Re: [time-nuts] Cable length calibration

2016-06-30 Thread Scott McGrath
If the nominal velocity of propagation is known and length is known delay is 
easily determined mathematically 

Time Domain Reflectrometry is the usual technique for finding cable length but 
even there the cables NVP is an essential parameter if you want to compute 
length but not essential in time nuts application because we are interested in 
delay which a TDR reads directly When using a TDR its best if cable is 
unterminated as the discontinuity at the end is helpful as a marker.   Also 
most TDRs like the Tek 1502 can put 100v or more on the cable which will blow 
most GPS antennas


> On Jun 29, 2016, at 4:08 PM, Brooke Clarke  wrote:
> 
> Hi Hal:
> 
> I think the cal process is essentially a time domain reflection measure of 
> cable length.  The GPS receiver and the cable cal hardware would be in the 
> antenna unit.
> The 1 PPS signal would be aligned at the output of the cable.
> 
> -- 
> Have Fun,
> 
> Brooke Clarke
> http://www.PRC68.com
> http://www.end2partygovernment.com/2012Issues.html
> The lesser of evils is still evil.
> 
>  Original Message 
>> bro...@pacific.net said:
>>> At one point they were looking into making a GPS time receiver where the
>>> cable length calibration would be built-in.
>> How would you do that?
>> 
>> The obvious way is to compare the time you get with a known-good time, but if
>> you had that, why would you want this new GPS with an unknown cable length.
>> 
>> You might be able to do it by measuring the DC drop.  Getting enough accuracy
>> seems tough.
> 
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] HP 5071A second life

2016-06-30 Thread Matus Michael
I have an old HP 5071 Cs-standard here on my desk. The oven of the (second) 
tube has lost its Cs completely so the instrument is no longer working. Since 
the electronic is quite old we acquired a complete new 5071A (now from 
Symmetricom).

I tried to "reactivate" the HP as a fancy desktop clock without success. Tried 
different commands on the serial bus but was not able to switch off the fatal 
error states (which is ok for the proper use). It is not possible to set Time 
and Date, nor perform a synchronisation. The display just stays dark. Also the 
reference oscillator is driven to one end causing a horrible frequency 
stability.

Does somebody know a procedure to override the error states so that the display 
works with a depleted CBT by programming? I do not want to modify the hardware. 
And to state it explicitly: I do not need a actual time/frequency standard just 
a mockup for my desk.

Michael

Dr. Michael Matus
-
BEV - Bundesamt fuer Eich- und Vermessungswesen
Leiter des Referates Dimensionelle Groessen, Frequenz, Zeit
(Head of Section, Dimensional Quantities, Frequency, Time)
Tel. +43 1 21110-6540, Fax +43 1 21110-996000
1160 Wien, Arltgasse 35, Austria
michael.ma...@bev.gv.at
www.bev.gv.at
-
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] MSF 60 kHz off the air during the day for annual maintenance: 4 – 20 July 2016

2016-06-30 Thread David J Taylor
MSF 60 kHz off the air during the day for annual maintenance: 4 – 20 July 
2016


See:
 
http://www.npl.co.uk/science-technology/time-frequency/products-and-services/time/msf-outages

Cheers,
David
--
SatSignal Software - Quality software written to your requirements
Web: http://www.satsignal.eu
Email: david-tay...@blueyonder.co.uk
Twitter: @gm8arv 


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 5071A second life

2016-06-30 Thread Tom Van Baak
>From the front panel: Top -> Clock -> Set -> Display:ON
The same menu page also allows you to set the HH:MM:SS time.
This can be done within a few seconds of power up, regardless if the CBT is 
warming up, dead, or alive.
The amber attention LED will remain lit, but a dead CBT should not affect the 
red LED time display.

/tvb

- Original Message - 
From: "Matus Michael" 
To: 
Sent: Thursday, June 30, 2016 5:51 AM
Subject: [time-nuts] HP 5071A second life


>I have an old HP 5071 Cs-standard here on my desk. The oven of the (second) 
>tube has lost its Cs completely so the instrument is no longer working. Since 
>the electronic is quite old we acquired a complete new 5071A (now from 
>Symmetricom).
> 
> I tried to "reactivate" the HP as a fancy desktop clock without success. 
> Tried different commands on the serial bus but was not able to switch off the 
> fatal error states (which is ok for the proper use). It is not possible to 
> set Time and Date, nor perform a synchronisation. The display just stays 
> dark. Also the reference oscillator is driven to one end causing a horrible 
> frequency stability.
> 
> Does somebody know a procedure to override the error states so that the 
> display works with a depleted CBT by programming? I do not want to modify the 
> hardware. And to state it explicitly: I do not need a actual time/frequency 
> standard just a mockup for my desk.
> 
> Michael
> 
> Dr. Michael Matus
> -
> BEV - Bundesamt fuer Eich- und Vermessungswesen
> Leiter des Referates Dimensionelle Groessen, Frequenz, Zeit
> (Head of Section, Dimensional Quantities, Frequency, Time)
> Tel. +43 1 21110-6540, Fax +43 1 21110-996000
> 1160 Wien, Arltgasse 35, Austria
> michael.ma...@bev.gv.at
> www.bev.gv.at
> -
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] buying a time interval counter

2016-06-30 Thread Stéphane Rey

Hi,
I've ordered a SR620 with the option 01 (higher stability standard)
Should be there  in August
Thanks
Stephane

-- Message d'origine --
De : "Brooke Clarke" 
À : "Stéphane Rey" ; "Discussion of precise time 
and frequency measurement" 

Envoyé 29/06/2016 19:20:22
Objet : Re: [time-nuts] buying a time interval counter


Hi Stephane:

I traded my HP 53132A counter for an SR 620.  The 53132 has what I'd 
call a user hostile interface, so if you are manually controlling the 
counter the SR 620 has a huge advantage.
I also like the long display on the 620 which can be read from across 
the room.


PS Stanford Research is a company founded by physicists and makes some 
really high quality stuff.  In fact some of the products 
HP/Agilent/Keysight sells are repackaged SR instruments.

http://www.prc68.com/I/TandFTE.shtml#SR620

The claim to fame for the HP 53132A is that it can make a frequency 
(not time interval) measurement to 1E12 in a second.  Here's how to get 
that same result with the SR620:

http://www.prc68.com/I/FTS4060.shtml#SR620Fast

On the down side the printing functions on the 620 require an Epsom 
printer.  Does anyone have a solution for that?


PS SR also makes a 10 MHz crystal oscillator that has options trading 
stability for aging as well as the EFC tuning polarity and range so as 
to match other OCXOs.

http://prc68.com/I/TandFTE.shtml#SC10

At one point they were looking into making a GPS time receiver where 
the cable length calibration would be built-in.


-- Have Fun,

Brooke Clarke
http://www.PRC68.com
http://www.end2partygovernment.com/2012Issues.html
The lesser of evils is still evil.

 Original Message 

Hello there,

I'm planning to buy a such instrument in order to do some frequency 
stability measurement at work. The SR620 seems to be discontinued. 
What model still distributed would you think is good for that at the 
moment ?


Thanks & cheers
Stephane
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

and follow the instructions there.





___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 5071A second life

2016-06-30 Thread paul swed
Michael
Thats going to be quite the clock. Very unfortunate but thats what happens
the Cs dry up. I seriously messed with a much older CS and was able to get
a few more C's out of the tube. The 5071 is smart so not sure any of what I
did makes any sense. Basically heating the oven to a slightly higher
temperature.
The other point that may not be of interest is that even dead 5071s seem to
have a great deal of value.
Regards
Paul
WB8TSL

On Thu, Jun 30, 2016 at 8:51 AM, Matus Michael 
wrote:

> I have an old HP 5071 Cs-standard here on my desk. The oven of the
> (second) tube has lost its Cs completely so the instrument is no longer
> working. Since the electronic is quite old we acquired a complete new 5071A
> (now from Symmetricom).
>
> I tried to "reactivate" the HP as a fancy desktop clock without success.
> Tried different commands on the serial bus but was not able to switch off
> the fatal error states (which is ok for the proper use). It is not possible
> to set Time and Date, nor perform a synchronisation. The display just stays
> dark. Also the reference oscillator is driven to one end causing a horrible
> frequency stability.
>
> Does somebody know a procedure to override the error states so that the
> display works with a depleted CBT by programming? I do not want to modify
> the hardware. And to state it explicitly: I do not need a actual
> time/frequency standard just a mockup for my desk.
>
> Michael
>
> Dr. Michael Matus
> -
> BEV - Bundesamt fuer Eich- und Vermessungswesen
> Leiter des Referates Dimensionelle Groessen, Frequenz, Zeit
> (Head of Section, Dimensional Quantities, Frequency, Time)
> Tel. +43 1 21110-6540, Fax +43 1 21110-996000
> 1160 Wien, Arltgasse 35, Austria
> michael.ma...@bev.gv.at
> www.bev.gv.at
> -
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cable length calibration

2016-06-30 Thread David
The Tektronix 1502 uses a tunnel diode pulser to produce a 50
picosecond output step of about 200 millivolts.  There is a misprint
in the theory section of the service manual which says "400 V" instead
of "400mV".

On Thu, 30 Jun 2016 04:19:43 -0400, you wrote:

>If the nominal velocity of propagation is known and length is known delay is 
>easily determined mathematically 
>
>Time Domain Reflectrometry is the usual technique for finding cable length but 
>even there the cables NVP is an essential parameter if you want to compute 
>length but not essential in time nuts application because we are interested in 
>delay which a TDR reads directly When using a TDR its best if cable is 
>unterminated as the discontinuity at the end is helpful as a marker.   Also 
>most TDRs like the Tek 1502 can put 100v or more on the cable which will blow 
>most GPS antennas
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] Information request

2016-06-30 Thread Peter Torry


I have been given a 10MHz oscillator that appears to be in good physical 
condition but I have no further information on it hence this request.


It is marked as follows:-Quartzkeramic 2006/10SCD-531
10.000MHz   12v=
K 837383

Any information such as connection details, performance etc. will be 
greatly appreciated.

Regards
Peter
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] buying a time interval counter

2016-06-30 Thread Jerry
How does the Fluke PM6690 (same as Pendulum CNT-90) compare to the SR620?  A 
neighbor is selling one in perfect condition (per him) for $900

Jerry
NY2KW

-Original Message-
From: time-nuts [mailto:time-nuts-boun...@febo.com] On Behalf Of Stéphane Rey
Sent: Thursday, June 30, 2016 10:07 AM
To: Brooke Clarke; Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] buying a time interval counter

Hi,
I've ordered a SR620 with the option 01 (higher stability standard) Should be 
there  in August Thanks Stephane

-- Message d'origine --
De : "Brooke Clarke"  À : "Stéphane Rey" 
; "Discussion of precise time and frequency measurement" 
 Envoyé 29/06/2016 19:20:22 Objet : Re: [time-nuts] buying 
a time interval counter

>Hi Stephane:
>
>I traded my HP 53132A counter for an SR 620.  The 53132 has what I'd 
>call a user hostile interface, so if you are manually controlling the 
>counter the SR 620 has a huge advantage.
>I also like the long display on the 620 which can be read from across 
>the room.
>
>PS Stanford Research is a company founded by physicists and makes some 
>really high quality stuff.  In fact some of the products 
>HP/Agilent/Keysight sells are repackaged SR instruments.
>http://www.prc68.com/I/TandFTE.shtml#SR620
>
>The claim to fame for the HP 53132A is that it can make a frequency 
>(not time interval) measurement to 1E12 in a second.  Here's how to get 
>that same result with the SR620:
>http://www.prc68.com/I/FTS4060.shtml#SR620Fast
>
>On the down side the printing functions on the 620 require an Epsom 
>printer.  Does anyone have a solution for that?
>
>PS SR also makes a 10 MHz crystal oscillator that has options trading 
>stability for aging as well as the EFC tuning polarity and range so as 
>to match other OCXOs.
>http://prc68.com/I/TandFTE.shtml#SC10
>
>At one point they were looking into making a GPS time receiver where 
>the cable length calibration would be built-in.
>
>-- Have Fun,
>
>Brooke Clarke
>http://www.PRC68.com
>http://www.end2partygovernment.com/2012Issues.html
>The lesser of evils is still evil.
>
> Original Message 
>>Hello there,
>>
>>I'm planning to buy a such instrument in order to do some frequency 
>>stability measurement at work. The SR620 seems to be discontinued.
>>What model still distributed would you think is good for that at the 
>>moment ?
>>
>>Thanks & cheers
>>Stephane
>>___
>>time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
>>https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>and follow the instructions there.
>>
>

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] buying a time interval counter

2016-06-30 Thread Stéphane Rey

Ji Jerry,
The Fluke exhibit a 100ps resolution whereas the SR620 has 25ps.
This is the main difference I see from datasheets.

cheers
Stephane

-- Message d'origine --
De : "Jerry" 
À : "'Stéphane Rey'" ; "'Discussion of precise 
time and frequency measurement'" ; "'Brooke Clarke'" 


Envoyé 30/06/2016 17:01:50
Objet : RE: [time-nuts] buying a time interval counter

How does the Fluke PM6690 (same as Pendulum CNT-90) compare to the 
SR620?  A neighbor is selling one in perfect condition (per him) for 
$900


Jerry
NY2KW

-Original Message-
From: time-nuts [mailto:time-nuts-boun...@febo.com] On Behalf Of 
Stéphane Rey

Sent: Thursday, June 30, 2016 10:07 AM
To: Brooke Clarke; Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] buying a time interval counter

Hi,
I've ordered a SR620 with the option 01 (higher stability standard) 
Should be there  in August Thanks Stephane


-- Message d'origine --
De : "Brooke Clarke"  À : "Stéphane Rey" 
; "Discussion of precise time and frequency 
measurement"  Envoyé 29/06/2016 19:20:22 Objet : 
Re: [time-nuts] buying a time interval counter



Hi Stephane:

I traded my HP 53132A counter for an SR 620.  The 53132 has what I'd
call a user hostile interface, so if you are manually controlling the
counter the SR 620 has a huge advantage.
I also like the long display on the 620 which can be read from across
the room.

PS Stanford Research is a company founded by physicists and makes some
really high quality stuff.  In fact some of the products
HP/Agilent/Keysight sells are repackaged SR instruments.
http://www.prc68.com/I/TandFTE.shtml#SR620

The claim to fame for the HP 53132A is that it can make a frequency
(not time interval) measurement to 1E12 in a second.  Here's how to 
get

that same result with the SR620:
http://www.prc68.com/I/FTS4060.shtml#SR620Fast

On the down side the printing functions on the 620 require an Epsom
printer.  Does anyone have a solution for that?

PS SR also makes a 10 MHz crystal oscillator that has options trading
stability for aging as well as the EFC tuning polarity and range so as
to match other OCXOs.
http://prc68.com/I/TandFTE.shtml#SC10

At one point they were looking into making a GPS time receiver where
the cable length calibration would be built-in.

-- Have Fun,

Brooke Clarke
http://www.PRC68.com
http://www.end2partygovernment.com/2012Issues.html
The lesser of evils is still evil.

 Original Message 

Hello there,

I'm planning to buy a such instrument in order to do some frequency
stability measurement at work. The SR620 seems to be discontinued.
What model still distributed would you think is good for that at the
moment ?

Thanks & cheers
Stephane
___
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.





___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

and follow the instructions there.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] The Nature of Noise

2016-06-30 Thread Bob Stewart
With my recent Cs problems, I've been wondering about the subject of noise 
generation and measurement.  Specifically, my question is this:  Let's say that 
I have 3 disciplined oscillators: A, B, and C.  So, I use the same 5370 to 
create a 1000 second ADEV and discover that the 1s noise value between A and B 
is low, while the 1s noise between A and C, as well as B and C is high.  Can I 
then say that both A and B are low noise devices?  Or is it possible that even 
though I'm measuring 1000 points, both A and B are high noise devices, but 
somehow are noisy in the same exact way, and it's actually C that's the low 
noise oscillator?

Bob - AE6RV 
---
GFS GPSDO list:
groups.yahoo.com/neo/groups/GFS-GPSDOs/info
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] The Nature of Noise

2016-06-30 Thread paul swed
Bob,
I am about to get myself into a lot of trouble. If the noise is coherent in
some way then its not really noise anymore its a signal or its filtered in
some way as compared to C.
There you go I feel the hot emails on the way already.
Should be a good read.
Regards
Paul
WB8TSL

On Thu, Jun 30, 2016 at 2:30 PM, Bob Stewart  wrote:

> With my recent Cs problems, I've been wondering about the subject of noise
> generation and measurement.  Specifically, my question is this:  Let's say
> that I have 3 disciplined oscillators: A, B, and C.  So, I use the same
> 5370 to create a 1000 second ADEV and discover that the 1s noise value
> between A and B is low, while the 1s noise between A and C, as well as B
> and C is high.  Can I then say that both A and B are low noise devices?  Or
> is it possible that even though I'm measuring 1000 points, both A and B are
> high noise devices, but somehow are noisy in the same exact way, and it's
> actually C that's the low noise oscillator?
>
> Bob -
> AE6RV 
> ---
> GFS GPSDO list:
> groups.yahoo.com/neo/groups/GFS-GPSDOs/info
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] The Nature of Noise

2016-06-30 Thread Attila Kinali
On Thu, 30 Jun 2016 18:30:35 + (UTC)
Bob Stewart  wrote:

> With my recent Cs problems, I've been wondering about the subject of noise
> generation and measurement.  Specifically, my question is this:  Let's say
> that I have 3 disciplined oscillators: A, B, and C.  So, I use the same 5370
> to create a 1000 second ADEV and discover that the 1s noise value between A
> and B is low, while the 1s noise between A and C, as well as B and C is
> high.  Can I then say that both A and B are low noise devices?  Or is it
> possible that even though I'm measuring 1000 points, both A and B are high
> noise devices, but somehow are noisy in the same exact way, and it's
> actually C that's the low noise oscillator?

Short answer: yes
Long answer: Yes, but unlikely

:-)

You can see the output of the oscillators as random processes and model
them as such (MIT opencoursware has some good lectures on that topic).
If the processes are all independent, then you can assume that A and B
will have the lowest noise and estimate the ADEV values of C and the
second nosiest oscillator with good confidence, and of the lowest noise
oscillator with some confidence (keyword: "three cornered hat").
In reality you will never have totally independent oscillators, there
will be always some coupling through power supplies (even mains power)
or through RF coupling. Heck, even your measurement instrument might
couple the oscillators if you are not very carefull. So it really might
be that A and B are just bad oscillators, that injection locked to
each other, and thus seemed more stable/less noisy than A+C and B+C.
But generally, high stability/low noise oscillators are usually made
such, that the output and the power supply are decoupled from the
oscillator itself as much as possible, to make the effect of coupling
(which is always present) so small that it get swamped out by all
other noise sources.

Attila Kinali
-- 
Malek's Law:
Any simple idea will be worded in the most complicated way.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cable length calibration

2016-06-30 Thread Dr. David Kirkby (Kirkby Microwave Ltd)
On 30 June 2016 at 09:19, Scott McGrath  wrote:

> If the nominal velocity of propagation is known and length is known delay
> is easily determined mathematically
>

Except that coax does not have a uniform impedance or velocity factor. Both
will vary as a function of position and frequency. How relevant this is
depends on the accuracy you require, but since it is time-nuts, it is
reasonable to assume that such a simplistic method is not of the standard
expected on time-nuts.

Dave
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cable length calibration

2016-06-30 Thread Charles Steinmetz

Scott wrote:


When using a TDR its best if cable is unterminated as the discontinuity at the 
end is helpful as a marker.


Someone else previously asked how you're expected to do TDR on a GPS 
antenna cable since everything is matched (therefore, presumably, no 
reflections).


There are ALWAYS reflections.  As one message in this thread noted, TDR 
easily shows sharp bends in coax.  Plus, a GPS antenna will be "matched" 
only at a relatively narrow band of frequencies centered on the 
frequency(ies) it receives (L1, L1+L2, etc.).  Also note that GPS 
antennas specified as having 50 ohm outputs are often (if not usually) 
used with 75 ohm cable.  (Trimble actually advises this.)


You can do the TDR situationally (when the receiver is powered up, and 
whenever an antenna fault is cleared), or repetitively during operation 
(e.g., once a minute).  I see no reason why either strategy would not 
work well on a GPS antenna cable.


There is no need to use signals that would damage a GPS antenna preamp.

Best regards,

Charles


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] The Nature of Noise

2016-06-30 Thread Bob Camp
Hi

If you are measuring noise, then C is the bad one in the group. If you are 
measuring something else, then it is 
possible that you are getting bad information. This is a classic argument about 
comparing devices from the 
same lot of parts. They might both have a very similar warmup curve or 
temperature or pressure coefficient. 

Bob

> On Jun 30, 2016, at 2:30 PM, Bob Stewart  wrote:
> 
> With my recent Cs problems, I've been wondering about the subject of noise 
> generation and measurement.  Specifically, my question is this:  Let's say 
> that I have 3 disciplined oscillators: A, B, and C.  So, I use the same 5370 
> to create a 1000 second ADEV and discover that the 1s noise value between A 
> and B is low, while the 1s noise between A and C, as well as B and C is high. 
>  Can I then say that both A and B are low noise devices?  Or is it possible 
> that even though I'm measuring 1000 points, both A and B are high noise 
> devices, but somehow are noisy in the same exact way, and it's actually C 
> that's the low noise oscillator?
> 
> Bob - AE6RV 
> ---
> GFS GPSDO list:
> groups.yahoo.com/neo/groups/GFS-GPSDOs/info
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cable length calibration

2016-06-30 Thread Michael Wouters
As a practical matter, in the lab we seldom need a cable delay
measured to better than +/- 0.5 ns, which we usually do as a time
interval measurement, with a 1 pps into a tee on channel A of a TIC
and then the cable from the tee to channel B.

For cables up to 40 m or so, just measuring the physical length is  as
accurate (experimentally determined!).

A while back, EURAMET ran a pilot where a few lengths of cable were
circulated amongst a number of NMIs who then had to measure the
delays. The reported scatter was about +/- 1 ns, as I recall. This
mainly came down to differences in test signals, trigger levels etc.

Cheers
Michael


On Fri, Jul 1, 2016 at 7:03 AM, Dr. David Kirkby (Kirkby Microwave
Ltd)  wrote:
> On 30 June 2016 at 09:19, Scott McGrath  wrote:
>
>> If the nominal velocity of propagation is known and length is known delay
>> is easily determined mathematically
>>
>
> Except that coax does not have a uniform impedance or velocity factor. Both
> will vary as a function of position and frequency. How relevant this is
> depends on the accuracy you require, but since it is time-nuts, it is
> reasonable to assume that such a simplistic method is not of the standard
> expected on time-nuts.
>
> Dave
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cable length calibration

2016-06-30 Thread Scott McGrath
I'd say it depends on the Time Nut 😀.  NVP X Distance will get you close and 
for a beginning time nut is a worthwhile exercise 

To improve delay calculations now you need instrumentation that not all 
especially beginning time nuts own.I've got a 20 Ghz Agilent scope/TDR 
along with a 110 Ghz network analyzer and just putting a cable on the network 
analyzer and handling it you can see the characteristics changeSo yes you 
are correct that the simple NVPxDistance is not suitable for advanced time 
nuttery.   In fact the delay will change with temperature and barometric 
pressure unless you are using hardline and that's subject to humidity unless 
filled with dry nitrogen under pressure.

Content by Scott
Typos by Siri

> On Jun 30, 2016, at 5:03 PM, Dr. David Kirkby (Kirkby Microwave Ltd) 
>  wrote:
> 
>> On 30 June 2016 at 09:19, Scott McGrath  wrote:
>> 
>> If the nominal velocity of propagation is known and length is known delay
>> is easily determined mathematically
> 
> Except that coax does not have a uniform impedance or velocity factor. Both
> will vary as a function of position and frequency. How relevant this is
> depends on the accuracy you require, but since it is time-nuts, it is
> reasonable to assume that such a simplistic method is not of the standard
> expected on time-nuts.
> 
> Dave
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cable length calibration

2016-06-30 Thread Scott McGrath
This is highly dependent on the TDR especially ones designed for long twisted 
pair runs where a high voltage pulse is used to overcome resistive losses

Content by Scott
Typos by Siri

> On Jun 30, 2016, at 10:59 AM, David  wrote:
> 
> The Tektronix 1502 uses a tunnel diode pulser to produce a 50
> picosecond output step of about 200 millivolts.  There is a misprint
> in the theory section of the service manual which says "400 V" instead
> of "400mV".
> 
>> On Thu, 30 Jun 2016 04:19:43 -0400, you wrote:
>> 
>> If the nominal velocity of propagation is known and length is known delay is 
>> easily determined mathematically 
>> 
>> Time Domain Reflectrometry is the usual technique for finding cable length 
>> but even there the cables NVP is an essential parameter if you want to 
>> compute length but not essential in time nuts application because we are 
>> interested in delay which a TDR reads directly When using a TDR its best if 
>> cable is unterminated as the discontinuity at the end is helpful as a 
>> marker.   Also most TDRs like the Tek 1502 can put 100v or more on the cable 
>> which will blow most GPS antennas
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cable length calibration

2016-06-30 Thread Tim Shoppa
We are all time nuts, so there's an obvious answer: What you do, is raise
the GPS up to a height the same as the cable length. You then drop it,
measure the time until it hits the ground, and use d = 0.5 a * t * t to
calculate d. Then you correct for the velocity factor.

Tim N3QE

On Wed, Jun 29, 2016 at 3:28 PM, Hal Murray  wrote:

>
> bro...@pacific.net said:
> > At one point they were looking into making a GPS time receiver where the
> > cable length calibration would be built-in.
>
> How would you do that?
>
> The obvious way is to compare the time you get with a known-good time, but
> if
> you had that, why would you want this new GPS with an unknown cable length.
>
> You might be able to do it by measuring the DC drop.  Getting enough
> accuracy
> seems tough.
>
>
> --
> These are my opinions.  I hate spam.
>
>
>
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] The Nature of Noise

2016-06-30 Thread Bob Stewart
Thanks Bob and Attila,
This problem with the Cs kind of brought the whole question of noise and what 
ADEVs mean into focus for me.  Like most newbies, I had read about the ever 
increasing stability of time standards and wondered how they could actually 
know that they were better.  I've seen the term 3-cornered hat bandied about, 
but I didn't really grock it until now.
Bob 
---
GFS GPSDO list:
groups.yahoo.com/neo/groups/GFS-GPSDOs/info

  From: Bob Camp 
 To: Bob Stewart ; Discussion of precise time and frequency 
measurement  
 Sent: Thursday, June 30, 2016 4:27 PM
 Subject: Re: [time-nuts] The Nature of Noise
   
Hi

If you are measuring noise, then C is the bad one in the group. If you are 
measuring something else, then it is 
possible that you are getting bad information. This is a classic argument about 
comparing devices from the 
same lot of parts. They might both have a very similar warmup curve or 
temperature or pressure coefficient. 

Bob

> On Jun 30, 2016, at 2:30 PM, Bob Stewart  wrote:
> 
> With my recent Cs problems, I've been wondering about the subject of noise 
> generation and measurement.  Specifically, my question is this:  Let's say 
> that I have 3 disciplined oscillators: A, B, and C.  So, I use the same 5370 
> to create a 1000 second ADEV and discover that the 1s noise value between A 
> and B is low, while the 1s noise between A and C, as well as B and C is high. 
>  Can I then say that both A and B are low noise devices?  Or is it possible 
> that even though I'm measuring 1000 points, both A and B are high noise 
> devices, but somehow are noisy in the same exact way, and it's actually C 
> that's the low noise oscillator?
> 
> Bob - AE6RV 
> ---
> GFS GPSDO list:
> groups.yahoo.com/neo/groups/GFS-GPSDOs/info
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.


  
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cable length calibration

2016-06-30 Thread jimlux

On 6/30/16 5:52 PM, Tim Shoppa wrote:

We are all time nuts, so there's an obvious answer: What you do, is raise
the GPS up to a height the same as the cable length. You then drop it,
measure the time until it hits the ground, and use d = 0.5 a * t * t to
calculate d. Then you correct for the velocity factor.



or rig it as a pendulum, and correct for the gravitational pull of the 
moon and sun


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cable length calibration

2016-06-30 Thread Charles Steinmetz
The 1503 was Tek's long-line (10km) TDR analyzer.  IIRC, it put 
haversine pulses of about 5v onto the line under test (~10v open 
circuit).  The 1502 (~600m line length) put pulses of about 200mV onto 
the line under test (~400mV open circuit).


Best regards,

Charles


Scott wrote:


This is highly dependent on the TDR especially ones designed for long twisted 
pair runs where a high voltage pulse is used to overcome resistive losses


David had written:

The Tektronix 1502 uses a tunnel diode pulser to produce a 50
picosecond output step of about 200 millivolts.  There is a misprint
in the theory section of the service manual which says "400 V" instead
of "400mV".


Scott had written:

Time Domain Reflectrometry is the usual technique for finding cable length but 
even there the cables NVP is an essential parameter if you want to compute 
length but not essential in time nuts application because we are interested in 
delay which a TDR reads directly When using a TDR its best if cable is 
unterminated as the discontinuity at the end is helpful as a marker.   Also 
most TDRs like the Tek 1502 can put 100v or more on the cable which will blow 
most GPS antennas



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.