RE: Conducted Emissions PK/AVG

2004-12-20 Thread owner-emc-p...@ieee.org
I assume the time-constant of the AVG Detector is specific in CISPR 16
and that this standard is called out in FCC Part 15.

I think if the VBW was placed before the LOG amplifier in the signal
chain, it would yield the correct response.  If placed after the LOG
amp, the BW limit would not provide the desired effect.

The UL EMC software we use has tick-boxes for AVG detection, but they
have no noticeable affect on the receiver actions nor on the screen
plot.  Updated versions of this program from UL have also failed to
provide a AVG detection function.  It could be that this is not
available on our 8568B. So, in the rather rare instance where we need to
do an AVG conducted scan, we simply reduce the VBW in the test setup
file to 10Hz.

Ralph McDiarmid, ASc
Compliance Engineering Group
Xantrex Technology Inc.


From: Brent DeWitt [mailto:bdew...@ix.netcom.com]
Sent: December 18, 2004 8:02 PM
To: Ken Javor; Ralph McDiarmid; ieee pstc list
Subject: RE: Conducted Emissions PK/AVG


Your point is well taken Ken, and this certainly is interesting.  My
approach to the concept of average does indeed run afoul of the
limitations of measurement devices and the potential for instrument
dependent results.  To the best of my knowledge, the FCC is the most
ill-defined with respect to the averaging time constant.  That is what
initially lead me to my stated empirical method.  Other thoughts are
welcomed!

Brent

 -Original Message-
 From: owner-emc-p...@ieee.org [mailto:owner-emc-p...@ieee.org]On
 Behalf Of Ken Javor
 Sent: Friday, December 17, 2004 8:14 PM
 To: bdew...@ix.netcom.com; Ralph McDiarmid; ieee pstc list
 Subject: Re: Conducted Emissions PK/AVG


 Now this is getting interesting.  If you arbitrarily reduce the VBW
 until you see no time dependent amplitude variation, I think that
 leads to non-repeatability.  That is, if my spectrum analyzer has a 1
 Hz VBW, and yours has a 10 Hz VBW, we could get different answers.  At

 least at one time, there was a specified time interval over which the
 averaging was to occur, which is another way of saying the the video
 bandwidth was specified.

 I think the idea of average detection is to be able to get an accurate

 assessment of the value of a cw or near cw (AM) signal, possibly in
 the presence of broadband noise.  Considering that pulsed cw was
 considered by the military to be a narrow-band signal, it would seem
 that the specific averaging time period would be very important.  On
 the other hand, MIL-STD-461 has for almost forty years required peak
 detection of all signals, whether NB or BB.  Go figure.


 I'm rambling on, but the point I am trying to make is that I believe
 average detection should have an associated time constant, somehow
 relating to the information content of the communication link
 protected by the radiated emission limit in question.


  From: Brent DeWitt bdew...@ix.netcom.com
  Reply-To: bdew...@ix.netcom.com
  Date: Fri, 17 Dec 2004 18:39:11 -0800
  To: Ralph McDiarmid ralph.mcdiar...@xantrex.com, ieee pstc
  list emc-p...@ieee.org
  Subject: RE: Conducted Emissions PK/AVG
 
  Hi Ralph,
 
  The HP 8568 and 8566 have a video averaging function, which
  averages multiple digital traces (up to 100 if I remember
  correctly).
 It can be used
  in conjunction with a narrow VBW to get very close to a true
  average measurement.  I've always been rather empirical about the
  averaging criterion.  On a spectrum analyzer in zero span and linear

  detection,  I simply reduce the VBW and possibly add video averaging

  until I
 don't see any
  more variation in the screen trace and call it done.
 
  Brent G DeWitt
  Laboratory Manager
  CKC Laboratories
  Redmond, WA
  email: brent.dew...@ckc.com
  phone: 425-883-4757
  cell: 425-417-8228
 
 
  -Original Message-
  From: owner-emc-p...@ieee.org [mailto:owner-emc-p...@ieee.org]On
  Behalf Of Ralph McDiarmid
  Sent: Friday, December 17, 2004 11:51 AM
  To: ieee pstc list
  Subject: RE: Conducted Emissions PK/AVG
 
 
  I suggest there is an important distinction between video
  averaging and average detection using limited VBW.
 
  If this is correct, can someone enlighten those of us who are
  unsure?
 
  Ralph McDiarmid, ASc
  Compliance Engineering Group
  Xantrex Technology Inc.
 
 
  -Original Message-
  From: owner-emc-p...@ieee.org [mailto:owner-emc-p...@ieee.org]
 On Behalf
  Of Cortland Richmond
  Sent: December 16, 2004 1:04 PM
  To: Alex McNeil; ieee pstc list
  Subject: Re: Conducted Emissions PK/AVG
 
 
  Alex McNeil asked:
  Is the problem that the Test House may have used AVG BW 30KHz
  and this
  should have been AVG BW 9KHz the problem 
 
  Sure looks like it from what you posted. There've been discussions
  here
  -- it comes up every so often -- how to to do averaging on a SA.
  Without knowing how they did, it's hard to say for sure, but it
looks
  like you are close enough to hit the target with a hand grenade
anyway.
  But are you sure it doesn't say

RE: Conducted Emissions PK/AVG

2004-12-19 Thread owner-emc-p...@ieee.org
Sorry to post a second response, but I hit send before I finished my
thoughts.

Ken's stated concept of the rational of averaging is a bit different from
mine.  This may be a MIL vs. FCC background related thing (me being the
latter).  I've always viewed the FCC's switch to average detection above
1GHz as more of a rough approximation to a spectral density measurement.  If
my approach is correct, the idea is that receivers of potential interference
above 1GHz have relatively wide resolution bandwidths, and that something
closer to the 1MHz RBW in combination with an average detector could measure
the annoyance factor.

Again, other thoughts are welcomed!

Brent

 -Original Message-
 From: owner-emc-p...@ieee.org [mailto:owner-emc-p...@ieee.org]On Behalf
 Of Ken Javor
 Sent: Friday, December 17, 2004 8:14 PM
 To: bdew...@ix.netcom.com; Ralph McDiarmid; ieee pstc list
 Subject: Re: Conducted Emissions PK/AVG


 Now this is getting interesting.  If you arbitrarily reduce the VBW until
 you see no time dependent amplitude variation, I think that leads to
 non-repeatability.  That is, if my spectrum analyzer has a 1 Hz VBW, and
 yours has a 10 Hz VBW, we could get different answers.  At least at one
 time, there was a specified time interval over which the averaging was to
 occur, which is another way of saying the the video bandwidth was
 specified.

 I think the idea of average detection is to be able to get an accurate
 assessment of the value of a cw or near cw (AM) signal, possibly in the
 presence of broadband noise.  Considering that pulsed cw was
 considered by
 the military to be a narrow-band signal, it would seem that the specific
 averaging time period would be very important.  On the other hand,
 MIL-STD-461 has for almost forty years required peak detection of all
 signals, whether NB or BB.  Go figure.


 I'm rambling on, but the point I am trying to make is that I
 believe average
 detection should have an associated time constant, somehow relating to the
 information content of the communication link protected by the radiated
 emission limit in question.


  From: Brent DeWitt bdew...@ix.netcom.com
  Reply-To: bdew...@ix.netcom.com
  Date: Fri, 17 Dec 2004 18:39:11 -0800
  To: Ralph McDiarmid ralph.mcdiar...@xantrex.com, ieee pstc list
  emc-p...@ieee.org
  Subject: RE: Conducted Emissions PK/AVG
 
  Hi Ralph,
 
  The HP 8568 and 8566 have a video averaging function, which averages
  multiple digital traces (up to 100 if I remember correctly).
 It can be used
  in conjunction with a narrow VBW to get very close to a true average
  measurement.  I've always been rather empirical about the averaging
  criterion.  On a spectrum analyzer in zero span and linear detection,  I
  simply reduce the VBW and possibly add video averaging until I
 don't see any
  more variation in the screen trace and call it done.
 
  Brent G DeWitt
  Laboratory Manager
  CKC Laboratories
  Redmond, WA
  email: brent.dew...@ckc.com
  phone: 425-883-4757
  cell: 425-417-8228
 
 
  -Original Message-
  From: owner-emc-p...@ieee.org [mailto:owner-emc-p...@ieee.org]On Behalf
  Of Ralph McDiarmid
  Sent: Friday, December 17, 2004 11:51 AM
  To: ieee pstc list
  Subject: RE: Conducted Emissions PK/AVG
 
 
  I suggest there is an important distinction between video averaging
  and average detection using limited VBW.
 
  If this is correct, can someone enlighten those of us who are unsure?
 
  Ralph McDiarmid, ASc
  Compliance Engineering Group
  Xantrex Technology Inc.
 
 
  -Original Message-
  From: owner-emc-p...@ieee.org [mailto:owner-emc-p...@ieee.org]
 On Behalf
  Of Cortland Richmond
  Sent: December 16, 2004 1:04 PM
  To: Alex McNeil; ieee pstc list
  Subject: Re: Conducted Emissions PK/AVG
 
 
  Alex McNeil asked:
  Is the problem that the Test House may have used AVG BW 30KHz and
  this
  should have been AVG BW 9KHz the problem 
 
  Sure looks like it from what you posted. There've been discussions here
  -- it comes up every so often -- how to to do averaging on a SA.
  Without knowing how they did, it's hard to say for sure, but it looks
  like you are close enough to hit the target with a hand grenade anyway.
  But are you sure it doesn't say 30 *Hz*? Narrowing the bandwidth 'way
  down is a method used to get averaging on a SA. (I will speculate that
  someone might have learned to -- incorrectly -- use video
 averaging in
  dB mode then increase BW to compensate for the error that causes.)
 
  Cortland Richmond
 
  
  This message is from the IEEE Product Safety Engineering Society
  emc-pstc discussion list.Website:  http://www.ieee-pses.org/
 
  To post a message to the list, send your e-mail to emc-p...@ieee.org
 
  Instructions:
 http://listserv.ieee.org/listserv/request/user-guide.html
 
  List rules: http://www.ieee-pses.org/listrules.html
 
  For help, send mail to the list administrators:
 
  Scott Douglas

RE: Conducted Emissions PK/AVG

2004-12-19 Thread owner-emc-p...@ieee.org
Your point is well taken Ken, and this certainly is interesting.  My
approach to the concept of average does indeed run afoul of the
limitations of measurement devices and the potential for instrument
dependent results.  To the best of my knowledge, the FCC is the most
ill-defined with respect to the averaging time constant.  That is what
initially lead me to my stated empirical method.  Other thoughts are
welcomed!

Brent

 -Original Message-
 From: owner-emc-p...@ieee.org [mailto:owner-emc-p...@ieee.org]On Behalf
 Of Ken Javor
 Sent: Friday, December 17, 2004 8:14 PM
 To: bdew...@ix.netcom.com; Ralph McDiarmid; ieee pstc list
 Subject: Re: Conducted Emissions PK/AVG


 Now this is getting interesting.  If you arbitrarily reduce the VBW until
 you see no time dependent amplitude variation, I think that leads to
 non-repeatability.  That is, if my spectrum analyzer has a 1 Hz VBW, and
 yours has a 10 Hz VBW, we could get different answers.  At least at one
 time, there was a specified time interval over which the averaging was to
 occur, which is another way of saying the the video bandwidth was
 specified.

 I think the idea of average detection is to be able to get an accurate
 assessment of the value of a cw or near cw (AM) signal, possibly in the
 presence of broadband noise.  Considering that pulsed cw was
 considered by
 the military to be a narrow-band signal, it would seem that the specific
 averaging time period would be very important.  On the other hand,
 MIL-STD-461 has for almost forty years required peak detection of all
 signals, whether NB or BB.  Go figure.


 I'm rambling on, but the point I am trying to make is that I
 believe average
 detection should have an associated time constant, somehow relating to the
 information content of the communication link protected by the radiated
 emission limit in question.


  From: Brent DeWitt bdew...@ix.netcom.com
  Reply-To: bdew...@ix.netcom.com
  Date: Fri, 17 Dec 2004 18:39:11 -0800
  To: Ralph McDiarmid ralph.mcdiar...@xantrex.com, ieee pstc list
  emc-p...@ieee.org
  Subject: RE: Conducted Emissions PK/AVG
 
  Hi Ralph,
 
  The HP 8568 and 8566 have a video averaging function, which averages
  multiple digital traces (up to 100 if I remember correctly).
 It can be used
  in conjunction with a narrow VBW to get very close to a true average
  measurement.  I've always been rather empirical about the averaging
  criterion.  On a spectrum analyzer in zero span and linear detection,  I
  simply reduce the VBW and possibly add video averaging until I
 don't see any
  more variation in the screen trace and call it done.
 
  Brent G DeWitt
  Laboratory Manager
  CKC Laboratories
  Redmond, WA
  email: brent.dew...@ckc.com
  phone: 425-883-4757
  cell: 425-417-8228
 
 
  -Original Message-
  From: owner-emc-p...@ieee.org [mailto:owner-emc-p...@ieee.org]On Behalf
  Of Ralph McDiarmid
  Sent: Friday, December 17, 2004 11:51 AM
  To: ieee pstc list
  Subject: RE: Conducted Emissions PK/AVG
 
 
  I suggest there is an important distinction between video averaging
  and average detection using limited VBW.
 
  If this is correct, can someone enlighten those of us who are unsure?
 
  Ralph McDiarmid, ASc
  Compliance Engineering Group
  Xantrex Technology Inc.
 
 
  -Original Message-
  From: owner-emc-p...@ieee.org [mailto:owner-emc-p...@ieee.org]
 On Behalf
  Of Cortland Richmond
  Sent: December 16, 2004 1:04 PM
  To: Alex McNeil; ieee pstc list
  Subject: Re: Conducted Emissions PK/AVG
 
 
  Alex McNeil asked:
  Is the problem that the Test House may have used AVG BW 30KHz and
  this
  should have been AVG BW 9KHz the problem 
 
  Sure looks like it from what you posted. There've been discussions here
  -- it comes up every so often -- how to to do averaging on a SA.
  Without knowing how they did, it's hard to say for sure, but it looks
  like you are close enough to hit the target with a hand grenade anyway.
  But are you sure it doesn't say 30 *Hz*? Narrowing the bandwidth 'way
  down is a method used to get averaging on a SA. (I will speculate that
  someone might have learned to -- incorrectly -- use video
 averaging in
  dB mode then increase BW to compensate for the error that causes.)
 
  Cortland Richmond
 
  
  This message is from the IEEE Product Safety Engineering Society
  emc-pstc discussion list.Website:  http://www.ieee-pses.org/
 
  To post a message to the list, send your e-mail to emc-p...@ieee.org
 
  Instructions:
 http://listserv.ieee.org/listserv/request/user-guide.html
 
  List rules: http://www.ieee-pses.org/listrules.html
 
  For help, send mail to the list administrators:
 
  Scott Douglas emcp...@ptcnh.net
 
  For policy questions, send mail to:
 
  Richard Nute:   ri...@ieee.org
  Jim Bacher: j.bac...@ieee.org
 
  All emc-pstc postings are archived and searchable on the web at:
 
  http

Re: Conducted Emissions PK/AVG

2004-12-18 Thread owner-emc-p...@ieee.org
Now this is getting interesting.  If you arbitrarily reduce the VBW until
you see no time dependent amplitude variation, I think that leads to
non-repeatability.  That is, if my spectrum analyzer has a 1 Hz VBW, and
yours has a 10 Hz VBW, we could get different answers.  At least at one
time, there was a specified time interval over which the averaging was to
occur, which is another way of saying the the video bandwidth was specified.

I think the idea of average detection is to be able to get an accurate
assessment of the value of a cw or near cw (AM) signal, possibly in the
presence of broadband noise.  Considering that pulsed cw was considered by
the military to be a narrow-band signal, it would seem that the specific
averaging time period would be very important.  On the other hand,
MIL-STD-461 has for almost forty years required peak detection of all
signals, whether NB or BB.  Go figure.


I'm rambling on, but the point I am trying to make is that I believe average
detection should have an associated time constant, somehow relating to the
information content of the communication link protected by the radiated
emission limit in question.


 From: Brent DeWitt bdew...@ix.netcom.com
 Reply-To: bdew...@ix.netcom.com
 Date: Fri, 17 Dec 2004 18:39:11 -0800
 To: Ralph McDiarmid ralph.mcdiar...@xantrex.com, ieee pstc list
 emc-p...@ieee.org
 Subject: RE: Conducted Emissions PK/AVG

 Hi Ralph,

 The HP 8568 and 8566 have a video averaging function, which averages
 multiple digital traces (up to 100 if I remember correctly).  It can be used
 in conjunction with a narrow VBW to get very close to a true average
 measurement.  I've always been rather empirical about the averaging
 criterion.  On a spectrum analyzer in zero span and linear detection,  I
 simply reduce the VBW and possibly add video averaging until I don't see any
 more variation in the screen trace and call it done.

 Brent G DeWitt
 Laboratory Manager
 CKC Laboratories
 Redmond, WA
 email: brent.dew...@ckc.com
 phone: 425-883-4757
 cell: 425-417-8228


 -Original Message-
 From: owner-emc-p...@ieee.org [mailto:owner-emc-p...@ieee.org]On Behalf
 Of Ralph McDiarmid
 Sent: Friday, December 17, 2004 11:51 AM
 To: ieee pstc list
 Subject: RE: Conducted Emissions PK/AVG


 I suggest there is an important distinction between video averaging
 and average detection using limited VBW.

 If this is correct, can someone enlighten those of us who are unsure?

 Ralph McDiarmid, ASc
 Compliance Engineering Group
 Xantrex Technology Inc.


 -Original Message-
 From: owner-emc-p...@ieee.org [mailto:owner-emc-p...@ieee.org] On Behalf
 Of Cortland Richmond
 Sent: December 16, 2004 1:04 PM
 To: Alex McNeil; ieee pstc list
 Subject: Re: Conducted Emissions PK/AVG


 Alex McNeil asked:
 Is the problem that the Test House may have used AVG BW 30KHz and
 this
 should have been AVG BW 9KHz the problem 

 Sure looks like it from what you posted. There've been discussions here
 -- it comes up every so often -- how to to do averaging on a SA.
 Without knowing how they did, it's hard to say for sure, but it looks
 like you are close enough to hit the target with a hand grenade anyway.
 But are you sure it doesn't say 30 *Hz*? Narrowing the bandwidth 'way
 down is a method used to get averaging on a SA. (I will speculate that
 someone might have learned to -- incorrectly -- use video averaging in
 dB mode then increase BW to compensate for the error that causes.)

 Cortland Richmond

 
 This message is from the IEEE Product Safety Engineering Society
 emc-pstc discussion list.Website:  http://www.ieee-pses.org/

 To post a message to the list, send your e-mail to emc-p...@ieee.org

 Instructions:  http://listserv.ieee.org/listserv/request/user-guide.html

 List rules: http://www.ieee-pses.org/listrules.html

 For help, send mail to the list administrators:

 Scott Douglas emcp...@ptcnh.net

 For policy questions, send mail to:

 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

 All emc-pstc postings are archived and searchable on the web at:

 http://www.ieeecommunities.org/emc-pstc

 
 This message is from the IEEE Product Safety Engineering Society
 emc-pstc discussion list.Website:  http://www.ieee-pses.org/

 To post a message to the list, send your e-mail to emc-p...@ieee.org

 Instructions:  http://listserv.ieee.org/listserv/request/user-guide.html

 List rules: http://www.ieee-pses.org/listrules.html

 For help, send mail to the list administrators:

 Scott Douglas emcp...@ptcnh.net

 For policy questions, send mail to:

 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

 All emc-pstc postings are archived and searchable on the web at:

 http://www.ieeecommunities.org/emc-pstc

RE: Conducted Emissions PK/AVG

2004-12-17 Thread owner-emc-p...@ieee.org
Hi Ralph,

The HP 8568 and 8566 have a video averaging function, which averages
multiple digital traces (up to 100 if I remember correctly).  It can be used
in conjunction with a narrow VBW to get very close to a true average
measurement.  I've always been rather empirical about the averaging
criterion.  On a spectrum analyzer in zero span and linear detection,  I
simply reduce the VBW and possibly add video averaging until I don't see any
more variation in the screen trace and call it done.

Brent G DeWitt
Laboratory Manager
CKC Laboratories
Redmond, WA
email: brent.dew...@ckc.com
phone: 425-883-4757
cell: 425-417-8228


 -Original Message-
 From: owner-emc-p...@ieee.org [mailto:owner-emc-p...@ieee.org]On Behalf
 Of Ralph McDiarmid
 Sent: Friday, December 17, 2004 11:51 AM
 To: ieee pstc list
 Subject: RE: Conducted Emissions PK/AVG


 I suggest there is an important distinction between video averaging
 and average detection using limited VBW.

 If this is correct, can someone enlighten those of us who are unsure?

 Ralph McDiarmid, ASc
 Compliance Engineering Group
 Xantrex Technology Inc.


 -Original Message-
 From: owner-emc-p...@ieee.org [mailto:owner-emc-p...@ieee.org] On Behalf
 Of Cortland Richmond
 Sent: December 16, 2004 1:04 PM
 To: Alex McNeil; ieee pstc list
 Subject: Re: Conducted Emissions PK/AVG


 Alex McNeil asked:
  Is the problem that the Test House may have used AVG BW 30KHz and
  this
 should have been AVG BW 9KHz the problem 

 Sure looks like it from what you posted. There've been discussions here
 -- it comes up every so often -- how to to do averaging on a SA.
 Without knowing how they did, it's hard to say for sure, but it looks
 like you are close enough to hit the target with a hand grenade anyway.
 But are you sure it doesn't say 30 *Hz*? Narrowing the bandwidth 'way
 down is a method used to get averaging on a SA. (I will speculate that
 someone might have learned to -- incorrectly -- use video averaging in
 dB mode then increase BW to compensate for the error that causes.)

 Cortland Richmond

 
 This message is from the IEEE Product Safety Engineering Society
 emc-pstc discussion list.Website:  http://www.ieee-pses.org/

 To post a message to the list, send your e-mail to emc-p...@ieee.org

 Instructions:  http://listserv.ieee.org/listserv/request/user-guide.html

 List rules: http://www.ieee-pses.org/listrules.html

 For help, send mail to the list administrators:

  Scott Douglas emcp...@ptcnh.net

 For policy questions, send mail to:

  Richard Nute:   ri...@ieee.org
  Jim Bacher: j.bac...@ieee.org

 All emc-pstc postings are archived and searchable on the web at:

 http://www.ieeecommunities.org/emc-pstc

 
 This message is from the IEEE Product Safety Engineering Society
 emc-pstc discussion list.Website:  http://www.ieee-pses.org/

 To post a message to the list, send your e-mail to emc-p...@ieee.org

 Instructions:  http://listserv.ieee.org/listserv/request/user-guide.html

 List rules: http://www.ieee-pses.org/listrules.html

 For help, send mail to the list administrators:

  Scott Douglas emcp...@ptcnh.net

 For policy questions, send mail to:

  Richard Nute:   ri...@ieee.org
  Jim Bacher: j.bac...@ieee.org

 All emc-pstc postings are archived and searchable on the web at:

 http://www.ieeecommunities.org/emc-pstc


This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/listserv/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas emcp...@ptcnh.net

For policy questions, send mail to:

 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc



RE: Conducted Emissions PK/AVG

2004-12-17 Thread owner-emc-p...@ieee.org
I suggest there is an important distinction between video averaging
and average detection using limited VBW.

If this is correct, can someone enlighten those of us who are unsure?

Ralph McDiarmid, ASc
Compliance Engineering Group
Xantrex Technology Inc.



From: owner-emc-p...@ieee.org [mailto:owner-emc-p...@ieee.org] On Behalf
Of Cortland Richmond
Sent: December 16, 2004 1:04 PM
To: Alex McNeil; ieee pstc list
Subject: Re: Conducted Emissions PK/AVG


Alex McNeil asked:
 Is the problem that the Test House may have used AVG BW 30KHz and
 this
should have been AVG BW 9KHz the problem 

Sure looks like it from what you posted. There've been discussions here
-- it comes up every so often -- how to to do averaging on a SA.
Without knowing how they did, it's hard to say for sure, but it looks
like you are close enough to hit the target with a hand grenade anyway.
But are you sure it doesn't say 30 *Hz*? Narrowing the bandwidth 'way
down is a method used to get averaging on a SA. (I will speculate that
someone might have learned to -- incorrectly -- use video averaging in
dB mode then increase BW to compensate for the error that causes.)

Cortland Richmond


This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/listserv/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas emcp...@ptcnh.net

For policy questions, send mail to:

 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc


This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/listserv/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas emcp...@ptcnh.net

For policy questions, send mail to:

 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc



Re: Conducted Emissions PK/AVG

2004-12-16 Thread owner-emc-p...@ieee.org
Alex McNeil asked:
 Is the problem that the Test House may have used AVG BW 30KHz and this
should have been AVG BW 9KHz the problem 

Sure looks like it from what you posted. There've been discussions here --
it comes up every so often -- how to to do averaging on a SA.  Without
knowing how they did, it's hard to say for sure, but it looks like you are
close enough to hit the target with a hand grenade anyway. But are you sure
it doesn't say 30 *Hz*? Narrowing the bandwidth 'way down is a method used
to get averaging on a SA. (I will speculate that someone might have
learned to -- incorrectly -- use video averaging in dB mode then increase
BW to compensate for the error that causes.)

Cortland Richmond


This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/listserv/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas emcp...@ptcnh.net

For policy questions, send mail to:

 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc



Re: Conducted Emissions PK/AVG

2004-12-16 Thread owner-emc-p...@ieee.org
scale set to LOG, 
 
OOPS, that should have been LINEAR  Sorry.


Bob Richards b...@toprudder.com wrote:

Alex,
 
It is very common that the emissions plots will be made with the RBW and VBW
that you mentioned. The actual QP and AV measurements should have the
instrument set with the appropriate detector and bandwidths. The RBW should
always be 9kHz, and for the average measurement the VBW should be
significantly lower than the pulse repetition frequency, the detector should
be SAMPLE, scale set to LOG, etc. Also, the frequency span should be set to
zero for a compliance measurement. I've seen commercial automated software
that does a very poor job of this.
 
Bob Richards, NCT
Square D.

Alex McNeil alex.mcn...@ingenico.co.uk wrote:

Hi Guys,

I had a product tested that failed the conducted emissions. The switch mode
power supply was the problem. It runs at approx. 200KHz. The failed readings
were in the 150KHz to 600KHz band. The peaks were over the Avg and below the
QPk, so the Avg test was done. This resulted in the failures. However, after
close examination of the Test lab results, the printout showed IF BW 9KHz AND
AVG BW 30KHz below the graphical and tabulated results. The tests were done on
a SA. I have a receiver here and my test results for PK are the same as the
Test Lab but for the Avg my test results are significantly lower than the Test
House.

 

Is the problem that the Test House may have used “AVG BW 30KHz” and this
should have been “AVG BW 9KHz” the problem or am I missing something here?

 

Thank you for your time, it is always appreciated.

 

Kind Regards

ALEX

 This message
is from the IEEE Product Safety Engineering Society emc-pstc discussion list.
Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to emc-p...@ieee.org 


Instructions: http://listserv.ieee.org/listserv/request/user-guide.html 


List rules: http://www.ieee-pses.org/listrules.html 


For help, send mail to the list administrators: 


Scott Douglas emcp...@ptcnh.net 


For policy questions, send mail to: 


Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org 


All emc-pstc postings are archived and searchable on the web at: 


http://www.ieeecommunities.org/emc-pstc

 This message
is from the IEEE Product Safety Engineering Society emc-pstc discussion list.
Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to emc-p...@ieee.org 


Instructions: http://listserv.ieee.org/listserv/request/user-guide.html 


List rules: http://www.ieee-pses.org/listrules.html 


For help, send mail to the list administrators: 


Scott Douglas emcp...@ptcnh.net 


For policy questions, send mail to: 


Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org 


All emc-pstc postings are archived and searchable on the web at: 


http://www.ieeecommunities.org/emc-pstc




Re: Conducted Emissions PK/AVG

2004-12-16 Thread owner-emc-p...@ieee.org
Alex,
 
It is very common that the emissions plots will be made with the RBW and VBW
that you mentioned. The actual QP and AV measurements should have the
instrument set with the appropriate detector and bandwidths. The RBW should
always be 9kHz, and for the average measurement the VBW should be
significantly lower than the pulse repetition frequency, the detector should
be SAMPLE, scale set to LOG, etc. Also, the frequency span should be set to
zero for a compliance measurement. I've seen commercial automated software
that does a very poor job of this.
 
Bob Richards, NCT
Square D.

Alex McNeil alex.mcn...@ingenico.co.uk wrote:

Hi Guys,

I had a product tested that failed the conducted emissions. The switch mode
power supply was the problem. It runs at approx. 200KHz. The failed readings
were in the 150KHz to 600KHz band. The peaks were over the Avg and below the
QPk, so the Avg test was done. This resulted in the failures. However, after
close examination of the Test lab results, the printout showed IF BW 9KHz AND
AVG BW 30KHz below the graphical and tabulated results. The tests were done on
a SA. I have a receiver here and my test results for PK are the same as the
Test Lab but for the Avg my test results are significantly lower than the Test
House.

 

Is the problem that the Test House may have used “AVG BW 30KHz” and this
should have been “AVG BW 9KHz” the problem or am I missing something here?

 

Thank you for your time, it is always appreciated.

 

Kind Regards

ALEX

 This message
is from the IEEE Product Safety Engineering Society emc-pstc discussion list.
Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to emc-p...@ieee.org 


Instructions: http://listserv.ieee.org/listserv/request/user-guide.html 


List rules: http://www.ieee-pses.org/listrules.html 


For help, send mail to the list administrators: 


Scott Douglas emcp...@ptcnh.net 


For policy questions, send mail to: 


Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org 


All emc-pstc postings are archived and searchable on the web at: 


http://www.ieeecommunities.org/emc-pstc




RE: Conducted Emissions PK/AVG

2004-12-16 Thread owner-emc-p...@ieee.org
Alex,
 
You are on the right track. However, even a VBW of 9 kHz will not yield a true
average measurement in this frequency range.
 
First of all, it is not likely that the spectrum analyzer meets CISPR
requirements as a test receiver. Second, using video averaging as a substitute
for an actual average detector requires a VBW that is significantly less than
the RBW, and the amplitude scale must be linear instead of log (the average of
the log is different than the average). This is a highlight of the most
important settings - CISPR 16 will have the detailed requirements from which
suitable and/or reasonable instrument settings can be determined.
 
It is not surprising to hear of these differences - the receiver measurement
definitely takes precedence over the spectrum analyzer.
 
Thanks,
Mike

From: owner-emc-p...@ieee.org [mailto:owner-emc-p...@ieee.org]On Behalf Of
Alex McNeil
Sent: Thursday, December 16, 2004 8:19 AM
To: emc-p...@ieee.org
Subject: Conducted Emissions PK/AVG


Hi Guys,
I had a product tested that failed the conducted emissions. The switch mode
power supply was the problem. It runs at approx. 200KHz. The failed readings
were in the 150KHz to 600KHz band. The peaks were over the Avg and below the
QPk, so the Avg test was done. This resulted in the failures. However, after
close examination of the Test lab results, the printout showed IF BW 9KHz AND
AVG BW 30KHz below the graphical and tabulated results. The tests were done on
a SA. I have a receiver here and my test results for PK are the same as the
Test Lab but for the Avg my test results are significantly lower than the Test
House.
 
Is the problem that the Test House may have used “AVG BW 30KHz” and this
should have been “AVG BW 9KHz” the problem or am I missing something here?
 
Thank you for your time, it is always appreciated.
 
Kind Regards
ALEX


This email has been scanned for all known viruses and appropriate content by
the Messagelabs mail service.
 This message
is from the IEEE Product Safety Engineering Society emc-pstc discussion list.
Website: http://www.ieee-pses.org/ 
To post a message to the list, send your e-mail to emc-p...@ieee.org 

Instructions: http://listserv.ieee.org/listserv/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators: 

Scott Douglas emcp...@ptcnh.net 

For policy questions, send mail to: 

Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org 

All emc-pstc postings are archived and searchable on the web at: 

http://www.ieeecommunities.org/emc-pstc 
 This message
is from the IEEE Product Safety Engineering Society emc-pstc discussion list.
Website: http://www.ieee-pses.org/ 
To post a message to the list, send your e-mail to emc-p...@ieee.org 

Instructions: http://listserv.ieee.org/listserv/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators: 

Scott Douglas emcp...@ptcnh.net 

For policy questions, send mail to: 

Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org 

All emc-pstc postings are archived and searchable on the web at: 

http://www.ieeecommunities.org/emc-pstc 




Re: Conducted Emissions PK/AVG

2004-12-16 Thread owner-emc-p...@ieee.org
Are you sure it says AVG BW 30kHz, and not AVG BW 30 Hz? Is this a typo?

Also, quasi-peak should always be at or below the peak measurement. This is
why a peak measurement can be made (quickly) and compared to the quasi-peak
limit line, and if all the peaks are below the quasi-peak limit, the
measurement is a pass.

If you have quasi-peak measurements higher than the peaks measurements,
then either something is wrong with the test equipment/software, or the
emissions of the EUT changed between the time the two measurements were
made.

Don Borowski
Schweitzer Engineering Labs
Pullman, WA, USA

Alex McNeil alex.mcn...@ingenico.co.uk wrote on 12/16/2004 08:19:08 AM:

 Hi Guys,
 I had a product tested that failed the conducted emissions. The
 switch mode power supply was the problem. It runs at approx. 200KHz.
 The failed readings were in the 150KHz to 600KHz band. The peaks
 were over the Avg and below the QPk, so the Avg test was done. This
 resulted in the failures. However, after close examination of the
 Test lab results, the printout showed IF BW 9KHz AND AVG BW 30KHz
 below the graphical and tabulated results. The tests were done on a
 SA. I have a receiver here and my test results for PK are the same
 as the Test Lab but for the Avg my test results are significantly
 lower than the Test House.

 Is the problem that the Test House may have used “AVG BW 30KHz” and
 this should have been “AVG BW 9KHz” the problem or am I missing
 something here?

 Thank you for your time, it is always appreciated.

 Kind Regards
 ALEX


This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/listserv/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas emcp...@ptcnh.net

For policy questions, send mail to:

 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc