Hi

I think the truth of the matter is that these gizmos have been cheap enough 
(and complicated enough) that they don't get repaired a lot. Certainly the VCXO 
gets tweaked, and there are a couple of caps that explode / get replaced. Past 
that either they don't break, or they get replaced.

Bob

On Feb 18, 2013, at 1:52 AM, Magnus Danielson mag...@rubidium.dyndns.org 
wrote:

 On 02/18/2013 04:34 AM, Bob Camp wrote:
 Hi
 
 There *must* be an alignment procedure that sets these things up. They 
 didn't put all those test points in there just for the fun of it. I'm sure 
 that these units would work a bit better if we knew how to tweak them back 
 to the original alignment specs.
 
 There is a document which gives some of it, but doing a reverse-engineering 
 project is primarily so it can be properly understood for trimming and repair 
 purposes in my mind.
 
 There is a 2x4 connectors which can connect in a test-rig while the 2x5 
 connector is hooked up. There is two different connectors inside that also 
 should be relevant, and then is at least one jumper field which would be good 
 to know how it works.
 
 Considering how common these are, I was surprised that more effort had not 
 gone into them, compared to how the group (and hams) tend to deal with all 
 this.
 
 Cheers,
 Magnus
 
 Bob
 
 On Feb 16, 2013, at 6:10 PM, Magnus Danielsonmag...@rubidium.dyndns.org  
 wrote:
 
 Fellow time-nuts,
 
 Considering that LPROs is pretty popular, I am a bit surprised that I have 
 not seen any major reverse-engineering effort on the LPROs. I have the 
 self-compiled LPRO service document, which collects parts of schematics 
 from patents, but still.
 
 My main reasons for asking is that I want to get a little better overview 
 of how they work, how I can tune them up and what signals is available 
 where. Naturally, always figuring if there is some interesting tweaks to be 
 done.
 
 LRPO is just a traditional analogue rubidium, in compact format, sure, but 
 never the less.
 
 I have noticed that different FPGAs have been used over time. Curious about 
 the various jumpers and connectors in it.
 
 Cheers,
 Magnus
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Reverse-engineering LPRO

2013-02-18 Thread paul swed
That was the intent. Break/toss / we get to buy'em after a pile stacks up.

From what I have run into peaking may or may not be the right answer is my
only real addition to the discussion.
The RF oscillator on the bulb peak if even tunable.
The hamonic multiplier. Peaking that could easily lead to the wrong
multiple killing the unit. Not really but hard to find the right multiple
without a spectrum analyzer. Its stuff like that which introduces risk.
But hey when they are dead/weak to begin with all is fair. Thats how I
learned at $25 a piece.
Regards
Paul

On Mon, Feb 18, 2013 at 8:20 AM, Bob Camp li...@rtty.us wrote:

 Hi

 I think the truth of the matter is that these gizmos have been cheap
 enough (and complicated enough) that they don't get repaired a lot.
 Certainly the VCXO gets tweaked, and there are a couple of caps that
 explode / get replaced. Past that either they don't break, or they get
 replaced.

 Bob

 On Feb 18, 2013, at 1:52 AM, Magnus Danielson mag...@rubidium.dyndns.org
 wrote:

  On 02/18/2013 04:34 AM, Bob Camp wrote:
  Hi
 
  There *must* be an alignment procedure that sets these things up. They
 didn't put all those test points in there just for the fun of it. I'm sure
 that these units would work a bit better if we knew how to tweak them back
 to the original alignment specs.
 
  There is a document which gives some of it, but doing a
 reverse-engineering project is primarily so it can be properly understood
 for trimming and repair purposes in my mind.
 
  There is a 2x4 connectors which can connect in a test-rig while the 2x5
 connector is hooked up. There is two different connectors inside that also
 should be relevant, and then is at least one jumper field which would be
 good to know how it works.
 
  Considering how common these are, I was surprised that more effort had
 not gone into them, compared to how the group (and hams) tend to deal with
 all this.
 
  Cheers,
  Magnus
 
  Bob
 
  On Feb 16, 2013, at 6:10 PM, Magnus Danielson
 mag...@rubidium.dyndns.org  wrote:
 
  Fellow time-nuts,
 
  Considering that LPROs is pretty popular, I am a bit surprised that I
 have not seen any major reverse-engineering effort on the LPROs. I have the
 self-compiled LPRO service document, which collects parts of schematics
 from patents, but still.
 
  My main reasons for asking is that I want to get a little better
 overview of how they work, how I can tune them up and what signals is
 available where. Naturally, always figuring if there is some interesting
 tweaks to be done.
 
  LRPO is just a traditional analogue rubidium, in compact format, sure,
 but never the less.
 
  I have noticed that different FPGAs have been used over time. Curious
 about the various jumpers and connectors in it.
 
  Cheers,
  Magnus
  ___
  time-nuts mailing list -- time-nuts@febo.com
  To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
  and follow the instructions there.
 
  ___
  time-nuts mailing list -- time-nuts@febo.com
  To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
  and follow the instructions there.
 
  ___
  time-nuts mailing list -- time-nuts@febo.com
  To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
  and follow the instructions there.

 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FRK-L Rubidium

2013-02-18 Thread Garren Davis

Bob,

Just wanted to let you know your advice to let the FRK run for a while was spot 
on.
The lock voltage is down to 9.2 volts.

I was able to get my thunderbolt with the bad oscillator oven working by 
heating the
oscillator with a power resistor. After it locked I could see on an 
oscilloscope that
it and the FRK were both exactly the same. I plan on putting the thunderbolt 
oscillator
in an oven I'm building that will have proportional control and see how that 
works.

Thanks for the help.
Garren


-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On Behalf 
Of Bob Camp
Sent: Saturday, February 09, 2013 10:14 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] FRK-L Rubidium

Hi

Let it run for a couple of days. If it's still up at   11V, I'd tweak it down 
a bit.

Bob

On Feb 9, 2013, at 12:38 PM, Garren Davis garren.da...@qlogic.com wrote:

 Found my problem with the FRK. R31 on the Osc board was burned and open. This 
 was caused by a shorted C16. Replaced and it is now locked. The lock voltage 
 is 12.7v. Is this good or should it be lower?

 Garren




 On Feb 9, 2013, at 6:10 AM, Bob Camp li...@rtty.us wrote:

 Hi

 How old is the FRK? Does it look like it's been run without a heat sink for 
 very long? They tend to get flaky if run for a while (many months) without 
 heat sinking. There's nothing mysterious about it. The MTBF of the parts 
 gets noticeably worse as the unit heats up.

 Bob

 On Feb 9, 2013, at 1:09 AM, Garren Davis garren.da...@qlogic.com wrote:


 Well for some reason the 10 Mhz stopped working on the FRK. Don't know why. 
 Started up the thunderbolt. It acquired satellites but then the DAC voltage 
 went to -5 volts. It's been there for an hour. Will this change after the 
 unit stabilizes? Going to bed. Will check it tomorrow.

 Garren




 On Feb 8, 2013, at 8:25 PM, Ed Palmer ed_pal...@sasktel.net wrote:

 Hi Garren,

 I suggest that you get the Thunderbolt working first.  Without a known 10 
 MHz source to compare to, you're flying blind.  Once the Tbolt is running, 
 you should be able to check the frequency of the FRK by feeding both into 
 your scope.  Trigger on the Tbolt and watch what the FRK does.  You should 
 see the trace scrolling in one direction, then slow down, then stop, then 
 scroll the other direction.  The 'stop' point is at 10 MHz.  The frequency 
 sweeps a total of 20-30 Hz so it's easy to see.  If you don't see the 
 'stop' point, the FRK isn't getting to 10 MHz.  Now use the best frequency 
 counter you've got to measure the Tbolt.  Regardless of the calibration of 
 your counter, the number your counter gives you becomes your 'new' 10 MHz. 
  Now measure the FRK to see if it's running fast or slow.

 You should check the temperature of the lamp.  It's easy to get at by 
 removing the cover in the center of the heat sink.  Probably best to 
 remove the cover and then power down before you go poking around inside!  
 The temperature of the cavity is also important, but getting to it is more 
 of a hassle - don't go there if you don't have to.

 Of course, check for the normal things like internal power supply 
 voltages, ripple, current drain (both initial and steady-state), etc.

 Regarding your second message, yes, the adjustment under the heat sink 
 near the edge is the C-field.  That won't help you at this point.  The 
 adjustment in the center of one side is the VCO.  You could try adjusting 
 it, but like I said, you're flying blind at this point.  You won't know if 
 you're adjusting closer to 10 MHz or further away.

 Good luck,

 Ed


 On 2/8/2013 6:12 PM, Garren Davis wrote:
 Been lurking on the list for a while and finally started playing
 with a FRK-L  rubidium frequency standard. I've had this thing for
 a while and decided to power it up and see what it would do. I do
 not get a lock. What I see is the lamp voltage at 8.54 volts which
 I think is good but the xtal control voltage swings from 2 volts
 to 15 volts and back to 2 volts and keeps cycling like that. I
 don't have a good frequency counter but I have a 3 Ghz 40 G/sample
 scope and it shows that the 10 MHz signal is there. I just don't
 know how accurate it is. Has anyone seen a problem like this? Can
 anyone point me to a place to start debugging this? I have the
 schematics and test tools. I am a test engineer so I'm not afraid
 to poke around in the guts of this thing. Hopefully I can get this
 thing running. I also have a thunderbolt that I'll get running
 this weekend. I don't know how deep I'll get into this time-nuts
 thing but I have this nice scope and a Wavecrest sitting in my
 garage

 a
 n
 d
 I'd lik
 e to put them to use. Any help would be appreciated.

 Garren

 ___
 time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.


 

[time-nuts] OCXO DC Current Question

2013-02-18 Thread Ed Palmer
I know that when making AC measurements on various OCXOs of the same 
type, you have to expect wide variations in the results. e.g.  TVB's 
Allan Deviation measurements on a selection of 10811A oscillators at  
http://www.leapsecond.com/pages/z3801a-osc .  But what about DC current 
measurements?  How much variability should you expect?


I recently bought 4 MTI 260 oscillators with thoughts of doing some 
3-cornered hat experiments.  I thought I'd use the best 3 of 4.  One 
test I always do on an OCXO is to measure the DC current drain as it 
warms up.  Nothing radical - I have an HP 6622A GPIB-equipped linear 
power supply.  I just do GPIB queries as fast as I can and log the 
results.  I get about 6 readings per second.  More than enough for my needs.


This time, I was surprised by the results of this test.  The attached 
picture shows why.  I've offset the traces horizontally and vertically 
for clarity so I deleted the axes.  The horizontal lines are 200 ma 
apart, but the position of each trace is arbitrary.  All four 
oscillators start at a current-limited value of ~1 Amp and have a 
steady-state current drain of ~230 mA.  The length of the graph is ~20 
minutes.


Although the family resemblance is obvious, I was surprised by the 
different noise levels.  I let one of the noisy units  run for a day to 
see if it would settle down, but there was no improvement.  Are these 
results reasonable, or do I have one oscillator with a good oven (blue 
trace), one marginal (purple), and two rather poor ones (red and 
green)?  I'm thinking that the noise on the oven could affect the Allan 
Deviation due to either or both of the thermal inconsistencies or 
varying load on the power supply.


Any thoughts?

Thanks,
Ed

attachment: image001.png___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

[time-nuts] Allan deviation and modified Allan deviation

2013-02-18 Thread cdelect
Ok I've been plotting my oscillators for years using Allan Deviation.

That way all my records can be compared easily.

Is there any advantage to using Modified Allan Deviation?

It seems to give a better stability plot but that just seems to be
cheating!

If I plot both ways what do the differences mean?

Thanks,

Corby
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] OCXO DC Current Question

2013-02-18 Thread Chuck Harris

It's not really noise, the feedback loop is marginally
stable.  What you are seeing is how the stability changes
due to minor differences in the parts used in the loop,
and the thermal masses and insulation of the oven.

-Chuck Harris

Ed Palmer wrote:

I know that when making AC measurements on various OCXOs of the same type, you 
have
to expect wide variations in the results. e.g.  TVB's Allan Deviation 
measurements on
a selection of 10811A oscillators at http://www.leapsecond.com/pages/z3801a-osc 
.
But what about DC current measurements?  How much variability should you expect?

I recently bought 4 MTI 260 oscillators with thoughts of doing some 3-cornered 
hat
experiments.  I thought I'd use the best 3 of 4.  One test I always do on an 
OCXO is
to measure the DC current drain as it warms up.  Nothing radical - I have an HP 
6622A
GPIB-equipped linear power supply.  I just do GPIB queries as fast as I can and 
log
the results.  I get about 6 readings per second.  More than enough for my needs.

This time, I was surprised by the results of this test.  The attached picture 
shows
why.  I've offset the traces horizontally and vertically for clarity so I 
deleted the
axes.  The horizontal lines are 200 ma apart, but the position of each trace is
arbitrary.  All four oscillators start at a current-limited value of ~1 Amp and 
have
a steady-state current drain of ~230 mA.  The length of the graph is ~20 
minutes.

Although the family resemblance is obvious, I was surprised by the different 
noise
levels.  I let one of the noisy units  run for a day to see if it would settle 
down,
but there was no improvement.  Are these results reasonable, or do I have one
oscillator with a good oven (blue trace), one marginal (purple), and two rather 
poor
ones (red and green)?  I'm thinking that the noise on the oven could affect the 
Allan
Deviation due to either or both of the thermal inconsistencies or varying load 
on the
power supply.

Any thoughts?

Thanks,
Ed

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] OCXO DC Current Question

2013-02-18 Thread Bob Camp
Hi

Like any control loop, gain, bandwidth and noise are all related. In a DOCXO 
you have two control loops and they do interact. That said, there's nothing 
grossly wrong with the four OCXO's. The noisy parts have a bit more gain in the 
controller. The quiet parts have a bit less gain. The easy way to see this is 
the ringing after the current changes. 

Without ADEV numbers there's no way to know which one is good (or better) and 
which one is bad (or worse). The noisy parts may be responding to the ambient 
temperature rumble (thus correcting for it) and the quiet ones may be ignoring 
it (allowing it to hit the crystal). It's also possible that the noisy ones 
have an electrical issue in the loop that generates the noise. 

If all the ovens started from the same temperature, there is a variance in the 
oven set points. Some take longer to warm up than others. You don't mention if 
they started the same, so that may or may not be significant. 

To further complicate things, in a double oven, you can have a noisy outer oven 
that gets suppressed by the inner oven. Are your specific 260's double ovens? 
No way to be sure without tearing one open. The second step in the current 
plot could be an inner oven cutting back.

Bob


On Feb 18, 2013, at 2:40 PM, Ed Palmer ed_pal...@sasktel.net wrote:

 I know that when making AC measurements on various OCXOs of the same type, 
 you have to expect wide variations in the results. e.g.  TVB's Allan 
 Deviation measurements on a selection of 10811A oscillators at  
 http://www.leapsecond.com/pages/z3801a-osc .  But what about DC current 
 measurements?  How much variability should you expect?
 
 I recently bought 4 MTI 260 oscillators with thoughts of doing some 
 3-cornered hat experiments.  I thought I'd use the best 3 of 4.  One test I 
 always do on an OCXO is to measure the DC current drain as it warms up.  
 Nothing radical - I have an HP 6622A GPIB-equipped linear power supply.  I 
 just do GPIB queries as fast as I can and log the results.  I get about 6 
 readings per second.  More than enough for my needs.
 
 This time, I was surprised by the results of this test.  The attached picture 
 shows why.  I've offset the traces horizontally and vertically for clarity so 
 I deleted the axes.  The horizontal lines are 200 ma apart, but the position 
 of each trace is arbitrary.  All four oscillators start at a current-limited 
 value of ~1 Amp and have a steady-state current drain of ~230 mA.  The length 
 of the graph is ~20 minutes.
 
 Although the family resemblance is obvious, I was surprised by the different 
 noise levels.  I let one of the noisy units  run for a day to see if it would 
 settle down, but there was no improvement.  Are these results reasonable, or 
 do I have one oscillator with a good oven (blue trace), one marginal 
 (purple), and two rather poor ones (red and green)?  I'm thinking that the 
 noise on the oven could affect the Allan Deviation due to either or both of 
 the thermal inconsistencies or varying load on the power supply.
 
 Any thoughts?
 
 Thanks,
 Ed
 
 image001.png___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] FE-5680A DDS Board/PIC Code

2013-02-18 Thread Herbert Poetzl

I'm new to the time-nuts community, so I simply start 
with a short info on how I got into this situation :)
(skip forward to ONTOPIC if not interested)

Not long ago, I decided to build a reasonably good
frequency counter for my personal use and maybe if
the result is simple and elegant, I'll publish the
details so that everybody can build one ...

It was clear to me, that it had to be able to count
up to at least 1GHz and thus show at least nine, 
better ten significant digits, so a precise time base
is required.

After some online searches and investigations, my
best options seemed to get a very stable oscillator
and a high quality time reference to sync with, which 
in turn brought me to the idea to use a cheap rubidium 
normal and somehow tune/measure/sync it via GPS or 
DCF-77/MSF-60.

Reading a lot of documentation and blogs from all over 
the world (sometimes in translation :) shed some light 
on the rubidium normal requirements, which I defined as:

 - has to have a 10MHz output not just the 1PPS
 - has to be programmable (i.e. can be tuned)
 - must be cheap

I quickly found two different RB standard models, 
readily available on ebay for a reasonable price, 
namely the Efratom FRS-C and the FEI FE-5680A.
I finally decided to go with the FE-5680A, mainly
because I liked the package. A seller was quickly
found offering something titled:

 'FE-5680A Rubidium Atomic Frequency Standard 
  Oscillator Transceivers 10Mhz Out'
 'Programmable from 1Hz to 20MHz'

Little did I know what that actually meant ...

When the units (I ordered two of them) arrived, I
couldn't wait to test if they actually work and get
a lock, so I quickly wired them up (according to the
pinout) and provided them with the advised 15V at
up to 2A each. To my astonishment, they heated up
rather quickly and got a lock in a little under two
minutes, so I happily got my scope out to check the
10MHz signal, just to find that there is no such
signal available on the 9pin D-sub connector.

Measuring pins against ground (pin 2) and 15Vx (pin 1)
I figured that neither pin 7 (10MHz) nor pin 8/9
(the rs232 interface for programming) was connected.
and to my great disappointment, pin 6 (1PPS) didn't 
output much either (I later discovered that this was
due to a defective unit, which is now being replaced)
 
After contacting the seller, I opened up the units
to investigate my options (and of course, because
I wanted to take a look inside :), which in turn led
to a number of high resolution scans and photos of
all the bits and pieces.

A (this time) more thorough search on the internet
resulted in a deeper understanding of the various 
options the FE-5680A can have (or usually doesn't 
have) and the inner workings of the different 
FE-5680A models (of course, all labeled FE-5680A :)

The DDS board, which actually can be programmed to
output certain frequencies derived from the 'locked'
1:136 frequency of the rubidium 6.8GHz transition,
caught my attention, as it has both, the '10Mhz'
output and the programming interface, so I decided
to analyze it further ...

ONTOPIC

The central part on this specific DDS board [1] is 
the AD9830A a Direct Digital Synthesizer (DDS) which
basically produces a sine wave at a well defined
multiple and phase of a given reference frequency.

Besides some other components, this board also 
includes an RS-232C line driver (Sipex SP233A) a
PIC16F84 microcontroller and two 74HC595 8bit shift
registers, with buffered outputs.

I read somewhere, that the blue buttons on that DDS
board can be used to adjust the output frequency,
this should be avoided, mainly because every button
press is an update and will cause a write to the
EEPROM data wearing it out.

Now as I've played with PIC microcontrollers for
a long time, I wanted to know what this specific
controller is doing and how I could use that for my
purposes ...

The chip was quickly removed and the program as well
as configuration memory retrieved (luckily FEI didn't
utilize the code/data protection) and together with 
high resolution scans and photos, a documented and 
verified assembler listing [2] reverse engineered.

Here are the (IMHO) quite interesting findings:

 - both FREQx registers can be adjusted
 - the PHASE0 register can be adjusted
 - none of the changes is permanent,
   unless you explicitely save the settings
 - there are only a few commands, without
   any plausibility checks and/or protection
 - and yes, the buttons increment/decrement
   the FREQx settings and trigger a write to
   the EEPROM after every update.
 - the serial interface is done in software
 - the DDS control words are shifted into
   the 74HC595, buffered and written 


; SCR STATUS 
;   R=50255057.012932Hz F=2ABB5040
;   OK
;
; F=CRFREQxREG (set divider)
;   OK
;
; G=CRPHASE (set phase register)
;   OK
;
; R=YYCR  RUBIDIUM (set calibrated freq)
;   OK
;   
; 

Re: [time-nuts] FE-5680A DDS Board/PIC Code

2013-02-18 Thread paul swed
Herbert
Thanks for the assembly listing. Someone else on time-nuts had done quite
the job of reverse engineering the schematics and other information. Seems
like the two of you could collaborate. Sorry do not recall the name.
Welcome to time-nuts.
Regards
Paul
WB8TSL

On Mon, Feb 18, 2013 at 4:32 PM, Herbert Poetzl herb...@13thfloor.atwrote:


 I'm new to the time-nuts community, so I simply start
 with a short info on how I got into this situation :)
 (skip forward to ONTOPIC if not interested)

 Not long ago, I decided to build a reasonably good
 frequency counter for my personal use and maybe if
 the result is simple and elegant, I'll publish the
 details so that everybody can build one ...

 It was clear to me, that it had to be able to count
 up to at least 1GHz and thus show at least nine,
 better ten significant digits, so a precise time base
 is required.

 After some online searches and investigations, my
 best options seemed to get a very stable oscillator
 and a high quality time reference to sync with, which
 in turn brought me to the idea to use a cheap rubidium
 normal and somehow tune/measure/sync it via GPS or
 DCF-77/MSF-60.

 Reading a lot of documentation and blogs from all over
 the world (sometimes in translation :) shed some light
 on the rubidium normal requirements, which I defined as:

  - has to have a 10MHz output not just the 1PPS
  - has to be programmable (i.e. can be tuned)
  - must be cheap

 I quickly found two different RB standard models,
 readily available on ebay for a reasonable price,
 namely the Efratom FRS-C and the FEI FE-5680A.
 I finally decided to go with the FE-5680A, mainly
 because I liked the package. A seller was quickly
 found offering something titled:

  'FE-5680A Rubidium Atomic Frequency Standard
   Oscillator Transceivers 10Mhz Out'
  'Programmable from 1Hz to 20MHz'

 Little did I know what that actually meant ...

 When the units (I ordered two of them) arrived, I
 couldn't wait to test if they actually work and get
 a lock, so I quickly wired them up (according to the
 pinout) and provided them with the advised 15V at
 up to 2A each. To my astonishment, they heated up
 rather quickly and got a lock in a little under two
 minutes, so I happily got my scope out to check the
 10MHz signal, just to find that there is no such
 signal available on the 9pin D-sub connector.

 Measuring pins against ground (pin 2) and 15Vx (pin 1)
 I figured that neither pin 7 (10MHz) nor pin 8/9
 (the rs232 interface for programming) was connected.
 and to my great disappointment, pin 6 (1PPS) didn't
 output much either (I later discovered that this was
 due to a defective unit, which is now being replaced)

 After contacting the seller, I opened up the units
 to investigate my options (and of course, because
 I wanted to take a look inside :), which in turn led
 to a number of high resolution scans and photos of
 all the bits and pieces.

 A (this time) more thorough search on the internet
 resulted in a deeper understanding of the various
 options the FE-5680A can have (or usually doesn't
 have) and the inner workings of the different
 FE-5680A models (of course, all labeled FE-5680A :)

 The DDS board, which actually can be programmed to
 output certain frequencies derived from the 'locked'
 1:136 frequency of the rubidium 6.8GHz transition,
 caught my attention, as it has both, the '10Mhz'
 output and the programming interface, so I decided
 to analyze it further ...

 ONTOPIC

 The central part on this specific DDS board [1] is
 the AD9830A a Direct Digital Synthesizer (DDS) which
 basically produces a sine wave at a well defined
 multiple and phase of a given reference frequency.

 Besides some other components, this board also
 includes an RS-232C line driver (Sipex SP233A) a
 PIC16F84 microcontroller and two 74HC595 8bit shift
 registers, with buffered outputs.

 I read somewhere, that the blue buttons on that DDS
 board can be used to adjust the output frequency,
 this should be avoided, mainly because every button
 press is an update and will cause a write to the
 EEPROM data wearing it out.

 Now as I've played with PIC microcontrollers for
 a long time, I wanted to know what this specific
 controller is doing and how I could use that for my
 purposes ...

 The chip was quickly removed and the program as well
 as configuration memory retrieved (luckily FEI didn't
 utilize the code/data protection) and together with
 high resolution scans and photos, a documented and
 verified assembler listing [2] reverse engineered.

 Here are the (IMHO) quite interesting findings:

  - both FREQx registers can be adjusted
  - the PHASE0 register can be adjusted
  - none of the changes is permanent,
unless you explicitely save the settings
  - there are only a few commands, without
any plausibility checks and/or protection
  - and yes, the buttons increment/decrement
the FREQx settings and trigger a write to
the EEPROM after every update.
  - the serial 

Re: [time-nuts] FE-5680A DDS Board/PIC Code

2013-02-18 Thread Bob Camp
Hi

There's quite a bit less in the PIC code than I would have expected. If that's 
everything that's there, the PIC does very little heavy lifting. Nice job on 
the dis-assembly. 

Many of the 10 MHz out FE Rb's are actually the 1 pps version that has been 
converted to 10 MHz after the fact. The 10 MHz is not very clean on the FE's in 
any case. 

Bob

On Feb 18, 2013, at 4:32 PM, Herbert Poetzl herb...@13thfloor.at wrote:

 
 I'm new to the time-nuts community, so I simply start 
 with a short info on how I got into this situation :)
 (skip forward to ONTOPIC if not interested)
 
 Not long ago, I decided to build a reasonably good
 frequency counter for my personal use and maybe if
 the result is simple and elegant, I'll publish the
 details so that everybody can build one ...
 
 It was clear to me, that it had to be able to count
 up to at least 1GHz and thus show at least nine, 
 better ten significant digits, so a precise time base
 is required.
 
 After some online searches and investigations, my
 best options seemed to get a very stable oscillator
 and a high quality time reference to sync with, which 
 in turn brought me to the idea to use a cheap rubidium 
 normal and somehow tune/measure/sync it via GPS or 
 DCF-77/MSF-60.
 
 Reading a lot of documentation and blogs from all over 
 the world (sometimes in translation :) shed some light 
 on the rubidium normal requirements, which I defined as:
 
 - has to have a 10MHz output not just the 1PPS
 - has to be programmable (i.e. can be tuned)
 - must be cheap
 
 I quickly found two different RB standard models, 
 readily available on ebay for a reasonable price, 
 namely the Efratom FRS-C and the FEI FE-5680A.
 I finally decided to go with the FE-5680A, mainly
 because I liked the package. A seller was quickly
 found offering something titled:
 
 'FE-5680A Rubidium Atomic Frequency Standard 
  Oscillator Transceivers 10Mhz Out'
 'Programmable from 1Hz to 20MHz'
 
 Little did I know what that actually meant ...
 
 When the units (I ordered two of them) arrived, I
 couldn't wait to test if they actually work and get
 a lock, so I quickly wired them up (according to the
 pinout) and provided them with the advised 15V at
 up to 2A each. To my astonishment, they heated up
 rather quickly and got a lock in a little under two
 minutes, so I happily got my scope out to check the
 10MHz signal, just to find that there is no such
 signal available on the 9pin D-sub connector.
 
 Measuring pins against ground (pin 2) and 15Vx (pin 1)
 I figured that neither pin 7 (10MHz) nor pin 8/9
 (the rs232 interface for programming) was connected.
 and to my great disappointment, pin 6 (1PPS) didn't 
 output much either (I later discovered that this was
 due to a defective unit, which is now being replaced)
 
 After contacting the seller, I opened up the units
 to investigate my options (and of course, because
 I wanted to take a look inside :), which in turn led
 to a number of high resolution scans and photos of
 all the bits and pieces.
 
 A (this time) more thorough search on the internet
 resulted in a deeper understanding of the various 
 options the FE-5680A can have (or usually doesn't 
 have) and the inner workings of the different 
 FE-5680A models (of course, all labeled FE-5680A :)
 
 The DDS board, which actually can be programmed to
 output certain frequencies derived from the 'locked'
 1:136 frequency of the rubidium 6.8GHz transition,
 caught my attention, as it has both, the '10Mhz'
 output and the programming interface, so I decided
 to analyze it further ...
 
 ONTOPIC
 
 The central part on this specific DDS board [1] is 
 the AD9830A a Direct Digital Synthesizer (DDS) which
 basically produces a sine wave at a well defined
 multiple and phase of a given reference frequency.
 
 Besides some other components, this board also 
 includes an RS-232C line driver (Sipex SP233A) a
 PIC16F84 microcontroller and two 74HC595 8bit shift
 registers, with buffered outputs.
 
 I read somewhere, that the blue buttons on that DDS
 board can be used to adjust the output frequency,
 this should be avoided, mainly because every button
 press is an update and will cause a write to the
 EEPROM data wearing it out.
 
 Now as I've played with PIC microcontrollers for
 a long time, I wanted to know what this specific
 controller is doing and how I could use that for my
 purposes ...
 
 The chip was quickly removed and the program as well
 as configuration memory retrieved (luckily FEI didn't
 utilize the code/data protection) and together with 
 high resolution scans and photos, a documented and 
 verified assembler listing [2] reverse engineered.
 
 Here are the (IMHO) quite interesting findings:
 
 - both FREQx registers can be adjusted
 - the PHASE0 register can be adjusted
 - none of the changes is permanent,
   unless you explicitely save the settings
 - there are only a few commands, without
   any plausibility checks and/or protection
 - and yes, the buttons 

Re: [time-nuts] OCXO DC Current Question

2013-02-18 Thread Ed Palmer

Hi Bob,

On 2/18/2013 2:42 PM, Bob Camp wrote:

Hi

Like any control loop, gain, bandwidth and noise are all related. In a DOCXO 
you have two control loops and they do interact. That said, there's nothing 
grossly wrong with the four OCXO's. The noisy parts have a bit more gain in the 
controller. The quiet parts have a bit less gain. The easy way to see this is 
the ringing after the current changes.


Yes, when I zoom in on the ringing after the first step change, they all 
show a period of ~9 seconds.  The blue trace (the quiet one) has a nice 
smooth damped sine wave while the other ones have varying amounts of 
noise superimposed on the sine wave.  The amplitude of the ringing on 
the noisy ones is greater than the amplitude of the quiet one.  In one 
case, more than twice as big.



Without ADEV numbers there's no way to know which one is good (or better) and 
which one is bad (or worse). The noisy parts may be responding to the ambient 
temperature rumble (thus correcting for it) and the quiet ones may be ignoring 
it (allowing it to hit the crystal).


I hadn't thought of that possibility.  Thanks for that!


It's also possible that the noisy ones have an electrical issue in the loop 
that generates the noise.


I was thinking of maybe a dead filter capacitor - hence the noise.


If all the ovens started from the same temperature, there is a variance in the 
oven set points. Some take longer to warm up than others. You don't mention if 
they started the same, so that may or may not be significant.


They all started at the same temperature.  Any apparent difference in 
warmup time is mostly due to the arbitrary offset of the traces. The 
offset was just so you could see all four traces clearly. However, isn't 
it also likely that each oscillator has had it's oven tweaked to match 
that particular crystal?



To further complicate things, in a double oven, you can have a noisy outer oven that gets 
suppressed by the inner oven. Are your specific 260's double ovens? No way to be sure 
without tearing one open. The second step in the current plot could be an 
inner oven cutting back.


I wondered about that second step.  The period of the ringing on the 
second step is about 1 sec. longer than the ringing on the first step so 
it's apparently a different circuit.  The 260 series datasheet does NOT 
say anything about it being a double oven.  They talk about customizing 
the unit to suit the customer, but that seems a bit extreme.


By the way, I purchased these oscillators from our favorite auction 
site.  If anyone's interested, the 260 series includes both AT and SC 
crystals.  Since these start out ~150 Hz low, they appear to be SC cut.  
The output is a 5 MHz sine wave @ ~+7 dBm into 50 ohms. They work fine 
with a 12V power supply.  They have no mechanical frequency adjustment 
(unless it's under a suspicious spot of solder on the case)  but my 4 
were all adjustable to 5 MHz via EFC.  They were apparently removed from 
Z3811 GPS receivers.  They each have a sticker on the side that says 
Z3811-80010. There doesn't seem to be much info available on the 
Z3811.  I have no relationship to the vendor - just a customer.


So it sounds like the proper thing to do is file this information and 
carry on.  After making some Allan Deviation measurements, review 
everything to see how, or if, oven 'noise' correlates to Allan Deviation 
results.


Thanks Bob,

Ed


Bob


On Feb 18, 2013, at 2:40 PM, Ed Palmer ed_pal...@sasktel.net wrote:


I know that when making AC measurements on various OCXOs of the same type, you 
have to expect wide variations in the results. e.g.  TVB's Allan Deviation 
measurements on a selection of 10811A oscillators at  
http://www.leapsecond.com/pages/z3801a-osc .  But what about DC current 
measurements?  How much variability should you expect?

I recently bought 4 MTI 260 oscillators with thoughts of doing some 3-cornered 
hat experiments.  I thought I'd use the best 3 of 4.  One test I always do on 
an OCXO is to measure the DC current drain as it warms up.  Nothing radical - I 
have an HP 6622A GPIB-equipped linear power supply.  I just do GPIB queries as 
fast as I can and log the results.  I get about 6 readings per second.  More 
than enough for my needs.

This time, I was surprised by the results of this test.  The attached picture 
shows why.  I've offset the traces horizontally and vertically for clarity so I 
deleted the axes.  The horizontal lines are 200 ma apart, but the position of 
each trace is arbitrary.  All four oscillators start at a current-limited value 
of ~1 Amp and have a steady-state current drain of ~230 mA.  The length of the 
graph is ~20 minutes.

Although the family resemblance is obvious, I was surprised by the different 
noise levels.  I let one of the noisy units  run for a day to see if it would 
settle down, but there was no improvement.  Are these results reasonable, or do 
I have one oscillator with a good oven (blue trace), one marginal 

Re: [time-nuts] FE-5680A DDS Board/PIC Code

2013-02-18 Thread GandalfG8
In a message dated 18/02/2013 22:45:12 GMT Standard Time,  
paulsw...@gmail.com writes:

Herbert
Thanks for the assembly listing. Someone else on  time-nuts had done quite
the job of reverse engineering the schematics and  other information. Seems
like the two of you could collaborate. Sorry do  not recall the name.
Welcome to  time-nuts.
Regards
Paul
WB8TSL

 
Was just thinking the same thing so checked through my archives.
 

That was Elio, _eliocor@gmail.com_ (mailto:elio...@gmail.com) ,  messages 
posted here about a year ago.
 

The messages will be in the list archives anyway but below is the text of  
two messages I saved.
 

Regards
 

Nigel
GM8PZR

--
 
From:  elio...@gmail.com
Reply-to: time-nuts@febo.com
To:  time-nuts@febo.com
Sent: 15/02/2012 01:31:32 GMT Standard Time
Subj:  [time-nuts] FE-5680A Schematics (v0.1)


at the following  address:

http://www.rhodiatoce.com/pics/time-nuts/FE-5680A/FE-5680A_schematics_v0.1.p
df

you  will find the 0.1 release of the FE-5680A schematics.

New additions:
-  pinout of the unpopulated (24 pin) connector (J9) behind the DB9
-  connections between CPU/MAX3232 and the *TWO* serial ports!!!
- component  values of the 10MHz filter + option on PCB to output a square
10MHz wave on  DB9
- connections between CPU and MAX1246 (A/D converter)
- connections  between XC9572 and MAX392 (quad analog switch)
- unknown pins marked as  '?'

As you can see, it seems FE-5680A fully supports 2 serial  ports:
one on DB9 and the other one on the unpopulated connector (J9) behind  DB9.

Any comments/suggestions are welcome.

.   ciao
_Elio.
_
 

 
From:  elio...@gmail.com
Reply-to: time-nuts@febo.com
To:  time-nuts@febo.com
Sent: 24/02/2012 01:03:56 GMT Standard Time
Subj:  [time-nuts] FE-5680A Schematics and scans


Thanks to Ignacio (EB4APL) and the generosity of Mike  Harrison, today I
received a PCB of a disassembled FE5680A.
I provided to  make some scans of the board at 2400DPI resolution: the
picture size is about  7700x11500 pixel (9MB)

Top face (with and without  ICs):
http://www.rhodiatoce.com/pics/time-nuts/FE-5680A/FE-5680A_Top001.jpg
http://www.rhodiatoce.com/pics/time-nuts/FE-5680A/FE-5680A_Top001_noIC.jpg

Bottom  face (with and without  ICs):
http://www.rhodiatoce.com/pics/time-nuts/FE-5680A/FE-5680A_Bottom001.jpg

http://www.rhodiatoce.com/pics/time-nuts/FE-5680A/FE-5680A_Bottom001_noIC.jp
g


Being  able to remove the ICs, I was able to correct some flaws in the
previous  schematics:

http://www.rhodiatoce.com/pics/time-nuts/FE-5680A/FE-5680A_schematics_v0.2.p
df


During  the next week I will continue to write down the logical part of the
circuit  and I hope will be able to dump the 8032 firmware

_  Elio.
_

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Allan deviation and modified Allan deviation

2013-02-18 Thread Tom Van Baak (lab)
I often plot both ADEV and MDEV, too. The difference between them indicates the 
amount of white phase noise there is in the measurement (the sum of DUT and REF 
and phase meter). You'll see for some standards, especially at longer tau, 
there is little or no difference. 

Right, MDEV is almost always less, since it artificially removes noise. Of 
course the noise is still there in the electrons, but it is not reported by 
this statistic. However, the sometimes legitimate reason for doing this is 
two-fold:

First, an MDEV plot allows one to distinguish white from flicker phase noise 
types; something plain ADEV cannot do (both noise types show up as slope -1 
with ADEV). As such, MDEV is a more powerful noise diagnostic than ADEV. But 
you can't necessarily treat the numbers as true, live performance. In order for 
MDEV to do its magic it must (mathematically) remove noise that you know is 
there.

Second, there are some cases where the user plans to use the oscillator in a 
system that removes white noise by averaging against an equal or better 
standard. In this case it's nice to know ahead of time how much white noise 
there is to remove; so using MDEV is more informative than ADEV.

For example, if you plan to compare two cesium time standards by collecting 
long-term data, then post-averaging, and plotting time stability once an hour 
or day, then MDEV is fair game. But if you plan to use the standard for any 
sort of real-time system then ADEV comes closer to representing the actual 
frequency stability that you will see.

MDEV tells you how little instability you would have left after you squeeze out 
all the white noise. Every orange has juice inside, but to measure juice or 
pulp you destroy the original orange.

Note a similar thing happens with ADEV and HDEV. If you plan to use an 
oscillator in a system that methodically removes linear frequency drift (like a 
smart GPSDO) then HDEV is a better (cheaper) indicator than ADEV.

In summary, use the measurement statistic (and bandwidth) in the lab that best 
matches how the oscillator will actually be used in real life. 
 
I too would like to compare records made over many years. One problem is that 
phase meters can have different bandwidths and so, especially at short term, 
it's hard to know how much of the reported [in]stability is due to the DUT or 
REF or the filtering before, in, or after the phase detector itself. Most 
meters have a tau- or noise- dependent bias.

/tvb (iPhone4) SF

On Feb 18, 2013, at 11:59 AM, cdel...@juno.com wrote:

 Ok I've been plotting my oscillators for years using Allan Deviation.
 
 That way all my records can be compared easily.
 
 Is there any advantage to using Modified Allan Deviation?
 
 It seems to give a better stability plot but that just seems to be
 cheating!
 
 If I plot both ways what do the differences mean?
 
 Thanks,
 
 Corby
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] OCXO DC Current Question

2013-02-18 Thread Bob Camp
Hi

The -Z3811 is indeed a part out of a HP GPSDO. HP is a big enough customer that 
the part they get *may* have little relation to the standard part.

If they all started out at the same temperature, then the noisy ones finish 
warmup quicker than the quiet ones. Take a look at when the final glitch 
happens in the warmup process. It's definitely happening earlier on the noisy 
parts. If warmup finishes quicker - the part is running at a lower temperature. 
That of course is only the time between the cutback from full current to the 
glitch. It may not mean it's running cooler.

Bob

On Feb 18, 2013, at 5:32 PM, Ed Palmer ed_pal...@sasktel.net wrote:

 Hi Bob,
 
 On 2/18/2013 2:42 PM, Bob Camp wrote:
 Hi
 
 Like any control loop, gain, bandwidth and noise are all related. In a DOCXO 
 you have two control loops and they do interact. That said, there's nothing 
 grossly wrong with the four OCXO's. The noisy parts have a bit more gain in 
 the controller. The quiet parts have a bit less gain. The easy way to see 
 this is the ringing after the current changes.
 
 Yes, when I zoom in on the ringing after the first step change, they all show 
 a period of ~9 seconds.  The blue trace (the quiet one) has a nice smooth 
 damped sine wave while the other ones have varying amounts of noise 
 superimposed on the sine wave.  The amplitude of the ringing on the noisy 
 ones is greater than the amplitude of the quiet one.  In one case, more than 
 twice as big.
 
 Without ADEV numbers there's no way to know which one is good (or better) 
 and which one is bad (or worse). The noisy parts may be responding to the 
 ambient temperature rumble (thus correcting for it) and the quiet ones may 
 be ignoring it (allowing it to hit the crystal).
 
 I hadn't thought of that possibility.  Thanks for that!
 
 It's also possible that the noisy ones have an electrical issue in the loop 
 that generates the noise.
 
 I was thinking of maybe a dead filter capacitor - hence the noise.
 
 If all the ovens started from the same temperature, there is a variance in 
 the oven set points. Some take longer to warm up than others. You don't 
 mention if they started the same, so that may or may not be significant.
 
 They all started at the same temperature.  Any apparent difference in warmup 
 time is mostly due to the arbitrary offset of the traces. The offset was just 
 so you could see all four traces clearly. However, isn't it also likely that 
 each oscillator has had it's oven tweaked to match that particular crystal?
 
 To further complicate things, in a double oven, you can have a noisy outer 
 oven that gets suppressed by the inner oven. Are your specific 260's double 
 ovens? No way to be sure without tearing one open. The second step in the 
 current plot could be an inner oven cutting back.
 
 I wondered about that second step.  The period of the ringing on the second 
 step is about 1 sec. longer than the ringing on the first step so it's 
 apparently a different circuit.  The 260 series datasheet does NOT say 
 anything about it being a double oven.  They talk about customizing the unit 
 to suit the customer, but that seems a bit extreme.
 
 By the way, I purchased these oscillators from our favorite auction site.  If 
 anyone's interested, the 260 series includes both AT and SC crystals.  Since 
 these start out ~150 Hz low, they appear to be SC cut.  The output is a 5 MHz 
 sine wave @ ~+7 dBm into 50 ohms. They work fine with a 12V power supply.  
 They have no mechanical frequency adjustment (unless it's under a suspicious 
 spot of solder on the case)  but my 4 were all adjustable to 5 MHz via EFC.  
 They were apparently removed from Z3811 GPS receivers.  They each have a 
 sticker on the side that says Z3811-80010. There doesn't seem to be much 
 info available on the Z3811.  I have no relationship to the vendor - just a 
 customer.
 
 So it sounds like the proper thing to do is file this information and carry 
 on.  After making some Allan Deviation measurements, review everything to see 
 how, or if, oven 'noise' correlates to Allan Deviation results.
 
 Thanks Bob,
 
 Ed
 
 Bob
 
 
 On Feb 18, 2013, at 2:40 PM, Ed Palmer ed_pal...@sasktel.net wrote:
 
 I know that when making AC measurements on various OCXOs of the same type, 
 you have to expect wide variations in the results. e.g.  TVB's Allan 
 Deviation measurements on a selection of 10811A oscillators at  
 http://www.leapsecond.com/pages/z3801a-osc .  But what about DC current 
 measurements?  How much variability should you expect?
 
 I recently bought 4 MTI 260 oscillators with thoughts of doing some 
 3-cornered hat experiments.  I thought I'd use the best 3 of 4.  One test I 
 always do on an OCXO is to measure the DC current drain as it warms up.  
 Nothing radical - I have an HP 6622A GPIB-equipped linear power supply.  I 
 just do GPIB queries as fast as I can and log the results.  I get about 6 
 readings per second.  More than enough for my needs.
 

Re: [time-nuts] FE-5680A DDS Board/PIC Code

2013-02-18 Thread WB6BNQ
Hi Herbert,

Nice work !  A few comments that may help you in your quest.  There are two
primary methodologies used in the FEI-5680A Rubidium.

It appears, from your picture, what you have is the older style that is analog
for control of the Rubidium frequency control loop.  A VERY GOOD THING !  Why,
because the analog C field pot is active in such a unit, permitting adjustment
allowing you to put it right on the mark.  This easily allows you to extend,
externally, the C field control and be able to put it in an external loop with
a GPS without much trouble.

In the first methodology, the original designed was, totally, an analog control.
The micro controller is used to talk to only the DDS which is outside of the
Rubidium control loop.  The Rubidium control loop starts with a 50.+ MHz
oscillator frequency multiplied up to the Rubidium and the feedback from the
photo multiplier is fed back to control the 50.+ MHz oscillator.  This 50.+ MHz
is the reference frequency for the DDS on the board in your picture.

The second methodology is a bit more complicated because FEI switched to digital
control of the Rubidium frequency loop.  Here, an internal 60. MHz 
oscillator
is used as a base signal.  A DDS, whose reference is from the same 60 MHz
oscillator generates the required offset and mixed with the 60 MHz to produced
the final multiplied signal to the Rubidium filter.  The photo multiplier is
sensed and eventually runs a DAC to control the 60 MHz oscillator.  The problem
with this approach, due to the computerization and digital stepping, is you can
get close but not necessarily right spot on for frequency.  Depending upon your
house standards, that may or may not be an issue.  To put it in perspective, I
think the adjustment process in the newer methodology is either parts in the
10E-12 or -13.

The output frequencies that the customer sees are derived via circuitry outside
the loop like the old method except they have switched to using either an ASIC 
or
FPGA of some sort making it quite difficult to determine what is going on.
Equally so is the fact that the ASIC/FPGA's are controlled by software.  Of
course, FEI is not going to be forthcoming on the internals, nor do they want to
tell how to use their computer link to control the Rubidium unless you are a
customer.

As a hobbyist, you are left with a much harder unit to integrate into external
control mechanisms.  Even though the C field pot is still there it is not
active in the unit.  Someone or more have been investigating methods to find a
proper point with which to inject an outside DC signal to attempt a C field
like control.  I do not know what their success has been.

FEI's part number system is really two different processes.  The FEI-5680A 
number
is the generic case and layout style, etc. and another order number is 
attached
that tells what is actually arranged inside.

So, the upshot is the older arrangement is easier to play with.  Hope this has
been helpful,

BillWB6BNQ



Herbert Poetzl wrote:

 I'm new to the time-nuts community, so I simply start
 with a short info on how I got into this situation :)
 (skip forward to ONTOPIC if not interested)

 Not long ago, I decided to build a reasonably good
 frequency counter for my personal use and maybe if
 the result is simple and elegant, I'll publish the
 details so that everybody can build one ...

 It was clear to me, that it had to be able to count
 up to at least 1GHz and thus show at least nine,
 better ten significant digits, so a precise time base
 is required.

 After some online searches and investigations, my
 best options seemed to get a very stable oscillator
 and a high quality time reference to sync with, which
 in turn brought me to the idea to use a cheap rubidium
 normal and somehow tune/measure/sync it via GPS or
 DCF-77/MSF-60.

 Reading a lot of documentation and blogs from all over
 the world (sometimes in translation :) shed some light
 on the rubidium normal requirements, which I defined as:

  - has to have a 10MHz output not just the 1PPS
  - has to be programmable (i.e. can be tuned)
  - must be cheap

 I quickly found two different RB standard models,
 readily available on ebay for a reasonable price,
 namely the Efratom FRS-C and the FEI FE-5680A.
 I finally decided to go with the FE-5680A, mainly
 because I liked the package. A seller was quickly
 found offering something titled:

  'FE-5680A Rubidium Atomic Frequency Standard
   Oscillator Transceivers 10Mhz Out'
  'Programmable from 1Hz to 20MHz'

 Little did I know what that actually meant ...

 When the units (I ordered two of them) arrived, I
 couldn't wait to test if they actually work and get
 a lock, so I quickly wired them up (according to the
 pinout) and provided them with the advised 15V at
 up to 2A each. To my astonishment, they heated up
 rather quickly and got a lock in a little under two
 minutes, so I happily got my scope out to check the
 10MHz signal, just to find that there is no such

Re: [time-nuts] FE-5680A DDS Board/PIC Code

2013-02-18 Thread EB4APL

Hi,

Probably Elio will jump in and tell us more on this, but I must advice 
that the unit where he did his forensic examination was a FE-5680A of 
the 10 MHz fixed frequency output  variety (not accounting for the small 
adjustment margin)and also has a fixed 1 PPS.  This unit doesn't have 
the two blue buttons and is essentially different from the 
programmable units.  Its DDS is used in a very different way and the 
PIC program should be totally different.


Regards,
Ignacio, EB4APL


On 18/02/2013 23:54, gandal...@aol.com wrote:

In a message dated 18/02/2013 22:45:12 GMT Standard Time,
paulsw...@gmail.com writes:

Herbert
Thanks for the assembly listing. Someone else on  time-nuts had done quite
the job of reverse engineering the schematics and  other information. Seems
like the two of you could collaborate. Sorry do  not recall the name.
Welcome to  time-nuts.
Regards
Paul
WB8TSL


Was just thinking the same thing so checked through my archives.


That was Elio, _eliocor@gmail.com_ (mailto:elio...@gmail.com) ,  messages
posted here about a year ago.


The messages will be in the list archives anyway but below is the text of
two messages I saved.


Regards


Nigel
GM8PZR

--

From:  elio...@gmail.com
Reply-to: time-nuts@febo.com
To:  time-nuts@febo.com
Sent: 15/02/2012 01:31:32 GMT Standard Time
Subj:  [time-nuts] FE-5680A Schematics (v0.1)


at the following  address:

http://www.rhodiatoce.com/pics/time-nuts/FE-5680A/FE-5680A_schematics_v0.1.p
df



you  will find the 0.1 release of the FE-5680A schematics.

New additions:
-  pinout of the unpopulated (24 pin) connector (J9) behind the DB9
-  connections between CPU/MAX3232 and the *TWO* serial ports!!!
- component  values of the 10MHz filter + option on PCB to output a square
10MHz wave on  DB9
- connections between CPU and MAX1246 (A/D converter)
- connections  between XC9572 and MAX392 (quad analog switch)
- unknown pins marked as  '?'

As you can see, it seems FE-5680A fully supports 2 serial  ports:
one on DB9 and the other one on the unpopulated connector (J9) behind  DB9.

Any comments/suggestions are welcome.

.   ciao
_Elio.
_



From:  elio...@gmail.com
Reply-to: time-nuts@febo.com
To:  time-nuts@febo.com
Sent: 24/02/2012 01:03:56 GMT Standard Time
Subj:  [time-nuts] FE-5680A Schematics and scans


Thanks to Ignacio (EB4APL) and the generosity of Mike  Harrison, today I
received a PCB of a disassembled FE5680A.
I provided to  make some scans of the board at 2400DPI resolution: the
picture size is about  7700x11500 pixel (9MB)

Top face (with and without  ICs):
http://www.rhodiatoce.com/pics/time-nuts/FE-5680A/FE-5680A_Top001.jpg
http://www.rhodiatoce.com/pics/time-nuts/FE-5680A/FE-5680A_Top001_noIC.jpg

Bottom  face (with and without  ICs):
http://www.rhodiatoce.com/pics/time-nuts/FE-5680A/FE-5680A_Bottom001.jpg

http://www.rhodiatoce.com/pics/time-nuts/FE-5680A/FE-5680A_Bottom001_noIC.jp
g




Being  able to remove the ICs, I was able to correct some flaws in the
previous  schematics:

http://www.rhodiatoce.com/pics/time-nuts/FE-5680A/FE-5680A_schematics_v0.2.p
df




During  the next week I will continue to write down the logical part of the
circuit  and I hope will be able to dump the 8032 firmware

_  Elio.
_

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] Future of UTC 2013

2013-02-18 Thread Steve Allen
In late 2011 we held a meeting on the Future of UTC in Exton Pennsylvania.
The 400 pages of proceedings from the presentations at that meeting
have been published on paper and in electronic form.

The sequel to that meeting is planned for late May of this year.
Requirements for UTC and Civil Timekeeping on Earth
A Colloquium Addressing a Continuous Time Standard to be held at
the University of Virginia, Charlottesville, VA May 29-31, 2013.

The subject of the meeting notes the ITU-R process regarding possible
changes to radio broadcast time signals (and the associated internet
equivalents).  Full details of the upcoming meeting are available via
http://futureofutc.org/
At that site are also links to the proceedings from the 2011 meeting.

The co-chairs are P. Kenneth Seidelmann, John Seago, Rob Seaman and me.
The period for submitting abstracts remains open until the end of this
week.  We invite contributions which address the issues described in
the Call for Papers at the website.

Folks on the time-nuts list know how time gets into operational
systems.  We would be glad to see some of that expertise at the
meeting.

--
Steve Allen s...@ucolick.orgWGS-84 (GPS)
UCO/Lick Observatory--ISB   Natural Sciences II, Room 165Lat  +36.99855
1156 High StreetVoice: +1 831 459 3046   Lng -122.06015
Santa Cruz, CA 95064http://www.ucolick.org/~sla/ Hgt +250 m
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] OCXO DC Current Question

2013-02-18 Thread Ed Palmer

Hi Bob,

On 2/18/2013 6:19 PM, Bob Camp wrote:

Hi

The -Z3811 is indeed a part out of a HP GPSDO. HP is a big enough customer that 
the part they get *may* have little relation to the standard part.


True, but it doesn't look like there were many Z3811s made.  The order 
wouldn't have been that large.  Maybe there were enough other orders 
from HP that they wanted to keep them happy.



If they all started out at the same temperature, then the noisy ones finish warmup 
quicker than the quiet ones. Take a look at when the final glitch happens in 
the warmup process. It's definitely happening earlier on the noisy parts. If warmup 
finishes quicker - the part is running at a lower temperature. That of course is only the 
time between the cutback from full current to the glitch. It may not mean it's running 
cooler.


Between possible temperature differences due to crystal tuning and 
component tolerances that affect the internal voltage regulators ( and 
therefore affect the maximum current available to the oven ), I don't 
know what, if any, conclusions can be drawn for such a small number of 
samples.  But when I rechecked the startup current, it looks like the 
noisy oscillators draw a few milliamps ( up to 10 ) more than the quiet 
one.  That might allow them to warm up quicker.


Thanks,
Ed


Bob

On Feb 18, 2013, at 5:32 PM, Ed Palmer ed_pal...@sasktel.net wrote:


Hi Bob,

On 2/18/2013 2:42 PM, Bob Camp wrote:

Hi

Like any control loop, gain, bandwidth and noise are all related. In a DOCXO 
you have two control loops and they do interact. That said, there's nothing 
grossly wrong with the four OCXO's. The noisy parts have a bit more gain in the 
controller. The quiet parts have a bit less gain. The easy way to see this is 
the ringing after the current changes.

Yes, when I zoom in on the ringing after the first step change, they all show a 
period of ~9 seconds.  The blue trace (the quiet one) has a nice smooth damped 
sine wave while the other ones have varying amounts of noise superimposed on 
the sine wave.  The amplitude of the ringing on the noisy ones is greater than 
the amplitude of the quiet one.  In one case, more than twice as big.


Without ADEV numbers there's no way to know which one is good (or better) and 
which one is bad (or worse). The noisy parts may be responding to the ambient 
temperature rumble (thus correcting for it) and the quiet ones may be ignoring 
it (allowing it to hit the crystal).

I hadn't thought of that possibility.  Thanks for that!


It's also possible that the noisy ones have an electrical issue in the loop 
that generates the noise.

I was thinking of maybe a dead filter capacitor - hence the noise.


If all the ovens started from the same temperature, there is a variance in the 
oven set points. Some take longer to warm up than others. You don't mention if 
they started the same, so that may or may not be significant.

They all started at the same temperature.  Any apparent difference in warmup 
time is mostly due to the arbitrary offset of the traces. The offset was just 
so you could see all four traces clearly. However, isn't it also likely that 
each oscillator has had it's oven tweaked to match that particular crystal?


To further complicate things, in a double oven, you can have a noisy outer oven that gets 
suppressed by the inner oven. Are your specific 260's double ovens? No way to be sure 
without tearing one open. The second step in the current plot could be an 
inner oven cutting back.

I wondered about that second step.  The period of the ringing on the second 
step is about 1 sec. longer than the ringing on the first step so it's 
apparently a different circuit.  The 260 series datasheet does NOT say anything 
about it being a double oven.  They talk about customizing the unit to suit the 
customer, but that seems a bit extreme.

By the way, I purchased these oscillators from our favorite auction site.  If anyone's 
interested, the 260 series includes both AT and SC crystals.  Since these start out ~150 
Hz low, they appear to be SC cut.  The output is a 5 MHz sine wave @ ~+7 dBm into 50 
ohms. They work fine with a 12V power supply.  They have no mechanical frequency 
adjustment (unless it's under a suspicious spot of solder on the case)  but my 4 were all 
adjustable to 5 MHz via EFC.  They were apparently removed from Z3811 GPS receivers.  
They each have a sticker on the side that says Z3811-80010. There doesn't 
seem to be much info available on the Z3811.  I have no relationship to the vendor - just 
a customer.

So it sounds like the proper thing to do is file this information and carry on. 
 After making some Allan Deviation measurements, review everything to see how, 
or if, oven 'noise' correlates to Allan Deviation results.

Thanks Bob,

Ed


Bob


On Feb 18, 2013, at 2:40 PM, Ed Palmer ed_pal...@sasktel.net wrote:


I know that when making AC measurements on various OCXOs of the same type, you 
have to expect wide variations in 

Re: [time-nuts] Allan deviation and modified Allan deviation

2013-02-18 Thread Magnus Danielson

Corby,

On 18/02/13 20:59, cdel...@juno.com wrote:

Ok I've been plotting my oscillators for years using Allan Deviation.

That way all my records can be compared easily.

Is there any advantage to using Modified Allan Deviation?

It seems to give a better stability plot but that just seems to be
cheating!


It's not.


If I plot both ways what do the differences mean?


When the Modified Allan Deviation was introduced, two things happen:

1) It fixes the bug in the Allan Deviation which can't make useful 
differentiation between white phase modulation and flicker phase 
modulation noises. This is achieved by using a tau-long averager prior 
to the Allan Deviation calculation. If you are in a region where these 
noises exist, use MDEV to separate them. Dave Allan himself is eager to 
push MDEV over ADEV.


2) The traditional (non-overlapping) Allan Deviation estimator has poor 
usage of the samples taken. The MDEV estimator given uses the 
overlapping strategy to increase (but not maximize) the use of samples, 
especially for longer taus. This gives improved confidence intervals.


These steps improved the state of art processing significantly in the 
early 80thies. Since then have the total and Theo strategies for even 
further improved use of samples to achieve good confidence intervals.


Using the traditional non-overlapping Allan deviation estimator for the 
Allan Deviation measure is throwing a lot of useful data in the trash 
and suffer by bad precision or long measurement-times.


Allan Deviation is frequency drift limited in its upper end, where 
linear (and higher order) drift will obscure the output and hide the 
noise processes that Allan Deviation is meant to analyse. Drift is a 
systematic component that needs to be taken out of the data in order for 
the noise processes to be visible.


The Allan Deviation has a cousin in the Hadamard Deviation, both is 
normalized to the same scale (i.e. same as RMS white frequency 
modulation noise). The Hadamard beats Allan for real-time processing, as 
it removes the first degree frequency drift. However, for 
post-processing a better drift model can be used, so drift removal on 
data prior to Allan Deviation is preferred. Then, both Allan and 
Hadamard Deviations can exist in their normal and modified variation, 
with the tau-long averager. For degrees of freedom treatment, you can do 
non-overlapping, overlapping, TOTAL and Theo processing, where TOTAL and 
Theo is the best.


There is also FFT based methods to process data, but I need to refresh 
myself on that.


So, it sounds like you can do a lot more with your data without cheating.

Cheers,
Magnus - from LA
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Future of UTC 2013

2013-02-18 Thread DaveH
Love the graphic on the website.

The use of a Foucault Pendulum with Earth as the bob is inspired... 

 -Original Message-
 From: time-nuts-boun...@febo.com 
 [mailto:time-nuts-boun...@febo.com] On Behalf Of Steve Allen
 Sent: Monday, February 18, 2013 20:18
 To: time-nuts@febo.com
 Subject: [time-nuts] Future of UTC 2013
 
 In late 2011 we held a meeting on the Future of UTC in Exton 
 Pennsylvania.
 The 400 pages of proceedings from the presentations at that meeting
 have been published on paper and in electronic form.
 
 The sequel to that meeting is planned for late May of this year.
   Requirements for UTC and Civil Timekeeping on Earth
 A Colloquium Addressing a Continuous Time Standard to be held at
 the University of Virginia, Charlottesville, VA May 29-31, 2013.
 
 The subject of the meeting notes the ITU-R process regarding possible
 changes to radio broadcast time signals (and the associated internet
 equivalents).  Full details of the upcoming meeting are available via
 http://futureofutc.org/
 At that site are also links to the proceedings from the 2011 meeting.
 
 The co-chairs are P. Kenneth Seidelmann, John Seago, Rob 
 Seaman and me.
 The period for submitting abstracts remains open until the end of this
 week.  We invite contributions which address the issues described in
 the Call for Papers at the website.
 
 Folks on the time-nuts list know how time gets into operational
 systems.  We would be glad to see some of that expertise at the
 meeting.
 
 --
 Steve Allen s...@ucolick.org
 WGS-84 (GPS)
 UCO/Lick Observatory--ISB   Natural Sciences II, Room 165
 Lat  +36.99855
 1156 High StreetVoice: +1 831 459 3046   
 Lng -122.06015
 Santa Cruz, CA 95064http://www.ucolick.org/~sla/ 
 Hgt +250 m
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] repairing General Technology (Tracor) 304-B rubidium standard

2013-02-18 Thread Stewart Cobb
Guys,

I'm repairing a 1960's vintage lab-grade rubidium standard, General
Technology Corporation model 304-B.  Apparently Tracor bought GTC soon
after this unit was made, because references to this as a Tracor 304-B
seem to be more common.  I've made some progress, but now it seems like
time to consult the hive mind.

The unit appears clean, but it doesn't lock.  I've read through old
comments on the list regarding this unit, and I've downloaded a copy of the
manual and schematics available at

*http://sundry.i2phd.com/ServiceManual_304b.pdf*

That file seems to contain a complete copy of the manual text, but some
schematics are missing.  In particular, the schematics for the
sweep/acquisition board (A8) and the three boards inside the physics
package (the lamp oscillator (A13), the SRD driver (A12), and the photocell
preamp (A11)) are not shown.  Does anyone know where to find copies of
those schematics?

The main power supply voltage on my unit seems to have been deliberately
adjusted lower than spec (18.54 V actual, versus 20 +/- 0.1V specified in
the manual).  Replacing a resistor on the regulator board (that had smoked
from overload due to the low voltage) didn't change the voltage much.  I
had to crank the trimmer across half of its range to get the voltage back
within spec.  Nothing in the regulator circuitry seemed to have drifted
enough to change the setpoint that much.  Is there a reason why a tech
would have deliberately set this voltage lower than spec, or did it just
drift down over the years?

A frequency counter (GPSDO reference) shows that the crystal oven warms up
as expected.  The output can be centered on 5 MHz and the sweep circuit
covers a symmetrical range around 5 MHz as expected.  The ovens for the
lamp and filter cell appear to warm up properly as well, judging from test
points available on the A1 oven controller board.  The test point voltages
don't quite match the ones in the PDF manual, but it looks like those
readings were typed into each individual manual after being read off the
particular unit that came with that manual.

The test point on the A5 board shows that 155 Hz resonance detector
modulation is within spec.  The A6 filter-amplifier board test points show
the system attempting (and failing) to detect 155 Hz and 310 Hz resonance
signals coming back from the photocell.

The manual says that the A7 RF pre-driver board (the x14 multiplier) should
be supplying 70 MHz at +13 dBm to the SRD driver inside the physics
package.  That would be about 2.8Vpp, assuming a 50-ohm system.  Instead,
it's supplying a clean 70 MHz at about 100mV into a 50-ohm load.  My best
guess is that the final amplifier transistor on that board is blown,
possibly from being operated with only a scope probe as a load (infinite
VSWR).  Replacement transistors are on order.  Any other thoughts?

Obviously, the box won't lock until the RF input is the right level.  But
it also requires the Rb lamp to light.  Corby Dawson posted to the list
back on 12 November 2009:

Tracor bulbs fail with a different mechanism and last maybe 10 years.

Anyone know what that different failure mechanism is?  Is it repairable
in an ordinary lab, like the heat-gun trick for LPRO bulbs?  If not, is it
feasible to build a Frankenstein replacement using something like an LPRO
or FEI bulb?

Is it possible to tell whether the lamp is lit without opening the physics
package?  If not, are there any tricks to opening the physics package?  Any
precautions to take before doing so?

Any other comments on how to get this box working again?

Cheers!
--Stu

Side note:  This unit was built during the era of elastic seconds
(roughly, the 1960's).  It contains a board (A9) which digitally offsets
the output frequency in increments of roughly 7E-10, without changing the
rubidium resonance frequency or the C-field.  There's also a note in the
manual saying that annual changes to the definition of the second may
require replacing the rubidium resonance cell in the physics package with a
new cell calibrated for the new second in the new year.  Leap seconds bring
their own problems, but compared to dismantling your lab instruments every
year, they're a breeze.