Re: [time-nuts] Raw phase data of super-5065

2018-06-05 Thread tim...@timeok.it


   Hi,

   certainly the environmental effects are significant and it is precisely for 
this reason that
   acquisitions have been made simultaneously on two 5065, one modified and the 
other not.
   To verify the results the same measures were taken with another pair of 
5065, one modified and the other not.
   This second test was carried out at another site 300Km from the first one.
   Both tests showed a better and more consistent medium/long term ADEV than 
the unmodified HP5065A.
   I want to clarify that these measures have the objective of verifying the 
medium / long term stability and not the short term stability of the Super 
version I agree it is improved..
   Luciano


   Da "Bob kb8tq" kb...@n1k.org
   A "Luciano Paramithiotti" tim...@timeok.it,"Discussion of precise time and 
frequency measurement" time-nuts@febo.com
   Cc
   Data Tue, 5 Jun 2018 10:05:34 -0400
   Oggetto Re: [time-nuts] Raw phase data of super-5065
   Hi

   It is pretty well known that without “help” ( beyond the filter stuff on the 
super) the 5065 has fairly poor temperature
   stability and somewhat variable response to pressure and humidity. Taking 
two samples right off the production line
   likely would have given you different long term data as a result.

   If you are going to the trouble of watching these devices for a long time, 
monitor temperature, pressure and humidity.
   I suspect you will find correlations ….

   Bob

   > On Jun 5, 2018, at 5:44 AM, tim...@timeok.it wrote:
   >
   >
   > Hi Attila,
   > I am doing the long therm stability test on a couple of HP5065, one 
original A type, the second B or "super" compared with my best long therm 
stability, a GPSDO HPz3816A.
   > This test will be 30 days long. It is a second test, with a second couple 
of HP5065A/B, to verify the long therm stability of both the version.
   >
   > Up to now the data available for version B or "super" only concerned the 
improvement of phase noise.
   >
   > From the first results it seems that the HP5065A is more stable within 
30days than the modified version with the filter.
   > It must be considered that the short-term stability of 5065 is better than 
the GPSDO / TICC up to 10-80k sec then, in this test configuration the valid 
data will be beyond these values.
   > The results will be ready within 28 days and I will be happy to share them 
in order to have everyone's opinion.
   > Regards,
   > Luciano
   >
   >
   > Da "time-nuts" time-nuts-boun...@febo.com
   > A "Discussion of precise time and frequency measurement" time-nuts@febo.com
   > Cc
   > Data Mon, 4 Jun 2018 10:11:50 +0200
   > Oggetto [time-nuts] Raw phase data of super-5065
   > Hoi zäme!
   >
   > Would one (preferably multiple) of the owners of a super-5065 be so kind
   > and share the raw phase data of an super-5065 against a stable reference
   > with me? The longer the data trace, the better.
   >
   > Thanks in advance!
   >
   > Attila Kinali
   >
   > --
   > It is upon moral qualities that a society is ultimately founded. All
   > the prosperity and technological sophistication in the world is of no
   > use without that foundation.
   > -- Miss Matheson, The Diamond Age, Neil Stephenson
   > ___
   > time-nuts mailing list -- time-nuts@febo.com
   > To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
   > and follow the instructions there.
   > ___
   > time-nuts mailing list -- time-nuts@febo.com
   > To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
   > and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] nuts about position

2018-06-05 Thread Poul-Henning Kamp

In message 
,
 Mark Sims writes:

>The effects of things like antenna position changing a couple of cm due to 
>expansion and humidity effects are swamped by the receiver accuracy and noise.

Short term: Yes.  Long term:  No.


-- 
Poul-Henning Kamp   | UNIX since Zilog Zeus 3.20
p...@freebsd.org | TCP/IP since RFC 956
FreeBSD committer   | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] nuts about position

2018-06-05 Thread Mark Sims
The typical receiver default self-surveys of 30 minutes to 2 hours are not 
ideal.  They will not include a lot of satellites or the effects of multi-path 
over time.   A survey of at least 12 hours is needed to include all the sats.   
 24-48 hours is even better.

The effects of things like antenna position changing a couple of cm due to 
expansion and humidity effects are swamped by the receiver accuracy and noise.

-

>  That said, there are certainly a lot of issues with how the self-survey
position should be calculated *and* maintained, because it really needs
to change over time for optimal performance.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] nuts about position

2018-06-05 Thread Poul-Henning Kamp

In message 
,
 Mark Sims writes:

>The next thing I want to try is a receiver self-survey vs the post processed 
>results...

I fell into that trap many years ago as well :-)

The post processed result eliminates and compensates for a lot of
effects which are invisible to the receiver, that is the entire
point of post processing.

But if your goal is to get good 1PPS out of your receiver, forcing
it to hold a "perfect" position, means asking it to deal with effects
it cannot measure, and as a result you get *worse* 1PPS performance
than if you use the "wrong" self-survey position.

That said, there are certainly a lot of issues with how the self-survey
position should be calculated *and* maintained, because it really needs
to change over time for optimal performance.

Almost any suboptimal antenna-position means that you should vary
the position-hold coordinates over the orbital period (but see
below), but over longer periods also environmental factors come
into play.

For instance if you put the antenna on a wood-construction, be it a
pole or a house, it coordinates will vary at the cm level over the
seasons as the wood expands and contracts.

The good news is that the receiver gives you the input data to work
and model these biases, the bad news you don't get a lot of data.

I did some experiments when ten Oncore M12+T's passed through my lab
many years ago:  I plotted the reported "residuals" vs.  satellite
position, and I tried to update the held position on a daily basis
using the trend, and the results were measurable at the several ns
level:

http://phk.freebsd.dk/raga/sneak/

I didn't persue this further, because my latitude is particular
unfavourable to the simple (N/S+E/W) approach (no birds north of
me) and with only a single receiver the amount of input data is too
low for such a crude algorithm.

A more complex geometrical model could probably get more mileage
out of the sparse data, as would tracking more birds than the
12 the M12 could track.

-- 
Poul-Henning Kamp   | UNIX since Zilog Zeus 3.20
p...@freebsd.org | TCP/IP since RFC 956
FreeBSD committer   | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] nuts about position

2018-06-05 Thread Mark Sims
The tripod is a survey grade tripod on a limestone terrace (in a horrible 
location for an antenna).

Heather's precision survey uses the receiver's reported position data.   It 
does not take advantage of carrier phase/pseduorange/doppler data and 
post-processing.   Unlike the simple averaging self-surveys that most receivers 
do it processes the average of medians over 1 minute/1 hour/24 hour intervals.  
 It typically produces results 1-2m better than the simple self-surveys.   The 
next thing I want to try is a receiver self-survey vs the post processed 
results...  I may be able to kludge up a test that does a self-survey / 
precision survey / post processed survey and compare the results of the methods 
on the same data set.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] nuts about position

2018-06-05 Thread Bob kb8tq
Hi

Hmmm …… that’s a very big difference between the RINEX and the “precision 
survey”. 

How do each compare to other receivers on the same antenna ( and super duper 
stable mount) ?
Even with a tripod on the grass, meter level variations likely are not the 
tripod’s fault. :)

Bob

> On Jun 5, 2018, at 1:39 PM, Mark Sims  wrote:
> 
> I did Lady Heather's 48-hour precision survey on an NVS-08 receiver and 
> collected RINEX data at the same time.  The NVS was tracking GPS and SBAS 
> satellites.
> 
> The RINEX result had lat/lon/alt error estimates of .175/.153/.396 meters.
> 
> The difference between Heather's precision survey results and the RINEX 
> position were:
> lat/lon/alt:  -.638m/1.190m/1.526m
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] nuts about position

2018-06-05 Thread Mark Sims
I did Lady Heather's 48-hour precision survey on an NVS-08 receiver and 
collected RINEX data at the same time.  The NVS was tracking GPS and SBAS 
satellites.

The RINEX result had lat/lon/alt error estimates of .175/.153/.396 meters.

The difference between Heather's precision survey results and the RINEX 
position were:
lat/lon/alt:  -.638m/1.190m/1.526m
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] Raw phase data of super-5065

2018-06-05 Thread cdelect
Attila,

I will send you some raw data.

Attached is a combined plot of a recent successful filter installation.

Only 20 to 25% of units will be improved with the filter mod.

Two things to look for.

1- The short term stability at 1 Sec (Allan Deviation) should drop from
the 2X10-12th range down to the mid -13ths.

2- The Allan Deviation plot should drop into the -14ths between 32 and 64
Seconds.

#1 is easily tested but not many have a reference to test #2 against.

Also as Bob indicated longer term plots need to be in a benign
environmental setting.

Cheers,

Corby___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Possible new run of TNS-BUF High Isolation, Low Noise Buffer Amp boards

2018-06-05 Thread John Ackermann N8UR
Good news -- we had enough orders to justify doing a small production 
run of the TNS-BUF isolation amplifiers.  We'll be placing the order 
with our contract manufacturer this week, and expect delivery in six to 
eight weeks.


There will be a few extra boards available, so if you haven't ordered 
yet, there's still time.


John


On 05/08/2018 03:09 PM, John Ackermann N8UR wrote:
Back in October, 2016, TAPR did a build-to-order offering of the TNS-BUF 
low noise, high isolation buffer amplifier.  The original 2016 post is 
attached below.  All of those units sold out, and recently I've had a 
few questions about whether more will be available.


TAPR will do another one-off production run if we can get orders for at 
least 25 boards (the minimum to make manufacturing viable) by *June 1*. 
You can place an order by going to http://tapr.org/kits_tns-buf.  The 
price will be the same as the previous production -- $119 plus shipping.


If we receive orders for at least 25 boards by June 1, we will charge 
credit cards and place the production order with our contract 
manufacturer.  If we don't get 25 orders, we'll cancel the project and 
your card will not be charged.


We expect to ship finished TNS-BUF boards about 60 days after placing 
the production order.


Best regards,
John

 Forwarded Message 
Subject: [time-nuts] TNS-BUF High Isolation, Low Noise Buffer Amp Available
Date: Wed, 5 Oct 2016 10:15:46 -0400
From: John Ackermann N8UR 

I've previously mentioned a high performance buffer amplifier called the 
"TNS-BUF" that I built based on a design by Dr. Bruce Griffiths with 
further input from John Miles. Key numbers are:


*  Phase noise -140dBc/Hz at 1 Hertz offset, noise floor -175dBc/Hz.
     (PN plot attached)
*  Reverse isolation greater than 100dB; low enough that I can't make a
     trustworthy measurement.
*  Gain from -10 to +7 dB from 1 to 30 MHz; maximum output >18dBm.
*  Nominal 18VDC operation, but works down to 12V with lowered
     maximum output level.

There's information, including performance data and schematic, at
http://www.febo.com/pages/TNS-BUF

There seems to be some interest in an amplifier like this, so TAPR has 
decided to do a limited production run.  The amp is built with surface 
mount parts, so we thought an assembled and tested board was better than 
a kit.  The price will be $119 each.  But we have no idea how much 
interest there is, and we need to build a minimum of 25 units to make 
production feasible.


So, here's the deal:  you can order your TNS-BUF at

http://tapr.org/kits_tns-buf

through *October 20*.  If we receive orders for at least 25 boards by 
then, we will charge credit cards and place the production order with 
our contract manufacturer.  If we don't get 25 orders, we'll cancel the 
project and credit cards will not be charged.  There's no guarantee that 
boards will be available for later order.


We expect about 60 days between placing the manufacturing order and 
receipt of the boards at TAPR.  We'll ship to customers ASAP after 
receipt.  So that means you can expect to receive your order shortly 
after January 1.


So, go to http://tapr.org/kits_tns-buf now to place your order before 
the deadline!


John




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Raw phase data of super-5065

2018-06-05 Thread Bob kb8tq
Hi

It is pretty well known that without “help” ( beyond the filter stuff on the 
super) the 5065 has fairly poor temperature 
stability and somewhat variable response to pressure and humidity. Taking two 
samples right off the production line
likely would have given you different long term data as a result. 

If you are going to the trouble of watching these devices for a long time, 
monitor temperature, pressure and humidity.
I suspect you will find correlations ….

Bob

> On Jun 5, 2018, at 5:44 AM, tim...@timeok.it wrote:
> 
> 
>   Hi Attila,
>   I am doing the long therm stability test on a couple of HP5065, one 
> original A type, the second B or "super" compared with my best long therm 
> stability, a GPSDO HPz3816A.
>   This test will be 30 days long. It is a second test, with a second couple 
> of HP5065A/B, to verify the long therm stability of both the version.
> 
>   Up to now the data available for version B or "super" only concerned the 
> improvement of phase noise.
> 
>   From the first results it seems that the HP5065A is more stable within 
> 30days than the modified version with the filter.
>   It must be considered that the short-term stability of 5065 is better than 
> the GPSDO / TICC up to 10-80k sec then, in this test configuration the valid 
> data will be beyond these values.
>   The results will be ready within 28 days and I will be happy to share them 
> in order to have everyone's opinion.
>   Regards,
>   Luciano
> 
> 
>   Da "time-nuts" time-nuts-boun...@febo.com
>   A "Discussion of precise time and frequency measurement" time-nuts@febo.com
>   Cc
>   Data Mon, 4 Jun 2018 10:11:50 +0200
>   Oggetto [time-nuts] Raw phase data of super-5065
>   Hoi zäme!
> 
>   Would one (preferably multiple) of the owners of a super-5065 be so kind
>   and share the raw phase data of an super-5065 against a stable reference
>   with me? The longer the data trace, the better.
> 
>   Thanks in advance!
> 
>   Attila Kinali
> 
>   --
>   It is upon moral qualities that a society is ultimately founded. All
>   the prosperity and technological sophistication in the world is of no
>   use without that foundation.
>   -- Miss Matheson, The Diamond Age, Neil Stephenson
>   ___
>   time-nuts mailing list -- time-nuts@febo.com
>   To unsubscribe, go to 
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>   and follow the instructions there.
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Raw phase data of super-5065

2018-06-05 Thread tim...@timeok.it


   Hi Attila,
   I am doing the long therm stability test on a couple of HP5065, one original 
A type, the second B or "super" compared with my best long therm stability, a 
GPSDO HPz3816A.
   This test will be 30 days long. It is a second test, with a second couple of 
HP5065A/B, to verify the long therm stability of both the version.

   Up to now the data available for version B or "super" only concerned the 
improvement of phase noise.

   From the first results it seems that the HP5065A is more stable within 
30days than the modified version with the filter.
   It must be considered that the short-term stability of 5065 is better than 
the GPSDO / TICC up to 10-80k sec then, in this test configuration the valid 
data will be beyond these values.
   The results will be ready within 28 days and I will be happy to share them 
in order to have everyone's opinion.
   Regards,
   Luciano


   Da "time-nuts" time-nuts-boun...@febo.com
   A "Discussion of precise time and frequency measurement" time-nuts@febo.com
   Cc
   Data Mon, 4 Jun 2018 10:11:50 +0200
   Oggetto [time-nuts] Raw phase data of super-5065
   Hoi zäme!

   Would one (preferably multiple) of the owners of a super-5065 be so kind
   and share the raw phase data of an super-5065 against a stable reference
   with me? The longer the data trace, the better.

   Thanks in advance!

   Attila Kinali

   --
   It is upon moral qualities that a society is ultimately founded. All
   the prosperity and technological sophistication in the world is of no
   use without that foundation.
   -- Miss Matheson, The Diamond Age, Neil Stephenson
   ___
   time-nuts mailing list -- time-nuts@febo.com
   To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
   and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] a newbie question: where can I purchase 794.7 nm VCSEL for building CPT rubidium clock?

2018-06-05 Thread Attila Kinali
On Mon, 4 Jun 2018 21:31:56 +0800
mimitech mimitech  wrote:

> I'm planning to build a CPT (coherent-population-trapping) rubidium clock
> as my next hobby project. The main purpose is to learn the principles
> behind CPT rubidium clock, and hopefully got similar or better performance
> than commercial miniature rubidium clock such as FE-5680A.

Building a CPT clock is slightly more involved than you might think
at first. The laser diode is only one part of it. You will most likely
be able to improve on the short-term stability of the FE-5680 (which
is rather poor). But I doubt you will be able to improve much on
the long term stability, which is where things actually become interesting,
if you use a naive approach.

Nevertheless, I have not seen many 794/795nm diodes around. The only
one that I have the datasheet of is the one from Vixar.
You might want want to consider going for the D2 line instead of the
D1 line, as 780nm diodes are more commonly available than 795nm. You will
also need to buy several of those and select the ones that come closest
to the wavelength at the desired opearating conditions (usuall spread
is +/-1nm to +/-10nm). Do not assume you can tune more than 0.1nm with
temperature and current (rule of thumb is that you get about 10GHz
per °C and mA). If you need more tuning range, you will need to add an
external cavity (can give you up to 5nm range), which then needs to be
tuned to the 3.45GHz (ie it's length needs to be approximately 8-9cm).

Alternatively, you can get two S1-0780-XXX from Sacher Laser
(cost IIRC 2500€ each) and keep them 6.9GHz apart (using an optical PLL).
If you have enough money to spend, I'd go for two Cateye diode laser CEL's
from Moglabs (cost AFAIK 5000€ each)

No matter what you choose, you will need some wavelength stabilization
scheme. You can either do that with the vapor cell itself or use
an additional cell and do a DVALL or a saturated absorption locking.
Note that this addtional cell will need to be without buffer gas.
An external cell will offer better stability and thus lower noise,
which directly translates into higher stability.

As polarisation scheme, I suggest using σ+/σ- as it seems to be more
robust than the lin/lin schemes.

Attila Kinali
-- 
It is upon moral qualities that a society is ultimately founded. All 
the prosperity and technological sophistication in the world is of no 
use without that foundation.
 -- Miss Matheson, The Diamond Age, Neil Stephenson