Ed,

Thank you for the tip on "warm-up" time -- or rather in this case, time for leakage to cool down. I will definitely give the 616 a chance by leaving it on for the next 24 hours.

Last time I did a measurement, it settled in the +20 fA range, and I confirmed this number by doing a dV/dT measurement in the charge mode. In this configuration, the input current plus leakage gets to charge one of the built-in capacitors (I selected the 1E-9 range) and the rate of voltage change (dV/dT) is a good indicator of current.

I once had a K610B electrometer, which as you know used a differential pair of 5886 tubes. It worked, but zero drift was very bothersome. I was also weary of using a tube-based instrument, which at the age of 50 could fail at any time with a dead tube. I finally sold it and bought the K616.

That's all for now, I'll write again tomorrow with measurement results !

Joel Setton

_______________________________________________
volt-nuts mailing list -- [email protected]
To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.

Reply via email to