Joe W. Walker, Jr. MS, SCT(ASCP)

Anatomical Pathology Manager

joewal...@rrmc.org<mailto:joewal...@rrmc.org>, 
https://urldefense.proofpoint.com/v2/url?u=http-3A__www.rrmc.org=DwIGaQ=e_HtEeZEQXP5NUOb33qoTj0AVvRFBS9_rhBTQcfkWoA=eCMS5E4UqfGKTlPknIMjMlCkk-KGKHflslRsFz3l5JE=2CdTK8PmO1-CXFcs6R-WWhnxgbCm1AHZ9wJ_1YOqDoA=JIavq_Wfz4S4TppCsMv8kQfnmD9wLQhRVNqSm1eJAO4=
 
<https://urldefense.proofpoint.com/v2/url?u=https-3A__nam02.safelinks.protection.outlook.com_-3Furl-3Dhttp-253A-252F-252Fwww.rrmc.org-26data-3D02-257C01-257Cjwwalker-2540rrmc.org-257C2403b15e19aa486985ac08d7596c3f45-257C0e55647d438e4a448437e959c3cf2240-257C0-257C0-257C637076193772807309-26sdata-3DVxp3O6jX4PVU4FDmXfFTwIQFG0oR03V3csN45z012Mg-253D-26reserved-3D0=DwIGaQ=e_HtEeZEQXP5NUOb33qoTj0AVvRFBS9_rhBTQcfkWoA=eCMS5E4UqfGKTlPknIMjMlCkk-KGKHflslRsFz3l5JE=2CdTK8PmO1-CXFcs6R-WWhnxgbCm1AHZ9wJ_1YOqDoA=Vn_yzz1gfScLUDaT695jKQgsKf5KhNHP41Y1ZT_Vz-Y=
 >



-Original Message-

From: Charles Riley via Histonet 
mailto:histonet@lists.utsouthwestern.edu>>

Sent: Friday, October 25, 2019 8:13 AM

To: Histo List 
mailto:histonet@lists.utsouthwestern.edu>>

Subject: [Histonet] Cell block processing



[External Email] This email originated from outside of the organization. Think 
before you click: Don’t click on links, open attachments or respond to requests 
for sensitive information if the email looks suspicious or you don’t recognize 
the sender.





Does anyone have any tips or suggestions on how to better process extremely

bloody FNA  specimens?Is there anyway to clear out some or all of the

blood without destroying the other tissues?



--



Charles Riley BS  HT, HTL(ASCP)CM



Histopathology Coordinator/ Mohs

___

Histonet mailing list

Histonet@lists.utsouthwestern.edu<mailto:Histonet@lists.utsouthwestern.edu>

https://urldefense.proofpoint.com/v2/url?u=https-3A__nam02.safelinks.protection.outlook.com_-3Furl-3Dhttps-253A-252F-252Furldefense.com-252Fv3-252F-5F-5Fhttp-253A-252F-252Flists.utsouthwestern.edu-252Fmailman-252Flistinfo-252Fhistonet-5F-5F-253B-215JlSGkcjda6Eqs5J-21rntZr6E7tVvCc2tzoJy-5FL8lFMF6EvHUTKJOsc3NTMLFSI35SgrcseW2ly3GyamY-2524-26amp-3Bdata-3D02-257C01-257Cjwwalker-2540rrmc.org-257C6e4cdd79edfa4e1ef23408d75944c458-257C0e55647d438e4a448437e959c3cf2240-257C0-257C0-257C637076024216531651-26amp-3Bsdata-3DxAxqyyCfRepB3DlAl-252Fw651nk3B5ViHQLjqdToa2iAhw-253D-26amp-3Breserved-3D0=DwIGaQ=e_HtEeZEQXP5NUOb33qoTj0AVvRFBS9_rhBTQcfkWoA=eCMS5E4UqfGKTlPknIMjMlCkk-KGKHflslRsFz3l5JE=2Cd

Re: [Histonet] Cell block processing

2019-10-25 Thread Joe W. Walker, Jr. via Histonet
As a cytotech, that wouldn’t be my first choice for collections and FNA 
specimens.  The main reason is that once fixed in 95% ETOH you are limited if 
you need to perform IHC stains on the cell block unless you have validated your 
IHCs on ETOH fixed specimens.  How do you process the FNA rinses that in in 
ETOH: Only Cell blocks or do you have another cytology liquid prep?

Without knowing your prep process, I’d suggest collecting the FNA needle rinses 
in Hank’s Balanced Salt solution.  After making the cell block, you could then 
formalin fix them.  I can send you a procedure that we utilize for this 
process.  The cell blocks cut great, look great, and you can perform IHC an 
molecular testing if needed.

Joe Walker

From: Charles Riley 
Sent: Friday, October 25, 2019 12:57 PM
To: Joe W. Walker, Jr. 
Cc: histonet@lists.utsouthwestern.edu
Subject: Re: [Histonet] Cell block processing

[External Email] This email originated from outside of the organization. Think 
before you click: Don’t click on links, open attachments or respond to requests 
for sensitive information if the email looks suspicious or you don’t recognize 
the sender.

Our tech said they use 95% alcohol to collect the specimen.

On Fri, Oct 25, 2019 at 12:23 PM Joe W. Walker, Jr. 
mailto:jwwal...@rrmc.org>> wrote:
Hi Charles,

What are you collecting the FNA into?  Cytorich? Cytolyt? Other?

Joe W. Walker, Jr. MS, SCT(ASCP)
Anatomical Pathology Manager
joewal...@rrmc.org<mailto:joewal...@rrmc.org>, 
www.rrmc.org<https://nam02.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.rrmc.org=02%7C01%7Cjwwalker%40rrmc.org%7C2403b15e19aa486985ac08d7596c3f45%7C0e55647d438e4a448437e959c3cf2240%7C0%7C0%7C637076193772807309=Vxp3O6jX4PVU4FDmXfFTwIQFG0oR03V3csN45z012Mg%3D=0>

-Original Message-
From: Charles Riley via Histonet 
mailto:histonet@lists.utsouthwestern.edu>>
Sent: Friday, October 25, 2019 8:13 AM
To: Histo List 
mailto:histonet@lists.utsouthwestern.edu>>
Subject: [Histonet] Cell block processing

[External Email] This email originated from outside of the organization. Think 
before you click: Don’t click on links, open attachments or respond to requests 
for sensitive information if the email looks suspicious or you don’t recognize 
the sender.


Does anyone have any tips or suggestions on how to better process extremely
bloody FNA  specimens?Is there anyway to clear out some or all of the
blood without destroying the other tissues?

--

Charles Riley BS  HT, HTL(ASCP)CM

Histopathology Coordinator/ Mohs
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu<mailto:Histonet@lists.utsouthwestern.edu>
https://nam02.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.com%2Fv3%2F__http%3A%2F%2Flists.utsouthwestern.edu%2Fmailman%2Flistinfo%2Fhistonet__%3B!5JlSGkcjda6Eqs5J!rntZr6E7tVvCc2tzoJy_L8lFMF6EvHUTKJOsc3NTMLFSI35SgrcseW2ly3GyamY%24data=02%7C01%7Cjwwalker%40rrmc.org%7C6e4cdd79edfa4e1ef23408d75944c458%7C0e55647d438e4a448437e959c3cf2240%7C0%7C0%7C637076024216531651sdata=xAxqyyCfRepB3DlAl%2Fw651nk3B5ViHQLjqdToa2iAhw%3Dreserved=0<https://nam02.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.com%2Fv3%2F__https%3A%2F%2Fnam02.safelinks.protection.outlook.com%2F%3Furl%3Dhttps*3A*2F*2Furldefense.com*2Fv3*2F__http*3A*2F*2Flists.utsouthwestern.edu*2Fmailman*2Flistinfo*2Fhistonet__*3B!5JlSGkcjda6Eqs5J!rntZr6E7tVvCc2tzoJy_L8lFMF6EvHUTKJOsc3NTMLFSI35SgrcseW2ly3GyamY*24%26data%3D02*7C01*7Cjwwalker*40rrmc.org*7C6e4cdd79edfa4e1ef23408d75944c458*7C0e55647d438e4a448437e959c3cf2240*7C0*7C0*7C637076024216531651%26sdata%3DxAxqyyCfRepB3DlAl*2Fw651nk3B5ViHQLjqdToa2iAhw*3D%26reserved%3D0__%3BJSUlJSUlJSUlJSUlJSUlJSUlJSUlJSU!5JlSGkcjda6Eqs5J!rGVYY5dOQ2xRCn-bsaKkh1bKH2qKCaDhf7LTZHmhPvEE8IgM-u6EOhcj-0W-HWE%24=02%7C01%7Cjwwalker%40rrmc.org%7C2403b15e19aa486985ac08d7596c3f45%7C0e55647d438e4a448437e959c3cf2240%7C0%7C0%7C637076193772817305=bd7V2hr4xJ0fe%2FUs39UFEPQaB4RKHS45Y42KOAaimIs%3D=0>
[https://www.rrmc.org/app/files/public/2633/2019_hyht_sig-_jan2019_final.jpg<https://nam02.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.rrmc.org%2Fapp%2Ffiles%2Fpublic%2F2633%2F2019_hyht_sig-_jan2019_final.jpg=02%7C01%7Cjwwalker%40rrmc.org%7C2403b15e19aa486985ac08d7596c3f45%7C0e55647d438e4a448437e959c3cf2240%7C0%7C0%7C637076193772817305=jp9HasHg%2BWY50DYjwo20Rv2X46JxVh3M6Oshnp%2B44H0%3D=0>]


--

Charles Riley BS  HT, HTL(ASCP)CM

Histopathology Coordinator/ Mohs


___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


Re: [Histonet] Cell block processing

2019-10-25 Thread Charles Riley via Histonet
Our tech said they use 95% alcohol to collect the specimen.

On Fri, Oct 25, 2019 at 12:23 PM Joe W. Walker, Jr. 
wrote:

> Hi Charles,
>
> What are you collecting the FNA into?  Cytorich? Cytolyt? Other?
>
> Joe W. Walker, Jr. MS, SCT(ASCP)
> Anatomical Pathology Manager
> joewal...@rrmc.org, www.rrmc.org
>
> -Original Message-
> From: Charles Riley via Histonet 
> Sent: Friday, October 25, 2019 8:13 AM
> To: Histo List 
> Subject: [Histonet] Cell block processing
>
> [External Email] This email originated from outside of the organization.
> Think before you click: Don’t click on links, open attachments or respond
> to requests for sensitive information if the email looks suspicious or you
> don’t recognize the sender.
>
>
> Does anyone have any tips or suggestions on how to better process extremely
> bloody FNA  specimens?Is there anyway to clear out some or all of the
> blood without destroying the other tissues?
>
> --
>
> Charles Riley BS  HT, HTL(ASCP)CM
>
> Histopathology Coordinator/ Mohs
> ___
> Histonet mailing list
> Histonet@lists.utsouthwestern.edu
>
> https://nam02.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.com%2Fv3%2F__http%3A%2F%2Flists.utsouthwestern.edu%2Fmailman%2Flistinfo%2Fhistonet__%3B!5JlSGkcjda6Eqs5J!rntZr6E7tVvCc2tzoJy_L8lFMF6EvHUTKJOsc3NTMLFSI35SgrcseW2ly3GyamY%24data=02%7C01%7Cjwwalker%40rrmc.org%7C6e4cdd79edfa4e1ef23408d75944c458%7C0e55647d438e4a448437e959c3cf2240%7C0%7C0%7C637076024216531651sdata=xAxqyyCfRepB3DlAl%2Fw651nk3B5ViHQLjqdToa2iAhw%3Dreserved=0
> [
> https://www.rrmc.org/app/files/public/2633/2019_hyht_sig-_jan2019_final.jpg
> ]
>


-- 

Charles Riley BS  HT, HTL(ASCP)CM

Histopathology Coordinator/ Mohs
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


Re: [Histonet] Cell block processing

2019-10-25 Thread Joe W. Walker, Jr. via Histonet
Hi Charles,

What are you collecting the FNA into?  Cytorich? Cytolyt? Other?

Joe W. Walker, Jr. MS, SCT(ASCP)
Anatomical Pathology Manager
joewal...@rrmc.org, www.rrmc.org

-Original Message-
From: Charles Riley via Histonet 
Sent: Friday, October 25, 2019 8:13 AM
To: Histo List 
Subject: [Histonet] Cell block processing

[External Email] This email originated from outside of the organization. Think 
before you click: Don’t click on links, open attachments or respond to requests 
for sensitive information if the email looks suspicious or you don’t recognize 
the sender.


Does anyone have any tips or suggestions on how to better process extremely
bloody FNA  specimens?Is there anyway to clear out some or all of the
blood without destroying the other tissues?

--

Charles Riley BS  HT, HTL(ASCP)CM

Histopathology Coordinator/ Mohs
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
https://nam02.safelinks.protection.outlook.com/?url=https%3A%2F%2Furldefense.com%2Fv3%2F__http%3A%2F%2Flists.utsouthwestern.edu%2Fmailman%2Flistinfo%2Fhistonet__%3B!5JlSGkcjda6Eqs5J!rntZr6E7tVvCc2tzoJy_L8lFMF6EvHUTKJOsc3NTMLFSI35SgrcseW2ly3GyamY%24data=02%7C01%7Cjwwalker%40rrmc.org%7C6e4cdd79edfa4e1ef23408d75944c458%7C0e55647d438e4a448437e959c3cf2240%7C0%7C0%7C637076024216531651sdata=xAxqyyCfRepB3DlAl%2Fw651nk3B5ViHQLjqdToa2iAhw%3Dreserved=0
[https://www.rrmc.org/app/files/public/2633/2019_hyht_sig-_jan2019_final.jpg]
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet