Hello,
 
Joe Geller once collected 3458A serial numbers, and concluded, that this 
instrument maybe sold to about 50k units in these 25 years. It's not clear, if 
it's still sold by high numbers, but the total turnover might have been around 
400M$, or 16M$/yr.
 
In the end, that should be enough to finance a redesign, or a model facelift.
This is urgently necessary, even if Keysight would not see a totally new or 
increasing market for 8.5 digits DMMs, but only wants to still offer this 
instrument 'as-is'.
Many components of the 3458A are already obsolete, or endangered by PTNs, not 
to speak about all these through-hole components. I've already seen pictures 
about a piggy-back solutions for several ICs, and maybe they have to use that 
already for the new production, especially the two fast comparators EL2010, 
U142 & U181, used with the A/D.
The 68HC000 is also obsolete in the DIL package, and the SMD package is 'not 
for new design' already.
 
Therefore, a complete redesign, including the software architecture, is more 
reasonable.
 
Keysight would need the budget for that, but they are meanwhile also lacking 
the brains, which have mostly left the company (Wayne C. Goeke, the inventor of 
the A/D, joined Keithley, and Ronald L. Swerlein, the God-father of the ACV 
processing, well he's retired, for some personal reasons, obviously).
 
Then, another big problem would arise, that is the verification/validation of 
the traceability of the 2 source / autocalibration feature.
It would be not so easy to again achieve the acceptance of the 'metrological 
community', if any of the crucial parts of this instrument would be touched.
I assume, this direct acceptance in 1989 was only due to the close cooperation 
with the NBS then, when they validated the ~0.02ppm linearity by means of the 
new JJ array.

I have read a lot about the history of the very similar FLUKE 5700/5720 
artefact calibration.
In contrast, it took FLUKE several years, before their instruments experienced 
the same reputation.
 
Well, the 3458A was designed for metrological use in 2nd instance only, due to 
the 55°C ambient operating temperature, and these many compromises they had to 
make concerning stability.. especially the LTZ1000A reference could have been 
optimized greatly (8x) with 20°C lower ambient requirements, and a bit more 
cleverness.
Regarding this aspect, please compare the stability specs to other real 
metrological instruments, like the FLUKE 732A/B, the 7001, and the 1281 / 8508A 
8.5digits DMM.
 
So, the 3458A was  mainly intended for military conditions, but also for harsh 
industrial application, e.g. end of line testing at the manufacturing line, 
where laboratory conditions can not be maintained.
 
I also think, that the mediocre / cheap (copy-and-paste) design of these new 
6.5 .. 7.5 digits DMMs still leaves a big field of other applications for 
precise 8.5 digits DMMs, as it always has been..
.
I used this instrument already in 1990, at university, for high SNR, low 
distortion digitizing @ 16bit/100kHz or 18bit/50kHz, down to -100dB / 0.001%, 
single shot.
The 3458A may still be benchmark in this category, probably also compared to 
modern delta sigma A/Ds, but for sure compared to the recent, new 6.5 and 7.5 
DMMs.
For my experiments, I also had the necessity to design and to adjust several 
precision current sources, DCI < 0.01%, ACI < 0.05%.
That's not yet a true 'metrological' application.. But if you study the 
specifications of these new DMMs, even the 7.5digits 344470A will still not 
manage that level of uncertainty, if you take the 90 days spec, or their T.C.s.
 
Generally, their crucial parameters do not fit their resolution.
All of them have an A/D (multislope IV), wich are linear to 1..3 ppm only.
A 7.5 digit instrument would instead require 0.1ppm linearity, otherwise the 
resolution is useless. For that reason also, the featured autocal function does 
not work like in the 3458A, not a quarter as good!
 
Same goes for the mid and long term stability and the T.C.s of the references 
and the ranges.. these are 2 or even 3 orders of magnitude beyond the claimed 
resolution.. that simply does not match.
 
The 3458A instead gives much better reliability, and comfortableness to the 
user, especially by it's unique autocal feature, which relevance can't be 
emphasized enough. 
Also, only the 0.02ppm linearity legitimates the 8.5digit resolution, and 
allows ultra precise ratio measurements.
 

A revised version should have lower ambient temperature specification, if used 
in metrology, more stable Volt and Ohm references, and better resistor 
networks, like in the 1281/8508A.
(It's a shame, that the 3458A does not even have Ohm transfer uncertainty 
specification, and only 12ppm uncertainty for 1KV DC.)
Then it would be prepared also for the upcoming uncertainty improvements on the 
electrical units in 2018, by the planned new definition of the SI. The 
uncertainty will jump from about 0.2ppm down to 0.001ppm for the mise en 
pratique, and to zero uncertainty for the definition.

I would not buy a revised instrument (3458B ?), as I hope that my existing 
3458A will survive another 15 years.
 
Maybe, I furthermore upgrade some key components like the 40k standard 
resistor, the DCI shunts, and part of the Ohm circuitry, instead..

Frank


---
Alle Postfächer an einem Ort. Jetzt wechseln und E-Mail-Adresse mitnehmen! 
Rundum glücklich mit freenetMail
_______________________________________________
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Reply via email to