Bob, thanks for clarification!

From: "Bob kb8tq" <kb...@n1k.org>
If you collect data over the entire second and average that down for a single point, then no, your ADEV will not be correct.

That probably explains why I got so nice (and suspicious) plots :)

There are a number of papers on this. What ADEV wants to see is a single phase “sample” at one second spacing.

After I read your answer I remembered some nice papers from prof. Rubiola, me bad - I was able to answer my question by myself. When we take a single phase "sample" at start and at end time of the "each tau" it is equivalent to summing all timestamps intervals I collect during that "tau", but by doing LR processing I calculate *weighted* sum, so the results will differ. So, it appears "ADEV" calculations is PDEV (parabolic) in reality, because of the current firmware processing.

I made a test with two plots for illustration - one is the classical ADEV (with 2.5ns time resolution), the second one with LR processed data (5e6 timestamps per second). Both plots are made from the same data. It is obvious the classical ADEV is limited by the counter resolution in the left part of the plot. It is interesting is it possible to use the 4999998 extra points per each second to improve counter resolution in ADEV measurements without affecting ADEV?

Thanks!
Oleg UR3IQO
_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to