Dear all,
does anybody know which safety standard should apply to a colposcope?
thanks
--
Paolo Peruzzi
Regulatory Medical RD
El.En. S.p.A.
Via Baldanzese, 17
50041 Calenzano (FI)
Italy
Tel. +39 055 8826807
FAX +39 055 8832884
standa...@elen.it mailto:standa...@elen.it
I want to share with you a potential pitfall in conducted emission testing of
IT equipment using an ISN (or alternative) on an Ethernet connection.
The pitfall will have to lead to adjustment of the test requirements for
IT equipment in this set-up, and/or to a change in requirements for
In message 491c02c6.6060...@elen.it, dated Thu, 13 Nov 2008, Paolo
Peruzzi standa...@elen.it writes:
does anybody know which safety standard should apply to a colposcope?
It's a medical device that contacts the patient. So IEC/EN
60601 applies. If you want to know which Section, look on the
A bit off topic, but list related:
I want to plea for allowing small attachments
and HTML messages to this list.
Most of the time my contributions are refused
because my company logo is recognized as a forbidden attachment.
Well, it is 2009 , not 1999 anymore, and small attachments
of
In message FCA549BE3ECF9D4CB8CB8576837EA489053330@ZEUS.cetest.local,
dated Thu, 13 Nov 2008, Gert Gremmen g.grem...@cetest.nl writes:
I want to plea for allowing small attachments
and HTML messages to this list.
I don't understand why you want to do that. I think that emails are like
Dear Paolo,
apparently there is no particular standard for a colposcope, so the general
standard is use for safety, EN 60601-1 or IEC 60601-1.
Best regards,
__
Eng. Silvia Díaz Monnier
NATIONAL INSTITUTE OF INDUSTRIAL TECHNOLOGY
R D Center of
Well John,
Indeed, it's not 1999 anymore.
Not seeing the need does not
imply others don't.
It would facilitate the use of tables,
small pictures (if compressed under 10k),
colors (to emphasize).
EMC-PSTC notes are not handwritten, but type-writer written.
If they were really hand written we
In message FCA549BE3ECF9D4CB8CB8576837EA489053331@ZEUS.cetest.local,
dated Thu, 13 Nov 2008, Gert Gremmen g.grem...@cetest.nl writes:
Well John,
Indeed, it's not 1999 anymore.
Not seeing the need does not
imply others don't.
I expressed my opinion. I did not say that everyone has to agree
Ron,
I suppose the questions from my post you understood as my assertions. I've
just asked based on what I should assume it is no 'apparatus'.
My product is in my opinion finished (no more assembly by producer or user
is planned) and will be marketed commercially.
I have in mind the open
In message 7861714BC9794A2C8FE39FAD9F42BC3C@MmPc21, dated Thu, 13 Nov
2008, Piotr Galka piotr.ga...@micromade.pl writes:
I have in mind the open (without case) boards question just from the
beginning of my interest in EMC (2 years before we joined EU in 2004)
because since 1992 we have in
- Original Message -
From: John Woodgate j...@jmwa.demon.co.uk
Then you use the procedure given in Annex III of the 2004/108/EC Directive
to get a formal acceptance of your solution by a Notified Body. You can
then, under Article 7, apply the CE mark.
And probably it is the only
In message B3AA818630C942DBA06B45F1E778B315@MmPc21, dated Thu, 13 Nov
2008, Piotr Galka piotr.ga...@micromade.pl writes:
Reading this Annex it looks that I first have to have the device and
then can ask if it is OK.
Well, what I would do is to determine how far away from the board the
Hello Group,
I have a question regarding the Electric Strength test in 60950-1 section 5.2.
It says in that section that you have a failure or the insulation broke down
when the current which flows as a result of the application of the test
voltage rapidly increases in an uncontrolled manner.
Hello Rob,
There is no current limit specified, and there is no way a limit could be
specified. Take a look at IEC 60950-1 section 5.1.7. You will see that there
are types of equipment where the touch current is allowed to exceed 3.5 mA.
This is typically equipment that is field wired or at
No specific limit for the di-electric withstand test. There is no
mention of the term short in the di-electric withstand type
test for I.T.
According to CSA engineering guidelines and some UL (Section
General) files, each individual model should have a
pre-determined limit.
From:
In message 491c5758.8090...@cclab.com, dated Thu, 13 Nov 2008, Robert
F. Keller r...@cclab.com writes:
I have a question regarding the Electric Strength test in 60950-1
section 5.2. It says in that section that you have a failure or the
insulation broke down when the current which flows as
Hi,
I need to run EMC tests on a piece of high-power medical equipment that is
used on an intermittent basis. The run time is about 15 seconds to 2 minutes
on-time, followed by about 10 minutes off-time. This is comparable to typical
system usage. This causes problems with EMC testing, since
In message
off6ed2562.359b1b29-on88257500.005e3ee6-88257500.00614...@slpower.com,
dated Thu, 13 Nov 2008, pat.law...@slpower.com writes:
I need to run EMC tests on a piece of high-power medical equipment that
is used on an intermittent basis. The run time is about 15 seconds to
2 minutes
From: emc-p...@ieee.org [mailto:emc-p...@ieee.org] On Behalf Of
pat.law...@slpower.com
Sent: Thursday, November 13, 2008 9:43 AM
To: emc-p...@ieee.org
Subject: Test setup for equipment operated intermittently
Hi Folks
From experience, I agree with what Ted says - especially w.r.t. his last
paragraph about setting a LOWER limit.
Failure to exceed the lower limit could also indicate a faulty EMI filter with
capacitors of lower values than specification - thus meaning you may have an
emissions issue
Maybe I should have said:
... since you normally need more than 2 minutes to complete an EMC test.
I was thinking about radiated emissions testing, where the spectrum is
manually scanned, or line surge testing, requiring dozens of pulses applied at
low repetition rates. RF immunity testing
In message
of8478fd46.7320072f-on88257500.00746b64-88257500.00766...@slpower.com,
dated Thu, 13 Nov 2008, pat.law...@slpower.com writes:
Maybe I should have said:
... since you normally need more than 2 minutes to complete an EMC
test.
I was thinking about radiated emissions testing, where
22 matches