We calibrate annually.
 
Minimum fit factor is 1000.
 
No margin added to minimum overall fit factor.
 
If Porta-count as-found condition is not within +/- 15%, we would investigate 
impact in accordance with our corrective action program. (If as found was low 
may be considered conservative, if as found was high could review fit test 
performed on that device and re-fit test as needed).
 

 
Denise DeGlopper
RPO Specialist
Respiratory Program Administrator 
DTE/Fermi 2
Work 734-586-1340
Pager 734-227-0218



[email protected] wrote: -----


To: "'[email protected]'" <[email protected]>
From: "Nadeau, Marc" <[email protected]>
Sent by: [email protected]
Date: 01/29/2015 12:26PM
Cc: "Perkins, David" <[email protected]>
Subject: Powernet: TSI Portacount


Seabrook Station recently sent a Portacount respirator fit test device to the 
vendor for calibration and it was found to be out of the vendor’s standard 
tolerance of ±15%.   So, we are benchmarking to understand how others maintain 
Portacount devices and are asking if you would please answer 4-benchmark 
questions? 
 
BENCHMARK QUESTIONS
If you use TSI Portacount devices to perform respirator fit testing;
 
1.       How frequently do you calibrate your Portacount device?
Why?
 
2.       What is your minimum overall fit factor?
 
3.       Do you add margin to your minimum overall fit factor in the event that 
the Portacount as-found condition is outside the vendor tolerance?
If so, what is the margin?
 
4.       What actions do you take if your Portacount device as-found condition 
is outside the ±15% vendor standard tolerance?
 
David Perkins
SR. RP Analyst
NextEra Energy –Seabrook Station
Ph:603-773-7079
 

Reply via email to