Re: [volt-nuts] Fluke 732A Questions

2014-01-01 Thread Bill Gold
Joe:

A correction on the dropout voltage when the In Cal light goes out, I
found my notes and it is 21 volts instead of 22 volts.  The way I did this
was to remove the battery pack, AC voltage and then just feed in a power
supply where the battery pack would have delivered it's voltage to the
back-plane mother board.  Then I just reduced the voltage until the light
turned off.  So that means that Fluke was following the battery
manufacturers specs of 5.25 volts at the end of the discharge curve.

The banana jacks on the battery pack have two jobs.  One is to provide a
very good chassis ground to the battery pack and the other is probably to
guide the pack accurately into the mother board.  One other thing I have
found out the hard way is that the holes in the top plate of the battery
pack where the battery terminals come through are very close to the battery
terminals and can short out to the chassis.  I have made the holes a LOT
bigger to avoid this problem.  This really depends upon the brand of
batteries you are using but after having to repair the regulator board I
would rather be safe than sorry.

I have 6 Pomona 1756-48 low thermal leads for when I need to be careful
with thermals.  I also have made my own low thermal(I think) using
Pomona 4892 gold plated double banana plugs and Belden 9272 twisted pair
twinax, using a Pomona 1825 for the ground lead.  If I need single banana
plugs to connect to the 732A I use Pomona 4897 gold plated.  I know that
this isn't the best possible low thermal because the wire is tin plated but
I have looked and I can't find a good source of bare copper twisted pair
shielded wire.  When I am making measurements with the home made cables I
just wait a few minutes for the thermals to settle down.  Why Belden 9272?
It was the best choice from what was available at work.  From what I have
seen Pomona was acquired by Fluke and makes special low thermal cables for
Fluke that are not available through the Pomona catalog.

When you find out from Fluke what services they offer and prices please
let all of us know.

Bill

- Original Message - 
From: J. L. Trantham jlt...@att.net
To: 'Discussion of precise voltage measurement' volt-nuts@febo.com
Sent: Tuesday, December 31, 2013 5:13 PM
Subject: Re: [volt-nuts] Fluke 732A Questions


 Bill,

 I have yet to put my 732A though it's paces but I could connect an
external
 power supply, disconnect the AC supply and the battery module and slowly
 decrease the supplied voltage and see what happens when.

 I'll try your measurement algorithm for the battery to see when the 'IN
CAL'
 LED goes out and what the current draws are.  If the measured voltage at
J10
 is 27 V, then the actual battery voltage should be about 27.6 V,
accounting
 for the diode in the circuit (unless the current draw during measurement
is
 so low that no voltage drop occurs across the diode).

 I found the NIST/NBS Tech Note easily.

 I think if I decide to send it out for calibration, I'll use Fluke, if
they
 still calibrate the unit.  I'll call them later this week of next week to
 see what services/accessories they might still have available.

 Yes, I wound up with the 735C.  It was easy to replace the NiCd battery
pack
 and bring it 'on scale' by adjusting the 'jumpers' and the front panel
 adjustment.  I now only have to deal with the periodic 'NO CAL' LED
 indication.  I need to study the 732A manual then take a close look at the
 735C to see how it works on it.

 This also brings up the question of 'low EMF cables' for accurate
 measurement.  I have a Fluke 720A K/V Divider on the shelf that I have yet
 to bring out to check to see if it is operational.  Another one of those
 'projects' for a later time.

 Where would one look for 'low EMF cables' to be used in precise
measurement
 experiments and what errors might one expect to find using Pomona gold
 plated Banana Plug cables (which is as good as I have at this time)?

 Thanks for the info.

 Happy New Year to all.

 Joe


___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Fluke 732A Questions

2014-01-01 Thread J. L. Trantham
Just noticed the 'banana plugs' mounted on the chassis to 'align' the
battery module.  Duh.  Makes perfect sense.

Also, thanks for the Pomona recommendations.

Joe

-Original Message-
From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
Behalf Of Bill Gold
Sent: Wednesday, January 01, 2014 10:38 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Fluke 732A Questions

Joe:

A correction on the dropout voltage when the In Cal light goes out, I
found my notes and it is 21 volts instead of 22 volts.  The way I did this
was to remove the battery pack, AC voltage and then just feed in a power
supply where the battery pack would have delivered it's voltage to the
back-plane mother board.  Then I just reduced the voltage until the light
turned off.  So that means that Fluke was following the battery
manufacturers specs of 5.25 volts at the end of the discharge curve.

The banana jacks on the battery pack have two jobs.  One is to provide a
very good chassis ground to the battery pack and the other is probably to
guide the pack accurately into the mother board.  One other thing I have
found out the hard way is that the holes in the top plate of the battery
pack where the battery terminals come through are very close to the battery
terminals and can short out to the chassis.  I have made the holes a LOT
bigger to avoid this problem.  This really depends upon the brand of
batteries you are using but after having to repair the regulator board I
would rather be safe than sorry.

I have 6 Pomona 1756-48 low thermal leads for when I need to be careful
with thermals.  I also have made my own low thermal(I think) using
Pomona 4892 gold plated double banana plugs and Belden 9272 twisted pair
twinax, using a Pomona 1825 for the ground lead.  If I need single banana
plugs to connect to the 732A I use Pomona 4897 gold plated.  I know that
this isn't the best possible low thermal because the wire is tin plated but
I have looked and I can't find a good source of bare copper twisted pair
shielded wire.  When I am making measurements with the home made cables I
just wait a few minutes for the thermals to settle down.  Why Belden 9272?
It was the best choice from what was available at work.  From what I have
seen Pomona was acquired by Fluke and makes special low thermal cables for
Fluke that are not available through the Pomona catalog.

When you find out from Fluke what services they offer and prices please
let all of us know.

Bill

- Original Message -
From: J. L. Trantham jlt...@att.net
To: 'Discussion of precise voltage measurement' volt-nuts@febo.com
Sent: Tuesday, December 31, 2013 5:13 PM
Subject: Re: [volt-nuts] Fluke 732A Questions


 Bill,

 I have yet to put my 732A though it's paces but I could connect an
external
 power supply, disconnect the AC supply and the battery module and slowly
 decrease the supplied voltage and see what happens when.

 I'll try your measurement algorithm for the battery to see when the 'IN
CAL'
 LED goes out and what the current draws are.  If the measured voltage at
J10
 is 27 V, then the actual battery voltage should be about 27.6 V,
accounting
 for the diode in the circuit (unless the current draw during measurement
is
 so low that no voltage drop occurs across the diode).

 I found the NIST/NBS Tech Note easily.

 I think if I decide to send it out for calibration, I'll use Fluke, if
they
 still calibrate the unit.  I'll call them later this week of next week to
 see what services/accessories they might still have available.

 Yes, I wound up with the 735C.  It was easy to replace the NiCd battery
pack
 and bring it 'on scale' by adjusting the 'jumpers' and the front panel
 adjustment.  I now only have to deal with the periodic 'NO CAL' LED
 indication.  I need to study the 732A manual then take a close look at the
 735C to see how it works on it.

 This also brings up the question of 'low EMF cables' for accurate
 measurement.  I have a Fluke 720A K/V Divider on the shelf that I have yet
 to bring out to check to see if it is operational.  Another one of those
 'projects' for a later time.

 Where would one look for 'low EMF cables' to be used in precise
measurement
 experiments and what errors might one expect to find using Pomona gold
 plated Banana Plug cables (which is as good as I have at this time)?

 Thanks for the info.

 Happy New Year to all.

 Joe


___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Fluke 732A Questions

2013-12-31 Thread J. L. Trantham
Charles,

I would like to find the mating plug for J10 for several reasons, being able
to get it to calibration being just one of them.  However, I was hoping it
would be good enough to make it to calibration on the internal batteries.

I seem to recall reading that the internal batteries were good for a few
hours but nothing like what the 732B would do, which sounded like the 732B
could make the trip on the internal batteries.

Where did you find the resistance of the thermistor 'as shipped'?  I saw
mention of being supplied to the original owner when shipped as well and I
have had the top cover off and the battery module out to replace the
batteries and found nothing to suggest a thermistor value.  I have also
lifted the top layer of insulation off the oven to move the cal-jumpers but
did not see anything that suggested a thermistor value.

My 'metrology level' standards are my HP 3458A's, calibrated by Agilent, and
Solartron (Ametek) 7081's calibrated by Ametek.  I have been thinking of
getting the 732A calibrated by Fluke but I didn't know if they would still
calibrate the unit.  I wonder if they would have the 'as shipped' thermistor
value on file for my specific serial number?

I, too, read about the stability of the thermistor being the most important
measurement.  That was in the 'change sheets' attached to the manual on
Didier's site.

The 'cal-jumper' was at '40' when I got the unit.  Per my 3458A, I could
only get it down from about +220 uV to about +100 uV with the front panel
adjustment.  By moving the jumper to '20', I could only get it up to about
-12 uV below 10 V by turning the adjustment all the way in the opposite
direction.  However, by connecting jumpers for both the '20' and '10', I was
able to get it at 10 V with the front panel adjustment at about center
position.  I have not 'logged' any values yet but every time I go to the
shop, it is about +/- 1 or 2 uV from 10 V.  I do the 'AUTO CAL' on the 3458A
each morning I go into the shop.  

I measure the resistance with a Fluke 8050A (simply because it is the
closest DMM to the standard) and it is 4229 to 4230 ohms each day.

Interestingly, when I got the 732A, undisturbed Fluke Calibration stickers
were in place over each of the adjustments on the front panel, on a back
screw on the top cover, and 'half' of a sticker on the back over the power
supply module, as if the other half was lost when someone removed the
battery module.  Somehow, I find that comforting, given that it now seems so
stable, making me think it is 'normal aging' that is responsible for the
values I measured rather than a problem with the oven, etc..  IIRC, the spec
is 6 ppm per year for 10 V, or 60 uV per year.  I'll have to look at the
power supply module to see if I can find some date codes but I suspect the
unit is 25 to 30 years old.  Unfortunately, there were no 'dates' on the
stickers to tell when the unit was last calibrated.

I was thinking of sending the 3458A's and the 732A out for calibration
simultaneously and seeing what I find when they get back.  

I'll call Fluke and ask about their Cal services, the plug, and the shipping
case/battery pack issues.

Thanks everyone for very useful information.

Joe

-Original Message-
From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
Behalf Of Charles Steinmetz
Sent: Monday, December 30, 2013 11:23 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Fluke 732A Questions

Joe wrote:

Also, how do you get your 732A calibrated?  Do the batteries last long 
enough to ship overnight to a facility?

No -- I assumed that was why you were looking for a battery connector.  You
install an external battery for shipment overnight.  Fluke can supply a
complete external battery with cable and plug ($375, last I knew), or you
can make your own.  I believe they can also supply a shipping container,
which has a dedicated space for the Fluke external battery.  (If you don't
use Fluke's container, build something very sturdy with wood and padding and
with handles that make it hard not to carry it and set it down right side
up.)

I wouldn't have anybody but Fluke calibrate a 732A.  There are 3 levels of
calibration, all with data, from about $700 to about $1500 (the last is
calibration by Fluke's Primary Standards Lab).

As far as thermistor values go, the relevant question is not what is the
value in absolute terms, but rather what is the value in relation to the as
manufactured value, which was provided by Fluke when the unit shipped.
This had been recorded by the original owners of each of my units, and they
are all within 2 ohms of the original value today.

If you do not know the original value, then your original measurement is a
going forward value and it cannot tell you anything about where you are in
relation to  the original manufactured value.  The fact that you had to
change cal jumpers suggests that either the oven temperature has shifted
since manufacture, the reference

Re: [volt-nuts] Fluke 732A Questions

2013-12-31 Thread R.Phillips

Joe
Reference the thermistor value and its recorded value - I asked Fluke this 
question and the answer was, they did supply the individual 732A's value in 
written form to the original purchaser, but unfortunately they did not 
maintain a separate record.

Roy

-Original Message- 
From: J. L. Trantham

Sent: Tuesday, December 31, 2013 1:22 PM
To: 'Discussion of precise voltage measurement'
Subject: Re: [volt-nuts] Fluke 732A Questions

Charles,

I would like to find the mating plug for J10 for several reasons, being able
to get it to calibration being just one of them.  However, I was hoping it
would be good enough to make it to calibration on the internal batteries.

I seem to recall reading that the internal batteries were good for a few
hours but nothing like what the 732B would do, which sounded like the 732B
could make the trip on the internal batteries.

Where did you find the resistance of the thermistor 'as shipped'?  I saw
mention of being supplied to the original owner when shipped as well and I
have had the top cover off and the battery module out to replace the
batteries and found nothing to suggest a thermistor value.  I have also
lifted the top layer of insulation off the oven to move the cal-jumpers but
did not see anything that suggested a thermistor value.

My 'metrology level' standards are my HP 3458A's, calibrated by Agilent, and
Solartron (Ametek) 7081's calibrated by Ametek.  I have been thinking of
getting the 732A calibrated by Fluke but I didn't know if they would still
calibrate the unit.  I wonder if they would have the 'as shipped' thermistor
value on file for my specific serial number?

I, too, read about the stability of the thermistor being the most important
measurement.  That was in the 'change sheets' attached to the manual on
Didier's site.

The 'cal-jumper' was at '40' when I got the unit.  Per my 3458A, I could
only get it down from about +220 uV to about +100 uV with the front panel
adjustment.  By moving the jumper to '20', I could only get it up to about
-12 uV below 10 V by turning the adjustment all the way in the opposite
direction.  However, by connecting jumpers for both the '20' and '10', I was
able to get it at 10 V with the front panel adjustment at about center
position.  I have not 'logged' any values yet but every time I go to the
shop, it is about +/- 1 or 2 uV from 10 V.  I do the 'AUTO CAL' on the 3458A
each morning I go into the shop.

I measure the resistance with a Fluke 8050A (simply because it is the
closest DMM to the standard) and it is 4229 to 4230 ohms each day.

Interestingly, when I got the 732A, undisturbed Fluke Calibration stickers
were in place over each of the adjustments on the front panel, on a back
screw on the top cover, and 'half' of a sticker on the back over the power
supply module, as if the other half was lost when someone removed the
battery module.  Somehow, I find that comforting, given that it now seems so
stable, making me think it is 'normal aging' that is responsible for the
values I measured rather than a problem with the oven, etc..  IIRC, the spec
is 6 ppm per year for 10 V, or 60 uV per year.  I'll have to look at the
power supply module to see if I can find some date codes but I suspect the
unit is 25 to 30 years old.  Unfortunately, there were no 'dates' on the
stickers to tell when the unit was last calibrated.

I was thinking of sending the 3458A's and the 732A out for calibration
simultaneously and seeing what I find when they get back.

I'll call Fluke and ask about their Cal services, the plug, and the shipping
case/battery pack issues.

Thanks everyone for very useful information.

Joe

-Original Message-
From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
Behalf Of Charles Steinmetz
Sent: Monday, December 30, 2013 11:23 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Fluke 732A Questions

Joe wrote:


Also, how do you get your 732A calibrated?  Do the batteries last long
enough to ship overnight to a facility?


No -- I assumed that was why you were looking for a battery connector.  You
install an external battery for shipment overnight.  Fluke can supply a
complete external battery with cable and plug ($375, last I knew), or you
can make your own.  I believe they can also supply a shipping container,
which has a dedicated space for the Fluke external battery.  (If you don't
use Fluke's container, build something very sturdy with wood and padding and
with handles that make it hard not to carry it and set it down right side
up.)

I wouldn't have anybody but Fluke calibrate a 732A.  There are 3 levels of
calibration, all with data, from about $700 to about $1500 (the last is
calibration by Fluke's Primary Standards Lab).

As far as thermistor values go, the relevant question is not what is the
value in absolute terms, but rather what is the value in relation to the as
manufactured value, which was provided by Fluke when the unit shipped.
This had

Re: [volt-nuts] Fluke 732A Questions

2013-12-31 Thread J. L. Trantham
I was afraid of that.

Joe

-Original Message-
From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
Behalf Of R.Phillips
Sent: Tuesday, December 31, 2013 7:53 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Fluke 732A Questions

Joe
Reference the thermistor value and its recorded value - I asked Fluke this
question and the answer was, they did supply the individual 732A's value in
written form to the original purchaser, but unfortunately they did not
maintain a separate record.
Roy

-Original Message-
From: J. L. Trantham
Sent: Tuesday, December 31, 2013 1:22 PM
To: 'Discussion of precise voltage measurement'
Subject: Re: [volt-nuts] Fluke 732A Questions

Charles,

I would like to find the mating plug for J10 for several reasons, being able
to get it to calibration being just one of them.  However, I was hoping it
would be good enough to make it to calibration on the internal batteries.

I seem to recall reading that the internal batteries were good for a few
hours but nothing like what the 732B would do, which sounded like the 732B
could make the trip on the internal batteries.

Where did you find the resistance of the thermistor 'as shipped'?  I saw
mention of being supplied to the original owner when shipped as well and I
have had the top cover off and the battery module out to replace the
batteries and found nothing to suggest a thermistor value.  I have also
lifted the top layer of insulation off the oven to move the cal-jumpers but
did not see anything that suggested a thermistor value.

My 'metrology level' standards are my HP 3458A's, calibrated by Agilent, and
Solartron (Ametek) 7081's calibrated by Ametek.  I have been thinking of
getting the 732A calibrated by Fluke but I didn't know if they would still
calibrate the unit.  I wonder if they would have the 'as shipped' thermistor
value on file for my specific serial number?

I, too, read about the stability of the thermistor being the most important
measurement.  That was in the 'change sheets' attached to the manual on
Didier's site.

The 'cal-jumper' was at '40' when I got the unit.  Per my 3458A, I could
only get it down from about +220 uV to about +100 uV with the front panel
adjustment.  By moving the jumper to '20', I could only get it up to about
-12 uV below 10 V by turning the adjustment all the way in the opposite
direction.  However, by connecting jumpers for both the '20' and '10', I was
able to get it at 10 V with the front panel adjustment at about center
position.  I have not 'logged' any values yet but every time I go to the
shop, it is about +/- 1 or 2 uV from 10 V.  I do the 'AUTO CAL' on the 3458A
each morning I go into the shop.

I measure the resistance with a Fluke 8050A (simply because it is the
closest DMM to the standard) and it is 4229 to 4230 ohms each day.

Interestingly, when I got the 732A, undisturbed Fluke Calibration stickers
were in place over each of the adjustments on the front panel, on a back
screw on the top cover, and 'half' of a sticker on the back over the power
supply module, as if the other half was lost when someone removed the
battery module.  Somehow, I find that comforting, given that it now seems so
stable, making me think it is 'normal aging' that is responsible for the
values I measured rather than a problem with the oven, etc..  IIRC, the spec
is 6 ppm per year for 10 V, or 60 uV per year.  I'll have to look at the
power supply module to see if I can find some date codes but I suspect the
unit is 25 to 30 years old.  Unfortunately, there were no 'dates' on the
stickers to tell when the unit was last calibrated.

I was thinking of sending the 3458A's and the 732A out for calibration
simultaneously and seeing what I find when they get back.

I'll call Fluke and ask about their Cal services, the plug, and the shipping
case/battery pack issues.

Thanks everyone for very useful information.

Joe

-Original Message-
From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
Behalf Of Charles Steinmetz
Sent: Monday, December 30, 2013 11:23 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Fluke 732A Questions

Joe wrote:

Also, how do you get your 732A calibrated?  Do the batteries last long 
enough to ship overnight to a facility?

No -- I assumed that was why you were looking for a battery connector.  You
install an external battery for shipment overnight.  Fluke can supply a
complete external battery with cable and plug ($375, last I knew), or you
can make your own.  I believe they can also supply a shipping container,
which has a dedicated space for the Fluke external battery.  (If you don't
use Fluke's container, build something very sturdy with wood and padding and
with handles that make it hard not to carry it and set it down right side
up.)

I wouldn't have anybody but Fluke calibrate a 732A.  There are 3 levels of
calibration, all with data, from about $700 to about $1500 (the last

Re: [volt-nuts] Fluke 732A Questions

2013-12-31 Thread T . Micallef
Joe,

I live in Florida and searched for cal labs here that could do the job of
calibrating references. Lockheed Martin in Orlando is certified, but may not
have the best uncertainties. Fluke shows 0.3uV/V where LM shows 0.35uV/V.
Tektronix in Orlando also offers calibration, but they ship it out and it
turns out to be 0.38uV/V at the Keithley Primary Standards lab in Ohio. You
would be better shipping it yourself to Fluke.

I think Fluke is the best choice, but at least there are other options.

Todd



___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Fluke 732A Questions

2013-12-31 Thread Charles Steinmetz

Joe wrote:


Where did you find the resistance of the thermistor 'as shipped'?  I saw
mention of being supplied to the original owner when shipped as well


It was recorded on the original factory cal certificate.  I have the 
original certificates for two of mine.  The others had rear panel 
stickers applied by their original owner (a cal lab).


A properly working, good example should hold 0.1 ppm or better year 
in, year out.  I've seen some that do not meet that, but lots of them 
do.  I'm not aware of any other non-JJ standard that can match it, 
including most 732Bs.  (Some folks think 732As are generally better 
than 732Bs because they are older and, thus, more stable because 
better aged.  I think there is more to it -- many 732Bs are plenty 
old enough to have settled down to that level if they are ever going 
to, but even the older 732Bs don't, in general, seem to exhibit quite 
the stability of the 732As.)


Best regards,

Charles



___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Fluke 732A Questions

2013-12-31 Thread jeffhook


Hi,  



Are you guys adjusting your 732As? 

I have a friend that works in a large CAL lab and he told me to never adjust my 
732As, just compare them. 

They send out their 732 Bs to Fluke but are never adjusted, just compared and 
sent back with current voltage reading ?? 



Thanks - Jeff  



- Original Message -


From: Charles Steinmetz csteinm...@yandex.com 
To: Discussion of precise voltage measurement volt-nuts@febo.com 
Sent: Tuesday, December 31, 2013 9:56:37 AM 
Subject: Re: [volt-nuts] Fluke 732A Questions 

Joe wrote: 

Where did you find the resistance of the thermistor 'as shipped'?  I saw 
mention of being supplied to the original owner when shipped as well 

It was recorded on the original factory cal certificate.  I have the 
original certificates for two of mine.  The others had rear panel 
stickers applied by their original owner (a cal lab). 

A properly working, good example should hold 0.1 ppm or better year 
in, year out.  I've seen some that do not meet that, but lots of them 
do.  I'm not aware of any other non-JJ standard that can match it, 
including most 732Bs.  (Some folks think 732As are generally better 
than 732Bs because they are older and, thus, more stable because 
better aged.  I think there is more to it -- many 732Bs are plenty 
old enough to have settled down to that level if they are ever going 
to, but even the older 732Bs don't, in general, seem to exhibit quite 
the stability of the 732As.) 

Best regards, 

Charles 



___ 
volt-nuts mailing list -- volt-nuts@febo.com 
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts 
and follow the instructions there. 
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] Fluke 732A Questions

2013-12-31 Thread Chuck Harris

The obvious reason for that suggestion is the adjustment
pot is likely to be noisy and is likely to introduce
uncertainty.

The big question that comes to my mind is if you care about
that level of uncertainty, why not replace the pot with a
voltage divider made of some low thermal coefficient fixed
value resistors?

-Chuck Harris

jeffh...@comcast.net wrote:



Hi,



Are you guys adjusting your 732As?

I have a friend that works in a large CAL lab and he told me to never adjust my
732As, just compare them.

They send out their 732 Bs to Fluke but are never adjusted, just compared and 
sent
back with current voltage reading ??



Thanks - Jeff

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Fluke 732A Questions

2013-12-31 Thread John Phillips
Call Agilent about 3458A calibration then tell them that you really want
your fluke calibrated and see what they will do for you.


On Tue, Dec 31, 2013 at 3:48 PM, John Phillips john.philli...@gmail.comwrote:

 Any thing you do other than record the current voltage reading will
 introduce more uncertainty and increase drift unless you have a defect.
 Replacing a defective part will start the aging process over again for that
 part. Do not change anything if you can help it.

 Agilent is a good cal lab to use as well. They will run it for a few days
 and take readings a few different times and give you a report.
 Most of the time the battery is not enough to keep the box hot until they
 plug it in  so they count on warming up a few days.

 Hope that helps.
 John


 On Tue, Dec 31, 2013 at 3:31 PM, Chuck Harris cfhar...@erols.com wrote:

 The obvious reason for that suggestion is the adjustment
 pot is likely to be noisy and is likely to introduce
 uncertainty.

 The big question that comes to my mind is if you care about
 that level of uncertainty, why not replace the pot with a
 voltage divider made of some low thermal coefficient fixed
 value resistors?

 -Chuck Harris

 jeffh...@comcast.net wrote:



 Hi,



 Are you guys adjusting your 732As?

 I have a friend that works in a large CAL lab and he told me to never
 adjust my
 732As, just compare them.

 They send out their 732 Bs to Fluke but are never adjusted, just
 compared and sent
 back with current voltage reading ??



 Thanks - Jeff

 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/
 mailman/listinfo/volt-nuts
 and follow the instructions there.




 --
 John Phillips




-- 
John Phillips
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Fluke 732A Questions

2013-12-31 Thread Bill Gold
Joe:

The NIST notes can be downloaded from this site.

https://archive.org/details/NISTTechNotes

Here is the battery check paragraph.  I check mine once and month and plot 
performance using Excel.

2-47. OPERATION ON BATTERY

2-48. Prior to operating the 732A on battery power, the following procedure 
should be followed.

1. Connect a voltmeter at the rear terminal POWER INPUT jack to monitor the 
battery voltage.

2. Note the voltage while the battery is on trickle charge (the AC PWR LED 
is on, but BTRY CHG off).
The voltage should be around 27V.

3. Disconnect the line cord, wait 10 minutes, and then monitor the battery 
voltage for one-half hour.
   The battery package may have to be replaced if the voltage decrease by 
more than 0.2V during
   this period and the load represented by the 732A is correct ( see 
Section 4, Battery Discharge).

2-49. It may be advisable to perform the above procedure periodically to ensure 
battery backup capability is maintained.
--

While the Fluke manual maintains that the battery life is 24 hours, the 
reality is that you can get maybe 12 to 15 hours on a fresh set of cells.  I 
think that the engineers didn't figure on the total heater current being around 
240 ma in the final design.  The batteries are rated at 4 AH and some simple 
math says that 4/.24 = 16.67 hours to a battery voltage of 5.25 V.  But the In 
Cal LED will turn off around 22 volts so maybe 15 hours at best.  The 
batteries will continue to supply heater current so it might keep the +18.6 
volt regulator working for a few more hours, but when the batteries are down 
that far it is hard to calculate remaining life.  This is why there is a Fluke 
732A-7003 Transport Case and Battery Charger and you can have up to 4 battery 
packs in the case.  This gives you a total of 72 hours transport time, 
according to Fluke.  I have seen people take a couple of 12 volt high capacity 
SLA and duct tape them to the top of the 732A to increase transport time.  I 
 am not too sure what the AH capacity is of the 12 volts cells.  I would guess 
that 12 AH might do the job but I have never gone that far to figure out.

The 24 hour spec is probably based upon the NIST Technical Note #1239 
(Solid-State Voltage Standard Performance and design Guidelines) which 
recommended at least 24 hours of battery life for transport.  You can get this 
from the same website as above.

I never had to worry about this (battery life) as Fluke used to have a 
repair and cal facility in Milpitas, CA and I worked in Morgan Hill, CA.  I 
would either hand carry the unit to be certified, or Fluke would pick it up and 
deliver it back, under power.  Then Fluke decided to sell the whole operation 
to FLW Service Corp who finally dropped the whole operation around 2005.  I 
also retired around that time so I didn't care to pay for all of this myself ( 
I was having the Company pay ).  Since then I have not had a Certification of 
any 732A units and just relied upon the inter-comparison method to hopefully 
keep a reasonable value of the Volt.  I have looked around the South SF Bay 
Area for a Cal Lab who could Certifiy one of my units and the only place I 
could find would be Simco in Sunnyvale.  I will have to ask them what they 
charge for a Certitication of a 732A.  The local Fluke place used to charge 
$500 but then that was using 732As that they had sent into Fluke in E
 verett WA.  But then since this is just a hobby anymore I don't need, and 
didn't need while I was working, the full blown services and accuracy of Fluke 
in Everett.  I guess a 732B would be a way to get all of this but who needs it 
right now, not me.  Of course I could always get my own Josephson Junction but 
then I would be a real crazy Volt Nut and certifiable at the same time.

I have never even heard about the Fluke 735C until I saw one on fleabay, 
and I assume you were the one who bought this when I saw you asking about 
schematics.

Happy New Year

Bill   

- Original Message - 
From: J. L. Trantham jlt...@att.net
To: 'Discussion of precise voltage measurement' volt-nuts@febo.com
Sent: Monday, December 30, 2013 6:28 PM
Subject: Re: [volt-nuts] Fluke 732A Questions


 Bill,
 
 Sounds like you are further along in the 'volt-nut' process than I am.  I'll
 have to chase down the NIST paper.  I have been very impressed by the
 stability of the 732A and the 735C.
 
 Unfortunately, I don't have any documentation on the 735C and only the early
 manual for the 732A.
 
 I would be interested in the specifics of the battery check procedure in
 your manual.  Unfortunately, my manual does not have a page 2-9.
 
 I note that there are two banana jacks on the front of the battery module,
 both black, one connected to ground and the other not connected to anything.
 Any idea of what

Re: [volt-nuts] Fluke 732A Questions

2013-12-31 Thread John Phillips
Any thing you do other than record the current voltage reading will
introduce more uncertainty and increase drift unless you have a defect.
Replacing a defective part will start the aging process over again for that
part. Do not change anything if you can help it.

Agilent is a good cal lab to use as well. They will run it for a few days
and take readings a few different times and give you a report.
Most of the time the battery is not enough to keep the box hot until they
plug it in  so they count on warming up a few days.

Hope that helps.
John


On Tue, Dec 31, 2013 at 3:31 PM, Chuck Harris cfhar...@erols.com wrote:

 The obvious reason for that suggestion is the adjustment
 pot is likely to be noisy and is likely to introduce
 uncertainty.

 The big question that comes to my mind is if you care about
 that level of uncertainty, why not replace the pot with a
 voltage divider made of some low thermal coefficient fixed
 value resistors?

 -Chuck Harris

 jeffh...@comcast.net wrote:



 Hi,



 Are you guys adjusting your 732As?

 I have a friend that works in a large CAL lab and he told me to never
 adjust my
 732As, just compare them.

 They send out their 732 Bs to Fluke but are never adjusted, just compared
 and sent
 back with current voltage reading ??



 Thanks - Jeff

 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/
 mailman/listinfo/volt-nuts
 and follow the instructions there.




-- 
John Phillips
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Fluke 732A Questions

2013-12-31 Thread J. L. Trantham
Jeff,

Too late. 

I have already moved the jumpers and adjusted all of the front panel 
adjustments to achieve the desired output voltages.

Since I don't have any 'historical' data about the performance of my specific 
unit, I will 'start from scratch' on the unit I now have.  Whether to send it 
out for calibration or just keep it, adjusted by the 3458A, and watch from now 
on is the issue.

I don't have enough data to decide about long term stability at this point but 
I am gaining valuable information about how long it will last on batteries and 
how to go about preparing it for shipment for calibration if I go that way.

Thanks for the input.

Joe

-Original Message-
From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf 
Of jeffh...@comcast.net
Sent: Tuesday, December 31, 2013 5:23 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Fluke 732A Questions



Hi,  



Are you guys adjusting your 732As? 

I have a friend that works in a large CAL lab and he told me to never adjust my 
732As, just compare them. 

They send out their 732 Bs to Fluke but are never adjusted, just compared and 
sent back with current voltage reading ?? 



Thanks - Jeff  



- Original Message -


From: Charles Steinmetz csteinm...@yandex.com
To: Discussion of precise voltage measurement volt-nuts@febo.com
Sent: Tuesday, December 31, 2013 9:56:37 AM
Subject: Re: [volt-nuts] Fluke 732A Questions 

Joe wrote: 

Where did you find the resistance of the thermistor 'as shipped'?  I 
saw mention of being supplied to the original owner when shipped as 
well

It was recorded on the original factory cal certificate.  I have the original 
certificates for two of mine.  The others had rear panel stickers applied by 
their original owner (a cal lab). 

A properly working, good example should hold 0.1 ppm or better year in, year 
out.  I've seen some that do not meet that, but lots of them do.  I'm not aware 
of any other non-JJ standard that can match it, including most 732Bs.  (Some 
folks think 732As are generally better than 732Bs because they are older and, 
thus, more stable because better aged.  I think there is more to it -- many 
732Bs are plenty old enough to have settled down to that level if they are ever 
going to, but even the older 732Bs don't, in general, seem to exhibit quite the 
stability of the 732As.) 

Best regards, 

Charles 



___
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there. 
___
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Fluke 732A Questions

2013-12-31 Thread J. L. Trantham
Interesting thought.  I love it when guys of that caliber compete.

Joe

-Original Message-
From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
Behalf Of John Phillips
Sent: Tuesday, December 31, 2013 5:51 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Fluke 732A Questions

Call Agilent about 3458A calibration then tell them that you really want
your fluke calibrated and see what they will do for you.


On Tue, Dec 31, 2013 at 3:48 PM, John Phillips
john.philli...@gmail.comwrote:

 Any thing you do other than record the current voltage reading will 
 introduce more uncertainty and increase drift unless you have a defect.
 Replacing a defective part will start the aging process over again for 
 that part. Do not change anything if you can help it.

 Agilent is a good cal lab to use as well. They will run it for a few 
 days and take readings a few different times and give you a report.
 Most of the time the battery is not enough to keep the box hot until 
 they plug it in  so they count on warming up a few days.

 Hope that helps.
 John


 On Tue, Dec 31, 2013 at 3:31 PM, Chuck Harris cfhar...@erols.com wrote:

 The obvious reason for that suggestion is the adjustment pot is 
 likely to be noisy and is likely to introduce uncertainty.

 The big question that comes to my mind is if you care about that 
 level of uncertainty, why not replace the pot with a voltage divider 
 made of some low thermal coefficient fixed value resistors?

 -Chuck Harris

 jeffh...@comcast.net wrote:



 Hi,



 Are you guys adjusting your 732As?

 I have a friend that works in a large CAL lab and he told me to 
 never adjust my 732As, just compare them.

 They send out their 732 Bs to Fluke but are never adjusted, just 
 compared and sent back with current voltage reading ??



 Thanks - Jeff

 ___
 volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to 
 https://www.febo.com/cgi-bin/ mailman/listinfo/volt-nuts and follow 
 the instructions there.




 --
 John Phillips




--
John Phillips
___
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Fluke 732A Questions

2013-12-31 Thread J. L. Trantham
Bill, 

I have yet to put my 732A though it's paces but I could connect an external
power supply, disconnect the AC supply and the battery module and slowly
decrease the supplied voltage and see what happens when.

I'll try your measurement algorithm for the battery to see when the 'IN CAL'
LED goes out and what the current draws are.  If the measured voltage at J10
is 27 V, then the actual battery voltage should be about 27.6 V, accounting
for the diode in the circuit (unless the current draw during measurement is
so low that no voltage drop occurs across the diode).

I found the NIST/NBS Tech Note easily.

I think if I decide to send it out for calibration, I'll use Fluke, if they
still calibrate the unit.  I'll call them later this week of next week to
see what services/accessories they might still have available.

Yes, I wound up with the 735C.  It was easy to replace the NiCd battery pack
and bring it 'on scale' by adjusting the 'jumpers' and the front panel
adjustment.  I now only have to deal with the periodic 'NO CAL' LED
indication.  I need to study the 732A manual then take a close look at the
735C to see how it works on it.

This also brings up the question of 'low EMF cables' for accurate
measurement.  I have a Fluke 720A K/V Divider on the shelf that I have yet
to bring out to check to see if it is operational.  Another one of those
'projects' for a later time.

Where would one look for 'low EMF cables' to be used in precise measurement
experiments and what errors might one expect to find using Pomona gold
plated Banana Plug cables (which is as good as I have at this time)?

Thanks for the info.

Happy New Year to all.

Joe

-Original Message-
From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
Behalf Of Bill Gold
Sent: Tuesday, December 31, 2013 6:26 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Fluke 732A Questions

Joe:

The NIST notes can be downloaded from this site.

https://archive.org/details/NISTTechNotes

Here is the battery check paragraph.  I check mine once and month and
plot performance using Excel.

2-47. OPERATION ON BATTERY

2-48. Prior to operating the 732A on battery power, the following procedure
should be followed.

1. Connect a voltmeter at the rear terminal POWER INPUT jack to monitor
the battery voltage.

2. Note the voltage while the battery is on trickle charge (the AC PWR
LED is on, but BTRY CHG off).
The voltage should be around 27V.

3. Disconnect the line cord, wait 10 minutes, and then monitor the
battery voltage for one-half hour.
   The battery package may have to be replaced if the voltage decrease
by more than 0.2V during
   this period and the load represented by the 732A is correct ( see
Section 4, Battery Discharge).

2-49. It may be advisable to perform the above procedure periodically to
ensure battery backup capability is maintained.
--

While the Fluke manual maintains that the battery life is 24 hours, the
reality is that you can get maybe 12 to 15 hours on a fresh set of cells.  I
think that the engineers didn't figure on the total heater current being
around 240 ma in the final design.  The batteries are rated at 4 AH and some
simple math says that 4/.24 = 16.67 hours to a battery voltage of 5.25 V.
But the In Cal LED will turn off around 22 volts so maybe 15 hours at
best.  The batteries will continue to supply heater current so it might keep
the +18.6 volt regulator working for a few more hours, but when the
batteries are down that far it is hard to calculate remaining life.  This is
why there is a Fluke 732A-7003 Transport Case and Battery Charger and you
can have up to 4 battery packs in the case.  This gives you a total of 72
hours transport time, according to Fluke.  I have seen people take a couple
of 12 volt high capacity SLA and duct tape them to the top of the 732A to
increase transport time.  I  am not too sure what the AH capacity is of the
12 volts cells.  I would guess that 12 AH might do the job but I have never
gone that far to figure out.

The 24 hour spec is probably based upon the NIST Technical Note #1239
(Solid-State Voltage Standard Performance and design Guidelines) which
recommended at least 24 hours of battery life for transport.  You can get
this from the same website as above.

I never had to worry about this (battery life) as Fluke used to have a
repair and cal facility in Milpitas, CA and I worked in Morgan Hill, CA.  I
would either hand carry the unit to be certified, or Fluke would pick it up
and deliver it back, under power.  Then Fluke decided to sell the whole
operation to FLW Service Corp who finally dropped the whole operation around
2005.  I also retired around that time so I didn't care to pay for all of
this myself ( I was having the Company pay ).  Since then I have not had a
Certification

Re: [volt-nuts] Fluke 732A Questions

2013-12-31 Thread jeffhook


Joe, 



I don't have any history on my 732As either. M y 3458As  2002  are out of cal 
too so I'm running blind with all my stuff :o) 

I'm hoping to find a cheap 732B and have my CAL lab buddy sneak it into work 
and do a poor mans cal on it. (hard to do with a 732A) 

He does this each year with my LN 4214 10K standards but he's getting close to 
retirement.    I wish Fluke had a cheap cal for the hobbyist - Jeff     



- Original Message -


From: J. L. Trantham jlt...@att.net 
To: Discussion of precise voltage measurement volt-nuts@febo.com 
Sent: Tuesday, December 31, 2013 7:01:32 PM 
Subject: Re: [volt-nuts] Fluke 732A Questions 

Jeff, 

Too late. 

I have already moved the jumpers and adjusted all of the front panel 
adjustments to achieve the desired output voltages. 

Since I don't have any 'historical' data about the performance of my specific 
unit, I will 'start from scratch' on the unit I now have.  Whether to send it 
out for calibration or just keep it, adjusted by the 3458A, and watch from now 
on is the issue. 

I don't have enough data to decide about long term stability at this point but 
I am gaining valuable information about how long it will last on batteries and 
how to go about preparing it for shipment for calibration if I go that way. 

Thanks for the input. 

Joe 
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] Fluke 732A Questions

2013-12-30 Thread R.Phillips

Hi Todd
I would appreciate a copy of your later issue manual for the 732A - please 
let me know when you make this available  -  Dider's site ?

Many thanks
Roy


-Original Message- 
From: T. Micallef

Sent: Monday, December 30, 2013 4:03 AM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Fluke 732A Questions

Joe,

I don't know if it makes any difference but my manual is for the 732A/AN p/n
788414

I found a couple of brochures that I scanned and uploaded too.

I forgot to mention, I have been collecting information regarding the
techniques used to do intercomparisons using scanners like the Data Proof
models. While searching a couple of weeks ago, I came across an old document
from a Measurement Science Conference back in 1983. The document contains an
article regarding the testing of the Fluke 735C. I thought you may have
little to no info regarding your reference and thought I would let you know
about it.

The document is easy to find using google. It is named MSC1983.pdf
I will upload it as well, but I was not sure if Didier wants to track
conference docs.

Todd





___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there. 



___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Fluke 732A Questions

2013-12-30 Thread Didier Juges
I may be unable to move manuals until the end of the week. However, you can 
always download a manual that was just uploaded by using ftp. Check the Upload 
button on the Manuals page.

Happy New Year everyone.

Didier KO4BB


R.Phillips phill...@btinternet.com wrote:
Hi Todd
I would appreciate a copy of your later issue manual for the 732A -
please 
let me know when you make this available  -  Dider's site ?
Many thanks
Roy


-Original Message- 
From: T. Micallef
Sent: Monday, December 30, 2013 4:03 AM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Fluke 732A Questions

Joe,

I don't know if it makes any difference but my manual is for the
732A/AN p/n
788414

I found a couple of brochures that I scanned and uploaded too.

I forgot to mention, I have been collecting information regarding the
techniques used to do intercomparisons using scanners like the Data
Proof
models. While searching a couple of weeks ago, I came across an old
document
from a Measurement Science Conference back in 1983. The document
contains an
article regarding the testing of the Fluke 735C. I thought you may have
little to no info regarding your reference and thought I would let you
know
about it.

The document is easy to find using google. It is named MSC1983.pdf
I will upload it as well, but I was not sure if Didier wants to track
conference docs.

Todd





___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there. 


___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

-- 
Sent from my Motorola Droid Razr 4G LTE wireless tracker while I do other 
things.
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Fluke 732A Questions

2013-12-30 Thread T . Micallef
My manual came from manualsplus. It was in excellent condition compared to
the ones on eBay. You do have to ask them for a picture of every manual
since they can vary in condition.

I have rescanned the whole manual, which I should have done anyway.
I have two missing pages and I am hopeful the seller will scan them and
forward them to me.

Todd

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Fluke 732A Questions

2013-12-30 Thread J. L. Trantham
I called them this AM and they only had a 'copy' of the manual and not an
original.

If I understand what you are saying, you got a copy as well.

If so, were the 'change sheets' part of the 'copy'?  When I asked about
that, I was told that there were no 'change sheets' in the copy she was
looking at when I called and asked about condition and change sheets.

Joe

-Original Message-
From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
Behalf Of T. Micallef
Sent: Monday, December 30, 2013 1:06 PM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Fluke 732A Questions

My manual came from manualsplus. It was in excellent condition compared to
the ones on eBay. You do have to ask them for a picture of every manual
since they can vary in condition.

I have rescanned the whole manual, which I should have done anyway.
I have two missing pages and I am hopeful the seller will scan them and
forward them to me.

Todd

___
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Fluke 732A Questions

2013-12-30 Thread J. L. Trantham
Charles,

I just assumed that it was obsolete.  I'll contact Fluke.

Thanks.

Joe

-Original Message-
From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
Behalf Of Charles Steinmetz
Sent: Monday, December 30, 2013 5:52 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Fluke 732A Questions

Robert wrote:

As Roy has already said, this is a hypertronics 
connector.*   *   *   Sorry, I don't have any, or a cheap source.

The plug that mates with the jack on the back of the newer 732A battery is
Fluke part number 2181497.  IIRC, I paid about $10 each (some years ago)
direct from Fluke.

Best regards,

Charles



___
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Fluke 732A Questions

2013-12-30 Thread T . Micallef
I received the manual with the plastic gray cover just like many of their
newer manuals. It appears to have sat on a shelf with little use. No user
notes have been found. The changes are part of the manual as an additional
section in the back.

I suspect that since it wasn't used much, no one noticed the missing pages.
Even the two brochures were in immaculate condition.

I am guessing that maybe their copies are from my original and are missing
the pages in question (4-9 through 4-12). Since I didn't get a response from
them, I will call in the morning since I am in need of another manual.

Todd

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Fluke 732A Questions

2013-12-29 Thread R.Phillips

Hi Joe
I also obtained a Fluke 732A back in October last, and I have fitted a new 
set of SLA batteries and it is working just fine - I am still very impressed 
by its accuracy and stability, and its build quality.
First, the external 3-pin power socket is a HYPERTRONICS 'D' type connector 
(female) type 100-166 - but I have not found one to date !
I also expected the temperature monitoring thermistor to give a 3 to 4 K 
ohms, as indicated in the manual P/N 645051 dated May 1983 - but I guess 
that you have a later model as I do, mine being serial # 4845***, some 
modifications were made,and I understand that this included a change of 
thermistor, so the standard reading is in keeping with your result. My unit 
gives a very stable 4.520 K ohms, being between 4.52080 and 4.52090 K ohms. 
Judging by your findings, you are also using the 1983 copy of the User 
Manual - we could both use a more up to date issue.
I am awaiting the repair of my 3458A - so I am having to rely upon a 
recently calibrated Keithley 2015 and my good old 3456A - but they both give 
very impressive results - The Keithley giving 1.07 volts for the 1 volt 
output, and 10.0 volts for the 10 volt output (how I miss that final 
digit).

Looking forward to your further results and any other owners comments.
Best regards

Roy


-Original Message- 
From: J. L. Trantham

Sent: Saturday, December 28, 2013 10:51 PM
To: volt-nuts@febo.com
Subject: [volt-nuts] Fluke 732A Questions

I have reviewed the prior postings on Volt-nuts (a very rewarding review,
BTW)) regarding the 732A and have two questions:



1.Has anyone determined the part number, or a source, for the mating
plug to J10, the external power connector for the 732A-7005 battery pack?  I
noted some postings recently about this but did not find a definitive
response regarding the identity of this plug.  The alternative would appear
to be a complete replacement of J10 and its plug, as long is it all fits in
the opening in the panel.  I hate to 'bore holes' in vintage equipment.

2.   The oven thermistor in my 'new to me' unit measures 4229 ohms (+/-
an ohm or two) after the unit has warmed up for a week or so.  I note the
manual refers to 3K to 4K for the value of this thermistor when the unit is
'stable' as well as other's posting values in the mid 3500's ohms for their
units.  Should I be concerned?  The unit seems to be stable to within about
2 uV over about a week (as measured by my 3458A - see below).  Should I open
the unit and try to measure the oven temperature or just be satisfied that
the unit seems to be working?



I had to replace the four 6V 4AH SLA batteries and they charged up
appropriately as judged by the front panel LED's.  I had to remove the
'jumper' for the '40' option on the A7 board and connect the jumper to the
'20' and '10' options (total of '30') in order to get the unit to adjust to
10.000 VDC on my 'Agilent In Cal'd' 3458A.



The need to change the jumpers, perhaps, could be just an ageing issue or,
on the other hand, a 'temperature' issue with the oven.



Should I open the unit and directly measure the temperature (supposedly
about 48 degrees C) or just be satisfied with what I have?



My recently added 735C also needed moving some 'jumpers' in order to get it
'on scale', as determined by my 3458A, although it's thermistor measures
about 3330 ohms (after being on for several weeks).



Thanks for everyone's help.



Joe







___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there. 



___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Fluke 732A Questions

2013-12-29 Thread J. L. Trantham
Roy,

Thanks for the info.

Looking at the Hypertronics catalog, it would appear that the panel mounted
socket (with male, solder cup, pins) is:

P/N D01EEB306MSUTH

And the mating plug would be:

P/N D01PB306FSUTAH

However, the dimensions are a bit off and the appearance of the panel
mounted socket is not the same as what is on my 732A, J10, on the back of
the 732A-7005 battery module.  My connector is exactly as shown in the
change sheets with a dimension of 0.475 inches side to side and 0.500 inches
top to bottom (as viewed from the back of the unit) rather than 0.512 x
0.512 as suggested in the catalog.

However, the diameter of the mounting hole and the presence of a 'notch'
appears to be as shown in the catalog, thus suggesting that it, indeed, is a
series D01 part.  The diameter of the threaded part of J10 is 0.410 inches
and the catalog calls for a hole diameter of 0.441 inches.

One other observation is the presence of a 'symbol' on the bottom edge of
the socket (as viewed from below) that looks like a stylized 'FRE',  The 'F'
perhaps could be an 'A' and the 'E' perhaps could be a 'B'.  The symbol
starts low and ends low with a 'peak' of the symbol in the middle of the
'R', sort of a 'roof' shape to the top of the symbol and flat on the bottom
of the symbol.

Not sure what that means but I have a picture of it if anyone thinks it
would help.  It does not look like the stylized 'H' symbol shown in the
Hpertronic catalog.  Perhaps the company was something else before they
became Hypertronics.

I looked through the change sheets for the 732A and could find no mention of
the value of the thermistor resistance at temperature, only comments about
what its stability should be.  Knowing that you have a value in the mid 4K
range is reassuring.  My S/N starts with 3.

Right now, it indicates 10.005 VDC on the 3458A and 4230 ohms on the
Fluke 8050A.

Thanks for your help.

Joe

-Original Message-
From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
Behalf Of R.Phillips
Sent: Sunday, December 29, 2013 8:40 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Fluke 732A Questions

Hi Joe
I also obtained a Fluke 732A back in October last, and I have fitted a new
set of SLA batteries and it is working just fine - I am still very impressed
by its accuracy and stability, and its build quality.
First, the external 3-pin power socket is a HYPERTRONICS 'D' type connector
(female) type 100-166 - but I have not found one to date !
I also expected the temperature monitoring thermistor to give a 3 to 4 K
ohms, as indicated in the manual P/N 645051 dated May 1983 - but I guess
that you have a later model as I do, mine being serial # 4845***, some
modifications were made,and I understand that this included a change of
thermistor, so the standard reading is in keeping with your result. My unit
gives a very stable 4.520 K ohms, being between 4.52080 and 4.52090 K ohms. 
Judging by your findings, you are also using the 1983 copy of the User
Manual - we could both use a more up to date issue.
I am awaiting the repair of my 3458A - so I am having to rely upon a
recently calibrated Keithley 2015 and my good old 3456A - but they both give
very impressive results - The Keithley giving 1.07 volts for the 1 volt
output, and 10.0 volts for the 10 volt output (how I miss that final
digit).
Looking forward to your further results and any other owners comments.
Best regards

Roy


-Original Message-
From: J. L. Trantham
Sent: Saturday, December 28, 2013 10:51 PM
To: volt-nuts@febo.com
Subject: [volt-nuts] Fluke 732A Questions

I have reviewed the prior postings on Volt-nuts (a very rewarding review,
BTW)) regarding the 732A and have two questions:



1.Has anyone determined the part number, or a source, for the mating
plug to J10, the external power connector for the 732A-7005 battery pack?  I
noted some postings recently about this but did not find a definitive
response regarding the identity of this plug.  The alternative would appear
to be a complete replacement of J10 and its plug, as long is it all fits in
the opening in the panel.  I hate to 'bore holes' in vintage equipment.

2.   The oven thermistor in my 'new to me' unit measures 4229 ohms (+/-
an ohm or two) after the unit has warmed up for a week or so.  I note the
manual refers to 3K to 4K for the value of this thermistor when the unit is
'stable' as well as other's posting values in the mid 3500's ohms for their
units.  Should I be concerned?  The unit seems to be stable to within about
2 uV over about a week (as measured by my 3458A - see below).  Should I open
the unit and try to measure the oven temperature or just be satisfied that
the unit seems to be working?



I had to replace the four 6V 4AH SLA batteries and they charged up
appropriately as judged by the front panel LED's.  I had to remove the
'jumper' for the '40' option on the A7 board and connect the jumper to the
'20

Re: [volt-nuts] Fluke 732A Questions

2013-12-29 Thread Robert Atkinson
Hi Joe,
As Roy has already said, this is a hypertronics connector. I've had one other 
piece of equipment with this type of connector, a nitrogen laser. See  
http://www.thinksrs.com/products/NL100.htm  The downloadable user manual has 
details. Sorry, I don't have any, or a cheap source.

Robert G8RPI.





 From: J. L. Trantham jlt...@att.net
To: volt-nuts@febo.com 
Sent: Saturday, 28 December 2013, 22:51
Subject: [volt-nuts] Fluke 732A Questions
 

I have reviewed the prior postings on Volt-nuts (a very rewarding review,
BTW)) regarding the 732A and have two questions:



1.        Has anyone determined the part number, or a source, for the mating
plug to J10, the external power connector for the 732A-7005 battery pack?  I
noted some postings recently about this but did not find a definitive
response regarding the identity of this plug.  The alternative would appear
to be a complete replacement of J10 and its plug, as long is it all fits in
the opening in the panel.  I hate to 'bore holes' in vintage equipment.

2.       The oven thermistor in my 'new to me' unit measures 4229 ohms (+/-
an ohm or two) after the unit has warmed up for a week or so.  I note the
manual refers to 3K to 4K for the value of this thermistor when the unit is
'stable' as well as other's posting values in the mid 3500's ohms for their
units.  Should I be concerned?  The unit seems to be stable to within about
2 uV over about a week (as measured by my 3458A - see below).  Should I open
the unit and try to measure the oven temperature or just be satisfied that
the unit seems to be working?



I had to replace the four 6V 4AH SLA batteries and they charged up
appropriately as judged by the front panel LED's.  I had to remove the
'jumper' for the '40' option on the A7 board and connect the jumper to the
'20' and '10' options (total of '30') in order to get the unit to adjust to
10.000 VDC on my 'Agilent In Cal'd' 3458A.  



The need to change the jumpers, perhaps, could be just an ageing issue or,
on the other hand, a 'temperature' issue with the oven.



Should I open the unit and directly measure the temperature (supposedly
about 48 degrees C) or just be satisfied with what I have?



My recently added 735C also needed moving some 'jumpers' in order to get it
'on scale', as determined by my 3458A, although it's thermistor measures
about 3330 ohms (after being on for several weeks).



Thanks for everyone's help.



Joe

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Fluke 732A Questions

2013-12-29 Thread T . Micallef
Joe,

I have the external battery pack (732A-7003) that I picked up a while ago.
It is indeed the D01 connector. It is printed clearly on the outside edge
along with the three letters that look like ARB.

The company's website states they are now owned by Smiths Connectors and the
connector is part of Hypertac. No list of distributors was found on the website.

I found a seller Astrex Electronics, but they do not have a price listed for
the female version, just the male. I searched for D01P306FSTAH. This is the
female plug with solder cups.

To date, I have reconditioned three 732A's and they have been operating fine
with little issues, except when I lost power for an extended time and two
batteries in one of the packs died. All four were replaced and they have
been working fine.

The two older 732A's have thermistor values in the 3k-4k range. The newer
unit (based on the newer heater rev) is around 4.25K.

All three had their potentiometers changed on the regulator and
pre-regulator boards. The datecodes on all appeared to be from 1985. The
parts list must have been updated as some of the pots did not have the same
value as the ones installed. I cannot remember which ones were different. I
did not see a mention in the change/errata.

Also, I had to remove the reference ovens to repair broken 10V
potentiometers, as someone had broken one (free spinning) and another unit
appeared to have a noisy wiper so it was replaced as well. The 1V and 1.018V
pots were not changed. The third 10V pot was replaced because I had success
with the other two and it was much easier the third time around. The jumpers
on the all the A7 boards were adjusted.

I highly recommend checking the inline fuse to the heater if you suspect it
is not working. You can find it tucked inside behind the oven. It should be
covered in heat shrink. 

Lastly, there was a rev done that does not appear in the downloadable users
manual on Didier's website.  It appears the heaters were separated into two
groups of two instead of one group of four. Some of the parts were mounted
under the oven assembly on the A8 Preheater PCA. If you look at the A4 board
from the top, you will see either one or two transistors mounted to the
frame (Q4,Q14). If you see only one (Q4), chances are you have the newer rev.

I bought the manual recently and just noticed these revs were in my book.
I will try and scan the extra info and upload them.

Todd




___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Fluke 732A Questions

2013-12-29 Thread J. L. Trantham
Todd,

Very good information.

Sounds like we are getting close to the correct P/N for the mating plug for
the socket.

My 732A has only a single transistor mounted to the frame from the A4
Regulator Assembly, in the position of Q4, and the resistance is 4230 ohms
after the unit is 'warmed up' and stable.

I recently bought a manual as well but it did not have any 'update sheets'.
I would love to have a copy of your 'update sheets'.

I did a 'clean up' of the update sheets on Didier's site and can send them
to anyone that wants one.

I printed the .PDF from the Fluke site and added the 'cleaned up' version of
the update sheets and it makes a very nice manual.  I would love to add the
other sheets to this.

I also have a 735C that is very similar to the 732A but lot's of mechanical
differences.  It has two transistors on the Regulator Assembly, presumably
A4 and Q4 and Q14.  I do not have a manual for the 735C. 

It's thermistor value is in the mid 3K range.  It, too, is very stable and
has many similarities to the 732A but has recently developed a 'new
problem'.  The 'NO CAL' LED comes on after a few days for no apparent
reason.  I need to open it up again and take a look at that circuit to see
what the problem is.

On the 732A, the 'IN CAL' LED is continuously lighted when there has been no
power interruption.  On the 735C, the 'NO CAL' LED illuminates when power is
restored after a power failure.  

I replaced the NiCd battery pack and all seems to be well but the 'NO CAL'
LED comes on after having been reset (by inserting a copper wire in the
'adjust' opening).  I wonder if my 'reset' process is the problem.  On the
732A, you connect to the negative terminal then touch the 'pad' inside the
'RESET' opening.  There is no such 'RESET' opening on the 735C.  Just
placing the wire (that I use to adjust the 10 V adjustment) inside the
adjustment opening and aiming to the side would reset the LED.

Happy to help with the scanning of the manual and update sheets if needed.

Thanks again for the info.

Joe

-Original Message-
From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
Behalf Of T. Micallef
Sent: Sunday, December 29, 2013 6:05 PM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Fluke 732A Questions

Joe,

I have the external battery pack (732A-7003) that I picked up a while ago.
It is indeed the D01 connector. It is printed clearly on the outside edge
along with the three letters that look like ARB.

The company's website states they are now owned by Smiths Connectors and the
connector is part of Hypertac. No list of distributors was found on the
website.

I found a seller Astrex Electronics, but they do not have a price listed for
the female version, just the male. I searched for D01P306FSTAH. This is the
female plug with solder cups.

To date, I have reconditioned three 732A's and they have been operating fine
with little issues, except when I lost power for an extended time and two
batteries in one of the packs died. All four were replaced and they have
been working fine.

The two older 732A's have thermistor values in the 3k-4k range. The newer
unit (based on the newer heater rev) is around 4.25K.

All three had their potentiometers changed on the regulator and
pre-regulator boards. The datecodes on all appeared to be from 1985. The
parts list must have been updated as some of the pots did not have the same
value as the ones installed. I cannot remember which ones were different. I
did not see a mention in the change/errata.

Also, I had to remove the reference ovens to repair broken 10V
potentiometers, as someone had broken one (free spinning) and another unit
appeared to have a noisy wiper so it was replaced as well. The 1V and 1.018V
pots were not changed. The third 10V pot was replaced because I had success
with the other two and it was much easier the third time around. The jumpers
on the all the A7 boards were adjusted.

I highly recommend checking the inline fuse to the heater if you suspect it
is not working. You can find it tucked inside behind the oven. It should be
covered in heat shrink. 

Lastly, there was a rev done that does not appear in the downloadable users
manual on Didier's website.  It appears the heaters were separated into two
groups of two instead of one group of four. Some of the parts were mounted
under the oven assembly on the A8 Preheater PCA. If you look at the A4 board
from the top, you will see either one or two transistors mounted to the
frame (Q4,Q14). If you see only one (Q4), chances are you have the newer
rev.

I bought the manual recently and just noticed these revs were in my book.
I will try and scan the extra info and upload them.

Todd




___
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts

Re: [volt-nuts] Fluke 732A Questions

2013-12-29 Thread T . Micallef
Joe,

I don't know if it makes any difference but my manual is for the 732A/AN p/n
788414

I found a couple of brochures that I scanned and uploaded too.

I forgot to mention, I have been collecting information regarding the
techniques used to do intercomparisons using scanners like the Data Proof
models. While searching a couple of weeks ago, I came across an old document
from a Measurement Science Conference back in 1983. The document contains an
article regarding the testing of the Fluke 735C. I thought you may have
little to no info regarding your reference and thought I would let you know
about it.

The document is easy to find using google. It is named MSC1983.pdf
I will upload it as well, but I was not sure if Didier wants to track
conference docs.

Todd 





___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts