Re: Size Strain in GSAS

2005-04-18 Thread Nicolae Popa
Leonid,
The lognormal distribution for particle size is not my modeling
(unfortunately), but if you insist, let see once again your equations.

D = Da + 0.25(DaDv)^0.5 and sigmaD = D(Dv/Da - 1/2)/2

For lognormal distribution first equation becomes:
2=(4/3)(1+c)**2+(1/4)sqrt[2*(1+c)**5]
For c=0.05 we obtain:  2=1.87,  for c=0.4,  2=3.43

The second equation becomes:
sqrt(c)=[(9/8)(1+c)-1/2]
For c=0.05,   0.22=0.68,   for c=0.4,  0.63=1.75

Well, taking account that the world is not ideal I'm ready to accept that,
then I think is time to close our discussion.

Best wishes,
Nicolae



- Original Message - 
From: Leonid Solovyov [EMAIL PROTECTED]
To: rietveld_l@ill.fr
Sent: Sunday, April 17, 2005 2:58 PM


 Dear Nicolae,

 I will comment only upon your last statement because the limitations of
 your modeling are clear.

  Well, I don't know where from you taken these formulae
  but I observe that for spheres of equal radius, then zero dispersion,
  you have:
  sigma(D)=5D/4,   different from zero!

 First of all, for spheres of equal radius and IDEAL definition
 of Dv and Da:
   sigmaD = D(Dv/Da - 1/2)/2 = D(9/8 - 1/2)/2 = 5D/16
 Yes it is not zero, but the expressions I derived work only for
 0.05  c  0.4 and I derived them not for IDEAL Dv and Da. If you
 perform WEIGHTED least-squares fitting of TCH p-V function to a profile
 simulated for spherical crystal and added by ~10% background level (to
 be closer to real Rietveld refinement) you will obtain the ratio of
 Dv/Da~3/4 not 9/8! This ratio is wrong but this is what we have from
 WEIGHTED least-squares fitting of TCH p-V to simulated data. In this
 case
   sigmaD = D(Dv/Da - 1/2)/2 = D(3/4 - 1/2)/2 = D/8,
 different from zero again, sorry, this world is not IDEAL.

 Best wishes,
 Leonid




 __
 Do you Yahoo!?
 Plan great trips with Yahoo! Travel: Now over 17,000 guides!
 http://travel.yahoo.com/p-travelguide





Re: Size Strain in GSAS

2005-04-18 Thread alan coelho
Title: Message




This is by far the best topic on 
this list for a long time as opposed to requests for Journal papers which as 
pointed out by someone else is inappropriate in the first place and illegal in 
the second.

Nicolae wrote:
(i) but a sum of two Lorentzians 
is not sharper than the sum of two pVs (Voigts)?

This I know, it should not matter 
what is used as long as the mapping of the 
function to a distribution is done accurately. Whether it is lognomal,gammaor 
something else does does not matter. 

Every thing we are talking about is 
additive meaning that the sum of what ever in 2Th space translates to the sum of 
what ever distributions. From the resulting distribution you are free to extract 
what ever parameter you choose.

The idea in Nick Armstrongs work of 
obtaining a distribution without knowing its functional form is a powerful one. 
But the Baysean approach without a 
functional formresults in large errors 
bars in the distribution, seehttp://nvl.nist.gov/pub/nistpubs/jres/109/1/cnt109-1.htm

What we should be looking 
forare cases where Voigt approximations are not possible. I have only ever seen one 
case where the actual Sinc type ripples are seen in a pattern. This was a 
pattern by Bob Cheary of gold columns. Other reports of ripples do exist in the 
literature (not available to me as I write). We must be careful not to include 
2Th independentbumps produced by 
long narrow Soller slits inserted 
inthe axial plane that limits 
horizontal divergence.

When sample related ripples are seen 
then you can throw Voigt based approximations out the window. In the case of the 
gold columns we fitted three Sinc functions added together. In other words the 
distribution wasreally a limited one.

The work of Nick et al is sound and 
approaches the problems from a different perspective; it does not however negate 
the need to determine a priori information. The question that is openin my opinion is whether a priori information is 
more easily incorporated into a least squares process or a Bayesian Maxent 
approach.

This 
discussion has reinvigoratedmy interest and 
like Bob, whom is now looking for an equation to approximate the log normal 
disribution, I will resurrectsome code myself that I did a while back which 
calculates profiles from a arbitrary distribution for the Sinc function. Instead 
I will include the equation for spherical crystallites and of course a user 
defined one which can be hkl dependent. 

As a hint to those who write such code the calculation 
of a profile for an arbitraty distribution operates at around 5000 profiles per 
second as I noticed over the weekend - not much slower that a gaussian 
Nicolae.Maybe there's no need for a pseuod-Voigt / Lorentzian 
basedapproximations after all.


all 
the best
Alan


  
  -Original Message-From: Nicolae Popa 
  [mailto:[EMAIL PROTECTED] Sent: Sunday, April 17, 2005 9:00 
  AMTo: rietveld_l@ill.frSubject: Re: Size Strain in 
  GSAS
  Alan,
  
  (i) but a sum of two Lorentzians is not sharper 
  than the sum of two pVs (Voigts)?
  
  (ii) We fitted the exact size profile 
  caused by the lognormal distribution by a pV (for low lognormal dispersion) or 
  by a sum of maximum 3 Lorenzians (for large lognormal 
dispersion).
  This is "cheaper" than the sum of 2 
  pVs.It involvesthe calculation ofmaximum 3 elementary 
  functions with 4 independent parameters (3 breadths + 2 mixing 
  parametersminus1 constraint =4)
  Sum of two pVs presumes 4 elementary function and 
  5 independent parameters (2 for one pV + 2 for the second one + a mixing 
  parameter).
  
  Best wishes,
  Nicolae
  
  
  


A pure peak fitting approach shows that two pVs (or two Voigts) 
when added with different FWHMs 
andintegrated intensities but similar peak positions and eta 
values can almost exactly fit Pearsons II functions that are sharper that 
Lorentzians. This is not surprising as both profiles comprise 6 parameters. 


Thus from my observations two pVs added together can fit a 
bimodal distributions quite easily. In fact my guess is that two pVs can fit 
a large range of crystallite size distributions. 


 

all the best
alan



Re: Size Strain in GSAS

2005-04-18 Thread Matteo Leoni
buna Nicolae,

 Not only arithmetic, I think is clear that both R and c were refined in a
 whole pattern least square fitting. A private program, not a popular
 Rietveld program because no one has inplemented the size profile caused by
 the lognormal distribution.

not sure no one did.. we're working with that kind of profiles at 
least since 2000 (published in 2001 Acta Cryst A57, 204), without the need 
for any approximation going through Voigts or Pseudo Voigts. Using FFT and 
some math tricks you can compute the true profile for a distribution of 
crystallites almost in the same time you calculate a Voigt curve, so why 
the need to use any approximate function? 
I think this agrees with what Alan just pointed out (well 5000 profiles 
per second if you do not include any hkl dependent broadening that has 
to be calculated for each of them (and perhaps for each subcomponent)... 
otherwise the speed reduces.. but yes few ms for each profile is the 
current speed for my WPPM code, implementing all this stuff within the 
WPPM frame). 

  But the most important disadvantage is the necessity to choose the
  exact type of size distribution. For Sample 1 (which, obviously, have
  certain distribution with certain R and c) you got quite different
  values of R and c for lognorm and gamma models, but the values of Dv
  and Da were nearly the same. Don't you feel that Dv and Da values
  contain more reliable information about R and c than those
  elaborate approximations described in chapter 6?
 
 Well, this is the general feature of the least square method. In the least
 square you must firstly to choose a parametrised model for something that
 you wish to fit.  Do you know another posibility with the least square than
 to priory choose the model? Without model is only the deconvolution, and
 even there, if you wish a stable solution you must use a deconvolution
 method that requres a prior, starting model (I presume you followed the
 disertation of Nick Armstrong on this theme).

also in this case it has ben shown possible to obtain a distribution 
without any prior information on its functional shape (J.Appl.Cryst (2004), 
37, 629) and without taking the maxent treatment into account. 
I'm currently using without much problems for the analysis of 
nanostructured materials... advantages with respect to maxent are the 
speed and the fact that it can coexist with other broadening models (still  
not possible with maxent and still have to see a specimen where strain  
broadening is absent) and it's able to recover also a polydisperse  
distribution if it's present Just need to test it against maxent (if 
data would be kindly provided to do so).
For the purists, just redo the calculation starting from different points 
and you can evaluate the error in the distribution using a 
MonteCarlo-like approach...

As for the TCH-pV, well, it is no more than a pV with the Scherrer 
trend (1/cos) and the differential of Bragg's law (tan term) plugged in.
This means it is ok as long as you consider a Williamson-Hall plot a good 
quantitative estimator for size and strain (IMHO).

Mat

PS I fully agree with Alan on the continuous request for Journals, but I 
bet the other Alan (the deus ex machina of the mailing list) should warn 
the members somehow...

--
Matteo Leoni
Department of Materials Engineering
and Industrial Technologies 
University of Trento
38050 Mesiano (TN)
ITALY






Re: Size Strain in GSAS

2005-04-18 Thread Von Dreele, Robert B.
Nic,
Thanks,it will take a while (as usual) to implement.
Bob

R.B. Von Dreele
IPNS Division
Argonne National Laboratory
Argonne, IL 60439-4814



-Original Message-
From: Nicolae Popa [mailto:[EMAIL PROTECTED] 
Sent: Sunday, April 17, 2005 1:27 AM
To: rietveld_l@ill.fr



Bob,

A nice math. description amenable to RR exists, take a look at
JAC(2002) 35, 338-346. Nice because the size profile is described by a
pV (at regular lognormal
dispersions) or by a sum of maximum three Lorentzians (at large
lognormal dispersions - those 3% that Alan spiked about). The breadths
and mixing parameters of pV or of the sum of Lorenzians are calculated
analytically from the two parameters of the lognormal distribution, R
and c.  You realize that if approximate the instrumental and the strain
profiles by Voigts, it results a sum of three Voigts for the whole
profile and for a fast computation, every Voigt can be replaced by our
loved TCH pV.

Best wishes,
Nic


 Nic,
 Well, I have been tempted from time to time to implement a log 
 normal
type distribution in on eof the profile functions. A nice math
description ameanable to RR would help.
 Bob








Re: Size Strain in GSAS

2005-04-17 Thread Nicolae Popa
Title: Message



Alan,

(i) but a sum of two Lorentzians is not sharper 
than the sum of two pVs (Voigts)?

(ii) We fitted the exact size profile caused 
by the lognormal distribution by a pV (for low lognormal dispersion) or by a sum 
of maximum 3 Lorenzians (for large lognormal dispersion).
This is "cheaper" than the sum of 2 
pVs.It involvesthe calculation ofmaximum 3 elementary 
functions with 4 independent parameters (3 breadths + 2 mixing 
parametersminus1 constraint =4)
Sum of two pVs presumes 4 elementary function and 5 
independent parameters (2 for one pV + 2 for the second one + a mixing 
parameter).

Best wishes,
Nicolae



  
  
  A pure peak fitting approach shows that two pV’s (or two Voigts) 
  when added with different FWHMs 
  andintegrated intensities but similar peak positions and eta 
  values can almost exactly fit Pearsons II functions that are sharper that 
  Lorentzians. This is not surprising as both profiles comprise 6 parameters. 
  
  
  Thus from my observations two pVs added together can fit a bimodal 
  distributions quite easily. In fact my guess is that two pVs can fit a large 
  range of crystallite size distributions. 
  
  
  
  all the best
  alan
  


Re: Size Strain in GSAS

2005-04-17 Thread Nicolae Popa

Bob,

A nice math. description amenable to RR exists, take a look at JAC(2002)
35, 338-346.
Nice because the size profile is described by a pV (at regular lognormal
dispersions) or by a sum of maximum three Lorentzians (at large lognormal
dispersions - those 3% that Alan spiked about). The breadths and mixing
parameters of pV or of the sum of Lorenzians are calculated analytically
from the two parameters of the lognormal distribution, R and c.  You
realize that if approximate the instrumental and the strain profiles by
Voigts, it results a sum of three Voigts for the whole profile and for a
fast computation, every Voigt can be replaced by our loved TCH pV.

Best wishes,
Nic


 Nic,
 Well, I have been tempted from time to time to implement a log normal
type distribution in on eof the profile functions. A nice math description
ameanable to RR would help.
 Bob





Re: Size Strain in GSAS

2005-04-17 Thread Leonid Solovyov
Dear Nicolae,

I will comment only upon your last statement because the limitations of
your modeling are clear.

 Well, I don't know where from you taken these formulae
 but I observe that for spheres of equal radius, then zero dispersion,
 you have:
 sigma(D)=5D/4,   different from zero!

First of all, for spheres of equal radius and IDEAL definition
of Dv and Da:
  sigmaD = D(Dv/Da - 1/2)/2 = D(9/8 - 1/2)/2 = 5D/16
Yes it is not zero, but the expressions I derived work only for
0.05  c  0.4 and I derived them not for IDEAL Dv and Da. If you
perform WEIGHTED least-squares fitting of TCH p-V function to a profile
simulated for spherical crystal and added by ~10% background level (to
be closer to real Rietveld refinement) you will obtain the ratio of
Dv/Da~3/4 not 9/8! This ratio is wrong but this is what we have from
WEIGHTED least-squares fitting of TCH p-V to simulated data. In this
case 
  sigmaD = D(Dv/Da - 1/2)/2 = D(3/4 - 1/2)/2 = D/8,
different from zero again, sorry, this world is not IDEAL.

Best wishes,
Leonid




__ 
Do you Yahoo!? 
Plan great trips with Yahoo! Travel: Now over 17,000 guides!
http://travel.yahoo.com/p-travelguide


Re: Size Strain in GSAS

2005-04-16 Thread Leonid Solovyov
 Indeed you missed something. I presume you have the paper.
 Then, take a look to the formula (15a). This is the size 
 profile for lognormal.
 There is the function PHI - bar of argument 2*pi*s*R.
 Replace this function PHI - bar from (15a) by the _expression
 (21a) with the argument x=2*pi*s*R.
 You get it? So, not only c but also R.

Dear Nicolae,

This arithmetic is clear, thanks, but since you did not specify this
exact way of R calculation in the paper it was not evident. There are
several other ways of deriving R, for instance: to calculate Dv from
the inverse integral breadth and then use eq. (12) or (17) etc.
Besides, you did not refine R for simulated data in chapter 6 - it
was fixed. When you apply this formalism to real data you refine both
R and c, they may correlate and the result of such correlation is not
apparent.

But the most important disadvantage is the necessity to choose the
exact type of size distribution. For Sample 1 (which, obviously, have
certain distribution with certain R and c) you got quite different
values of R and c for lognorm and gamma models, but the values of Dv
and Da were nearly the same. Don't you feel that Dv and Da values
contain more reliable information about R and c than those
elaborate approximations described in chapter 6?

In new version of DDM (see the following message) I included some
estimations of average crystallite diameter D and its dispersion
sigmaD based on empirical approximations derived from fitting TCH-pV
function to simulated profiles for the model of spherical  
crystallites with different size distribution dispersions. For
simulated data (which are supplied with the DDM package) these magic
expressions:

 D = Da + 0.25(DaDv)^0.5 and sigmaD = D(Dv/Da - 1/2)/2

allowed reproducing D and sigmaD with less than 10% deviation in
the interval of relative dispersions 0.05  c  0.4 for both gamma and
lognorm distributions. Of course, I don't think that these expressions
are perfect and I would be glad to see better estimations.

Best regards,
Leonid




__ 
Do you Yahoo!? 
Plan great trips with Yahoo! Travel: Now over 17,000 guides!
http://travel.yahoo.com/p-travelguide


Re: Size Strain in GSAS

2005-04-15 Thread Nicolae Popa
Title: Message



Dear Bob,

Perhaps I was not enough clear. Let me be more 
explicit.
It's about one sample of CeO2 (not that from 
round-robin) that we fitted in 4 ways.

(i) by GSAS with 
TCH-pV
(ii) by another pV resulted from gamma 
distribution of size
(iii) by Lorentz - (the limit of any 
"regular" pV - eta=1)

All these 3 variants given bad fits. For example 
(ii): Rw=0.144, similarly the rest. (In principle if one pV works (TCH for 
example) any other kind of pV must work.)

(iv) by the profile resulted from lognormal 
distribution of size; this time the fit was reasonably good: Rw=0.047. It 
resulted c=2.8, that means a "super Lorentzian" profile (I remember that the 
Lorentzian limit for lognormal is c~0.4 - JAC (2002) 35, 
338).
Attention, this "super Lorentzian" profile is not 
constructed as a pV with eta1.

Sure, such samples are rare, or, perhaps, not so 
rare. A Jap. group (Ida,, Toraya, JAC (2003) 36, 1107) reported "super 
Lorentzian"on a sample of SiC. They found c=1.37

Best wishes,
Nic Popa


  Nic,
  This 
  is true for the internal math but the TCH function was assembled to reproduce 
  the true Voigt over the entire range of differing Lorentz and Gauss FWHM 
  values so it works as if the two FWHM components are independent. As for your 
  question, I'm not aware that anyone has actually tried to do the fit both ways 
  on a "super Lorentzian" (eta1 for old psVoigt) sample to see if a) the fit 
  is the same and b) the eta1 was an artifact. Any takers to settle 
  this?
  Bob
  
  
  
  R.B. Von Dreele
  IPNS Division
  Argonne National Laboratory
  Argonne, IL 60439-4814
  
  

-Original Message-From: Nicolae Popa 
[mailto:[EMAIL PROTECTED] Sent: Thursday, April 14, 2005 9:11 
    AMTo: rietveld_l@ill.frSubject: Re: Size Strain in 
GSAS
Dear Bob,

If I understand well, you say that eta1 
(super Lorenzian) appeared only because eta was free parameter, but if TCH 
is used super Loreanzians do not occur?
Nevertheless, for that curious sample of cerium 
oxide wetried GSAS (with TCH) and the fit was very bad.
Best wishes,
Nicolae

PS. By the way, TCH also forces FWHM of the 
Gaussian and Lorenzian components to be equal, but indeed, eta is not free 
and cannot be greater than 1.


  
  Nic,
  I know about "super Lorentzians". Trouble is that many of those 
  older reports were from Rietveld refinements "pre TCH" and used a 
  formulation of the pseudo-Voigt which forced the FWHM of the Gaussian and 
  Lorentzian components to be equal and allowed the mixing coefficient (eta) 
  to be a free variable (n.b. it is not free in the TCH formulation). Thus, 
  these ought to be discounted in any discussion about the occurence of 
  super Lorentzian effects in real samples.
  Bob
  
  
  
  R.B. Von Dreele
  IPNS Division
  Argonne National Laboratory
  Argonne, IL 60439-4814
  
  

-Original Message-From: Nicolae 
Popa [mailto:[EMAIL PROTECTED] Sent: Thursday, April 14, 2005 
    8:10 AMTo: rietveld_l@ill.frSubject: Re: Size 
Strain in GSAS

Right, is rare, but we have meet once. A 
cerium oxide sample from a commercial company, c=2.8. I don't know if 
they did deliberately, probably not, otherwise the hard work to obtain 
such curiosity is costly and the company risks a bankruptcy. On the 
other hand superlorenzian profiles were reported from a long time, only 
were interpreted as coming from bimodal size distributions. And third, 
you see, people have difficulties to extract size distribution from the 
Rietveld codes as they areat this moment.

Nicolae Popa


  
  A word from a "provider" of a Rietveld code (please don't call 
  me a "programmer"). 
  "But if 
  c0.4 any pV fails" - OK, for what fraction of the universe of 
  "real world" samples is "c"0.4? I suspect, given the general 
  success of the TCH pseudoVoigt function, that it is exceedingly small 
  and only occurs when one works hard to deliberately make a sample like 
  that.
  
  
  
  R.B. Von Dreele
  IPNS Division
  Argonne National Laboratory
  Argonne, IL 60439-4814
  
  

-Original Message-From: 
Nicolae Popa [mailto:[EMAIL PROTECTED] Sent: Thursday, 
    April 14, 2005 7:14 AMTo: 
[EMAIL PROTECTED]Cc: 
rietveld_l@ill.frSubject: Re: Size Strain in 
GSAS
Dear Nicolae, Maybe ya ploho 
chitayu i ploho soobrazhayu, but even after yourexplanation 
 

Re: Size Strain in GSAS

2005-04-15 Thread alan coelho
Title: Message




Nicolae, Nick, Bob, 
Leonid,

I have looked at many patterns (recorded by others) and a few cases 
have shown profiles that are sharper that a Lorentzian; whereby sharper means 
that the integral breadth is smaller than that for a unit area lorentzian. 


To put a figure on it would be difficult but at a guess I would say 
 3% of patterns fall into this category in a noticeable manner. 


I have no doubt that the work of Nick Armstrong and co. is 
mathematically sound but a 
simulated data round robin as suggested by Leonid Solovyov may be useful  and I 
am not generally a fan of round robins but this s different as the data is 
simulated. 

A pure peak fitting approach shows that two pVs (or two Voigts) when 
added with different FWHMs 
andintegrated intensities but similar peak positions and eta values 
can almost exactly fit Pearsons II functions that are sharper that Lorentzians. 
This is not surprising as both profiles comprise 6 parameters. 


Thus from my observations two pVs added together can fit a bimodal 
distributions quite easily. In fact my guess is that two pVs can fit a large 
range of crystallite size distributions. 

Thus distinguishing whether a distribution is not 
monomodal is of course possible especially if the two pV approach is taken. 


Attempting to determine more than that however takes 
some convincing as two pVs seem to fit almost anything that I have seen that is 
symmetric. Thus introducing more pVs seems 
unnecessary.

Thus yes GSAS can determine if a distribution is not monmodal if you 
were to fit two identical phases to the pattern except for the TCH parameters. 
If the Rwp drops by .1% then I wont be convinced.

Forgive me Nick but I have not yet read all of your work 
and I am certain that it is sound. Outside of nano particles (and maybe even 
inside) my reservation are that we may well be analyzing noise and second order 
sample and instrumental effects.

Thus to show up my naive ness can you categorically say 
that there are real world distributions that two pseudo Voigts cannot fit 
because I have not come across such a pattern. 

Once you have done that 
then it would be time to concentrate on strain, micro strain, surface roughness 
and then disloactions

all the best
alan



  
  -Original Message-From: Nicolae Popa 
  [mailto:[EMAIL PROTECTED] Sent: Friday, April 15, 2005 9:30 
  AMTo: rietveld_l@ill.frSubject: Re: Size Strain in 
  GSAS
  Dear Bob,
  
  Perhaps I was not enough clear. Let me be more 
  explicit.
  It's about one sample of CeO2 (not that from 
  round-robin) that we fitted in 4 ways.
  
  (i) by GSAS with 
  TCH-pV
  (ii) by another pV resulted from 
  gamma distribution of size
  (iii) by Lorentz - (the limit of any 
  "regular" pV - eta=1)
  
  All these 3 variants given bad fits. For example 
  (ii): Rw=0.144, similarly the rest. (In principle if one pV works (TCH for 
  example) any other kind of pV must work.)
  
  (iv) by the profile resulted from lognormal 
  distribution of size; this time the fit was reasonably good: Rw=0.047. It 
  resulted c=2.8, that means a "super Lorentzian" profile (I remember that the 
  Lorentzian limit for lognormal is c~0.4 - JAC (2002) 35, 
  338).
  Attention, this "super Lorentzian" profile is not 
  constructed as a pV with eta1.
  
  Sure, such samples are rare, or, perhaps, not so 
  rare. A Jap. group (Ida,, Toraya, JAC (2003) 36, 1107) reported "super 
  Lorentzian"on a sample of SiC. They found c=1.37
  
  Best wishes,
  Nic Popa
  
  
Nic,
This is true for the internal math but the TCH function was assembled 
to reproduce the true Voigt over the entire range of differing Lorentz and 
Gauss FWHM values so it works as if the two FWHM components are independent. 
As for your question, I'm not aware that anyone has actually tried to do the 
fit both ways on a "super Lorentzian" (eta1 for old psVoigt) sample to 
see if a) the fit is the same and b) the eta1 was an artifact. Any 
takers to settle this?
Bob



R.B. Von Dreele
IPNS Division
Argonne National Laboratory
Argonne, IL 60439-4814


  
  -Original Message-From: Nicolae Popa 
  [mailto:[EMAIL PROTECTED] Sent: Thursday, April 14, 2005 9:11 
      AMTo: rietveld_l@ill.frSubject: Re: Size Strain in 
  GSAS
  Dear Bob,
  
  If I understand well, you say that eta1 
  (super Lorenzian) appeared only because eta was free parameter, but if TCH 
  is used super Loreanzians do not occur?
  Nevertheless, for that curious sample of 
  cerium oxide wetried GSAS (with TCH) and the fit was very 
  bad.
  Best wishes,
  Nicolae
  
  PS. By the way, TCH also forces FWHM of the 
  Gaussian and Lorenzian components to be equal, but indeed, eta is not free 
  and cannot be greater than 1.
  
  

Nic,
I kn

Re: Size Strain in GSAS

2005-04-15 Thread Jim Cline

Hi,
In response to some of this post:
There was a move by a bunch of us in the ICDD to hold a profile fitting
round robin ( which I think would by quite useful ). But it died
when we realized the prodigious level of resources that would be required
to make sense of the rather large matrix of data that would
arrive.
But with regards to a round robin on this question: seems to me some
qualified individual could simply do the work and publish a nice paper on
it.
Regards,
Jim
At 12:30 PM 4/15/2005 +0200, you wrote:
urn:schemas-microsoft-com:office:office

Nicolae, Nick, Bob, Leonid,

I have looked at many patterns (recorded by others)
and a few cases have shown profiles that are sharper that a Lorentzian;
whereby sharper means that the integral breadth is smaller than that for
a unit area lorentzian. 

To put a figure on it would be difficult but at a
guess I would say  3% of patterns fall into this category in a
noticeable manner. 

I have no doubt that the work of Nick Armstrong and
co. is mathematically sound but a simulated data round robin as
suggested by Leonid Solovyov may be useful and I am not generally a fan
of round robins but this s different as the data is simulated. 

A pure peak fitting approach shows that two pV s (or
two Voigts) when added with different FWHMs and integrated intensities
but similar peak positions and eta values can almost exactly fit Pearsons
II functions that are sharper that Lorentzians. This is not surprising as
both profiles comprise 6 parameters. 

Thus from my observations two pVs added together can
fit a bimodal distributions quite easily. In fact my guess is that two
pVs can fit a large range of crystallite size distributions. 

Thus distinguishing whether a distribution is not
monomodal is of course possible especially if the two pV approach is
taken. 

Attempting to determine more than that however takes
some convincing as two pVs seem to fit almost anything that I have seen
that is symmetric. Thus introducing more pVs seems unnecessary.

Thus yes GSAS can determine if a distribution is not
monmodal if you were to fit two identical phases to the pattern except
for the TCH parameters. If the Rwp drops by .1% then I wont be
convinced.

Forgive me Nick but I have not yet read all of your
work and I am certain that it is sound. Outside of nano particles (and
maybe even inside) my reservation are that we may well be analyzing noise
and second order sample and instrumental effects.

Thus to show up my naive ness can you categorically
say that there are real world distributions that two pseudo Voigts cannot
fit because I have not come across such a pattern. 

Once you have done that then it would be time to
concentrate on strain, micro strain, surface roughness and then
disloactions

all the best

alan


-Original Message-
From: Nicolae Popa
[mailto:[EMAIL PROTECTED]] 
Sent: Friday, April 15, 2005 9:30 AM
To: rietveld_l@ill.fr
Subject: Re: Size Strain in GSAS

Dear Bob,

Perhaps I was not enough clear. Let me be more
explicit.
It's about one sample of CeO2 (not that from round-robin) that we
fitted in 4 ways.

(i) by GSAS with
TCH-pV
(ii) by another pV resulted from gamma distribution of
size
(iii) by Lorentz - (the limit of any regular pV -
eta=1)

All these 3 variants given bad fits. For example (ii): Rw=0.144,
similarly the rest. (In principle if one pV works (TCH for example) any
other kind of pV must work.)

(iv) by the profile resulted from lognormal distribution of
size; this time the fit was reasonably good: Rw=0.047. It resulted c=2.8,
that means a super Lorentzian profile (I remember that the
Lorentzian limit for lognormal is c~0.4 - JAC (2002) 35,
338).
Attention, this super Lorentzian profile is not
constructed as a pV with eta1.

Sure, such samples are rare, or, perhaps, not so rare. A Jap. group
(Ida,, Toraya, JAC (2003) 36, 1107) reported super
Lorentzian on a sample of SiC. They found c=1.37

Best wishes,
Nic Popa

Nic,
This is true for the internal math but the TCH function was assembled
to reproduce the true Voigt over the entire range of differing Lorentz
and Gauss FWHM values so it works as if the two FWHM components are
independent. As for your question, I'm not aware that anyone has actually
tried to do the fit both ways on a super Lorentzian (eta1
for old psVoigt) sample to see if a) the fit is the same and b) the
eta1 was an artifact. Any takers to settle
this?
Bob



R.B. Von Dreele

IPNS Division

Argonne National Laboratory

Argonne, IL 60439-4814


-Original Message-
From: Nicolae Popa
[mailto:[EMAIL PROTECTED]] 
Sent: Thursday, April 14, 2005 9:11 AM
To: rietveld_l@ill.fr
Subject: Re: Size Strain in GSAS

Dear Bob,

If I understand well, you say that eta1 (super Lorenzian)
appeared only because eta was free parameter, but if TCH is used super
Loreanzians do not occur?
Nevertheless, for that curious sample of cerium oxide we tried GSAS
(with TCH) and the fit was very bad.
Best wishes,
Nicolae

PS. By the way, TCH also forces

Re: Size Strain in GSAS

2005-04-15 Thread Von Dreele, Robert B.
Nic,
Well, I have been tempted from time to time to implement a log normal type 
distribution in on eof the profile functions. A nice math description 
ameanable to RR would help.
Bob



From: Nicolae Popa [mailto:[EMAIL PROTECTED]
Sent: Fri 4/15/2005 2:30 AM
To: rietveld_l@ill.fr


Dear Bob,
 
Perhaps I was not enough clear. Let me be more explicit.
It's about one sample of CeO2 (not that from round-robin) that we fitted in 4 
ways.
 
(i)by GSAS with TCH-pV
(ii)   by another pV resulted from gamma distribution of size
(iii)  by Lorentz - (the limit of any regular pV - eta=1)
 
All these 3 variants given bad fits. For example (ii): Rw=0.144, similarly the 
rest. (In principle if one pV works (TCH for example) any other kind of pV must 
work.)
 
(iv)  by the profile resulted from lognormal distribution of size; this time 
the fit was reasonably good: Rw=0.047. It resulted c=2.8, that means a super 
Lorentzian profile (I remember that the Lorentzian limit for lognormal is 
c~0.4 -  JAC (2002) 35, 338).
Attention, this super Lorentzian profile is not constructed as a pV with 
eta1.
 
Sure, such samples are rare, or, perhaps, not so rare. A Jap. group (Ida,, 
Toraya, JAC (2003) 36, 1107) reported super Lorentzian on a sample of SiC. 
They found c=1.37
 
Best wishes,
Nic Popa
 

Nic,
This is true for the internal math but the TCH function was assembled 
to reproduce the true Voigt over the entire range of differing Lorentz and 
Gauss FWHM values so it works as if the two FWHM components are independent. As 
for your question, I'm not aware that anyone has actually tried to do the fit 
both ways on a super Lorentzian (eta1 for old psVoigt) sample to see if a) 
the fit is the same and b) the eta1 was an artifact. Any takers to settle this?
Bob
 
 

R.B. Von Dreele

IPNS Division

Argonne National Laboratory

Argonne, IL 60439-4814

 

-Original Message-
From: Nicolae Popa [mailto:[EMAIL PROTECTED] 
Sent: Thursday, April 14, 2005 9:11 AM
To: rietveld_l@ill.fr
Subject: Re: Size Strain in GSAS


Dear Bob,
 
If I understand well, you say that eta1 (super Lorenzian) 
appeared only because eta was free parameter, but if TCH is used super 
Loreanzians do not occur?
Nevertheless, for that curious sample of cerium oxide we tried 
GSAS (with TCH) and the fit was very bad.
Best wishes,
Nicolae
 
PS. By the way, TCH also forces FWHM of the Gaussian and 
Lorenzian components to be equal, but indeed, eta is not free and cannot be 
greater than 1.
 



Nic,
I know about super Lorentzians. Trouble is that many 
of those older reports were from Rietveld refinements pre TCH and used a 
formulation of the pseudo-Voigt which forced the FWHM of the Gaussian and 
Lorentzian components to be equal and allowed the mixing coefficient (eta) to 
be a free variable (n.b. it is not free in the TCH formulation). Thus, these 
ought to be discounted in any discussion about the occurence of super 
Lorentzian effects in real samples.
Bob
 
 

R.B. Von Dreele

IPNS Division

Argonne National Laboratory

Argonne, IL 60439-4814

 

-Original Message-
From: Nicolae Popa [mailto:[EMAIL PROTECTED] 
Sent: Thursday, April 14, 2005 8:10 AM
To: rietveld_l@ill.fr
Subject: Re: Size Strain in GSAS


 
Right, is rare, but we have meet once. A cerium 
oxide sample from a commercial company, c=2.8. I don't know if they did 
deliberately, probably not, otherwise the hard work to obtain such curiosity is 
costly and the company risks a bankruptcy. On the other hand superlorenzian 
profiles were reported from a long time, only were interpreted as coming from 
bimodal size distributions. And third, you see, people have difficulties to 
extract size distribution from the Rietveld codes as they are at this moment.
 
Nicolae Popa
 



A word from a provider of a Rietveld 
code (please don't call me

Re: Size Strain in GSAS

2005-04-15 Thread Von Dreele, Robert B.
Alan,
Ah - the rocks  dust model. It works well.
Bob



From: alan coelho [mailto:[EMAIL PROTECTED]
Sent: Fri 4/15/2005 5:30 AM
To: rietveld_l@ill.fr



Nicolae, Nick, Bob, Leonid,

 

I have looked at many patterns (recorded by others) and a few cases have shown 
profiles that are sharper that a Lorentzian; whereby sharper means that the 
integral breadth is smaller than that for a unit area lorentzian. 

 

To put a figure on it would be difficult but at a guess I would say  3% of 
patterns fall into this category in a noticeable manner. 

 

I have no doubt that the work of Nick Armstrong and co. is mathematically sound 
but  a simulated data round robin as suggested by Leonid Solovyov may be useful 
- and I am not generally a fan of round robins but this s different as the data 
is simulated. 

 

A pure peak fitting approach shows that two pV's (or two Voigts) when added 
with different FWHMs and integrated intensities but similar peak positions and 
eta values can almost exactly fit Pearsons II functions that are sharper that 
Lorentzians. This is not surprising as both profiles comprise 6 parameters. 

 

Thus from my observations two pVs added together can fit a bimodal 
distributions quite easily. In fact my guess is that two pVs can fit a large 
range of crystallite size distributions. 

 

Thus distinguishing whether a distribution is not monomodal is of course 
possible especially if the two pV approach is taken. 

 

Attempting to determine more than that however takes some convincing as two pVs 
seem to fit almost anything that I have seen that is symmetric. Thus 
introducing more pVs seems unnecessary.

 

Thus yes GSAS can determine if a distribution is not monmodal if you were to 
fit two identical phases to the pattern except for the TCH parameters. If the 
Rwp drops by .1% then I wont be convinced.

 

Forgive me Nick but I have not yet read all of your work and I am certain that 
it is sound. Outside of nano particles (and maybe even inside) my reservation 
are that we may well be analyzing noise and second order sample and 
instrumental effects.

 

Thus to show up my naive ness can you categorically say that there are real 
world distributions that two pseudo Voigts cannot fit because I have not come 
across such a pattern. 

 

Once you have done that then it would be time to concentrate on strain, micro 
strain, surface roughness and then disloactions

 

all the best

alan

 

 


-Original Message-
From: Nicolae Popa [mailto:[EMAIL PROTECTED] 
Sent: Friday, April 15, 2005 9:30 AM
To: rietveld_l@ill.fr
Subject: Re: Size Strain in GSAS


Dear Bob,
 
Perhaps I was not enough clear. Let me be more explicit.
It's about one sample of CeO2 (not that from round-robin) that we 
fitted in 4 ways.
 
(i)by GSAS with TCH-pV
(ii)   by another pV resulted from gamma distribution of size
(iii)  by Lorentz - (the limit of any regular pV - eta=1)
 
All these 3 variants given bad fits. For example (ii): Rw=0.144, 
similarly the rest. (In principle if one pV works (TCH for example) any other 
kind of pV must work.)
 
(iv)  by the profile resulted from lognormal distribution of size; this 
time the fit was reasonably good: Rw=0.047. It resulted c=2.8, that means a 
super Lorentzian profile (I remember that the Lorentzian limit for lognormal 
is c~0.4 -  JAC (2002) 35, 338).
Attention, this super Lorentzian profile is not constructed as a pV 
with eta1.
 
Sure, such samples are rare, or, perhaps, not so rare. A Jap. group 
(Ida,, Toraya, JAC (2003) 36, 1107) reported super Lorentzian on a sample 
of SiC. They found c=1.37
 
Best wishes,
Nic Popa
 

Nic,
This is true for the internal math but the TCH function was 
assembled to reproduce the true Voigt over the entire range of differing 
Lorentz and Gauss FWHM values so it works as if the two FWHM components are 
independent. As for your question, I'm not aware that anyone has actually tried 
to do the fit both ways on a super Lorentzian (eta1 for old psVoigt) sample 
to see if a) the fit is the same and b) the eta1 was an artifact. Any takers 
to settle this?
Bob
 
 

R.B. Von Dreele

IPNS Division

Argonne National Laboratory

Argonne, IL 60439-4814

 

-Original Message-
From: Nicolae Popa [mailto:[EMAIL PROTECTED] 
Sent: Thursday, April 14, 2005 9:11 AM
To: rietveld_l@ill.fr
Subject: Re: Size Strain in GSAS


Dear Bob

RE: Size Strain in GSAS

2005-04-14 Thread Leonid Solovyov
 It was shown in paragraph 6 of JAC 35 (2002) 338-346 that
 size-broadened
 profiles given by both lognormal and gamma distributions can be
 approximated
 by a weighted sum of Lorentz and Gauss functions for a broad range of
 distribution dispersions. Besides, round robins can sometimes be long
 adventures...

Yes, profiles can be approximated, but the question is not in
approximating profiles. The primary topic of the discussion is Size
Strain in GSAS. GSAS and most other Rietveld refinement programs use
TCH-pV profile function which provides the simplest and more or less
correct way for separating microstructural and instrumental broadening
contributions. Unfortunately, the microstructural parameters such as Dv
and Da sizes derived (classically) from TCH-pV deviate significantly
from reality for narrow and broad dispersions. That's why the
TCH-pV-based calculations of Dv, Da or average crystallite diameter
need to be modified and calibrated on, at least, simulated data for
various dispersions.
The formalisms presented in chapters 6 and 7 of JAC 35 (2002) 338-346
are more complex and they don't give a clear way for separating
microstructural from instrumental effects and, besides, for estimating
the values of Dv, Da or R. 

Leonid


__
Do You Yahoo!?
Tired of spam?  Yahoo! Mail has the best spam protection around 
http://mail.yahoo.com 


Re: Size Strain in GSAS

2005-04-14 Thread Nicolae Popa

...

 Yes, profiles can be approximated, but the question is not in
 approximating profiles. The primary topic of the discussion is Size
 Strain in GSAS. GSAS and most other Rietveld refinement programs use
 TCH-pV profile function which provides the simplest and more or less
 correct way for separating microstructural and instrumental broadening
 contributions. Unfortunately, the microstructural parameters such as Dv
 and Da sizes derived (classically) from TCH-pV deviate significantly
 from reality for narrow and broad dispersions. That's why the
 TCH-pV-based calculations of Dv, Da or average crystallite diameter
 need to be modified and calibrated on, at least, simulated data for
 various dispersions.
 The formalisms presented in chapters 6 and 7 of JAC 35 (2002) 338-346
 are more complex and they don't give a clear way for separating
 microstructural from instrumental effects and, besides, for estimating
 the values of Dv, Da or R.

 Leonid


Dear Leonid,

It is not exact what you say, ty ploho cital.
6  7 from JAC 35 (2002) 338-346 gives the size profile - formulae (15a)
combined with (21,22)
or (20a) combined with (23,24). If you look carefully, these profiles are
approximated by pseudo-Voigt or sums of 2 or 3 Lorentzians. These profiles
depends of 2 parameters, R and c, that are refinable and, once refined,
both Dv and Da can be calculated (formulae (12,13) or (17,18)).

NOW, if approximate the instrumental function by Voigt, that is possible in
very many cases (or sums of Voigts to account for asymmetry for example) and
also the strain effect by Gaussian or even Voigt, the resulted profile will
be a Voigt (or sum of Voigt), that is used as profile in the whole pattern
fitting, this profile including in principle all broadening effects
(isotropic).
You are claiming that it is not TCH-pseudoVoigt. Right, it is not, and can
not be, in general, because for c0.4 the size profile is no more
pseudo-Voigt. The size profile given in that paper cover a much wider range
of  c (for lognormal distribution), including superlorentzians. On the
other hand is a trivial matter for a programmer to include this profile in
any whole pattern fitting code (Rietveld included). (We did that in a
private whole pattern fitting program). But certainly not, if the
programmer wish to use exclusively TCH and nothing else. Why? I don't know.
Note that TCH is an empirical profile that reasonably approximate a Voigt
function (not the tails) that contains an empirical constraint: that FWHH of
Lorenz and Gauss components are equal one to another and equal with that of
the whole psudoVoigt.

Best wishes,
Nicolae Popa




Re: Size Strain in GSAS

2005-04-14 Thread Leonid Solovyov
 It is not exact what you say, ty ploho cital.
 6  7 from JAC 35 (2002) 338-346 gives the size profile - formulae 
 (15a)
 combined with (21,22)
 or (20a) combined with (23,24). If you look carefully, these profiles

 are
 approximated by pseudo-Voigt or sums of 2 or 3 Lorentzians. These 
 profiles
 depends of 2 parameters, R and c, that are refinable and, once 
 refined,
 both Dv and Da can be calculated (formulae (12,13) or (17,18)).


Dear Nicolae, 

Maybe ya ploho chitayu i ploho soobrazhayu, but even after your
explanation I can't see a way to calculate R from the results of
fitting described in chapters 6  7 of JAC 35 (2002) 338-346. From such
fitting you obtain only dispersion parameter c. Or I missed something?
Anyway, being Rietvelders we still have to deal with TCH-pV function
and we need to extract as much as possible correct information from it.
Hope we shall see more appropriate functions for microstructure
analysis in popular Rietveld programs.

Cheers,
Leonid




__ 
Do you Yahoo!? 
Read only the mail you want - Yahoo! Mail SpamGuard. 
http://promotions.yahoo.com/new_mail 


Re: Size Strain in GSAS

2005-04-14 Thread Nicolae Popa



Dear Nicolae, Maybe ya ploho chitayu i 
ploho soobrazhayu, but even after yourexplanation I can't see a way to 
calculate R from the results offitting described in chapters 6 
 7 of JAC 35 (2002) 338-346. From suchfitting you obtain only 
dispersion parameter c. Or I missed something?Anyway, being 
"Rietvelders" we still have to deal with TCH-pV functionand we need to 
extract as much as possible correct information from it.Hope we shall 
see more appropriate functions for microstructureanalysis in popular 
Rietveld programs.Cheers,Leonid
Dear Leonid,

Indeed you missed something. I presume you 
have the paper. Then, take a look to the formula (15a). This is the size profile 
for lognormal. There is the function PHI - bar of argument 
2*pi*s*R. Replace this function PHI - bar from (15a) by the 
_expression_ (21a)with theargument x=2*pi*s*R. You get it? So, 
not only "c" but also R.

"We are Rietvelders" means that we must be 
only "codes drivers", "cheffeurs des codes", "voditeli program"? Have we to 
accept the "Procust bed" of the Rietveld codes at a given moment? All Rietveld 
codes are improving in time, isn't it? 

In particular for the Round_Robin sample 
TCH-pV works because c=0.18. (Davor explained how Dv and Da are found). But if 
c0.4 any pV fails.

Best wishes,

Nicolae



RE: Size Strain in GSAS

2005-04-14 Thread Von Dreele, Robert B.
Title: Message



A word 
from a "provider" of a Rietveld code (please don't call me a "programmer"). 

"But if c0.4 any 
pV fails" - OK, for what fraction of the universe of "real world" samples is 
"c"0.4? I suspect, given the general success of the TCH pseudoVoigt 
function, that it is exceedingly small and only occurs when one works hard to 
deliberately make a sample like that.



R.B. Von Dreele
IPNS Division
Argonne National Laboratory
Argonne, IL 60439-4814


  
  -Original Message-From: Nicolae Popa 
  [mailto:[EMAIL PROTECTED] Sent: Thursday, April 14, 2005 7:14 
  AMTo: [EMAIL PROTECTED]Cc: 
  rietveld_l@ill.frSubject: Re: Size Strain in 
  GSAS
  Dear Nicolae, Maybe ya ploho chitayu i 
  ploho soobrazhayu, but even after yourexplanation I can't see a way to 
  calculate R from the results offitting described in chapters 6 
   7 of JAC 35 (2002) 338-346. From suchfitting you obtain only 
  dispersion parameter c. Or I missed something?Anyway, being 
  "Rietvelders" we still have to deal with TCH-pV functionand we need to 
  extract as much as possible correct information from it.Hope we shall 
  see more appropriate functions for microstructureanalysis in popular 
  Rietveld programs.Cheers,Leonid
  Dear Leonid,
  
  Indeed you missed something. I presume you 
  have the paper. Then, take a look to the formula (15a). This is the size 
  profile for lognormal. There is the function PHI - bar of argument 
  2*pi*s*R. Replace this function PHI - bar from (15a) by the 
  _expression_ (21a)with theargument x=2*pi*s*R. You get it? 
  So, not only "c" but also R.
  
  "We are Rietvelders" means that we must be 
  only "codes drivers", "cheffeurs des codes", "voditeli program"? Have we to 
  accept the "Procust bed" of the Rietveld codes at a given moment? All Rietveld 
  codes are improving in time, isn't it? 
  
  In particular for the Round_Robin sample 
  TCH-pV works because c=0.18. (Davor explained how Dv and Da are found). But if 
  c0.4 any pV fails.
  
  Best wishes,
  
  Nicolae
  


RE: Size Strain [ In GSAS ??? ]

2005-04-14 Thread Jim Cline
Hi all,
Well, I thought I'd weigh in on this with a discussion of an aforementioned 
SRM project:

We are in the final stages of preparing an SRM for determination of 
crystallite size from line profile analysis.  Through the course of his PhD 
work and NIST postdoctoral position, Nick Armstrong has developed a MaxEnt 
/ Bayesian method specifically for the certification.  The method can 
quantify, from the quality of the raw data, the probably that a proposed 
model for the crystallite size distribution is the true one.  Thus, the 
certified values of the standard will include a valid measure of their 
uncertainty that, in our humble opinion, would not be obtainable with 
alternative methods.  Details, and results of the method as applied to the 
RR CeO2, have been published.  We are working at the production of ~kg 
quantities of strain free CeO2 and ZnO for use as the SRM feedstock; no 
small challenge.

We expect two outcomes:  1) The community will have a standard by which 
results from mortal methods may be readily tested and compared.   2) A 
high-intensity squabble will ensue as to whether or not we got the right 
answer.

 With regards to the latter issue: Nick and I have been approached about 
another round robin.  Forgive me, but: round robins don't have anything to 
do with accuracy.  They test for uniformity of measurements in the field, 
the major premise being that, as a result of mature methodology, a narrow 
distribution is expected.  Note the highly successful, Rietveld, QPA, and 
instrument sensitivity round robins.  There is, however, no mature 
methodology here.  Indeed, only a small number of operations worldwide can 
perform a credible microstructure analysis, and their methods certainly 
differ.  It is our intention to make the data collected for the 
certification available to the community.  But with regards to model 
testing for the more advanced, physical model, methods (with the use of 
simulated, bi-model data for instance), I would suggest that more of a 
collaborative effort be organized.

Regards,
Jim
James P. Cline  [EMAIL PROTECTED]
Ceramics Division   Voice (301) 975 
5793
National Institute of Standards and Technology  FAX (301) 975 5334
100 Bureau Dr. stop 8520
Gaithersburg, MD 20899-8523USA



Re: Size Strain in GSAS

2005-04-14 Thread Nicolae Popa
Title: Message




Right, is rare, but we have meet once. A cerium 
oxide sample from a commercial company, c=2.8. I don't know if they did 
deliberately, probably not, otherwise the hard work to obtain such curiosity is 
costly and the company risks a bankruptcy. On the other hand superlorenzian 
profiles were reported from a long time, only were interpreted as coming from 
bimodal size distributions. And third, you see, people have difficulties to 
extract size distribution from the Rietveld codes as they areat this 
moment.

Nicolae Popa


  
  A 
  word from a "provider" of a Rietveld code (please don't call me a 
  "programmer"). 
  "But if c0.4 
  any pV fails" - OK, for what fraction of the universe of "real world" samples 
  is "c"0.4? I suspect, given the general success of the TCH pseudoVoigt 
  function, that it is exceedingly small and only occurs when one works hard to 
  deliberately make a sample like that.
  
  
  
  R.B. Von Dreele
  IPNS Division
  Argonne National Laboratory
  Argonne, IL 60439-4814
  
  

-Original Message-From: Nicolae Popa 
[mailto:[EMAIL PROTECTED] Sent: Thursday, April 14, 2005 7:14 
AMTo: [EMAIL PROTECTED]Cc: 
rietveld_l@ill.frSubject: Re: Size Strain in 
GSAS
Dear Nicolae, Maybe ya ploho chitayu i 
ploho soobrazhayu, but even after yourexplanation I can't see a way 
to calculate R from the results offitting described in 
chapters 6  7 of JAC 35 (2002) 338-346. From suchfitting you 
obtain only dispersion parameter c. Or I missed something?Anyway, 
being "Rietvelders" we still have to deal with TCH-pV functionand we 
need to extract as much as possible correct information from it.Hope 
we shall see more appropriate functions for microstructureanalysis 
in popular Rietveld 
programs.Cheers,Leonid
Dear Leonid,

Indeed you missed something. I presume 
you have the paper. Then, take a look to the formula (15a). This is the size 
profile for lognormal. There is the function PHI - bar of argument 
2*pi*s*R. Replace this function PHI - bar from (15a) by the 
_expression_ (21a)with theargument x=2*pi*s*R. You get it? 
So, not only "c" but also R.

"We are Rietvelders" means that we must 
be only "codes drivers", "cheffeurs des codes", "voditeli program"? Have we 
to accept the "Procust bed" of the Rietveld codes at a given moment? All 
Rietveld codes are improving in time, isn't it? 

In particular for the Round_Robin sample 
TCH-pV works because c=0.18. (Davor explained how Dv and Da are found). But 
if c0.4 any pV fails.

Best wishes,

Nicolae



RE: Size Strain in GSAS

2005-04-14 Thread Von Dreele, Robert B.
Title: Message



Nic,
I know 
about "super Lorentzians". Trouble is that many of those older reports were from 
Rietveld refinements "pre TCH" and used a formulation of the pseudo-Voigt which 
forced the FWHM of the Gaussian and Lorentzian components to be equal and 
allowed the mixing coefficient (eta) to be a free variable (n.b. it is not free 
in the TCH formulation). Thus, these ought to be discounted in any discussion 
about the occurence of super Lorentzian effects in real 
samples.
Bob



R.B. Von Dreele
IPNS Division
Argonne National Laboratory
Argonne, IL 60439-4814


  
  -Original Message-From: Nicolae Popa 
  [mailto:[EMAIL PROTECTED] Sent: Thursday, April 14, 2005 8:10 
  AMTo: rietveld_l@ill.frSubject: Re: Size Strain in 
  GSAS
  
  Right, is rare, but we have meet once. A cerium 
  oxide sample from a commercial company, c=2.8. I don't know if they did 
  deliberately, probably not, otherwise the hard work to obtain such curiosity 
  is costly and the company risks a bankruptcy. On the other hand superlorenzian 
  profiles were reported from a long time, only were interpreted as coming from 
  bimodal size distributions. And third, you see, people have difficulties to 
  extract size distribution from the Rietveld codes as they areat this 
  moment.
  
  Nicolae Popa
  
  

A 
word from a "provider" of a Rietveld code (please don't call me a 
"programmer"). 
"But if c0.4 
any pV fails" - OK, for what fraction of the universe of "real world" 
samples is "c"0.4? I suspect, given the general success of the TCH 
pseudoVoigt function, that it is exceedingly small and only occurs when one 
works hard to deliberately make a sample like that.



R.B. Von Dreele
IPNS Division
Argonne National Laboratory
Argonne, IL 60439-4814


  
  -Original Message-From: Nicolae Popa 
  [mailto:[EMAIL PROTECTED] Sent: Thursday, April 14, 2005 7:14 
  AMTo: [EMAIL PROTECTED]Cc: 
  rietveld_l@ill.frSubject: Re: Size Strain in 
  GSAS
  Dear Nicolae, Maybe ya ploho chitayu 
  i ploho soobrazhayu, but even after yourexplanation I can't see a 
  way to calculate R from the results offitting described in 
  chapters 6  7 of JAC 35 (2002) 338-346. From suchfitting you 
  obtain only dispersion parameter c. Or I missed something?Anyway, 
  being "Rietvelders" we still have to deal with TCH-pV functionand 
  we need to extract as much as possible correct information from 
  it.Hope we shall see more appropriate functions for 
  microstructureanalysis in popular Rietveld 
  programs.Cheers,Leonid
  Dear Leonid,
  
  Indeed you missed something. I presume 
  you have the paper. Then, take a look to the formula (15a). This is the 
  size profile for lognormal. There is the function PHI - bar of argument 
  2*pi*s*R. Replace this function PHI - bar from (15a) by the 
  _expression_ (21a)with theargument x=2*pi*s*R. You get 
  it? So, not only "c" but also R.
  
  "We are Rietvelders" means that we must 
  be only "codes drivers", "cheffeurs des codes", "voditeli program"? Have 
  we to accept the "Procust bed" of the Rietveld codes at a given moment? 
  All Rietveld codes are improving in time, isn't it? 
  
  In particular for the Round_Robin 
  sample TCH-pV works because c=0.18. (Davor explained how Dv and Da are 
  found). But if c0.4 any pV fails.
  
  Best wishes,
  
  Nicolae
  


Re: Size Strain in GSAS

2005-04-14 Thread Nicolae Popa
Title: Message



Dear Bob,

If I understand well, you say that eta1 (super 
Lorenzian) appeared only because eta was free parameter, but if TCH is used 
super Loreanzians do not occur?
Nevertheless, for that curious sample of cerium 
oxide wetried GSAS (with TCH) and the fit was very bad.
Best wishes,
Nicolae

PS. By the way, TCH also forces FWHM of the 
Gaussian and Lorenzian components to be equal, but indeed, eta is not free and 
cannot be greater than 1.


  
  Nic,
  I 
  know about "super Lorentzians". Trouble is that many of those older reports 
  were from Rietveld refinements "pre TCH" and used a formulation of the 
  pseudo-Voigt which forced the FWHM of the Gaussian and Lorentzian components 
  to be equal and allowed the mixing coefficient (eta) to be a free variable 
  (n.b. it is not free in the TCH formulation). Thus, these ought to be 
  discounted in any discussion about the occurence of super Lorentzian effects 
  in real samples.
  Bob
  
  
  
  R.B. Von Dreele
  IPNS Division
  Argonne National Laboratory
  Argonne, IL 60439-4814
  
  

-Original Message-From: Nicolae Popa 
[mailto:[EMAIL PROTECTED] Sent: Thursday, April 14, 2005 8:10 
AMTo: rietveld_l@ill.frSubject: Re: Size Strain in 
GSAS

Right, is rare, but we have meet once. A cerium 
oxide sample from a commercial company, c=2.8. I don't know if they did 
deliberately, probably not, otherwise the hard work to obtain such curiosity 
is costly and the company risks a bankruptcy. On the other hand 
superlorenzian profiles were reported from a long time, only were 
interpreted as coming from bimodal size distributions. And third, you see, 
people have difficulties to extract size distribution from the Rietveld 
codes as they areat this moment.

Nicolae Popa


  
  A word from a "provider" of a Rietveld code (please don't call me a 
  "programmer"). 
  "But if 
  c0.4 any pV fails" - OK, for what fraction of the universe of "real 
  world" samples is "c"0.4? I suspect, given the general success of the 
  TCH pseudoVoigt function, that it is exceedingly small and only occurs 
  when one works hard to deliberately make a sample like 
  that.
  
  
  
  R.B. Von Dreele
  IPNS Division
  Argonne National Laboratory
  Argonne, IL 60439-4814
  
  

-Original Message-From: Nicolae 
Popa [mailto:[EMAIL PROTECTED] Sent: Thursday, April 14, 2005 
7:14 AMTo: [EMAIL PROTECTED]Cc: 
rietveld_l@ill.frSubject: Re: Size Strain in 
GSAS
Dear Nicolae, Maybe ya ploho 
chitayu i ploho soobrazhayu, but even after yourexplanation I 
can't see a way to calculate R from the results 
offitting described in chapters 6  7 of JAC 35 (2002) 
338-346. From suchfitting you obtain only dispersion parameter 
c. Or I missed something?Anyway, being "Rietvelders" we still 
have to deal with TCH-pV functionand we need to extract as much 
as possible correct information from it.Hope we shall see more 
appropriate functions for microstructureanalysis in popular 
Rietveld programs.Cheers,Leonid
Dear Leonid,

Indeed you missed something. I 
presume you have the paper. Then, take a look to the formula (15a). This 
is the size profile for lognormal. There is the function PHI - bar of 
argument 2*pi*s*R. Replace this function PHI - bar from 
(15a) by the _expression_ (21a)with theargument 
x=2*pi*s*R. You get it? So, not only "c" but also 
R.

"We are Rietvelders" means that we 
must be only "codes drivers", "cheffeurs des codes", "voditeli program"? 
Have we to accept the "Procust bed" of the Rietveld codes at a given 
moment? All Rietveld codes are improving in time, isn't it? 


In particular for the Round_Robin 
sample TCH-pV works because c=0.18. (Davor explained how Dv and Da are 
found). But if c0.4 any pV fails.

Best wishes,

Nicolae



RE: Size Strain in GSAS

2005-04-14 Thread Von Dreele, Robert B.
Title: Message



Nic,
This 
is true for the internal math but the TCH function was assembled to reproduce 
the true Voigt over the entire range of differing Lorentz and Gauss FWHM values 
so it works as if the two FWHM components are independent. As for your question, 
I'm not aware that anyone has actually tried to do the fit both ways on a "super 
Lorentzian" (eta1 for old psVoigt) sample to see if a) the fit is the same 
and b) the eta1 was an artifact. Any takers to settle 
this?
Bob



R.B. Von Dreele
IPNS Division
Argonne National Laboratory
Argonne, IL 60439-4814


  
  -Original Message-From: Nicolae Popa 
  [mailto:[EMAIL PROTECTED] Sent: Thursday, April 14, 2005 9:11 
  AMTo: rietveld_l@ill.frSubject: Re: Size Strain in 
  GSAS
  Dear Bob,
  
  If I understand well, you say that eta1 
  (super Lorenzian) appeared only because eta was free parameter, but if TCH is 
  used super Loreanzians do not occur?
  Nevertheless, for that curious sample of cerium 
  oxide wetried GSAS (with TCH) and the fit was very bad.
  Best wishes,
  Nicolae
  
  PS. By the way, TCH also forces FWHM of the 
  Gaussian and Lorenzian components to be equal, but indeed, eta is not free and 
  cannot be greater than 1.
  
  

Nic,
I 
know about "super Lorentzians". Trouble is that many of those older reports 
were from Rietveld refinements "pre TCH" and used a formulation of the 
pseudo-Voigt which forced the FWHM of the Gaussian and Lorentzian components 
to be equal and allowed the mixing coefficient (eta) to be a free variable 
(n.b. it is not free in the TCH formulation). Thus, these ought to be 
discounted in any discussion about the occurence of super Lorentzian effects 
in real samples.
Bob



R.B. Von Dreele
IPNS Division
Argonne National Laboratory
Argonne, IL 60439-4814


  
  -Original Message-From: Nicolae Popa 
  [mailto:[EMAIL PROTECTED] Sent: Thursday, April 14, 2005 8:10 
  AMTo: rietveld_l@ill.frSubject: Re: Size Strain in 
  GSAS
  
  Right, is rare, but we have meet once. A 
  cerium oxide sample from a commercial company, c=2.8. I don't know if they 
  did deliberately, probably not, otherwise the hard work to obtain such 
  curiosity is costly and the company risks a bankruptcy. On the other hand 
  superlorenzian profiles were reported from a long time, only were 
  interpreted as coming from bimodal size distributions. And third, you see, 
  people have difficulties to extract size distribution from the Rietveld 
  codes as they areat this moment.
  
  Nicolae Popa
  
  

A word from a "provider" of a Rietveld code (please don't call me 
a "programmer"). 
"But if 
c0.4 any pV fails" - OK, for what fraction of the universe of "real 
world" samples is "c"0.4? I suspect, given the general success of 
the TCH pseudoVoigt function, that it is exceedingly small and only 
occurs when one works hard to deliberately make a sample like 
that.



R.B. Von Dreele
IPNS Division
Argonne National Laboratory
Argonne, IL 60439-4814


  
  -Original Message-From: Nicolae 
  Popa [mailto:[EMAIL PROTECTED] Sent: Thursday, April 14, 
  2005 7:14 AMTo: [EMAIL PROTECTED]Cc: 
  rietveld_l@ill.frSubject: Re: Size Strain in 
  GSAS
  Dear Nicolae, Maybe ya ploho 
  chitayu i ploho soobrazhayu, but even after yourexplanation I 
  can't see a way to calculate R from the results 
  offitting described in chapters 6  7 of JAC 35 (2002) 
  338-346. From suchfitting you obtain only dispersion parameter 
  c. Or I missed something?Anyway, being "Rietvelders" we still 
  have to deal with TCH-pV functionand we need to extract as 
  much as possible correct information from it.Hope we shall see 
  more appropriate functions for microstructureanalysis in 
  popular Rietveld 
  programs.Cheers,Leonid
  Dear Leonid,
  
  Indeed you missed something. I 
  presume you have the paper. Then, take a look to the formula (15a). 
  This is the size profile for lognormal. There is the function PHI - 
  bar of argument 2*pi*s*R. Replace this function PHI - 
  bar from (15a) by the _expression_ (21a)with theargument 
  x=2*pi*s*R. You get it? So, not only "c" but also 
  R.
  
  "We are Rietvelders" means that we 
  must be only "codes drivers", "cheffeurs des codes", "voditeli 
  program"? Have we to accept the "Procust bed" of the Rietveld code

Re: Size Strain in GSAS

2005-04-13 Thread Leonid Solovyov
 8. The simple modified TCH model (triple-Voigt), used in most major
 Rietveld programs these days, is surprisingly flexible. It works well
 for most of the samples (super-Lorentzian is an example when it
 fails, as well as many others, but this is less frequent that
 onewould expect) and gives some numbers for coherent domain size
 and strain. If we are lucky to know more about the sample (for
 instance, the information is available that a lognormal size
 distribution, certain type of dislocations, etc., is most likely to
 be prevalent for majority of grains in 
 the sample), those numbers will let us calculate real numbers that
 relate to the real physical parameters (say, the first moment and
 dispersion of the size distribution, etc.) in many cases, as
 discussed here previously.

Good conclusion, but before deriving real numbers that relate to the
real physical parameters one needs first to calibrate the
pseudo-Voigt-based calculation of those numbers (Dv and Da, or
L_volume and L_area in other notations)using at least simulated
profiles for VARIOUS dispersions.
I hope that the round robin on simulated data will be translated into
reality soon.

Leonid


__
Do You Yahoo!?
Tired of spam?  Yahoo! Mail has the best spam protection around 
http://mail.yahoo.com 


RE: Size Strain in GSAS

2005-04-13 Thread Davor Balzar
It was shown in paragraph 6 of JAC 35 (2002) 338-346 that size-broadened
profiles given by both lognormal and gamma distributions can be approximated
by a weighted sum of Lorentz and Gauss functions for a broad range of
distribution dispersions. Besides, round robins can sometimes be long
adventures...

Davor 

 -Original Message-
 From: Leonid Solovyov [mailto:[EMAIL PROTECTED] 
 Sent: Wednesday, April 13, 2005 12:11 AM
 To: rietveld_l@ill.fr
 Subject: Re: Size Strain in GSAS
 
  8. The simple modified TCH model (triple-Voigt), used in 
 most major
  Rietveld programs these days, is surprisingly flexible. It 
 works well
  for most of the samples (super-Lorentzian is an example when it
  fails, as well as many others, but this is less frequent that
  onewould expect) and gives some numbers for coherent domain size
  and strain. If we are lucky to know more about the sample (for
  instance, the information is available that a lognormal size
  distribution, certain type of dislocations, etc., is most likely to
  be prevalent for majority of grains in 
  the sample), those numbers will let us calculate real numbers that
  relate to the real physical parameters (say, the first moment and
  dispersion of the size distribution, etc.) in many cases, as
  discussed here previously.
 
 Good conclusion, but before deriving real numbers that relate to the
 real physical parameters one needs first to calibrate the
 pseudo-Voigt-based calculation of those numbers (Dv and Da, or
 L_volume and L_area in other notations)using at least simulated
 profiles for VARIOUS dispersions.
 I hope that the round robin on simulated data will be translated into
 reality soon.
 
 Leonid
 
 
 __
 Do You Yahoo!?
 Tired of spam?  Yahoo! Mail has the best spam protection around 
 http://mail.yahoo.com 
 



Re: Size Strain in GSAS

2005-04-12 Thread Davor Balzar
I guess, this discussion has already died down but I couldn't find a moment for 
reply soon enough:-)

As Prague was already mentioned, let me try to summarize what I think about 
this subject and have said there (let's hope I actually remember it:-):
1. A careful line broadening analysis (at this point in time) is better done 
outside Rietveld refinement
2. A physical model is better and preferred to a phenomenological model for 
analyzing line broadening

However, because we discuss the Size-Strain analysis in Rietveld here:
3. Rietveld obviously needs some kind of line-broadening modeling in order to 
at least correct for sample broadening effects (especially anisotropic ones) to 
extract correct integrated intensities for crystal-structure refinement. Thus, 
any model that works is good.
4. Rietveld needs to have a line-broadening model that works for an arbitrary 
crystal structure (up to triclinic) and arbitrary sample (i.e. many possible 
sources of broadening could be present in a given sample). Therefore, a 
phenomenological model is the only one available at this point, as physical 
models are still struggling with cubic (or hexagonal) structures and a very 
limited spectrum of physical sources causing broadening.

In conclusion:
5. I think that the work done by Nick Armstrong and others is definitely a way 
to go, but also a long way to go before we get to the level mentioned under 4 
(I certainly won't live to see it:-).
6. I also believe that (even when 5 is fulfilled) diffraction will often need 
some additional information provided by complementary characterization methods 
(i.e. TEM, SEM,...) to completely and accurately characterize defects in a 
sample, as we may calculate the most probable solution but won't often be able 
to discriminate between other very likely solutions, that is, the most probable 
is very often not significantly different from other physically plausible 
solutions (lognormal and gamma examples already mentioned).
7. Previous point implies that trying to do too much with only diffraction 
data might actually be dangerous. One can find too many dead-wrong numbers in 
the literature using some of the physical models (for instance, dislocation 
densities, etc.), as a real physical cause of broadening was probably different 
and/or there was a strong correlation between refinable parameters that depend 
on the diffraction angle in a similar way.

Considering the above:
8. The simple modified TCH model (triple-Voigt), used in most major Rietveld 
programs these days, is surprisingly flexible. It works well for most of the 
samples (super-Lorentzian is an example when it fails, as well as many 
others, but this is less frequent that one would expect) and gives some 
numbers for coherent domain size and strain. If we are lucky to know more 
about the sample (for instance, the information is available that a lognormal 
size distribution, certain type of dislocations, etc., is most likely to be 
prevalent for majority of grains in the sample), those numbers will let us 
calculate real numbers that relate to the real physical parameters (say, the 
first moment and dispersion of the size distribution, etc.) in many cases, as 
discussed here previously.

Davor
P.S:
9. The fact that a certain physical model does not yield a particular 
analytical function as a physically broadened profile does not mean that the 
function cannot successfully approximate that profile, as any such calculation 
includes many approximations of different kinds. There were numerous examples 
in literature showing that a simple Voigt function was able to approximate 
quite different cases. Of course, that is not true in general.


 -Original Message-
 From: Matteo Leoni [mailto:[EMAIL PROTECTED] 
 Sent: Tuesday, March 29, 2005 4:59 AM
 To: rietveld_l@ill.fr
 Subject: RE: Size Strain In GSAS
 
 Leonid (and others)
 
 just my 2 cents to the whole story (as this is a long 
 standing point of  
 discussion: Davor correct me if I'm wrong, but this was also 
 one of the 
 key points in the latest size-strain meeting in Prague, right?)
 
  Your recipe for estimating size distribution from the 
 parameters of a
  Voight-fitted profile is clear and straightforward, but I 
 wonder have
  you, or someone else, tested it on, say, simulated data for 
 the model
  of spherical crystallites having lognormal size distribution with
  various dispersions?
 
 done several times... if you start from a pattern synthesised from a 
 lognormal and you analyse it using a post-mortem LPA method 
 (i.e. extract 
 a width and a shape parameter and play with them to get some 
 microstructural information), you obtain a result which (in 
 most cases) 
 does not allow you to reconstruct the original data (the Fourier 
 transform of a Voigt and that of the function describing a lognormal 
 distribution of spherical domains are different).
 I would invite all people using ANY traditional line 
 profile analysis  
 method to do

Re: Size Strain in GSAS

2005-04-12 Thread Miguel Hesiquio-Garduño
Hi,
maybe I'm late in the discussion, but what about if we use a Rietveld ( or
whole pattern fitting) refinement in order to extract data for the profile
and use it to make the extraction of size and strain effects?
 thanks and greetings



Miguel Hesiquio-Garduño
Profesor Asociado C
Departamento de Ciencia de Materiales
Academia de Ciencias de la Ingeniería
ESFM-IPN
 I guess, this discussion has already died down but I couldn't find a
 moment for reply soon enough:-)

 As Prague was already mentioned, let me try to summarize what I think
 about this subject and have said there (let's hope I actually remember
 it:-):
 1. A careful line broadening analysis (at this point in time) is better
 done outside Rietveld refinement
 2. A physical model is better and preferred to a phenomenological model
 for analyzing line broadening

 However, because we discuss the Size-Strain analysis in Rietveld here:
 3. Rietveld obviously needs some kind of line-broadening modeling in order
 to at least correct for sample broadening effects (especially anisotropic
 ones) to extract correct integrated intensities for crystal-structure
 refinement. Thus, any model that works is good.
 4. Rietveld needs to have a line-broadening model that works for an
 arbitrary crystal structure (up to triclinic) and arbitrary sample (i.e.
 many possible sources of broadening could be present in a given sample).
 Therefore, a phenomenological model is the only one available at this
 point, as physical models are still struggling with cubic (or hexagonal)
 structures and a very limited spectrum of physical sources causing
 broadening.

 In conclusion:
 5. I think that the work done by Nick Armstrong and others is definitely a
 way to go, but also a long way to go before we get to the level mentioned
 under 4 (I certainly won't live to see it:-).
 6. I also believe that (even when 5 is fulfilled) diffraction will often
 need some additional information provided by complementary
 characterization methods (i.e. TEM, SEM,...) to completely and accurately
 characterize defects in a sample, as we may calculate the most probable
 solution but won't often be able to discriminate between other very likely
 solutions, that is, the most probable is very often not significantly
 different from other physically plausible solutions (lognormal and gamma
 examples already mentioned).
 7. Previous point implies that trying to do too much with only
 diffraction data might actually be dangerous. One can find too many
 dead-wrong numbers in the literature using some of the physical models
 (for instance, dislocation densities, etc.), as a real physical cause of
 broadening was probably different and/or there was a strong correlation
 between refinable parameters that depend on the diffraction angle in a
 similar way.

 Considering the above:
 8. The simple modified TCH model (triple-Voigt), used in most major
 Rietveld programs these days, is surprisingly flexible. It works well for
 most of the samples (super-Lorentzian is an example when it fails, as
 well as many others, but this is less frequent that one would expect) and
 gives some numbers for coherent domain size and strain. If we are lucky
 to know more about the sample (for instance, the information is available
 that a lognormal size distribution, certain type of dislocations, etc., is
 most likely to be prevalent for majority of grains in the sample), those
 numbers will let us calculate real numbers that relate to the real
 physical parameters (say, the first moment and dispersion of the size
 distribution, etc.) in many cases, as discussed here previously.

 Davor
 P.S:
 9. The fact that a certain physical model does not yield a particular
 analytical function as a physically broadened profile does not mean that
 the function cannot successfully approximate that profile, as any such
 calculation includes many approximations of different kinds. There were
 numerous examples in literature showing that a simple Voigt function was
 able to approximate quite different cases. Of course, that is not true in
 general.


 -Original Message-
 From: Matteo Leoni [mailto:[EMAIL PROTECTED]
 Sent: Tuesday, March 29, 2005 4:59 AM
 To: rietveld_l@ill.fr
 Subject: RE: Size Strain In GSAS

 Leonid (and others)

 just my 2 cents to the whole story (as this is a long
 standing point of
 discussion: Davor correct me if I'm wrong, but this was also
 one of the
 key points in the latest size-strain meeting in Prague, right?)

  Your recipe for estimating size distribution from the
 parameters of a
  Voight-fitted profile is clear and straightforward, but I
 wonder have
  you, or someone else, tested it on, say, simulated data for
 the model
  of spherical crystallites having lognormal size distribution with
  various dispersions?

 done several times... if you start from a pattern synthesised from a
 lognormal and you analyse it using a post-mortem LPA method
 (i.e. extract
 a width and a shape parameter and play

Re: Size Strain In GSAS

2005-04-07 Thread Nicolae Popa
Hi,
Long text but not fully convincing. At least concerning my questions (still
posted at the bottom). I'm risking a hurry reply without reading all
references (including to be published and PhD Thesis).
See comments below.


 that likelihood term is described by a goodness of fit, say chi-square
 function, which improves as the models become more complex or increase
 in the number of parameters. The Ockham's Razor term, on the other hand,
 penalizes a model for the number of parameters by including the a priori
 distribution and uncertainties in parameter values. Hence this term
 offsets the influence the likelihood term may have, thereby arriving
 at a choice of model where the number of parameters can be justified. In
 this case it is possible to use a uniform prior

Both, lognormal and gamma are physically based distributions, both have the
same numbers of parameters,
both give the same restored profile inside the noise, the same chi square
(and Rw-gamma even slightly better), the prior distributions are the same
(uniform?)
uncertainties in parameter values are comparable, what Razor term penalizes?


 space (see below). The solution with the greatest entropy relative to a
 priori model and experimental data is the solution with the least
 assumptions or the solution with the most amount of randomness.
 Solutions with a lower entropy are solutions where specific assumptions
 have been made which can not be justified. For example, we apply this
 method to determining the modal properties of a size distribution i.e.
 monomodal or bimodal size distributions. Say if we assume, distribution
 has a bimodal features, when in reality it is a monodal distribution,
 this assumption will results in a solution with a lower entropy. The
 same is also case for the converse problem (see [3]). Moreover, we can
 use this method to select between different distribution models such as
 lognormal or gamma distributions.

It means that for gamma distribution you found the entropy was lower
and then this model is not justified. Significantly lower?  Probably you
have a measure concerning the significance of the difference between the
entropies of the two models? On the other hand, why lognormal is the
solution with the least assumptions?

 What we are trying to do with the full Bayesian/MaxEnt method (see
 [1])is determine a free form solution or a non-parametric solution
 [5],f, where the  solutions is either a line profile or a distribution
 determine from the experimental data and knowledge of the instrumental,
 noise and background effects. By free formor non-parametric solution
 I mean a profile and/or distributions which does not assume a specific
 set of parameters, as defined by say a lognormal distribution or a Voigt
 line profile function. The a priori model, m, can be defined by a

And, certainly, the free form solution has the highest entropy, anyway
higher than the initial guess (lognormal). This is the optimal solution, if
I understood correctly. I wonder if this optimal solution from the point of
view of MaxEnt. is not one from the following solutions: w*Logn+(1-w)*Gamm.
To these solutions (of infinite multiplicity) the peak profiles are
indifferent.



 Dear All,
 I'm sorry for the delay in relying. I also want to pass on my thanks to
 Jim Cline for pointing out that wasn't around to response to some of the
 queries/issues. It has been interesting reading the discussion, since
 coming back to Sydney. I don't mean to add more fuel to the fire, but I
 do hope to outline/address some of the issues which have been raised,
 while  giving a more precise outline of the Bayesian/maximum entropy
 (MaxEnt) method as applied to line profile analysis.

 By way of background information and reference, the most recent
 publications which present the theory and application of the
 Bayesian/MaxEnt to analyzing size/shape broadened (simulated and
 experimental) data are given in [1-4]-- see below. None of this work
 would be possible without the core collaborators which include: Jim
 Cline (NIST), Walter Kalceff (UTS), Annette Dowd (UTS) and John Bonevich
 (NIST) in various combinations.

 In summary, [1] gives a full  and mathematical derivation of the
 Bayesian/MaxEnt method and is applied to simulated  and experimental
 data. In [2], the application of the full Bayesian/Maxent and Markov
 Chain Monte Carlo (MCMC) methods to size analysis. Ref.[3] shows how
 Bayesian/MaxEnt/MCMC methods can be applied to distinguish monomodal and
 bimodal size distributions i.e Bayesian model selection. Ref [4] is
 another application of Bayesian model selection  to distinguish between
 lognormal and gamma distributions which also includes full TEM
 data/analysis and also demonstrates how the method is sensitive to shape
 and microstrain effects in the line profile data.

 Let me first address  Nicolae's queries. The application of the
 Bayesian/MCMC method in [2] was simply to demonstrate how the method
 could be used to explore the 

RE: Size Strain In GSAS

2005-03-30 Thread Matteo Leoni
Leonid,

 Could you, please, give a reference to a study where Dv and Da sizes
 were derived from the parameters of pseudo-Voight or Voight fitted to
 simulated profiles for various size distribution dispersions?

I did something better (I hope).. at the end of the mesg you find xy 
data with a simulation:
Ceria (Fm-3m), CuKa 0.15406 nm (delta function, i.e. no emission 
profile aberration)), size broadening due to a lognormal distribution of  
spheres only, no background, no noise, no Lorentz-Polarization, no 
aberrations of any kind. 
Peaks present in the pattern:
111
200
220
311
222
400
331
420
422
333/511 fitted but not used in the analysis as they have same d. 

Simulation done using WPPM, or, simply, taking the FT of the formulae 
proposed in Acta Cryst (2002) A58, 190-200 for the lognorm.
The program used for the simulation allows fully recovery of the original 
parameters, starting from different initial values (self consistent 
check..).

Analysis done using traditional peak fitting Williamson-Hall and 
Warren-Averbach methods, applying the formulae found e.g. on Phil Mag 
(1998) A77 [3], 621-640 to obtain the lognorm from the Da and Dv 
values (should the ref be unavailable, they can be easily calculated, or  
I can also provide the formulae). 

I can send the simulation parameters and all plots/calculations I did to 
the interested members (just drop me a line). Results of WH and WA seems 
good but the distributions they provide are completely out with respect  
to the true one (unless I did some mistake, but that can be easily 
checked).
I could have attached the results file here, but I'm sure Alan (as list 
moderator) would haven't been quite happy about attachments.

Happy calculation, for those who wants to do the analysis by themselves 
without knowing the result in advance!

Mat


-
Matteo Leoni, PhD
Department of Materials Engineering
and Industrial Technologies 
University of Trento
38050 Mesiano (TN)
ITALY


SIMULATED DATA in xy (2theta Intensity) format.

1.80e+001   4.470421e+000
1.806000e+001   4.513074e+000
1.812000e+001   4.556395e+000
1.818000e+001   4.600401e+000
1.824000e+001   4.645108e+000
1.83e+001   4.690529e+000
1.836000e+001   4.736682e+000
1.842000e+001   4.783659e+000
1.848000e+001   4.831332e+000
1.854000e+001   4.879787e+000
1.86e+001   4.929045e+000
1.866000e+001   4.979123e+000
1.872000e+001   5.030047e+000
1.878000e+001   5.081826e+000
1.884000e+001   5.134484e+000
1.89e+001   5.188043e+000
1.896000e+001   5.242526e+000
1.902000e+001   5.297955e+000
1.908000e+001   5.354352e+000
1.914000e+001   5.411743e+000
1.92e+001   5.470151e+000
1.926000e+001   5.529602e+000
1.932000e+001   5.590123e+000
1.938000e+001   5.651740e+000
1.944000e+001   5.714484e+000
1.95e+001   5.778379e+000
1.956000e+001   5.843577e+000
1.962000e+001   5.909879e+000
1.968000e+001   5.977428e+000
1.974000e+001   6.046257e+000
1.98e+001   6.116400e+000
1.986000e+001   6.187892e+000
1.992000e+001   6.260770e+000
1.998000e+001   6.335082e+000
2.004000e+001   6.410849e+000
2.01e+001   6.488118e+000
2.016000e+001   6.566933e+000
2.022000e+001   6.647338e+000
2.028000e+001   6.729377e+000
2.034000e+001   6.813098e+000
2.04e+001   6.898548e+000
2.046000e+001   6.985779e+000
2.052000e+001   7.074842e+000
2.058000e+001   7.165793e+000
2.064000e+001   7.258687e+000
2.07e+001   7.353585e+000
2.076000e+001   7.450544e+000
2.082000e+001   7.549629e+000
2.088000e+001   7.650906e+000
2.094000e+001   7.754684e+000
2.10e+001   7.860571e+000
2.106000e+001   7.968865e+000
2.112000e+001   8.079643e+000
2.118000e+001   8.192986e+000
2.124000e+001   8.308978e+000
2.13e+001   8.427707e+000
2.136000e+001   8.549279e+000
2.142000e+001   8.673760e+000
2.148000e+001   8.801265e+000
2.154000e+001   8.931897e+000
2.16e+001   9.065765e+000
2.166000e+001   9.202984e+000
2.172000e+001   9.343672e+000
2.178000e+001   9.487954e+000
2.184000e+001   9.635962e+000
2.19e+001   9.787833e+000
2.196000e+001   9.943709e+000
2.202000e+001   1.010374e+001
2.208000e+001   1.026809e+001
2.214000e+001   1.043692e+001
2.22e+001   1.061041e+001
2.226000e+001   1.078874e+001
2.232000e+001   1.097211e+001
2.238000e+001   1.116071e+001
2.244000e+001   1.135477e+001
2.25e+001   1.155511e+001
2.256000e+001   1.176082e+001
2.262000e+001   1.197270e+001
2.268000e+001   1.219102e+001
2.274000e+001   1.241607e+001
2.28e+001   1.264814e+001
2.286000e+001   1.288756e+001
2.292000e+001   1.313466e+001
2.298000e+001   1.338978e+001
2.304000e+001   1.365333e+001
2.31e+001   1.392571e+001
2.316000e+001   1.420732e+001
2.322000e+001   1.449861e+001
2.328000e+001   1.480009e+001
2.334000e+001   1.511227e+001
2.34e+001   1.543569e+001
2.346000e+001   1.577095e+001
2.352000e+001   1.611867e+001
2.358000e+001   1.647952e+001
2.364000e+001   1.685423e+001
2.37e+001   1.724355e+001
2.376000e+001   1.764832e+001
2.382000e+001   1.806942e+001

RE: Size Strain In GSAS

2005-03-30 Thread Leonid Solovyov
Dear Matteo,

Thanks for the exercise.
From pseudo-Voight fitting I have got Dv=33A, Da=23A,
which gives the average size D=21A and the relative 
dispersion c=0.28 (c = [sigmaD/D]^2).
However, I suspect that the actual values you used for the simulation
were D~30A and c~0.25.
Do I win the F1 GP? :-)

Actually, I have also played with simulated profiles for various
dispersions and I am going to calibrate the calculations of Dv and Da
from pseudo-Voight. Hope I will include these calibrated calculations
to the next release of DDM.

Cheers,
Leonid

P.S. As for a Ferrari powered by a John Deere tractor engine, I believe
that none of us would win a F1 GP even with an original Ferrari engine.
I have better recipe: take a Russian tank T-72 (which is simpler and
cheaper than Ferrari), put it at the pole-position, turn the gun back
and enjoy driving - F1 GP is guaranteed.





__ 
Do you Yahoo!? 
Yahoo! Small Business - Try our new resources site!
http://smallbusiness.yahoo.com/resources/ 


RE: Size Strain In GSAS

2005-03-30 Thread apu
Dear Matteo,
Thanks for the problem.
I have used pseudo voigt function to fit the peaks and finally used the program 
BREADTH and obtained Dv=31 A, Da=18 A.

Please send me your simulation parameters, plots/calculations.

Regards,
Apu


/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/
Apu Sarkar
Research Fellow
Variable Energy Cyclotron Centre
Kolkata 700 064
phone: 91-33-2337-1230 (extn. 3190)
Fax:   91-33-2334-6871 
INDIA
/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/


- Original Message -
From: Matteo Leoni [EMAIL PROTECTED]
Date: Wednesday, March 30, 2005 4:39 pm

---BeginMessage---
Leonid,

 Could you, please, give a reference to a study where Dv and Da sizes
 were derived from the parameters of pseudo-Voight or Voight fitted to
 simulated profiles for various size distribution dispersions?

I did something better (I hope).. at the end of the mesg you find xy 
data with a simulation:
Ceria (Fm-3m), CuKa 0.15406 nm (delta function, i.e. no emission 
profile aberration)), size broadening due to a lognormal distribution of  
spheres only, no background, no noise, no Lorentz-Polarization, no 
aberrations of any kind. 
Peaks present in the pattern:
111
200
220
311
222
400
331
420
422
333/511 fitted but not used in the analysis as they have same d. 

Simulation done using WPPM, or, simply, taking the FT of the formulae 
proposed in Acta Cryst (2002) A58, 190-200 for the lognorm.
The program used for the simulation allows fully recovery of the original 
parameters, starting from different initial values (self consistent 
check..).

Analysis done using traditional peak fitting Williamson-Hall and 
Warren-Averbach methods, applying the formulae found e.g. on Phil Mag 
(1998) A77 [3], 621-640 to obtain the lognorm from the Da and Dv 
values (should the ref be unavailable, they can be easily calculated, or  
I can also provide the formulae). 

I can send the simulation parameters and all plots/calculations I did to 
the interested members (just drop me a line). Results of WH and WA seems 
good but the distributions they provide are completely out with respect  
to the true one (unless I did some mistake, but that can be easily 
checked).
I could have attached the results file here, but I'm sure Alan (as list 
moderator) would haven't been quite happy about attachments.

Happy calculation, for those who wants to do the analysis by themselves 
without knowing the result in advance!

Mat


-
Matteo Leoni, PhD
Department of Materials Engineering
and Industrial Technologies 
University of Trento
38050 Mesiano (TN)
ITALY


SIMULATED DATA in xy (2theta Intensity) format.

1.80e+001   4.470421e+000
1.806000e+001   4.513074e+000
1.812000e+001   4.556395e+000
1.818000e+001   4.600401e+000
1.824000e+001   4.645108e+000
1.83e+001   4.690529e+000
1.836000e+001   4.736682e+000
1.842000e+001   4.783659e+000
1.848000e+001   4.831332e+000
1.854000e+001   4.879787e+000
1.86e+001   4.929045e+000
1.866000e+001   4.979123e+000
1.872000e+001   5.030047e+000
1.878000e+001   5.081826e+000
1.884000e+001   5.134484e+000
1.89e+001   5.188043e+000
1.896000e+001   5.242526e+000
1.902000e+001   5.297955e+000
1.908000e+001   5.354352e+000
1.914000e+001   5.411743e+000
1.92e+001   5.470151e+000
1.926000e+001   5.529602e+000
1.932000e+001   5.590123e+000
1.938000e+001   5.651740e+000
1.944000e+001   5.714484e+000
1.95e+001   5.778379e+000
1.956000e+001   5.843577e+000
1.962000e+001   5.909879e+000
1.968000e+001   5.977428e+000
1.974000e+001   6.046257e+000
1.98e+001   6.116400e+000
1.986000e+001   6.187892e+000
1.992000e+001   6.260770e+000
1.998000e+001   6.335082e+000
2.004000e+001   6.410849e+000
2.01e+001   6.488118e+000
2.016000e+001   6.566933e+000
2.022000e+001   6.647338e+000
2.028000e+001   6.729377e+000
2.034000e+001   6.813098e+000
2.04e+001   6.898548e+000
2.046000e+001   6.985779e+000
2.052000e+001   7.074842e+000
2.058000e+001   7.165793e+000
2.064000e+001   7.258687e+000
2.07e+001   7.353585e+000
2.076000e+001   7.450544e+000
2.082000e+001   7.549629e+000
2.088000e+001   7.650906e+000
2.094000e+001   7.754684e+000
2.10e+001   7.860571e+000
2.106000e+001   7.968865e+000
2.112000e+001   8.079643e+000
2.118000e+001   8.192986e+000
2.124000e+001   8.308978e+000
2.13e+001   8.427707e+000
2.136000e+001   8.549279e+000
2.142000e+001   8.673760e+000
2.148000e+001   8.801265e+000
2.154000e+001   8.931897e+000
2.16e+001   9.065765e+000
2.166000e+001   9.202984e+000
2.172000e+001   9.343672e+000
2.178000e+001   9.487954e+000
2.184000e+001   9.635962e+000
2.19e+001   9.787833e+000
2.196000e+001   9.943709e+000
2.202000e+001   1.010374e+001
2.208000e+001   1.026809e+001
2.214000e+001   1.043692e+001
2.22e+001   1.061041e+001
2.226000e+001   1.078874e+001
2.232000e+001   1.097211e+001
2.238000e+001   1.116071e+001
2.244000e+001   1.135477e+001
2.25e+001   1.155511e+001
2.256000e+001   1.176082e+001
2.262000e+001   1.197270e+001

RE: Size Strain In GSAS

2005-03-29 Thread Matteo Leoni
Leonid (and others)

just my 2 cents to the whole story (as this is a long standing point of  
discussion: Davor correct me if I'm wrong, but this was also one of the 
key points in the latest size-strain meeting in Prague, right?)

 Your recipe for estimating size distribution from the parameters of a
 Voight-fitted profile is clear and straightforward, but I wonder have
 you, or someone else, tested it on, say, simulated data for the model
 of spherical crystallites having lognormal size distribution with
 various dispersions?

done several times... if you start from a pattern synthesised from a 
lognormal and you analyse it using a post-mortem LPA method (i.e. extract 
a width and a shape parameter and play with them to get some 
microstructural information), you obtain a result which (in most cases) 
does not allow you to reconstruct the original data (the Fourier 
transform of a Voigt and that of the function describing a lognormal 
distribution of spherical domains are different).
I would invite all people using ANY traditional line profile analysis  
method to do always this check. Davor already pointed out cases where it 
works and cases where it does not: according to my experience those 
belonging to the first category are just a few.

With a whole pattern approach and working directly with the profile 
arising from a distribution of domains, in most cases you're able to 
recostruct the original distribution without making any assumption on its  
functional shape (after all, most of the information to do so is 
contained in the whole pattern, even if it is well hidden).

Concerning the Beyesian/maxent method, well, it is always a great idea, 
but unfortunately right now it is not mature enough to cope with a simple 
problem of combined instrumental, size AND strain broadening (unless 
something has been done in the last year). So ok it gives you the best 
result compatible with your hypotheses, but beware that absence of any 
other source of broadening should be listed among them.. and I'm not sure 
this is always the case!

To put some water on the fire (otherwise it will burn all of us), I think 
the level of detail one needs on the microstructure, conditions the 
methods one's going to use to extract a result. No need to use highly 
sophisticated methods to roughly estimate a domain size (with an error up 
to +/- 50%) or to establish a trend within a homogeneous set of data, or 
also to obtain a better fit in the Rietveld method.

Conversely, if a very high level of detail is sought, then I'd forget 
about a traditional Rietveld refinement and start approaching the 
problem from the microstructure point of view (after all, if one is 
interested in winning a F1 GP, he'd certainly not go for a Ferrari  
powered by a John Deere tractor engine!).

cheers

Mat

-
Matteo Leoni, PhD
Department of Materials Engineering
and Industrial Technologies 
University of Trento
38050 Mesiano (TN)
ITALY



RE: Size Strain In GSAS

2005-03-29 Thread Leonid Solovyov

done several times...

With a whole pattern approach and working directly with the profile 
arising from a distribution of domains, in most cases you're able to 
recostruct the original distribution without making any assumption on 
its functional shape (after all, most of the information to do so is 
contained in the whole pattern, even if it is well hidden).

Dear Matteo (and others),

Could you, please, give a reference to a study where Dv and Da sizes
were derived from the parameters of pseudo-Voight or Voight fitted to
simulated profiles for various size distribution dispersions?

Thanks in advance,
Leonid




__ 
Do you Yahoo!? 
Yahoo! Small Business - Try our new resources site!
http://smallbusiness.yahoo.com/resources/ 


Re: Size Strain In GSAS

2005-03-28 Thread Nicolae Popa
Hi,
So, to resume your statements, by using Bayesian/Max.Entr. we can
distinguish between two distributions that can not be distinguished by
maximum likelihood (least square)?  Hard to swallow, once the restored peak
profiles are the same inside the noise. What other information than the
peak profile, instrumental profile and statistical noise we have that
Bayes/Max.ent. can use and the least square cannot?

prior distributions to be uniform - if I understand correctly you refer to
the distributions of  D0 and sigma of the lognormal (gamma) distribution
from which the least square chooses the solution, not to the distribution
itself (logn, gamm). Then, how is this prior distribution for Baye/Max.ent.?

Best,
Nick Popa


 Hi
 Sorry for the delay. The Bayesian results showed that the lognormal was
more probable. Yes, the problem is ill-condition which why you need to use
the Bayesian/Maximum entropy method. This method takes into account the
ill-conditioning of the problem. The idea being it determines the most
probable solutions from the set of solutions.  This solution can be shown to
be the most consistent solution or the solution with the least assumptions
given the experimental data, noise, instrument effects etc (see Skilling 
Bryan 1983; Skilling 1990; Sivia 1996). This is the role of entropy
function. There are many mathemaitcal proofs for this (see Jaynes' recent
book). The Bayesian analysis maps out the solution/model spaces.

 Also the least squares solution is simple a special case of a class of
deconvolution problems. This s well established result. It is not the least
ill-posed, since it assumes the prior distributions to be uniform (in a
Bayesian case. See Sivia and reference therein). In fact it's likely to be
the worst solution since it assumes a most ignorant state knowledge (ie.
uniform proir) and doesn't always take into consideration the surrounding
information. Moreover, it doesn't account for the underlying
physics/mathematics, that the probability distributions/line profiles are
positive  additive distributions (Skilling 1990; Sivia 1996).

 Best wishes, Nick


  Dr Nicholas Armstrong


  Hi, once again,
  Fine, I'm sure you did. And what is the most plausible, lognormal
  or gamma?
  From the tests specific for least square (pattern fitting) they are
  equallyplausible. And take a combination of the type  w*Log+(1-
  w)*Gam, that will be
  equally plausible.
  On the other hand, why should believe that the Baesian
  deconvolution (or any
  other sophisticated deconvolution method that can imagine) give the
  answermuch precisely? Both, the least square and deconvolution are
  ill-posed
  problems, but the least square is less ill-posed than the
  deconvolution. At
  least that say the  manuals for statistical mathematics.
 
  Best wishes,
  Nicolae Popa
 
 
 
 
 
   Hi,
   I pointed out that each model needs to be tested and their
  plausibilitydetermined.  This can be achieved by employing Bayesian
  analysis, which
  takes into account the diffraction data and underlying physics.
  
   I have carried out exactly same analysis for the round robin CeO2
  samplefor both size distributions using lognormal and gamma
  distributionfunctions, and similarly for dislocations: screw, edge
  and mixed. The
  plausibility of each model was quantified using Bayesian analysis,
  where the
  probability of each model was determined, from which the model with
  thegreatest probability was selected. This approach takes into
  account the
  assumptions of each model, parameters, uncertainties,  instrumental
  andnoise effects etc. See Sivia (1996)Data Analysis: A Bayesian
  Tutorial(Oxford Science Publications).
  
   Best wishes,
   Nick
  
Dr Nicholas Armstrong
 
  
 
  
Hi,
But the diffraction alone cannot  determine  uniquely the physical
model. An
example at hand: the CeO2 pattern from round-robin can be
  equally well
described by two different size distributions, lognormal and gamma
and by
any linear combinations of these two distributions. Is the
  situation  different with the strain profile caused by different
  types of
dislocations,possible mixed?
   
Best wishes,
Nicolae Popa
   
   
   
 Best approach is to develop physical models for the line profile
broadening and test them for their plausibility i.e. model
  selection.  
 Good luck.

 Best Regards, Nick


  Dr Nicholas Armstrong
   
   
   
  
  
   -- 
   UTS CRICOS Provider Code:  00099F
   DISCLAIMER: This email message and any accompanying attachments may
  contain
   confidential information.  If you are not the intended recipient,
  do not
   read, use, disseminate, distribute or copy this message or
  attachments.If
   you have received this message in error, please notify the sender
  immediately
   and delete this message. Any views expressed in this message are
  those of
  the
   individual sender, 

RE: Size Strain In GSAS

2005-03-28 Thread Aaron Celestian
Give me 10 min.





-Original Message-
From: Davor Balzar [mailto:[EMAIL PROTECTED] 
Sent: Monday, March 28, 2005 4:45 PM
To: rietveld_l@ill.fr

Dear Leonid:

Yes, but on real data: Journal of Applied Crystallography 35 (2002) 338-346
(the paper is available from the Web site below). Two examples are reported;
in the second, the Voigt approximation failed due to an extremely broad size
distribution (or bimodal distribution). However, Voigt size-broadened
profile should work for most real-world samples. The paper shows that either
pseudo-Voigt or Voigt are good approximations for both (moderately broad)
lognormal and (any physical) gamma size distribution of spherical
crystallites.

Davor


Davor Balzar
Department of Physics  Astronomy
University of Denver
2112 E Wesley Ave
Denver, CO 80208
Phone: 303-871-2137
Fax: 303-871-4405
Web: www.du.edu/~balzar



 -Original Message-
 From: Leonid Solovyov [mailto:[EMAIL PROTECTED] 
 Sent: Sunday, March 27, 2005 12:49 AM
 To: rietveld_l@ill.fr
 Subject: RE: Size Strain In GSAS
 
 On Friday 03/25 Davor Balzar wrote:
  Paragraph 3.3 of the article that you mentioned explains how were 
 size and strain values calculated. One can even obtain size 
  distribution by following the procedure that was posted to this 
  mailing list several months ago; 
 
 Dear Davor,
 
 Your recipe for estimating size distribution from the parameters of a
 Voight-fitted profile is clear and straightforward, but I wonder have
 you, or someone else, tested it on, say, simulated data for the model
 of spherical crystallites having lognormal size distribution with
 various dispersions?
 
 Best regards,
 Leonid
 
 
 
   
 __ 
 Do you Yahoo!? 
 Make Yahoo! your home page 
 http://www.yahoo.com/r/hs
 




Re: Size Strain In GSAS

2005-03-28 Thread Jim Cline
Hi,
Nick Armstrong has advised me he will in non-email-land for a week or so.
I'm sure he'll resume this discussion when he returns...
Jim
At 03:45 PM 3/28/2005 +0400, you wrote:
Hi,
So, to resume your statements, by using Bayesian/Max.Entr. we can
distinguish between two distributions that can not be distinguished by
maximum likelihood (least square)?  Hard to swallow, once the restored peak
profiles are the same inside the noise. What other information than the
peak profile, instrumental profile and statistical noise we have that
Bayes/Max.ent. can use and the least square cannot?
prior distributions to be uniform - if I understand correctly you refer to
the distributions of  D0 and sigma of the lognormal (gamma) distribution
from which the least square chooses the solution, not to the distribution
itself (logn, gamm). Then, how is this prior distribution for Baye/Max.ent.?
Best,
Nick Popa
 Hi
 Sorry for the delay. The Bayesian results showed that the lognormal was
more probable. Yes, the problem is ill-condition which why you need to use
the Bayesian/Maximum entropy method. This method takes into account the
ill-conditioning of the problem. The idea being it determines the most
probable solutions from the set of solutions.  This solution can be shown to
be the most consistent solution or the solution with the least assumptions
given the experimental data, noise, instrument effects etc (see Skilling 
Bryan 1983; Skilling 1990; Sivia 1996). This is the role of entropy
function. There are many mathemaitcal proofs for this (see Jaynes' recent
book). The Bayesian analysis maps out the solution/model spaces.

 Also the least squares solution is simple a special case of a class of
deconvolution problems. This s well established result. It is not the least
ill-posed, since it assumes the prior distributions to be uniform (in a
Bayesian case. See Sivia and reference therein). In fact it's likely to be
the worst solution since it assumes a most ignorant state knowledge (ie.
uniform proir) and doesn't always take into consideration the surrounding
information. Moreover, it doesn't account for the underlying
physics/mathematics, that the probability distributions/line profiles are
positive  additive distributions (Skilling 1990; Sivia 1996).

 Best wishes, Nick


  Dr Nicholas Armstrong
snip
James P. Cline  [EMAIL PROTECTED]
Ceramics Division   Voice (301) 975 
5793
National Institute of Standards and Technology  FAX (301) 975 5334
100 Bureau Dr. stop 8520
Gaithersburg, MD 20899-8523USA



RE: Size Strain In GSAS

2005-03-26 Thread Nicholas Armstrong
Dear Apu  All
Firstly, GSAS wasn't designed for line profile analysis.

More importantly, the line profiles resulting from the from nanocrystallites 
and dislocations, generally do not have the functional form described by 
functions, such as Voigt. I will also go as far as to say that there is no 
physical basis for these line profile functions for quantifying the 
microstructure of a sample, in terms of shape/size distribution of 
crystallites, and spatial distributions/type/density of dislocations. 

For example, the line profile arising from a lognormal distribution of 
spherical crystallites doesn't have the form of Voigt or Lorentzian or Gaussian 
line profile functions (i.e. see Scardi  Leoni (2001), Acta Cryst., A52, 
605-613.). Moreover, while Krivoglaz  Ryaboshapka (1963) (Fiz. metal., 
metalloved., 15(1),18-31) showed that Gaussian line profiles can arise from a 
crystallite containing screw dislocations, it resulted in the strain energy 
diverging  as the crystallite increased. This was only resolved by Wilkens 
(1970a,b,c) and Krivoglaz et al. (1983). These two latter cases produced 
general expressions for the Fourier coefficients/line profile which depended on 
the characteristics/density of the dislocations.

Best approach is to develop physical models for the line profile broadening and 
test them for their plausibility i.e. model selection. 

Good luck.

Best Regards, Nick


 Dr Nicholas Armstrong
NIST-UTS Research Fellow
***
(in Australia) UTS, Department of Applied Physics
***
University of Technology,Sydney* Location: Bld 1,Level 12,Rm1217
P.O Box 123* Ph:(+61-2) 9514-2203 
Broadway NSW 2007  * Fax:   (+61-2) 9514-2219 
Australia  * E-mail:[EMAIL PROTECTED]
***
(in USA) NIST, Ceramics Division
***
National Institute of Standards and Technology * Fax: (+1-301) 975-5334
100 Bureau Dr. stop 8520   *
Gaithersburg, MD 20899-8523USA *
***


- Original Message -
From: Davor Balzar [EMAIL PROTECTED]
Date: Friday, March 25, 2005 7:05 pm

 Hi Apu:
 
 As everybody pointed out, there are better ways (for now) to do the
 size/strain analysis, but GSAS can also be used if observed, size-
 broadenedand strain-broadened profiles can all be approximated with 
 Voigt functions. 
 
 Paragraph 3.3 of the article that you mentioned explains how were 
 size and
 strain values calculated. One can even obtain size distribution by 
 followingthe procedure that was posted to this mailing list several 
 months ago; see
 below.
 
 Best wishes,
 
 Davor
 
 
 Davor Balzar
 Department of Physics  Astronomy
 University of Denver
 2112 E Wesley Ave
 Denver, CO 80208
 Phone: 303-871-2137
 Fax: 303-871-4405
 Web: www.du.edu/~balzar
 
 
 National Institute of Standards and Technology (NIST)
 Division 853
 Boulder, CO 80305
 Phone: 303-497-3006
 Fax: 303-497-5030
 Web: www.boulder.nist.gov/div853/balzar
 
 
 
 
  -Original Message-
  From: Davor Balzar [EMAIL PROTECTED] 
  Sent: Monday, November 22, 2004 3:58 PM
  To: rietveld_l@ill.fr
  Subject: RE: Size distribution from Rietveld refinement
  
  Yes, one can determine size distribution parameters by using 
 Rietveld refinement. In particular, the lognormal size 
 distribution is 
  defined by two
  parameters (say, the average radius and the distribution 
  dispersion, see,
  for instance, (2) and (3) of JAC 37 (2004) 911, SSRR for 
  short here, or
  other references therein). It was first shown by Krill  
  Birringer that both
  volume-weighted (Dv) and area-weighted (Da) domain size (that 
  are normally
  evaluated in a diffraction experiment) can be related to the 
  average radius
  and dispersion of the lognormal distribution; one obtains 
  something like (5)
  in the paper SSRR. Therefore, if one can evaluate both Dv and 
  Da by Rietveld
  refinement, it would be possible to determine the parameters 
  of the size
  distribution, as two independent parameters are required to 
 define the
  lognormal or similar types of bell-shaped distributions. Note 
  here that a
  different distribution can be used, which will change the 
 relationship between Dv  Da and the parameters of the 
 distribution (for the gamma
  distribution, see JAC 35 (2002) 338, for the equations 
  equivalent to (5) in
  SSRR). The value that is normally evaluated through the 
  Rietveld

Re: Size Strain In GSAS

2005-03-25 Thread Luca Lutterotti
Dear Apu,
I know I will start up a good debate here, but size-strain analysis 
with GSAS is a non-sense. The program was not written with that purpose 
in mind and in fact it does not contains the instrumental aberration 
part of the broadening that is necessary for such computation.
Indeed it is possible to get at end some size-strain data, but quite 
hard as you have to do all correction later and out of the program. So 
it is like using GSAS for peak fitting, so better to use a peak fitting 
dedicated program.

Best wishes,
Luca Lutterotti
On Mar 25, 2005, at 7:15, [EMAIL PROTECTED] wrote:
Dear All,
I am trying to perform Rietveld refinement on a very simple system 
using GSAS. I have obtained a reasonable fit except the peak widths.

 I want to use the size and strain refinement option in GSAS to make 
the fit well.

Please tell me how to use the SIZE STRAIN refinement option in GSAS.
P.S. I am using the EXPGUI.
Thanks in advace.
Regards,
Apu
/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/
 Apu Sarkar
Research Fellow
Variable Energy Cyclotron Centre
Kolkata 700 064
phone: 91-33-2337-1230 (extn. 3190)
Fax: 91-33-2334-6871
INDIA
/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/



Re: Size Strain In GSAS

2005-03-25 Thread apu
Dear Prof. Lutterotti,
 I was also aware of the fact that GSAS is not made for Size Strain analysis. I 
got interested to use the Size strain refinement feature of GSAS only after 
going through the article :
Size-strain line broadening analysis of the ceria round-robin sample by Prof. 
D. Balzar et. al. Journal of Applied Crys. 37(2004)911-924.

In that round robin results they have reported the size strain obtained from 
GSAS.

I my case also when I am trying with GSAS, the diffraction pattern is fitting 
well except the peak braodening. I think this brodening is due to small domain 
size effect. I that case how will I obatin a good fit with GSAS.

Thanking you.

Best Regards,
Apu







/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/
Apu Sarkar
Research Fellow
Variable Energy Cyclotron Centre
Kolkata 700 064
phone: 91-33-2337-1230 (extn. 3190)
Fax:   91-33-2334-6871 
INDIA
/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/


- Original Message -
From: Luca Lutterotti [EMAIL PROTECTED]
Date: Friday, March 25, 2005 3:31 pm

 Dear Apu,
 
 I know I will start up a good debate here, but size-strain 
 analysis 
 with GSAS is a non-sense. The program was not written with that 
 purpose 
 in mind and in fact it does not contains the instrumental 
 aberration 
 part of the broadening that is necessary for such computation.
 Indeed it is possible to get at end some size-strain data, but 
 quite 
 hard as you have to do all correction later and out of the 
 program. So 
 it is like using GSAS for peak fitting, so better to use a peak 
 fitting 
 dedicated program.
 
   Best wishes,
   Luca Lutterotti
 
 On Mar 25, 2005, at 7:15, [EMAIL PROTECTED] wrote:
 
  Dear All,
  I am trying to perform Rietveld refinement on a very simple 
 system 
  using GSAS. I have obtained a reasonable fit except the peak widths.
 
   I want to use the size and strain refinement option in GSAS to 
 make 
  the fit well.
 
  Please tell me how to use the SIZE STRAIN refinement option in GSAS.
 
  P.S. I am using the EXPGUI.
 
 
  Thanks in advace.
 
  Regards,
 
  Apu
 
 
  /_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/
   Apu Sarkar
  Research Fellow
  Variable Energy Cyclotron Centre
  Kolkata 700 064
  phone: 91-33-2337-1230 (extn. 3190)
  Fax: 91-33-2334-6871
  INDIA
  /_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/
 
 
 




Re: Size Strain In GSAS

2005-03-25 Thread Andreas Leineweber
Dear all,



I think the statement that one cannot do line-profile analysis using GSAS is
too strong. In principle it is possible to do some 

size strain analysis using GSAS, if the instrumental profile is e.g.
sufficiently described previously

by the Thompson-Cox-Hastings (TCH) profile function (includes measuring
corresponding data on a suitable standard). I think, even involvement of the
Finger asymmetry correction does not introduce systematic errors. Then the
increase in the tantheta

and 1/costheta related Gaussian and Lorentzian line width components of the TCH
description upon Rietveld refinement on the basis of diffraction data
exhibiting physical line broadening can 

in principle be associated with microstrain- and size-related quantities.

This can also be extended by involving anisotropic size and microstrain models.
 



Thus, on this level of line broadening analysis, everything neccessary is
contained in GSAS. Of course, something like microstrain and size distributions
cannot be obtained using GSAS.



Of course there are problems:

0. Microstrain broadening must be proportional to tan(theta). This is not
neccessarily the case. There mustn't be further line broadening contributions
like stacking faults, complicating the situation.

1. If both size and strain contributions are present, one most be aware of the
correlation between the tantheta and 1/costheta dependent compontents.

2. One has to be aware to which average values of the size and microstrain
distributions the increases of the line width parameters can be associated
with.

This requires for GSAS a close analysis of the GSAS manual (how are the
line-width paramerters defined!) and line broadening literature, how such
pseudo-Voigt line width parameters can be related with averages microstructure
parameters.



Thus it is not made easy for the user to extract something like that from the
line width parameters. But perhaps it is better that way, because consequently
the user is forced to to deal himself with the required theory, rather than
just refining a parameter called size and one called microstrain, believing in
the results and publish the values 



Definitely there are much better procedures to analyse size and microstrain
than by GSAS.



So, going to the problem of Apu: If the instrumental profile is well described
before using TCH, refinement of (only) LX (and perhaps P) gives you quantities
whiuch you should be able to relate to size related quantities upon reading the
GSAS manual and some line broadening literature.



Best regards

Andreas Leineweber





 Dear Prof. Lutterotti,

  I was also aware of the fact that GSAS is not made for Size Strain analysis.
I got interested to use the Size strain refinement feature of GSAS only after
going through the article :

 Size-strain line broadening analysis of the ceria round-robin sample by
Prof. D. Balzar et. al. Journal of Applied Crys. 37(2004)911-924.

 

 In that round robin results they have reported the size strain obtained from
GSAS.

 

 I my case also when I am trying with GSAS, the diffraction pattern is fitting
well except the peak braodening. I think this brodening is due to small domain
size effect. I that case how will I obatin a good fit with GSAS.

 

 Thanking you.

 

 Best Regards,

 Apu

 

 

 

 

 

 

 

 /_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/

 Apu Sarkar

 Research Fellow

 Variable Energy Cyclotron Centre

 Kolkata 700 064

 phone: 91-33-2337-1230 (extn. 3190)

 Fax:   91-33-2334-6871 

 INDIA

 /_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/

 

 

 - Original Message -

 From: Luca Lutterotti [EMAIL PROTECTED]

 Date: Friday, March 25, 2005 3:31 pm

 

  Dear Apu,

  

  I know I will start up a good debate here, but size-strain 

  analysis 

  with GSAS is a non-sense. The program was not written with that 

  purpose 

  in mind and in fact it does not contains the instrumental 

  aberration 

  part of the broadening that is necessary for such computation.

  Indeed it is possible to get at end some size-strain data, but 

  quite 

  hard as you have to do all correction later and out of the 

  program. So 

  it is like using GSAS for peak fitting, so better to use a peak 

  fitting 

  dedicated program.

  

  Best wishes,

  Luca Lutterotti

  

  On Mar 25, 2005, at 7:15, [EMAIL PROTECTED] wrote:

  

   Dear All,

   I am trying to perform Rietveld refinement on a very simple 

  system 

   using GSAS. I have obtained a reasonable fit except the peak widths.

  

I want to use the size and strain refinement option in GSAS to 

  make 

   the fit well.

  

   Please tell me how to use the SIZE STRAIN refinement option in GSAS.

  

   P.S. I am using the EXPGUI.

  

  

   Thanks in advace.

  

   Regards,

  

   Apu

  

  

   /_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/

Apu Sarkar

   Research Fellow

   Variable Energy Cyclotron Centre

   Kolkata 700 064

Re: Size Strain In GSAS

2005-03-25 Thread Luca Lutterotti
Dear Apu,
difficult to say without seeing the pattern with your actual fitting. 
Broadening with small domain size is normally more easy to fit. It 
could be you didn't use the proper function or refines all necessary 
parameters, or there is an anisotropic broadening or faulting.
Every sample/analysis has his own story.
Try to put a picture of the fitting somewhere on the web and post here 
a link to the picture as attachments are not permitted in the list. 
This would be better as everyone can see it and give you some 
suggestions.

Luca
On Mar 25, 2005, at 12:00, [EMAIL PROTECTED] wrote:
Dear Prof. Lutterotti,
 I was also aware of the fact that GSAS is not made for Size Strain 
analysis. I got interested to use the Size strain refinement feature 
of GSAS only after going through the article :
Size-strain line broadening analysis of the ceria round-robin sample 
by Prof. D. Balzar et. al. Journal of Applied Crys. 37(2004)911-924.

In that round robin results they have reported the size strain 
obtained from GSAS.

I my case also when I am trying with GSAS, the diffraction pattern is 
fitting well except the peak braodening. I think this brodening is due 
to small domain size effect. I that case how will I obatin a good fit 
with GSAS.

Thanking you.
Best Regards,
Apu



/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/
Apu Sarkar
Research Fellow
Variable Energy Cyclotron Centre
Kolkata 700 064
phone: 91-33-2337-1230 (extn. 3190)
Fax:   91-33-2334-6871
INDIA
/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/
- Original Message -
From: Luca Lutterotti [EMAIL PROTECTED]
Date: Friday, March 25, 2005 3:31 pm
Dear Apu,
I know I will start up a good debate here, but size-strain
analysis
with GSAS is a non-sense. The program was not written with that
purpose
in mind and in fact it does not contains the instrumental
aberration
part of the broadening that is necessary for such computation.
Indeed it is possible to get at end some size-strain data, but
quite
hard as you have to do all correction later and out of the
program. So
it is like using GSAS for peak fitting, so better to use a peak
fitting
dedicated program.
Best wishes,
Luca Lutterotti
On Mar 25, 2005, at 7:15, [EMAIL PROTECTED] wrote:
Dear All,
I am trying to perform Rietveld refinement on a very simple
system
using GSAS. I have obtained a reasonable fit except the peak widths.
 I want to use the size and strain refinement option in GSAS to
make
the fit well.
Please tell me how to use the SIZE STRAIN refinement option in GSAS.
P.S. I am using the EXPGUI.
Thanks in advace.
Regards,
Apu
/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/
 Apu Sarkar
Research Fellow
Variable Energy Cyclotron Centre
Kolkata 700 064
phone: 91-33-2337-1230 (extn. 3190)
Fax: 91-33-2334-6871
INDIA
/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/






Re: Size Strain In GSAS

2005-03-25 Thread Jim Cline
Hi,
I wrote an article [ that appeared in Dean Smith's book ] some time back 
that describes how to use SRM 660a, LaB6, and the TCH function of GSAS for 
characterization of the IPF, and then refine the only the microstructure 
specific terms for an estimation of the size and strain in subsequent 
analyses.  The approach has its limitations for sure, but it is accessible 
with a small effort.

Regards,
Jim
At 02:10 PM 3/25/2005 +0100, you wrote:
Dear Andreas,
I didn't said it cannot be done. Only that was not made for and so it 
is not easy as using other tools. In principle every diffraction fitting 
program can be used for size-strain.
Few questions: have you ever tried to do such analysis with GSAS the right 
way using the instrumental profile correction?
Did you use only GSAS for that or you had to use other external 
tools/computations? (this to get a feeling about your statement: Thus, on 
this level of line broadening analysis, everything necessary is contained 
in GSAS; may be level should be clarified).
I would like to know what there is inside GSAS for crystallite size and 
microstrain analysis in particular.

Best regards,
Luca Lutterotti
On Mar 25, 2005, at 12:55, Andreas Leineweber wrote:
Dear all,

I think the statement that one cannot do line-profile analysis using GSAS is
too strong. In principle it is possible to do some
size strain analysis using GSAS, if the instrumental profile is e.g.
sufficiently described previously
by the Thompson-Cox-Hastings (TCH) profile function (includes measuring
corresponding data on a suitable standard). I think, even involvement of the
Finger asymmetry correction does not introduce systematic errors. Then the
increase in the tantheta
and 1/costheta related Gaussian and Lorentzian line width components of 
the TCH
description upon Rietveld refinement on the basis of diffraction data
exhibiting physical line broadening can

in principle be associated with microstrain- and size-related quantities.
This can also be extended by involving anisotropic size and microstrain 
models.


Thus, on this level of line broadening analysis, everything neccessary is
contained in GSAS. Of course, something like microstrain and size 
distributions
cannot be obtained using GSAS.


Of course there are problems:
0. Microstrain broadening must be proportional to tan(theta). This is not
neccessarily the case. There mustn't be further line broadening contributions
like stacking faults, complicating the situation.
1. If both size and strain contributions are present, one most be aware 
of the
correlation between the tantheta and 1/costheta dependent compontents.

2. One has to be aware to which average values of the size and microstrain
distributions the increases of the line width parameters can be associated
with.
This requires for GSAS a close analysis of the GSAS manual (how are the
line-width paramerters defined!) and line broadening literature, how such
pseudo-Voigt line width parameters can be related with averages 
microstructure
parameters.


Thus it is not made easy for the user to extract something like that from the
line width parameters. But perhaps it is better that way, because 
consequently
the user is forced to to deal himself with the required theory, rather than
just refining a parameter called size and one called microstrain, 
believing in
the results and publish the values


Definitely there are much better procedures to analyse size and microstrain
than by GSAS.

So, going to the problem of Apu: If the instrumental profile is well 
described
before using TCH, refinement of (only) LX (and perhaps P) gives you 
quantities
whiuch you should be able to relate to size related quantities upon 
reading the
GSAS manual and some line broadening literature.


Best regards
Andreas Leineweber


Dear Prof. Lutterotti,

 I was also aware of the fact that GSAS is not made for Size Strain 
analysis.
I got interested to use the Size strain refinement feature of GSAS only after
going through the article :
Size-strain line broadening analysis of the ceria round-robin sample by
Prof. D. Balzar et. al. Journal of Applied Crys. 37(2004)911-924.

In that round robin results they have reported the size strain obtained from
GSAS.

I my case also when I am trying with GSAS, the diffraction pattern is 
fitting
well except the peak braodening. I think this brodening is due to small 
domain
size effect. I that case how will I obatin a good fit with GSAS.


Thanking you.

Best Regards,

Apu




/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/

Apu Sarkar

Research Fellow

Variable Energy Cyclotron Centre

Kolkata 700 064

phone: 91-33-2337-1230 (extn. 3190)

Fax:   91-33-2334-6871

INDIA

/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/


- Original Message -

From: Luca Lutterotti [EMAIL PROTECTED]

Date: Friday, March 25, 2005 3:31 pm

Dear Apu,

I know I will start up a good debate here, but size-strain

analysis

with GSAS is a non-sense

RE: Size Strain In GSAS

2005-03-25 Thread Davor Balzar
Hi Apu:

As everybody pointed out, there are better ways (for now) to do the
size/strain analysis, but GSAS can also be used if observed, size-broadened
and strain-broadened profiles can all be approximated with Voigt functions. 

Paragraph 3.3 of the article that you mentioned explains how were size and
strain values calculated. One can even obtain size distribution by following
the procedure that was posted to this mailing list several months ago; see
below.

Best wishes,

Davor


Davor Balzar
Department of Physics  Astronomy
University of Denver
2112 E Wesley Ave
Denver, CO 80208
Phone: 303-871-2137
Fax: 303-871-4405
Web: www.du.edu/~balzar


National Institute of Standards and Technology (NIST)
Division 853
Boulder, CO 80305
Phone: 303-497-3006
Fax: 303-497-5030
Web: www.boulder.nist.gov/div853/balzar


 

 -Original Message-
 From: Davor Balzar [mailto:[EMAIL PROTECTED] 
 Sent: Monday, November 22, 2004 3:58 PM
 To: rietveld_l@ill.fr
 Subject: RE: Size distribution from Rietveld refinement
 
 Yes, one can determine size distribution parameters by using Rietveld
 refinement. In particular, the lognormal size distribution is 
 defined by two
 parameters (say, the average radius and the distribution 
 dispersion, see,
 for instance, (2) and (3) of JAC 37 (2004) 911, SSRR for 
 short here, or
 other references therein). It was first shown by Krill  
 Birringer that both
 volume-weighted (Dv) and area-weighted (Da) domain size (that 
 are normally
 evaluated in a diffraction experiment) can be related to the 
 average radius
 and dispersion of the lognormal distribution; one obtains 
 something like (5)
 in the paper SSRR. Therefore, if one can evaluate both Dv and 
 Da by Rietveld
 refinement, it would be possible to determine the parameters 
 of the size
 distribution, as two independent parameters are required to define the
 lognormal or similar types of bell-shaped distributions. Note 
 here that a
 different distribution can be used, which will change the relationship
 between Dv  Da and the parameters of the distribution (for the gamma
 distribution, see JAC 35 (2002) 338, for the equations 
 equivalent to (5) in
 SSRR). The value that is normally evaluated through the 
 Rietveld refinement
 is Dv, as the refinable parameters in the 
 Thompson-Cox-Hastings (TCH) model
 are based on the integral-breadth methods. This means that 
 one would have to
 use (9) and (15)-(18) in SSRR, to obtain Dv, which depends on 
 both P and X
 parameters. As the TCH model implicitly assumes Voigt 
 functions for both
 size and strain-broadened profiles (double-Voigt model), Da 
 can be also
 calculated, but from X only, as it depends only on the Lorentzian
 size-broadened integral breadth, Da=1/(2betaL) (this and 
 other consequences
 of a double-Voigt model were shown/discussed in JAC 26 (1993) 97).
 
 HOWEVER, as pointed out by others in previous messages, this 
 assumes that
 (i) Both observed and physically broadened profiles are Voigt 
 functions,
 which is implicit to the TCH model; (ii) Size distribution is 
 lognormal,
 gamma, or whatever we assume it to be. On the former, it is 
 easy to see if
 observed profiles can't be successfully fit 
 (super-Lorentzian peak shapes,
 for instance), which means that the TCH peak shape cannot be 
 used. However,
 an assumption that physically broadened profiles (size and 
 strain) are also
 Voigt function is more difficult to prove; if not and one 
 uses the equations
 described above, a systematic error will be introduced. On 
 the latter, a
 good fit in Rietveld means only that a lognormal or other assumed
 distribution is one POSSIBLE approximation of the real size 
 distribution in
 the sample. However, this equally applies to all the other parameters
 obtained through the Rietveld refinement and is not a special 
 deficiency of
 this model. Second, even if one obtains more information 
 about the actual
 size distribution via TEM, SEM, etc., sometimes it is very 
 difficult to
 discern between different bell-shaped size distributions, 
 especially if the
 size distribution is narrow.
 
 Davor


 -Original Message-
 From: [EMAIL PROTECTED] [mailto:[EMAIL PROTECTED] 
 Sent: Friday, March 25, 2005 4:01 AM
 To: rietveld_l@ill.fr
 Subject: Re: Size Strain In GSAS
 
 Dear Prof. Lutterotti,
  I was also aware of the fact that GSAS is not made for Size 
 Strain analysis. I got interested to use the Size strain 
 refinement feature of GSAS only after going through the article :
 Size-strain line broadening analysis of the ceria 
 round-robin sample by Prof. D. Balzar et. al. Journal of 
 Applied Crys. 37(2004)911-924.
 
 In that round robin results they have reported the size 
 strain obtained from GSAS.
 
 I my case also when I am trying with GSAS, the diffraction 