Thanks to all for very helpful responses. I have some great leads now that I'm looking into.

Neil, here's a phase noise spec for an HP/Keysight 8648. It's good enough to measure phase noise in moderate quality clocks, but a better setup can measure down to around -170 dBc/Hz at 100 kHz offset from the carrier.



Best,
Karl

On 8/21/2024 2:02 AM, salmon.na via casper@lists.berkeley.edu wrote:

Hi Karl,

It’s perhaps a naïve question, but how far can you get in measuring phase noise using a good spectrum analyser?

I’ve an old HP dial up oscillator up to 40 GHz tube source that has phase noise -107 dBc/Hz 100 kHz from the carrier.

Cheers, Neil

*From:*casper@lists.berkeley.edu <casper@lists.berkeley.edu> *On Behalf Of *Daniel Blakley
*Sent:* 21 August 2024 05:57
*To:* casper@lists.berkeley.edu
*Subject:* Re: [casper] Low cost phase noise analysis

Dear Karl,

   I definitely don't have a simple quick answer to your question, so this is a good place to ask others who may.

   I find and found Phase Noise analysis and its measurement to be very interesting.  As you know, fundamentally, phase noise and Alan Deviation are very closely related, as is the measurement of clock jitter.  It is significant to note that NIST (Boulder CO) historically has made significant contributions to Phase Noise Analysis, beginning long ago with the work of David Alan (to which Alan Deviation owes its namesake).  More recently (several years ago) again in significant work in phase noise measurement, NIST introduced a new, more accurate, phase noise measurement architecture and method.  Out of this work, came to pass several instruments which largely emulated or followed this new architecture that is evident in some of the Keysight phase noise offerings as well as other instruments from manufacturers such as Holzworth, Rhode & Schwartz, et al.

  -Daniel Blakley

On Tue, Aug 20, 2024 at 2:37 PM Karl Warnick <warn...@ee.byu.edu> wrote:

    Hi all,

    I've spent some time this summer as part of a radar project
    digging into
    calculating phase noise for highly stable tones. I have
    implemented what
    I think is a decent algorithm. My next steps are to look for test
    data
    sets and tips for the hardware.

    Do you have a file of samples of a stable tone? If anyone has a test
    data set consisting of samples of a pure tone that they would like to
    share as a test data set, I'd like to apply my codes to that and
    check
    the phase noise. Both the tone generator and the ADC sample clock
    should
    be phase stable to the order of a Keysight signal generator, or
    ideally
    better. The data set length should be a reasonable fraction of a
    second
    for ~1 Hz phase noise resolution. The frequency of the tone and the
    sample rate are fairly arbitrary as I'm mainly looking to
    benchmark the
    algorithm.

    How cheaply can stable samples be acquired? I'm looking for low cost
    hardware (a few $100s up to a few $k) that is stable enough to
    measure
    phase noise comparable to a Keysight source or better. Phase noise
    can
    be measured with an expensive phase noise analyzer, but I believe it
    should be possible to do this with a low cost digitizer with a
    suitably
    stable sample clock. The sample clock could (or perhaps must) be
    external. The sample rate should be around 80-100 Msps or higher
    and the
    platform should be able to store a burst of samples of length on the
    order of 1 sec. We have done this using a ZCU 216 and it seems to
    work,
    but that isn't really a low cost board. I've looked into Picoscope
    products, which might be ideal, but their support people don't know
    anything about the phase noise properties of their samplers.

    Thanks in advance to anyone whose interest is piqued enough to
    respond.

    Best,
    Karl

-- Karl F. Warnick
    Parkinson Engineering Research Professor
    Department of Electrical and Computer Engineering
    Brigham Young University
    450 Engineering Building
    Provo, UT 84602
    (801) 422-1732





-- You received this message because you are subscribed to the Google
    Groups "casper@lists.berkeley.edu" group.
    To unsubscribe from this group and stop receiving emails from it,
    send an email to casper+unsubscr...@lists.berkeley.edu
    <mailto:casper%2bunsubscr...@lists.berkeley.edu>.
    To view this discussion on the web visit
    
https://groups.google.com/a/lists.berkeley.edu/d/msgid/casper/8839ddb3-83fd-40be-8a9d-c90ae6f9678e%40ee.byu.edu.


--

-Daniel Blakley

--
You received this message because you are subscribed to the Google Groups "casper@lists.berkeley.edu" group. To unsubscribe from this group and stop receiving emails from it, send an email to casper+unsubscr...@lists.berkeley.edu. To view this discussion on the web visit https://groups.google.com/a/lists.berkeley.edu/d/msgid/casper/CAJa6%3DL1qhK7U6BZCpKxcVq6R_jiUyVtOF11Sy7rxaojsVt_OhQ%40mail.gmail.com <https://groups.google.com/a/lists.berkeley.edu/d/msgid/casper/CAJa6%3DL1qhK7U6BZCpKxcVq6R_jiUyVtOF11Sy7rxaojsVt_OhQ%40mail.gmail.com?utm_medium=email&utm_source=footer>.

--
You received this message because you are subscribed to the Google Groups "casper@lists.berkeley.edu" group. To unsubscribe from this group and stop receiving emails from it, send an email to casper+unsubscr...@lists.berkeley.edu. To view this discussion on the web visit https://groups.google.com/a/lists.berkeley.edu/d/msgid/casper/001a01daf3a0%2473c331e0%245b4995a0%24%40tiscali.co.uk <https://groups.google.com/a/lists.berkeley.edu/d/msgid/casper/001a01daf3a0%2473c331e0%245b4995a0%24%40tiscali.co.uk?utm_medium=email&utm_source=footer>.

--
Karl F. Warnick
Parkinson Engineering Research Professor
Department of Electrical and Computer Engineering
Brigham Young University
450 Engineering Building
Provo, UT 84602
(801) 422-1732




--
You received this message because you are subscribed to the Google Groups 
"casper@lists.berkeley.edu" group.
To unsubscribe from this group and stop receiving emails from it, send an email 
to casper+unsubscr...@lists.berkeley.edu.
To view this discussion on the web visit 
https://groups.google.com/a/lists.berkeley.edu/d/msgid/casper/7aa228cc-c79b-467a-afc0-3608b5fad565%40ee.byu.edu.

Reply via email to