thank you for your reply. Only a few comments still:

Stephen Quinto schrieb:
> 
> Hi, Guenther,
> 
> First, these micrographs were taken at 100,000x for comparative purposes.
> But we have pics also at 480,000x (the limit of our machine).  There is also
> an optical scope with 8x objective lenses that further magnifies the image
> on the plate, so theoretically we can see 3,840,000x.  But  resolution is
> super difficult since the additional range is an image of an image (like a
> copy of a copy).  However....

The resolution of a microscope is limited, roughly spoken, by the
wavelength of the light. (xmin=1.22xlambda/(nxsin(alpha/2); with
xmin=minimal distance, lambda=wavelength, n=refractive index between
object and lens, alpha=opening angle of the light cone between object
and lens).
For an optical microscope this is say 500nm. Magnification by a factor
of 1000 expand 500 nm objects to 0.5mm. Even if you magnify further you
can't reveal details of this image. They are not present.

> 
> You're right, the samples are dried first, ie. the water is evaporated
> because the vacuum in the scope must be at least atmosphere x 10 to the -6.
> 

The solution of this problem is probably the TEM with frozen objects,
which was mentioned by James Osbourne, Holmes in this thread. I am not
an expert on that.

snip
.......

Guenter


--
The silver-list is a moderated forum for discussion of colloidal silver.

To join or quit silver-list or silver-digest send an e-mail message to: 
[email protected]  -or-  [email protected]
with the word subscribe or unsubscribe in the SUBJECT line.

To post, address your message to: [email protected]
Silver-list archive: http://escribe.com/health/thesilverlist/index.html
List maintainer: Mike Devour <[email protected]>