A calibration indicates that the unit under test is withing manufactures specification. The equipment and procedure used has to be "good enough" (bad words in a cal lab) to have a high probability (nothing is 100%) of insuring the calibration documentation is valid. Things can can be a little looser if you are calibrating a 1% meter with a 10 ppm meter but it does not work the other way around.
On Mon, Aug 12, 2013 at 9:43 AM, Mike S <[email protected]> wrote: > On 8/12/2013 12:21 PM, Charles Steinmetz wrote: > >> No, you could not perform ANY traceable calibration with the 3458A >> itself, much less with any instrument you had calibrated with the 3458A, >> because *you* are not accredited >> > > That's simply not true. Some organizations may require a lab to be > accredited in order to accept their services, but it is not a requirement > in order to legitimately claim traceability. > > The NIST Traceability Policy is found here: > http://www.nist.gov/**traceability/nist_**traceability_policy_external.** > cfm<http://www.nist.gov/traceability/nist_traceability_policy_external.cfm> > > > ______________________________**_________________ > volt-nuts mailing list -- [email protected] > To unsubscribe, go to https://www.febo.com/cgi-bin/** > mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts> > and follow the instructions there. > -- John Phillips _______________________________________________ volt-nuts mailing list -- [email protected] To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.