Hi, Stephen, thank you for your reply. Only a few comments still: Stephen Quinto schrieb: > > Hi, Guenther, > > First, these micrographs were taken at 100,000x for comparative purposes. > But we have pics also at 480,000x (the limit of our machine). There is also > an optical scope with 8x objective lenses that further magnifies the image > on the plate, so theoretically we can see 3,840,000x. But resolution is > super difficult since the additional range is an image of an image (like a > copy of a copy). However....
The resolution of a microscope is limited, roughly spoken, by the wavelength of the light. (xmin=1.22xlambda/(nxsin(alpha/2); with xmin=minimal distance, lambda=wavelength, n=refractive index between object and lens, alpha=opening angle of the light cone between object and lens). For an optical microscope this is say 500nm. Magnification by a factor of 1000 expand 500 nm objects to 0.5mm. Even if you magnify further you can't reveal details of this image. They are not present. > > You're right, the samples are dried first, ie. the water is evaporated > because the vacuum in the scope must be at least atmosphere x 10 to the -6. > The solution of this problem is probably the TEM with frozen objects, which was mentioned by James Osbourne, Holmes in this thread. I am not an expert on that. snip ....... Guenter -- The silver-list is a moderated forum for discussion of colloidal silver. To join or quit silver-list or silver-digest send an e-mail message to: [email protected] -or- [email protected] with the word subscribe or unsubscribe in the SUBJECT line. To post, address your message to: [email protected] Silver-list archive: http://escribe.com/health/thesilverlist/index.html List maintainer: Mike Devour <[email protected]>