This message is in MIME format. Since your mail reader does not understand this format, some or all of this message may not be legible.
I agree with Doug and with John but, I have the same question. And I will add, Doug is speaking of the FDA-CDRH requirement for the US. But, to ascertain what the European mode is, this requires a single-fault condition for classification as described by John Juhasz. That, I understand. John, is there a free downloadable copy of ANSI Z136.1 and .2 specs?
I have used the calculations to determine the classification of a laser based on the no fault and a single fault mode. These lasers are operating in a pulsed mode at a 50% duty cycle. I have attached a sample calculation of a 1310nmn and a 1550nm laser that I used to determine that the lasers either do or do not fall within the Class 1 laser classification. Another question... Can the same label be used for FDA as for IEC? Has anyone had any feedback or problems with laser labels that deviated from recommended markings by the 21 CFR or IEC 825-1?
--- Begin Message ---Do the acid test type question ... During normal use, what are you exposing the end user to? For instance, with a laser pointer using a Class IIIb laser that emits unprotected from the pointer, the pointer is Class IIIb. If instead, you are using a Class IIIb laser for fiber optic communication and the entire beam is contained within the fiber, no lasing is emitted from the product during normal operation and/or service or maintainance, then you *could* declare the telco product as Class I. - Doug McKean ------------------------------------------- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on "browse" and then "emc-pstc mailing list"
--- End Message ---
Laser Class calculation.doc
Description: MS-Word document