Double copies

2003-11-30 Thread David Sproul
Dear group,
Is it just me or is anyone else receiving each message twice?
 
Best regards,
David Sproul.



FW: More CE mark issues...

2003-11-11 Thread David Sproul




From: David Sproul [mailto:david.spr...@alexanderlynn.co.uk]
Sent: 11 November 2003 22:35
To: John Allen
Subject: RE: More CE mark issues...


Hi John,
Yes you are absolutely correct.  The LVD doesn't have any exemptions for
military equipment.  I was still thinking EMC directive when I responded to
Dave's posting.

I have written to him to acknowledge my misinformation, and now also feel a
little foolish about my sarcastic comments in my original e-mail.  I didn't
know about Geoff Hoon's reinforcement of civil safety legislation for
military equipment, but I do now.

Thanks for keeping me right.

Best regards,
David Sproul

From: owner-emc-p...@majordomo.ieee.org
[mailto:owner-emc-p...@majordomo.ieee.org]On Behalf Of John Allen
Sent: 11 November 2003 15:28
To: 'David Sproul'
Cc: EMC-PSTC; Coleman, David
Subject: RE: More CE mark issues...



David S

Unfortunately I have to correct you with respect to the LVD - it currently
does not have an exemption for military equipment and neither does most
(including UK) national implementing legislation.

In the UK, the need for military equipment to comply with civilian
legislation is reinforced by the Secretary of State for Defence' (Geoff
Hoon) Safety and Environment policy statement in July 2000, which basically
says the military will follow civilian legislation and standards unless
there is an overriding operational need to deviate from the latter
requirements - and even then the military must operate precautions to ensure
a similar level of overall safety (and do the formal Safety Case).

Thus my opinion is CE Mark it - And it is easier overall than explaining
why you did not

John Allen
(ex Racal/Thales, but now)
Technical Consultant
EMC and Safety Engineering
ERA Technology Ltd.
Cleeve Road
Leatherhead
Surrey KT22 7SA
UK

Tel: +44-1372-367025 (Direct)
   +44-1372-367000 (Switchboard)
Fax: +44-1372-367102


From: David Sproul [mailto:david.spr...@alexanderlynn.co.uk]
Sent: 11 November 2003 13:21
To: Coleman, David
Cc: EMC-PSTC
Subject: RE: More CE mark issues...



Hi David,
I can only answer with my own opinion which will probably be both backed up
and refuted by others who may respond.
First of all, military equipment does not have to be CE marked, unless that
has been changed very recently.  However if the equipment has to go to
Germany, they apparently insist that electronic military equipment be CE
marked.

From your description it would appear that, to your customer at least, your
IA is the part of the ATE that they will access most.
Although you are not responsible for the safe design of the ATE, unless you
are selling it to the customer, (then you share responsibility with the
manufacturer) you are responsible for your IA.

You must therefore ensure that it is safe for your customers to use.
Supplies passing through your IA from the ATE to the customers equipment
must be handled in an acceptably safe manner.  Since you are familiar with
EN 61010 you will be aware of the potential hazards you should protect
against.  From discussions with individuals from within your organisation I
know that your company can be very thorough in their risk assessment, so
identifying those risk probably won't be a problem to you either.

As you will be aware, a recent string in this group has been discussing when
to CE mark and when not.  I will dare to suggest that since without the ATE
your IA is an expensive paperweight, you need not CE mark it.  However, this
does not release you from your obligations to ensure it is safe for your
customers to use.

I hope this has helped.  Undoubtedly confusion will still reign when others
tell you that they agree with me, while other will suggest that I should
find another profession.

Best regards,
David Sproul


From: owner-emc-p...@majordomo.ieee.org
[mailto:owner-emc-p...@majordomo.ieee.org]On Behalf Of Coleman, David
Sent: 11 November 2003 08:36
To: EMC-PSTC
Subject: More CE mark issues...



My company designs and manufactures ATE systems and components for the
Military, Aerospace and Functional test markets. Often what we are supplying
is just an interface adapter (IA) that sits between another suppliers ATE
and the customers unit under test (UUT).

An IA consists of a metal box mounted on the front of the ATE, with I/O
connectors on most panels. Whilst it is normally the case that the IA
consists entirely of passive components, they do route AC power through to
the UUT. From an EMC point of view, as they are passive and a component of a
larger system (the ATE), the EMC directive need not apply.

The problem is the LVD. Whilst we design and manufacture the item to meet
the relevant harmonised standard (EN61010), the safety of the IA is often
heavily dependant on the parent ATE (for which we are not the design
authority), due to the use of safety features in the ATE (i.e. interlocks
etc.)

Historically, we have not CE marked these IAs. Are we correct?

regards,
Dave Coleman
Racal Instruments Group

RE: More CE mark issues...

2003-11-11 Thread David Sproul

Hi David,
I can only answer with my own opinion which will probably be both backed up
and refuted by others who may respond.
First of all, military equipment does not have to be CE marked, unless that
has been changed very recently.  However if the equipment has to go to
Germany, they apparently insist that electronic military equipment be CE
marked.

From your description it would appear that, to your customer at least, your
IA is the part of the ATE that they will access most.
Although you are not responsible for the safe design of the ATE, unless you
are selling it to the customer, (then you share responsibility with the
manufacturer) you are responsible for your IA.

You must therefore ensure that it is safe for your customers to use.
Supplies passing through your IA from the ATE to the customers equipment
must be handled in an acceptably safe manner.  Since you are familiar with
EN 61010 you will be aware of the potential hazards you should protect
against.  From discussions with individuals from within your organisation I
know that your company can be very thorough in their risk assessment, so
identifying those risk probably won't be a problem to you either.

As you will be aware, a recent string in this group has been discussing when
to CE mark and when not.  I will dare to suggest that since without the ATE
your IA is an expensive paperweight, you need not CE mark it.  However, this
does not release you from your obligations to ensure it is safe for your
customers to use.

I hope this has helped.  Undoubtedly confusion will still reign when others
tell you that they agree with me, while other will suggest that I should
find another profession.

Best regards,
David Sproul


From: owner-emc-p...@majordomo.ieee.org
[mailto:owner-emc-p...@majordomo.ieee.org]On Behalf Of Coleman, David
Sent: 11 November 2003 08:36
To: EMC-PSTC
Subject: More CE mark issues...



My company designs and manufactures ATE systems and components for the
Military, Aerospace and Functional test markets. Often what we are supplying
is just an interface adapter (IA) that sits between another suppliers ATE
and the customers unit under test (UUT).

An IA consists of a metal box mounted on the front of the ATE, with I/O
connectors on most panels. Whilst it is normally the case that the IA
consists entirely of passive components, they do route AC power through to
the UUT. From an EMC point of view, as they are passive and a component of a
larger system (the ATE), the EMC directive need not apply.

The problem is the LVD. Whilst we design and manufacture the item to meet
the relevant harmonised standard (EN61010), the safety of the IA is often
heavily dependant on the parent ATE (for which we are not the design
authority), due to the use of safety features in the ATE (i.e. interlocks
etc.)

Historically, we have not CE marked these IAs. Are we correct?

regards,
Dave Coleman
Racal Instruments Group


**
IMPORTANT NOTICE

The information contained in this e-mail is confidential. It may also
be legally privileged. It is intended only for the stated addressee(s)
and access to it by any other person is unauthorised. If you are not
an addressee, you must not disclose, copy, circulate or in any other
way use or rely on the information contained in this e-mail. Such
unauthorised use may be unlawful.

If you have received this e-mail in error, please inform
Racal Instruments Group Ltd. immediately by
emailing  postmas...@racalinstrumentsgroup.co.uk
or
phoning +44 (0)1202 872800  (ask for the I.T. Dept.)
and delete it and all copies from your system.

www.racalinstrumentsgroup.co.uk

**



This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   emc_p...@symbol.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

Archive is being moved, we will announce when it is back on-line.
All emc-pstc postings are archived and searchable on the web at:
http://www.ieeecommunities.org/emc-pstc



This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   emc_p...@symbol.com

For policy questions, send mail to:
 Richard Nute

RE:FCC and Canada

2003-08-26 Thread David Sproul
Many thanks to all of you who replied.  You have all been very helpful, and
have taught me quite a bit.
 
Thanks again,
 
David Sproul,
Business Development Manager,
Alexander Lynn Approvals Management Services



RE: FCC and Canada

2003-08-21 Thread David Sproul
Before anyone asks,  some of somebody means that I changed my mind on the
wording of the message and omitted to delete all of the original word.
 
Hopefully this explanation can help avert the bizarre tangential discussions
involving some of the stranger senses of humour that participate in this
discussion group (amusing as they often are.)
 
Best regards,
David Sproul.
 



FW: FCC and Canada

2003-08-21 Thread David Sproul
Dear Group,
I apologise to for sending this out twice, but the copy of my original message
arrived back with no text. 
 
I hope the text will be included this time.
 
Best regards,
David Sproul.

From: David Sproul [mailto:david.spr...@alexanderlynn.co.uk]
Sent: 20 August 2003 11:27
To: EMC-PSTC
Subject: FCC and Canada


Dear Group,
Could some of somebody please remind me whether or not IT equipment compliant
with the FCC requirements for the US would also be eligible for sale to
Canada, or would we need to apply further CSA standards ?
 
(The safety assessment has been done by UL.)
 
Thanks to you all for your assistance,
 
Best regards,
David Sproul,
 



RE: Russian (GOST) equivalent standards.

2003-04-22 Thread David Sproul

Hello John,
I find your response to Neil's comments to be most interesting.  In my
original posting I said that I had used the generics to cover the emc
requirements.  I had guessed these from memory, but when I actually checked
I found that we had in fact used the EN 55103 standards that Neil had
referred to.  I also noticed that section G.2.2.4 (Entertainment lighting
control apparatus:) was highlighted in the my copy of the standard.

After an endless period of head scratching, I began to recall the lengthy
debates we had at the time with, our client and the test house.

There are 2 standards for EMC that generally cover lighting products, namely
EN 55015 for emissions and EN61547 for immunity. EN 61547 also includes in
it's scope entertainment lighting control equipment for professional
purpose.

On the other hand there are the EN 55103 standards which are specifically
named Product family standard for audio, video, audio-visual and
entertainment lighting control apparatus for professional use.  Since our
client made lighting control equipment for use in theatres, we saw the
latter standards as being the most appropriate ones to use.

EN 55015 did not seem to be the most appropriate standard for us.  Our
client's units have an on-board processor running at several MHz, which
monitors load currents and voltages, dimmer temperature and control
instructions for the 96 channels being run simultaneously.  With load cables
several 10's of metres long, we saw there was a high likelihood of radiated
emissions, which are not covered by EN55015.   This again is why we opted
for the EN55103 standards which do specify radiated emissions measurements
up to 1GHz.

As I stated earlier, we did also notice this peculiar clause in the
informative Annex G within EN55103-1 (Product family standard for
...entertainment lighting control apparatus for professional use) which
seemed to excluded professional lighting control equipment from the
professional lighting control equipment standard.

Perhaps incorrectly, as you would argue, we interpreted that clause in a way
that suited our case.

It says that the standard applies to control desks (but not dimmers or
luminaires, to which EN 55014 or EN 55015 applies)
If it had said  to which EN 55014 or EN 55015 apply,  that would have said
to me  that dimmers or luminaires are covered exclusively by these 2
standards.  But because it said applies we chose to read that as EN55103
does not apply to dimmers for which the scope of EN 55015 fully covers all
their EMC emissions protection requirements.   As far as we could see, EN
55015 did not address all the particular protection requirements for
emissions from our clients' equipment.

Nevertheless.  In the light of (no pun intended) the recent prosecution of a
hairdryer manufacturer for not ensuring their equipment was compliant with
all the EMC requirements for the intend working environment, we have
recommended to our client that they carry out Conducted Emission tests
between 9kHz and 150kHz, as required by EN55015, but excluded from
EN55103-1.

In this way we are now confident that our clients equipment now meets the
protection requirements for the environment in which it is designed to work.

I look forward (I think) to your comments,

Best regards,
David Sproul.



From: owner-emc-p...@majordomo.ieee.org
[mailto:owner-emc-p...@majordomo.ieee.org]On Behalf Of John Woodgate
Sent: 15 April 2003 08:31
To: emc-p...@majordomo.ieee.org
Subject: Re: Russian (GOST) equivalent standards.



I read in !emc-pstc that Barker, Neil neil.bar...@e2vtechnologies.com
wrote (in 4f826f960057d4118ec3009027e2453808a52...@whl17.eev.uk) about
'Russian (GOST) equivalent standards.' on Mon, 14 Apr 2003:
Although the technical requirements are somewhat similar, you should be
using EN 55103-1  EN 55103-2 for EMC as they are the product specific
standards for audio, video, and lighting control apparatus for professional
use. I would make sure that you meet these first. I don't agree with
earlier
advice to us EN 55015 as this applies to the luminaires rather than the
control equipment.

Well, I'm afraid you are wrong in respect of dimmers, as you would know
if you had read clause G.2.2.4 of EN 55103-1. The OP's 'dimmer/control
equipment' suggested to me that the control was integrated in the dimmer
rack, in which case EN 55015 applies to the whole box.
--
Regards, John Woodgate, OOO - Own Opinions Only. http://www.jmwa.demon.co.uk
Interested in professional sound reinforcement and distribution? Then go to
http://www.isce.org.uk
PLEASE do NOT copy news posts to me by E-MAIL!


This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p

Russian (GOST) equivalent standards.

2003-04-10 Thread David Sproul
Dear Group,
I am trying to find out what standards a theatre lighting dimmer / controller
would have to comply with before being eligible for sale in Russia.
In Europe we currently use EN 61000-6-1  -3 for EMC, and EN 60439-1, EN
61010-1 and EN 60950 to cover the safety requirements.  
Can anyone tell me if the Russians have direct equivalents of these standards,
or whether there is a product specific standard that they would use?
 
Are there any agency requirements still exist in Russia, such as GOST, and if
so what do we have to do to meet their requirements?. 
 
I look forward to your responses,
Best regards,
David Sproul,



G5-4 Harmonics Emissions Limits for Industrial Apparatus.

2003-04-01 Thread David Sproul
Dear group,
Has anyone heard of G5-4 which is apparently a new standard or amendment to an
existing standard within Europe, dealing with the Harmonic Current limits for
industrial equipment, presumably meaning equipment rated at more than 16A per
phase.
 
I would like to know when is it due to come into force, or when it was
ratified, and what exactly it applies to.
 
I thank you in advance for your responses.
 
Best regards,
David Sproul,
Business Development Manager,
Alexander Lynn Approvals Management Services,
 



EU standards applying to Furnaces

2003-03-31 Thread David Sproul
Dear all,
Can anyone advise what directive and standards we should be applying to
industrial furnaces.  I am quite confident in covering the safety requirements
of the associated control and electrics, but I have no idea on how to approach
a furnace with vacuum inside to remove impurities from their process.
 
Secondly, on some of the larger models, there is an overhead crane included in
the  plant to move the metal parts from one part of the process to another.
 
The company is confident that they have all the necessary safety requirements
covered, but they are looking for verification from an outside source.
 
Can anyone suggest which standards from the Machinery Directive and / or the
Low Voltage Directives should be applied.
 
I thank you in advance for your help.
 
Best regards,
David Sproul,
Business Development Manager,
Alexander Lynn Approvals Management Services
 



Sun Beds

2003-01-08 Thread David Sproul
Many thanks to all of you for your help with the Sun Bed Standards.  
 
Best regards,
David Sproul,
Business Development Manager,
Alexander Lynn Approvals Management Services
 
 



RE: EN 55103 declaration

2003-01-07 Thread David Sproul

Dear Mr Chan,
you may get an e-mail from BSI justifying what they do, but I find it
irritating (mainly because I wish I had thought of it first.)
If you look at most of the BS EN standards, you will see that they are the
EN Standards with a BSI front and back cover.  This makes it an official
British standard, since traditionally BSI set various standards within the
UK.

The rest of the standard is invariably identical to the EN standard.  What I
find irritating is that BSI then charge us a lot of money for putting these
extra pages onto the EN standard for you.  No-one (except this miserable
tight-fisted Scotsman) seems to mind paying them for this service.  Oh how I
wish I had thought of this first.

Best regards,
David Sproul.


From: owner-emc-p...@majordomo.ieee.org
[mailto:owner-emc-p...@majordomo.ieee.org]On Behalf Of KC CHAN [PDD]
Sent: 07 January 2003 02:42
To: emc-p...@majordomo.ieee.org; boconn...@t-yuden.com
Subject: RE: EN 55103 declaration



EN and BSI EN are the same, but I would suggest you use EN standards on your
DoC, instead of BS EN.  The publish date of the standards from EN and BS EN
might not be the same, so to avoid confusion, use EN, which the publish
date(actually the date is version) are same as those listed in OJ.

 boconn...@t-yuden.com 01/04/03 12:46am 
Sir

What is the rationale given for some people not accepting an EN standard
issued by BSI?
My employer typically buys EN stds from BSI. Is there something different
about BSI-printed standards?

thanks for info.
R/S,
Brian O'Connell
Taiyo Yuden (USA), Inc.



From: John Woodgate [mailto:j...@jmwa.demon.co.uk]
Sent: Friday, January 03, 2003 7:41 AM
To: emc-p...@majordomo.ieee.org
Subject: Re: EN 55103 declaration

76011...@gandalf.allen-heath.com) about 'EN 55103 declaration' on Fri,
3 Jan 2003:
As a UK manufacturer and European exporter of professional electronic
audio
products, we declare conformity with the relevant EMC product family
standards. For testing purposes, we use EN 55103:1996 Parts 1 2 but
our EU
Declaration of conformity states compliance with BS EN 55103:1997 Parts
1 
2. As the majority of our products are exported to mainland Europe
should we
specify compliance with the EN or BS EN version of the standards?

While it is quite reasonable to cite the BS EN, some people may not
think so, thus you should cite the EN.



This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list



This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list



Sun Beds

2003-01-06 Thread David Sproul
Dear Group,
A Happy and Prosperous New Year to you all. 
Could someone please advise me on the appropriate EMC and Safety Standards to
apply to domestic sun-beds for sale in Europe and the US.
 
Thank you all in anticipation,
 
Best regards,
David Sproul,
Business Development Manager,
Alexander Lynn Approvals Management Services
 



Vehicle Mounted PCB

2002-12-09 Thread David Sproul
Thanks to all those who responded to my posting.

Some responses were helpful, some were amusing, and some were both, but all
were gladly received.

Best regards,
David Sproul,



Vehicle Mounted PCB's

2002-12-05 Thread David Sproul
Dear group,
A customer has been asked to redesign a board to stop it malfunctioning due
to vibration, temperature and radio interference.  They have been given a
control PCB which they are told is the main culprit for the malfunctions.

They circuit is mounted in an armoured vehicle as part of a security system
which sprays bank notes with red ink when it believes a robbery is taking
place.  Apparently it sprays ink if it gets too hot, too cold, gets bumped,
or if a radio or mobile phone is operated too close to it.

The bad news is that this is on the market and being used by security firms
within the UK and mainland Europe, with no real evidence of previous
compliance with anything.

Should this comply with any of Vehicle directive requirements?
Are there any peculiar EMC requirements other than the usual 61000 series?
What standard is likely to cover this device for use in vehicles?
Although they haven't asked, what would be the most appropriate standard to
cover safety requirements for this device.  (Although is runs of only 12V, I
am concerned about a fault causing a short across the battery, for example)?

There was mention of selling it in the US too.  If any has thoughts on what
such a device should comply with there, all comments would be gratefully
received.

If you are thinking of writing back and suggesting the device be thrown in
the bin, then I'm sorry to say that someone else has beaten you to it.

Best regards,
David Sproul,



RE: Filtered D Types.

2002-10-14 Thread David Sproul
Many thanks for all your help.

Best regards,
David Sproul.


Filtered D-Types.

2002-10-12 Thread David Sproul
Hello Group,
does anyone knee where I can get either filtered D-type connectors or filter
cards that be inserted inside the shell to filter the signals?  I was at a
seminar where these were shown being used in a Japanese video, (with English
voice over) and I  was asked if I could source them.

Can anyone help?

Best regards,
David Sproul,



RE: Creepage

2002-10-12 Thread David Sproul

Peter Tarver and all,
Yes I know that these aren't harmonised standards.  My original query I
guess was why independent bodies should decide upon such varying creepage
distance requirements.  I had naively thought that these distances would
have been determined by some universally recognised formula, that would have
given the same results for the same circumstances wherever in the world it
was applied.
It has now be made clear to me  (thanks to all who helped me understand)
that this wasn't the case, and that throughout the world some very different
methods are used to determine appropriate creepage distances.

Perhaps one day there will be method accepted by all disciplines of product
design around the world, like everyone using IEC 60664 for instance. But
until that day comes I now at least have an answer for our customers when
they ask about the different requirements.

Thanks you all again.

Best regards,
David Sproul.

-Original Message-
From: owner-emc-p...@majordomo.ieee.org
[mailto:owner-emc-p...@majordomo.ieee.org]On Behalf Of Peter L. Tarver
Sent: 11 October 2002 17:27
To: EMC-PSTC
Subject: RE: Creepage



David -

These are not harmonized standards.  UL 891 bases it's
spacings on US electrical distribution standards, while
EN60439-1 most likely relies on IEC60664.


Regards,

Peter L. Tarver, PE
Product Safety Manager
Sanmina-SCI Homologation Services
San Jose, CA
peter.tar...@sanmina-sci.com

 -Original Message-
 From: David Sproul
 Sent: Friday, October 11, 2002 6:15 AM

 Sorry Brian,
 I missed your response until now.  The standards
 I was using was EN 60439-1
 for Europe and UL 891 in America. These were the
 standards recommended by
 the customers association and UL.

 Best regards,
 David Sproul.


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list


RE: David Sproul...UL creepage limits ;~)

2002-10-11 Thread David Sproul

Ted,
thank you for your response.  I do not claim to be an expert, but I cannot
accept that creepage has anything to to with the current flowing in a
circuit.  Surely it is the voltage across the material and the CTI of that
material which determines the likelihood of tracking across the material to
take place.

As for your car battery melting story, cars must be wired differently in the
US than in the UK, because I have connected negative to negative and
positive to positive on many occaisions, and never had anything anymore
exciting happen than the second car starts.

Best regards,
David Sproul.

-Original Message-
From: owner-emc-p...@majordomo.ieee.org
[mailto:owner-emc-p...@majordomo.ieee.org]On Behalf Of Ted Rook
Sent: 09 October 2002 15:28
To: 
Subject: David Sproul...UL creepage limits ;~)



This is because when you double the voltage the power is proportional to a
quarter of the current squared. In America the 120V power is at lower
voltage but the current is twice as much and so the creepage is twice as
well.

Very high voltage circuits hardly creep at all whereas low voltages creep
the most. That is why you should never join the two negative terminals when
you jump start a car, the car battery charging circuits have so much
creepage they can melt the battery.

I though everybody knew that...



---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list


RE: Creepage

2002-10-08 Thread David Sproul

Can anyone give a rational reason why UL ask for around twice the creepage
distances for 120V that EN standards ask for 230V.

I realise this is a bit vague, but I can dig out some examples if any needs
to be convinced.  I have been asked this by customers on a number of
occasions, and have always meant to find out so I would no longer have to
apologise for not knowing.

Thanks Richard for jogging my memory.

Best regards,
David Sproul.



-Original Message-
From: owner-emc-p...@majordomo.ieee.org
[mailto:owner-emc-p...@majordomo.ieee.org]On Behalf Of
richwo...@tycoint.com
Sent: 07 October 2002 19:46
To: emc-p...@majordomo.ieee.org
Subject: Creepage



What is a good source that explains the rational for the values for creepage
distances?

Richard Woods
Sensormatic Electronics
Tyco International


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list


RE: Safety Symbol / Hot surfaces

2002-09-26 Thread David Sproul


Hello David / Neil,
I wasn't sure if either of you were in the UK or on the other side of the
pond.  In the UK I have been supplied with these labels from a company
called HTE Controls.  Their phone number is +44 (0) 1355 238641.  They have
many different labels in different sizes , printed on self adhesive plastic.

I have no connection with this company other than to occasionally buy labels
from them.

Best regards,
David Sproul,
Alexander Lynn Approvals Management Services

-Original Message-
From: owner-emc-p...@majordomo.ieee.org
[mailto:owner-emc-p...@majordomo.ieee.org]On Behalf Of Neil Helsby
Sent: 26 September 2002 08:34
To: emc-p...@majordomo.ieee.org
Subject: Hot surfaces



The hot symbol is also found at ISO 7000 pattern 0535 and BS 6217
pattern 5041.

We have recently introduced this symbol on some of our products. The
problem with this, and other safety required symbols, is that we have
been unable to locate a source that supplies these as standard off the
shelf items. It has therefore been necessary to pay for artworks to be
designed and this adds considerable cost to the product. Once the artwork
has been paid for, the cost between printing 100 and 1000 labels s
generally minimal in comparison so our batch of labels should last a long
time.

Does anyone know of a source of these types of labels or is this a
business opening for someone?

Regards,

Neil Helsby


**
This email and any files transmitted with it are confidential and
intended solely for the use of the individual or entity to whom they
are addressed. If you have received this email in error please notify
the system manager.

This footnote also confirms that this email message has been swept by
MIMEsweeper for the presence of computer viruses.

www.mimesweeper.com
**


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list


RE: Earthing through screws.

2002-09-17 Thread David Sproul

Hello Doug,
I believe you are right. If I had taken the trouble to read EN60950 para
2.6.5.7, I would have saved my self from asking stupid questions.

Thanks also to all those who responded to my orignial posting.

regards,
David Sproul.


-Original Message-
From: owner-emc-p...@majordomo.ieee.org
[mailto:owner-emc-p...@majordomo.ieee.org]On Behalf Of Doug McKean
Sent: 16 September 2002 18:58
To: EMC-PSTC
Subject: Re: Earthing through screws.



I wasn't aware that primary grounding securements
could be used for another purpose.

Regards, Doug McKean


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list


Earthing through screws.

2002-09-14 Thread David Sproul
Dear group,

Could someone please remind me of the marking requirements for screws used
for safety earthing purposes.  The screws in question are also used to
secure a small drawer into a grounded metal rack.  Would it be sufficient to
include a warning in the installation manual that the screws are also
required for earthing purposes, or would they have to be visibly identified
as part of the safety earthing?

Your comment would be gratefully received.
Best regards,
David Sproul,



RE: ISO 9k/2k relevance

2002-08-09 Thread David Sproul

Hello Brian,
I don't have any helpful answers for you.  I used to sell ISO 9000 and the
majority of times I heard its benefits praised was by my bosses and the
sales information. I do remember a small number of customers saying that,
for them to be considered in bids, they had to have ISO 9000.  However, I
have also work for ISO 9000 companies whose day to day working practices
were a mess, but come audit day they would come up smelling of roses.  I
suspect that ISO 9000 / 2000 will improve this situation, and perhaps
through time, if customer companies see marked improvements in their own
operations, more of them will start to insist on their suppliers having the
same accreditation.

From personal experience, the companies I have had dealings with are still
in a period of transition from the old ISO 9000, and therefore any benefits
of the new system are still to be realised.

Best regards,
David Sproul

-Original Message-
From: owner-emc-p...@majordomo.ieee.org
[mailto:owner-emc-p...@majordomo.ieee.org]On Behalf Of Brian O'Connell
Sent: 08 August 2002 15:36
To: Product Safety Technical Committee
Subject: ISO 9k/2k relevance



Good People of the PSTC:

I've had some conversations with our Component Engineers, Sales and QA
people. I could not identify any customer that placed an order based on our
ISO 9k and/or 2k certification.  Nor could I identify any component
specified and/or purchased that was based on whether a supplier has ISO
certification.

Is the ISO paper mill relevant? Is there empirical evidence that ISO
certification results in better stuff?  Is ISO certification a requirement
for your purchasing policies? Has ISO certification been a determining or
contributing factor for selection of your company's products?

At this point, I am not being critical of the ISO process; I am attempting
to understand its ROI and relevance to product quality.

I speak only for myself; nothing said here represents my employer's
policies.

R/S,
Brian


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list


RE: HALT and HAAS

2002-08-01 Thread David Sproul

Hello group,

I would actually disagree that HALT  is misnamed, as Brian suggests.  I have
seen clients find mechanical and electronic weakness in their designs within
days of testing that may otherwise have taken many weeks of conventional
temperature and vibration testing, or even years in the field, before they
were discovered.

By fixing these problems these companies were able to strengthen their
product still further and greatly reduce, and in some cases eradicate,
repairs or returns from customers.

I may be wrong, but is it not the case that components destined for non
military use do not come with MTBF figures, thus making it impossible to
calculate accurate MTBF figures for products they are used in?  I must
stress that I this is more supposition that known fact.  If it is so, then
perhaps HALT could be seen as a commercial alternative to MTBF.

I think HAAS  may actually be HASS, which stands for highly accelerated
stress screening.  This I'm afraid I've had very little experience of, but
It seems to take the product lessons learned from your HALT testing and
apply them, in a less destructive manner, to the manufacturing process.

HALT and HASS are very expensive and must be weighed against perceived
potential savings from reduced field repairs or replacements.  I know it is
not for all companies.  I used to work for a large British manufacturer of
military electronics who made as much money (if not more) from spares and
repairs, as they did from the original product sales.

I hope this has helped.

Best regards,
David Sproul.
-Original Message-
From: owner-emc-p...@majordomo.ieee.org
[mailto:owner-emc-p...@majordomo.ieee.org]On Behalf Of Brian O'Connell
Sent: 31 July 2002 14:12
To: emc-p...@ieee.org
Subject: RE: HALT and HAAS



Geez, just jump to Google. HALT- misnamed: Highly Accelerated Life Testing.
Actually has noting to do with Life-time or MTBF-type longevity analysis.

Look at the Calmer web site; they were one of the pioneers of HALT. Also, GM
has published an engineering standard, GMW8287, that provides a decent
overview of HALT and HASS/HASA processes.

There may be minimal ROI for simple digital stuff, but mixed-signal and/or
more complex mechanical constructions should be subject to HALT, and perhaps
HASS follow-up.

If your manufacturing and design engineers take the time to actually READ
your HALT report and take corrective action, you will probably find that RMA
problems are greatly reduced.

In any case, playing with liquid nitrogen is great fun...

Brian

-Original Message-
From: Peter Merguerian [mailto:pmerguer...@itl.co.il]
Sent: Wednesday, July 31, 2002 3:38 AM
To: EMC-PSTC (E-mail) 
Subject: HALT and HAAS



Dear All,

1. Has anyone heard of HALT and HAAS?

2. What are some overseas labs (in Europe, North America and  Asia) testing
for HALT and HAAS?

3. What does a manufacturer testing to HALT and HAAS gain?



---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list


RE: Multiple Product Variants under EN60439.

2002-06-28 Thread David Sproul

Scott,
from reading what you have written, I guess you must be about 900 years old
to have acquired such wisdom.

Thank you for your encouragement.  Now I will feel a little less intimidated
by experts.

Thanks  best regards,

David Sproul.

-Original Message-
From: Scott Douglas [mailto:dougl...@naradnetworks.com]
Sent: 28 June 2002 13:34
To: David Sproul; Nick Williams
Cc: EMC-PSTC
Subject: RE: Multiple Product Variants under EN60439.


Nick,

In my years of being a compliance type, I have found many cases where I
could not get an agency or test house to make a bold statement about this
or that. Often one can accept that and understand why they will not commit.
For one reason, they do not want to be the one sued by some disgruntled
customer because something happened (or didn't happen). But just as often,
I have found it extremely frustrating that I cannot get a straight answer
from the same people.

What I have come to realize is this. Yes, I always thought the agency /
test house / whatever should be the experts and should always be able to
give concrete answers. Why do I think this? Because they eat sleep and
breathe standards every day for a living. So they should be able to answer
our sometimes really hard questions. As for me, I am a compliance guy with
19 years on the job. But you know what? I do not live with standards every
day, some days I actually do other stuff not compliance related. That is
why I do not always know my standards as well as I might need to. So I turn
to the experts. And sometimes get the same frustrations as you.

Another thing I have come to realize is that the people in our field of
compliance are just that, people. Wherever we work, be it an agency, test
house, or manufacturer, we are all human. And that means that some of us
are really good at what we do, some of us are okay at the job, and some of
us should be in another field. I won't go in to percentages in any
category, we know who we are.

So how does this help you? I can tell you that I have learned to always
question everything. Especially if it does not sound or feel right.
Compliance is really about common sense with a little bit of overkill
thrown in for good measure. When someone tells you that you must do it this
way, ask why, ask to see it written in the standard, ask them to justify
what they are saying. Done reasonably, no one should take offense at your
questioning them. You are only trying to learn after all. To become better
at your job. And, as often as not, I have seen lights go on in many an
agent or test house person when they had to explain themselves.

As I said earlier, we are all human. And for every one you ask a question
of, you will as often get that many answers. In the end, it is always up to
us as individuals to do what we think is right in order to comply with the
spirit and intent of the standard. Sometimes that means gritting our teeth
and doing something that is not likeable or pleasant. Sometimes that means
not doing specifically what is told to us to do. And sometimes that means
negotiating an acceptable compromise between you and the teller of what you
don't agree with.

So, never accept anything on blind faith because the other person is the
expert. Sometimes you put it in mind to accept it, but always keep that
question there as well. Have faith in yourself that you know what you are
doing. But always ask to be shown how you are wrong about some particular
point. Then, if they convince you, admit your error, change your way, and
move on. And if they don't convince you, then make that hard decision to do
what you feel is right and be prepared to support or defend that position.

Anyway, off the soap box now.

Regards,
Scott






At 10:28 PM 6/27/02 +0100, David Sproul wrote:

Thanks Nick,
I knew that what was happening with the CB wasn't right, but I was afraid
to
challenge them since they would presumably know more than me, and therefore
in some way may have been able to justify what they were (are) doing.

As for the EN60439 issue, I have found that certain labs will not make
engineering judgements, depending on which body has accredited them.  They
will only report on facts derived from testing.  Even when the engineers
say
off the record it will be OK, they still won't put it in a report and say
It is our considered opinion that this or that change is unlikely to
effect
the compliance of this or that product.  I do find the whole thing so very
frustrating, made even more so because, to me EN60439 is an awkward
standard
to follow.  Perhaps it is just me.

Thanks again,

David Sproul.

-Original Message-
From: Nick Williams [mailto:nick.willi...@conformance.co.uk]
Sent: 27 June 2002 21:47
To: David Sproul
Cc: EMC-PSTC
Subject: Re: Multiple Product Variants under EN60439.


David,

Replies in the text of your questions...


At 17:25 +0100 26/6/02, David Sproul wrote:
 Dear Group,
 As you will all know, under the EMC Directive a range of product
 variants can

RE: Multiple Product Variants under EN60439.

2002-06-27 Thread David Sproul

Thanks Nick,
I knew that what was happening with the CB wasn't right, but I was afraid to
challenge them since they would presumably know more than me, and therefore
in some way may have been able to justify what they were (are) doing.

As for the EN60439 issue, I have found that certain labs will not make
engineering judgements, depending on which body has accredited them.  They
will only report on facts derived from testing.  Even when the engineers say
off the record it will be OK, they still won't put it in a report and say
It is our considered opinion that this or that change is unlikely to effect
the compliance of this or that product.  I do find the whole thing so very
frustrating, made even more so because, to me EN60439 is an awkward standard
to follow.  Perhaps it is just me.

Thanks again,

David Sproul.

-Original Message-
From: Nick Williams [mailto:nick.willi...@conformance.co.uk]
Sent: 27 June 2002 21:47
To: David Sproul
Cc: EMC-PSTC
Subject: Re: Multiple Product Variants under EN60439.


David,

Replies in the text of your questions...


At 17:25 +0100 26/6/02, David Sproul wrote:
Dear Group,
As you will all know, under the EMC Directive a range of product
variants can be covered by a Technical Construction File.

A company who manufactures control panels for large industrial plant
has to re-submit every control panel it makes the a Competent Body
because it was impossible for them to build a super panel with every
conceivable component included.

I have 3 questions:
1/  Is there any other way they can have their range cover by a TCF
without having to re-submit to the Competent Body?

Firstly, I suggest they find a new CB - it sounds like the one they
have got is more interested in making money at their expense than in
solving a problem for their clients. Secondly, it's highly likely
they don't need to be using a TCF anyway.

2/  Under the LVD, do these panels have to be type tested to EN60439
by a Certified Body?

No. And 60439 may not be the right standard anyway.

3/ Is there anyway they can have their panel range reviewed to
EN60439 and be sure that their Safety Technical File will cover
future variants, without having to return to the Certified Body for
each new Panel.


Yes. They need to be trained to know how to do it for themselves.

The industry they work in is very competitive, so they cannot raise
their prices too much.  Therefore each visit to the Competent Body
and the Certified Body is eating into their meagre profits.


Most of their competitors don't bother with independent verification
for either safety or EMC.

Are there any significantly knowledgeable people who could bring
this company some good news?


It's what we do for a living. To say more would be an abuse of the
EMC-PSTC forum.

Best regards,
David Sproul,
Alexander Lynn Approvals Management,
Tel/fax +44 (0) 1383 85
Mobile +44 (0) 7950 744466



Regards

Nick.



Conformance Ltd - Product safety, approvals and CE-marking consultants
Tel. + 44 1298 873800, Fax. +44 1298 873801
Registered in England, Company No. 3478646


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list


Multiple Product Variants under EN60439.

2002-06-26 Thread David Sproul
Dear Group,
As you will all know, under the EMC Directive a range of product variants
can be covered by a Technical Construction File.

A company who manufactures control panels for large industrial plant has to
re-submit every control panel it makes the a Competent Body because it was
impossible for them to build a super panel with every conceivable component
included.

I have 3 questions:
1/  Is there any other way they can have their range cover by a TCF without
having to re-submit to the Competent Body?
2/  Under the LVD, do these panels have to be type tested to EN60439 by a
Certified Body?
3/ Is there anyway they can have their panel range reviewed to EN60439 and
be sure that their Safety Technical File will cover future variants, without
having to return to the Certified Body for each new Panel.

The industry they work in is very competitive, so they cannot raise their
prices too much.  Therefore each visit to the Competent Body and the
Certified Body is eating into their meagre profits.

Are there any significantly knowledgeable people who could bring this
company some good news?

Best regards,
David Sproul,
Alexander Lynn Approvals Management,
Tel/fax +44 (0) 1383 85
Mobile +44 (0) 7950 744466





RE: Slotted Busbars.

2002-06-26 Thread David Sproul
Dear Group,
Thank you all again for you help in this issue.  I have since found out that
one of the Engineers I initially contacted was in fact working on a
customer's site.  For this reason he was unable to respond to my initial
request for information.

It was therefore perhaps unjustified, not to mention unproffessional, for me
to imply that his company had given me a poor service.

I would therefore like to take this opportunity to apologise for including
this unfair remark in my original e-mail.

Best regards,
David Sproul,
Alexander Lynn Approvals Management,
Tel/fax +44 (0) 1383 85
Mobile +44 (0) 7950 744466


Slotted Busbars.

2002-06-24 Thread David Sproul
Dear Group,
thank you all for you speedy but varied answers.

Best regards,
David Sproul,




Slotted Busbars

2002-06-21 Thread David Sproul
Dear Group,
I have a client who wishes to use open slots instead of closed holes for the
connecting bolts used in joining two 400A busbar within their product.  This
decision was taken to make installation and maintenance easier.  However,
their installation Manager has objected to this without giving a reason.  I
have e-mailed my contacts within TUV and UL, and after some 48 hours they
have not responded.

Are there any sufficiently knowledgeable people who can tell me whether or
not this is permissible, and why.

Best regards,
David Sproul.