Re: [volt-nuts] Fluke 732A schematic re-drawn?
David: I will be guessing that you are discussing the lack of Q14 that is chassis mounted on some versions of A4, driven by Q13. I have sent you by PM a pdf of a 732A manual that I probably got from FLUKE. If you look at the schematic for A4 this one has the chassis mounted Q14 shown and listed in the parts list for A4. The interesting thing here is that a manual I sent you is P/N 645051 dated May 1983, while another manual P/N 788414 dated May 1986 does NOT show Q14. So for some reason known only to the people at FLUKE Q14 must have been dropped by May 1986. Go figure. I don't have any other manuals except those two P/Ns. The best I can figure out at this time is that Q14 was dropped between A4 Rev A and A4 Rev D. According to the parts list in the May 1986 manual Q13 is a 2N3906. The Ic maximum is 200 ma. I know that the overall current draw from the batteries when no AC power applied is about 250ma, but I don't know how much of that is for the heaters. Bill - Original Message - From: "David C. Partridge"To: "'Discussion of precise voltage measurement'" Sent: Friday, January 19, 2018 4:16 AM Subject: [volt-nuts] Fluke 732A schematic re-drawn? > The schematic for the A5 Reference and Oven Controller Board has to be one > of the worst schematics I've seen in a while, breaking pretty much all the > rules on making a schematic understandable. I suspect this was done > deliberately. > > Has anyone ever re-drafted this (to include the A4 pass transistor > components) to make sense of it? > > May your volt always be stable > Dave > > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Get 10K resistor calibrated in UK
Frans: Mine which was built in 1979, and corrected for the 1990 change in the ohm, and calibrated last in 2009 has drifted upwards by 1.61 ppm total. So about 0.054 ppm/year. I don't have any data from those 30 years so this is the beginning and end points. These all seem to have a spectacular performance to say the least! Regards Bill - Original Message - From: "Electronics and Books via volt-nuts"To: "Discussion of precise voltage measurement" Sent: Wednesday, November 29, 2017 6:25 AM Subject: Re: [volt-nuts] Get 10K resistor calibrated in UK > can anyone tell me howmuch a SR104 10k resistor drift > > Met vriendelijke groeten > Regards > Frans > > electronicsandbo...@yahoo.com > http://ElectronicsAndBooks.com > Netherlands > > Discere ne cesses > > > On Wed, 11/8/17, David C. Partridge wrote: > > Subject: Re: [volt-nuts] Get 10K resistor calibrated in UK > To: "'Discussion of precise voltage measurement'" > Date: Wednesday, November 8, 2017, 4:08 PM > > I guess that depends on your > definition of cheap :( > > Dave > > -Original Message- > From: volt-nuts [mailto:volt-nuts-boun...@febo.com] > On Behalf Of Alan Ambrose > Sent: 08 November > 2017 14:31 > To: volt-nuts@febo.com > Subject: Re: [volt-nuts] Get 10K resistor > calibrated in UK > > NPL i.e. > the UK national lab, are also surprising flexible and cheap. > Looking at their schedule for '15/'16 (the one I > happen to have to hand) they had 10K ohm cal 30-Nov-15 to > 11-Jan-16 @ £510. They also had voltage cal October and > March at ~£420. > > A. > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Update on 720A
David: I had the same problem as you when I got my used 720A years ago. Both the "A" and "B" decade had a few resistors that had changed over the years. I don't think as bad as you experienced but still they would not balance within the range of the adjustment pots. So I "adjusted" those that needed correcting with either parallel or series resistors as necessary. I used the best tolerance resistors I could get at that time. I figured that the "adjustment" resistors were such a small amount of the resistors I was trying to adjust, that the TC wouldn't be much of a problem. However I did have to go through these "adjustments" several times as I noticed that correcting one would then affect others in the string. This is probably what you are experiencing with needing to "readjust" others again. A few passes in each string finally got the adjustments into a good range and they have stayed there for years now. As far as I am concerned this is all "good enough" for hobbyists or VoltNuts who don't want to spend $5,000 or more to get FLUKE to refurbish your instrument, if FLUKE is even interested in doing the job and has the parts. Notice that some resistors in the "A" and "B" decades are selected at the factory anyhow. If you have the money, then be a "purest". It might be cheaper just to buy a new 720A. As you realize now the 720A input resistance is 100k. So you will get a small voltage drop from the voltage source to the input of the 720A, depending upon the size of the wire you are using. Of course the proper way to insure that there is no drop is to use Kelvin connections from the voltage source to the input of the 720A and then set the voltage source to external Kelvin sensing. What I usually do is to set the ratio dials to .99X and then set the voltage supply to get exactly 10.000 on my 3458A. Then you have the best chance to get an idea of the accuracy of the 720A ratio at .8, .7, .6 and so on. Don't let the 3458A autorange when down to the 0.1 ratio! Then to minimize the effects of the "B" decade, use .89X, .79X down to .09X. The only problem with this method is that you don't get to check the .9 ratio. So what you can do is connect the voltage source to the "1.1" input, set the ratio dials to 1.000 and then set the voltage source to give you 10.000 volts on the DVM. Now you can check the .900, .800 and so on ratios. I am sure that I did the R203 adjustment as described in paragraph 4-25 of the manual. It has been years now. All IMHO of course. You are now down into measurement "mud" so to speak. This is probably why the 3458A development team used the JJA at Loveland to check the overall linearity of the 3458A A/D. The 720A (or equivalent) wasn't good enough at 0.1 ppm! Bill - Original Message - From: "David C. Partridge"To: "'Discussion of precise voltage measurement'" Sent: Tuesday, August 08, 2017 7:39 AM Subject: Re: [volt-nuts] Update on 720A > Chuck, > > I totally get your point that ideally you should use any KVD in null-balance mode. I do note however that you can load it up to 11mA without damage though I've not once gone anywhere near that (worst case insult has been about 0.1mA). However it does also say that to avoid loading errors you should load with an impedance >= 1TeraOhm (10**12). > > Clearly a 3458A on the 10V and lower ranges is merely > 10GOhm and will show loading effects. > > So, yes, I do recognise I was "doing it wrong"! > > I don't believe I've inflicted any damage, and have managed to restore it to a useable state from a fairly sorry condition: When I got it, two oil bath resistors were *way* off value (202 and 100 ohms low respectively), and the S2 shunt resistors were about 1.3 ohms high which was far enough off to prevent S2 calibration. There were were also sundry other problems like two open circuits in the final decade, a badly worn trim pot, and sundry wires broken at solder joints. > > Ideally I'd like to replace the KVD "Bridge Balance" pot as that's a bit noisy but shudder to think what that might cost if I could even find one. > > Cheers > Dave > -Original Message- > From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Chuck Harris > Sent: 08 August 2017 13:34 > To: Discussion of precise voltage measurement > Subject: Re: [volt-nuts] Update on 720A > > Hi Dave, > > Fundamentally, any input impedance from your 3458 is going to throw your 720A out of wack. the 720A is not supposed to have even picoamps of loading nothing, nada, zip. > > I would suggest that before you damage your 720A further, you read the manual for the 720A and follow the instructions to the letter. An HP3458A is not a substitute for a null detector and a voltage calibrator. > > The calibration of a 720A requires that no current, none, zip, be drawn from the 720A on any port. It can only be done by using a voltage
Re: [volt-nuts] Update on 720A
David: Ignore my previous post. In thinking about this while out shopping this morning I know the answer. The output resistance of the 720A is about 66k ohms. Your voltmeter has an input resistance of 10 megohms. You are simply loading the output of the 720A. The reason that the 0.1 ratio has about the expected voltage is because the 10 Volt range your voltmeter probably has a very high input resistance. You are on "autorange" and when the output is set for 0.2 ratio the voltmeter is now on the 20 volt range which is now 10 megohms. I get the same results as you do when measuring 0.9 ratio on my setup. 89.920 volts. Check your 720A by applying 10 volts and then do your ratio checks on the first decade. I just haven't used a 720A for a while and I didn't remember about the output resistance. When you get to become a "well aged citizen" it takes a while to recall things. Bill - Original Message - From: "David C. Partridge"To: "'Discussion of precise voltage measurement'" Sent: Monday, August 07, 2017 1:49 AM Subject: [volt-nuts] Update on 720A > I've added resistors in series with R1008 and R1044 to bring the resistance > for those parts of the A divider to a value that it would calibrate OK (two > of the resistor sets in the oil bath very low in value). I also had to > remove five turns (about 1.3 Ohms) from R1051 as the combined value of R313 > and R314 was too high to allow me to calibrate S2. I also replaced R1041 > as it was worn out (almost impossible to get a null). > > I've successfully calibrated the C decade as per the maintenance part of the > manual. > > Sadly the beast isn't giving me a nice linear voltage sequence when using > just the A decade. > > Input voltage 100.001,4V fed to 0/1.0 Input terminals All Decades set to > zero except Decade A. > > Decade A Output V > 0.1 9.999,88 > 0.2 19.967,8 > 0.3 29.936,7 > 0.4 39.903,8 > 0.5 49.874,7 > 0.6 59.855,6 > 0.7 69.852,4 > 0.8 79.871,3 > 0.9 89.919,6 > > See also attached graph. > > Any suggestions on where to look for trouble will be most welcome ... > > All sections of the A decade attenuator now measure about 9.999,2 ohms. > > It was suggested on eevblog that the next decade (or a subsequent one) was > loading it, but with all the other decades set to 0 I don't quite "get" how. > > Thanks > Dave > > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Update on 720A
David: First suggestion would be to see what the output measures, with the input still set to 100.0014 volts (same as before), when setting the dials to: 0.9X(equivalent to setting the "A" decade to "0.1") 1.9X(equivalent to setting the "A" decade to "0.2") 2.9X(equivalent to setting the "A" decade to "0.3") 3.9X(equivalent to setting the "A" decade to "0.4") and so on up to 8.9X(equivalent to setting the "A" decade to "0.9") What is the output when the dials are set to 9.9X? It should be 100.0014 (or very, very close) just like the input voltage. Also what do you mean by your statement "Sadly the beast isn't giving me a nice linear voltage sequence when using just the A decade." Do you mean that other settings when decade B, C, D and so on are not "0", produce ok results? Bill - Original Message - From: "David C. Partridge"To: "'Discussion of precise voltage measurement'" Sent: Monday, August 07, 2017 1:49 AM Subject: [volt-nuts] Update on 720A > I've added resistors in series with R1008 and R1044 to bring the resistance > for those parts of the A divider to a value that it would calibrate OK (two > of the resistor sets in the oil bath very low in value). I also had to > remove five turns (about 1.3 Ohms) from R1051 as the combined value of R313 > and R314 was too high to allow me to calibrate S2. I also replaced R1041 > as it was worn out (almost impossible to get a null). > > I've successfully calibrated the C decade as per the maintenance part of the > manual. > > Sadly the beast isn't giving me a nice linear voltage sequence when using > just the A decade. > > Input voltage 100.001,4V fed to 0/1.0 Input terminals All Decades set to > zero except Decade A. > > Decade A Output V > 0.1 9.999,88 > 0.2 19.967,8 > 0.3 29.936,7 > 0.4 39.903,8 > 0.5 49.874,7 > 0.6 59.855,6 > 0.7 69.852,4 > 0.8 79.871,3 > 0.9 89.919,6 > > See also attached graph. > > Any suggestions on where to look for trouble will be most welcome ... > > All sections of the A decade attenuator now measure about 9.999,2 ohms. > > It was suggested on eevblog that the next decade (or a subsequent one) was > loading it, but with all the other decades set to 0 I don't quite "get" how. > > Thanks > Dave > > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Fluke 332B
David: The 332D and the 332B are identical except for the accuracy specs for the stability of the output voltages. 332B spec for 10 Volts is +/- .002% (20 ppm) for 90 days while the 332D spec for 10 Volts is +/- .001% (10 ppm) for 60 days. The monthly drift specs are also slightly better on the D than the B. I suspect that FLUKE simply aged and selected a more stable reference device along with matching the resistors in the sampling string for less drift. For all of that you paid US$2,595 for the 332B and US$2,995 for the 332D in 1973. I seem to remember someone replacing the chopper assembly with something a little more modern. Don't know where I saw that. I have a B and a D and both seemed to be very stable with the existing chopper assembly. One thing I did find is that the electrolitic caps have a high tendency to go bad. Not the big ones on the raw supply but those on the various plug in assemblies. I seem to remember that they were in a range of 1 ufd to 47 ufd or so. So check them all or better yet just recap. You will save yourself a LOT of troubleshooting time in the long run. There is also a discussion of this subject either on VoltNuts or EEVBLOG. Bill - Original Message - From: "David C. Partridge"To: "'Discussion of precise voltage measurement'" Sent: Saturday, July 08, 2017 5:09 AM Subject: [volt-nuts] Fluke 332B > Amongst the stuff I recently acquired is a Fluke 332B (sn 4151). > > I found the manual for the 332D on the Fluke website, but can't find one > specifically for the 332B apart from an incomplete US Army version. > > I have some questions: > > 1) Does anyone know where I can find the matching manual > > 2) If not, how useful is the 332D manual? > > 3) It was last calibrated in 2010, and appears to be consistently low in > output (about 5mV at 100V, 25mV at 500V, 50mV at 1000V) after being on for > an hour. Is that the sort of error I should expect after that period? > > 4) I'm leery of the chopper amp, did anyone (Dr. Frank?) do a retrofit using > a modern chopper op-amp or similar? > > > Thanks > Dave > > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Initial Turn-On Fluke 332D?
Good point. I have found that checking all the electrolytic caps on the PC Boards is a good idea. Many have failed over the years and the unit will not work properly, including oscillations on the output. Usually they have little or no capacitance. I have done this step to 4 units over the years and have found a lot of bad electrolytic caps in that time. Bill - Original Message - From: "Mark Sims"To: Sent: Saturday, November 26, 2016 10:02 AM Subject: [volt-nuts] Initial Turn-On Fluke 332D? > You will probably need to replace a lot of the electrolytic caps on the circuit board... > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Initial Turn-On Fluke 332D?
Brooke: The 332D has no internal batteries, CPU or other internal memory. Accuracy is dependent upon the internal reference on the reference card and the adjustments of the first 3 decade resistors, the three range resistors, the reference resistor and the three zero range resistor adjustments. Make sure that there are jumpers installed between the "output" and "sense" terminals (usually two flat blade jumpers). Turn on the 332D to "standby/reset" and then wait for a few seconds. You should hear a relay click, then the switch can be set to "opr". For intial settings set the "range" switch to 10 V, the "readout" dials to all 0s, the meter switch to "output voltage", the "voltage trip" switch to 10 V, turn the "vernier" control CW to full, and the "current limit" control to about 1/2 CW or so. Then, if everything is working, you can connect a voltmeter to the outputs and turn up the first decade dial to a voltage. Have fun, Bill - Original Message - From: "Brooke Clarke"To: "Discussion of precise voltage measurement" Sent: Friday, November 25, 2016 10:46 AM Subject: [volt-nuts] Initial Turn-On Fluke 332D? > Hi: > > I just received a Fluke 332D from Fair Radio, but before I do anything is there some initial checks or a battery that > needs attention? > Just started reading the manual. > > -- > Have Fun, > > Brooke Clarke > http://www.PRC68.com > http://www.end2partygovernment.com/2012Issues.html > The lesser of evils is still evil. > > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Fluke DVMP Care Plan
Ken: I had my 732B calibrated at FLUKE in Everett WA last year. I had the Z540 with before and after data cal done. The price was $695 which included FLUKE shipping the 732B back to me, one day service. FLUKE wanted me to ship it to them overnight delivery by 10 AM which cost $250 UPS. But after that I decided that I would ship it either overnight afternoon delivery (which is cheaper) or 2 day delivery (which is even more reasonable). Hindsight is always better, just like Monday morning quarterbacking. I know that my battery can last 3 days real easy by discharging it for 3 days and monitoring the battery voltage. I think that FLUKE is just trying to make sure that some units are not shipped with old, marginal batteries. Then the battery would go dead and the "before" data could not be supplied to the customer. The one surprise was that my unit had the handle missing and FLUKE installed a new handle and included that in the total cal cost. The stupid handle retails for $120 and the hardware is another $27. Bottom line here is that I don't care if the "calibration stickers" are inplace or not when FLUKE receives the unit, so if the battery wouldn't hold a 3 day charge I would replace it and test it before I send it in again. I would use the 2 day service. The uncertainty would be how well the bank of 4 732Bs that they compared my unit with at the Cal Lab. But since those don't leave FLUKE and are certified every 90 days (probably to the FLUKE JVS), according to the certificate, my guess is that the voltage would be within .3 PPM depending upon how much drift your unit would have during transit. Since I have 4 ea 732As, collected over the years, and I had checked the stability before I sent the 732B in, I was pretty sure that my 732B was very stable over 13 days, less than 0.1ppm. When I got it back I began comparing it again to the bank of 732As and it remained within 0.1ppm for 90 days. But after that it has started to drift steadly upwards around 0.7 ppm per year. Since I have a very good history on the bank of 732As I can see that the 732B is most of the drift against the average of the bank. Of course all IMHO and the best data reductions I can make. The final test of my data keeping would be to send the 732B again and see if I am close to "The Volt" as defined by FLUKE. Someday when I am curious. The whole turn around time was 10 days portal to portal to me. Bill - Original Message - From: "Ken Peek"To: "volt-nuts" Sent: Thursday, October 20, 2016 11:26 AM Subject: [volt-nuts] Fluke DVMP Care Plan > I own one 732B that needs calibration. Fluke wants about $500 for the > ISO-9001 calibration, but it would be about $250 EACH WAY to ship it > to/from them with the proper insurance (so, about $1000 total). > > Mitch Van Ochten mentioned that the Fluke DVMP program might be a good way > to do this. [http://download.flukecal.com/pub/literature/2001531_A_w.pdf] > > Does anyone have an idea about how much this costs to calibrate just one > 732B, and what kind of uncertainty I might expect? > > -Ken > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] fluke 731b battery pack
Eric: I have been using NiMH AA batteries in my 731B for at least 15 years now. I am sure that the specs say don't use them in a circuit designed for NiCd but I haven't had any problems with over heating or overcharging. You have to do some mods to the battery mounting to get them in there but once done you are good for a lifetime. I have never checked to see how long the NiMH batteries would run for but my guess would be longer than the original NiCds. Bill - Original Message - From: "Eric Garner"To: "Discussion of precise voltage measurement" Sent: Wednesday, August 03, 2016 10:27 AM Subject: [volt-nuts] fluke 731b battery pack > So I managed to get one of the Fluke 731b's mentioned on the list a while > back. It was a little more chewed up than the description in the auction > led me to believe: > > https://plus.google.com/u/0/112871219575380932919/posts/iDWyzwYYRdj?pid=6314634305601690930=112871219575380932919 > > but it appears to be functional. > > The NiCd pack appears to have leaked some time in the past and was removed, > which is fine since I was going to replace the batteries anyway. > > Looking through the archives, replacement battery packs for the 731b has > been a popular topic but I have some questions that remain open. > > 1. NiMH cells seem to work in the 731b so I'd prefer to use them over NiCd > because of the higher capacity and lower toxicity of potential leaks. Is > there a downside to doing this? > > 2. If I go with NiMH, is there a manufacturer that's prefered? reading > reviews of NiMH cells online leads me to believe that some manufacturers > (Tenergy) have lower quality and are less reliable. > > 3. If I go with NiCd would you expect problems if I used larger cells (4/5A > vs 2/3A) > > > > > thanks > > --Eric > _ > Eric Garner > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] “A Solid-State Reference System”. Author: John R. Pickering
Alan: The best information I have ever been ever to dig up on this subject is contained in US patent #6,342,780. I will sent this to you by PM as well as trying to attach it to this post. Bill - Original Message - From: "Alan Ambrose"To: Sent: Wednesday, November 11, 2015 1:16 PM Subject: [volt-nuts] “A Solid-State Reference System”. Author: John R. Pickering > Hi, > > Anyone have a copy of this paper/article? > > [2] “A Solid-State Reference System”. Author: John R. Pickering, > Metron Designs Ltd., Norwich, UK. > > referenced in the Fluke 700 doc here: > > http://www.elcal.ch/files/11749-eng-01-a.pdf > > TIA, Alan > > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. 6342780_Zener_diode_reference_voltage_st.pdf Description: Adobe PDF document ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] any voltnuts in the Chicagoland area?
Randy: Ok so does Morgan Hill qualify? Or is that too far away for you? Just kidding as usual. Bill - Original Message - From: "Randy Evans"To: "Discussion of precise voltage measurement" Sent: Thursday, October 01, 2015 8:18 AM Subject: Re: [volt-nuts] any voltnuts in the Chicagoland area? > I would like to know if there are any volt-nuts in the San Francisco bay > area; e.g., San Jose area where I live. > > Randy Evans AE6YG > > On Wed, Sep 30, 2015 at 4:29 PM, Pete Lancashire > wrote: > > > Interesting idea a voltnuts location database > > On Sep 30, 2015 12:34 PM, wrote: > > > > > I am looking for fellow voltnuts in the Chicagoland area to share > > > calibration, voltage standards, etc... > > > > > > This hobby can get quite expensive when the costs don't get shared... > > > > > > Please reply to the list or direct via e-mail... > > > > > > Thanks > > > ___ > > > volt-nuts mailing list -- volt-nuts@febo.com > > > To unsubscribe, go to > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > > and follow the instructions there. > > > > > ___ > > volt-nuts mailing list -- volt-nuts@febo.com > > To unsubscribe, go to > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > and follow the instructions there. > > > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] 540b Arrival...........
David: Sounds like you got a good find. Clean, batteries hold a charge and has been updated by someone. The lithium batteries are probably the power for the modification that solves the problem that the Mercury Battery for the Reference Supply being "unobtainium" anymore. I have a pdf of the modification including schematic. I can send it to you by PM if you are interested. Bill - Original Message - From: "David Garrido"To: Sent: Wednesday, September 30, 2015 6:22 AM Subject: [volt-nuts] 540b Arrival... > Hello All, > > My new to me 540b arrived last evening. I have never used one of these so I put it on the bench, opened it up and inspected it looking for leaks primarily. EVERYTHING is spotless. Fluke factory cal stickers everywhere. AND.after plugging her in over night, all of the batteries take a charge and test in the middle of the BATT scale. > > CRAZY! > > All of the batteries in the 20 cell pack are individually shrink wrapped in light blue and the two others on top are AA size lithiums. Two of them side-by-side. Was this a factory mod? All the work looks to be factory, but who knows if the PO was just REALLY good at clean work. > > Anyhoo………thanks for the bandwidth. > > Cheers, > > David > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Fluke 720a self calibration repair
David: I just looked inside my 720A at the 20 turn 5K trimpots. They are Bourns 3005P-502 printed on them, cermets, standard 3/4", tempco 50 ppm. So the replacement is available directly from stock and nothing fancy. Mine are date code 8412 so my unit was built somewhere in the mid '80s. Not too sure where the idea of a wirewound pot came from here. They are more than a little pricey at $13.98 ea from Allied. Looks like you need 24 to replace them all in the "A" and "B" decades. So that would be $335.52 unless you buy 25 or more at which time the price drops to $12.51 which totals $312.75. They come in tubes of 25 each which accounts for the price drop. Digikey is $14.35 and Mouser is $14.35. Wow! Expensive little devils. On the other hand (left) Digikey shows a unit from "TT Electronics/BI Technologies" for $1.40 ea. Interesting. Why such a massive difference in price? These are 100 ppm but as I have noted in another posting the tempco isn't very important. It's the resistors in the oil filled can that set the overall stability. Bill - Original Message - From: "David Garrido"To: Sent: Saturday, September 26, 2015 5:05 AM Subject: Re: [volt-nuts] Fluke 720a self calibration repair > Thank you for all of the input on this folks. > > Does anyone have a copy of the datasheet for the original Ramo-Eltra P/N 3800P-502? > > The reason I ask is that the price of these parts is driven in part by the tempco and I found info that indicated 20ppm/C on one website, but that was not a DS, just a listing of the info. If they can be 50ppm/C or even 100ppm/C, then the price drops exponentially. > > I will do whatever is necessary to maintain or (LOL) improve the 720’s legendary performance. > > Cheers, > > David > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Fluke 720a self calibration repair
David: Run the numbers. Assuming that the 5K-20 turn pot R1005 is set in the middle of it's travel (2.5K), that all of the other resistors, in series and parallel ( R1006,R1007,R1008 and R302), that make up one step of the "A" decade have a "0" TC, only R1005 has a maximum of +/- 100 ppm per degree C, then the maximum change on the total resistance between step 0.9 and 1.0 of the "A" decade would only be 0.00410 % change over the entire +/- TC of R1005. That is 0.00410 ppm. The "Absolute Linearity" of the "A" decade is specified at 0.1 ppm at +/- 1 degree C, without a temp correction chart. Of course all of the other resistors mentioned have a TC also, which could be + or - to the total nominal of 10,000. Therefore, unless I am mistaken, the less expensive +/- 100ppm pots would work just fine and you could not observe or measure the changes. This is also why I didn't worry about my "trimming" R1008, the "factory selected" resistor, slightly to get the 20 turn pots to adjust to the middle of their ranges, when necessary. IMHO Bill - Original Message - From: "David Garrido"To: Sent: Saturday, September 26, 2015 5:05 AM Subject: Re: [volt-nuts] Fluke 720a self calibration repair > Thank you for all of the input on this folks. > > Does anyone have a copy of the datasheet for the original Ramo-Eltra P/N 3800P-502? > > The reason I ask is that the price of these parts is driven in part by the tempco and I found info that indicated 20ppm/C on one website, but that was not a DS, just a listing of the info. If they can be 50ppm/C or even 100ppm/C, then the price drops exponentially. > > I will do whatever is necessary to maintain or (LOL) improve the 720’s legendary performance. > > Cheers, > > David > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Sometimes you just gotta gloat............
David: I don't suppose you would give us the address of this Equipment Surplus Center of yours I like those places where they don't have an IDEA of what they have and what it is worth!!! Great catch to say the least. At least you will have something to do for a while. What do those PCA interconnect cables look like? Gold plated? Banana, spade etc.? Might be low thermal? Bill - Original Message - From: David Garrido d.garr...@me.com To: volt-nuts@febo.com Sent: Tuesday, August 18, 2015 8:59 AM Subject: [volt-nuts] Sometimes you just gotta gloat Hello All, I have been on the lookout for a variety of different FLUKE equipment. And today, it all came together in a wonderful way. I went to my favorite local Equipment Surplus Center on the hunt for an HP 6260B DC PSU. Found one, 40 dollars. DONE!!! Or so I thought, until I decided to take one more quick look before leaving with my new to me DC PSU. While cruising around I spotted what looked like a FLUKE 7105A Calibration System. You know the one.all hidden back in a dark corner. Well, sure enough it was a 7105A with its full compliment of test gear. 335D, 845AR, 720A, 721A, and a BUNCH of interconnects. All of the equipment looks to be in fine shape and VERY clean. No busted or bent posts, all knobs and switches intact. The best part..when I asked how much he would sell it for he told me, “If and ONLY if you take both the HP 6260B and the 7105A, I will give it to you for $250 cash total.” Needles to say, I could not get my pocket opened fast enoughout flew the cash and the gear was even loaded into the back of my truck for me. It has been a VERY good morning gentleman, a very good morning indeed. I do have a question regarding all of the FLUKE cables that came with the 7105A: These are the part no’s: PCA-0110 PCA-0116 PCA-0126 PCA-0136 PCA-0227 PCA-0536 PCA-0627 PCA-0636 I have multiples of these and cannot seem to locate any info on these anywhere. I have skimmed through the manuals for all of the gear and they seem to have no reference. What SPECIFICALLY are these for and where would one find the info for them and their use? Cheers, David ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Test Leads
Randy: From my own experiences and what I have done over the years about low thermal leads. I have made my own leads from the Pomona #4892 double gold plated banana plugs along with using Pomona #5406 single gold plated banana plugs, where I needed to spread the leads futher apart than the standard 3/4 inch, such as plugging into the Fluke 732A. I have been using Belden #9272 shielded twinax as my cable material only because it was available at the time to me. It is a twisted pair of 20 ga. tin plated wires inside of a braided shield. The purests will say that the tin plating will cause Thermals when screwed down on the Gold plated banana plugs. I realize that and I now have 100 feet ( 50 red, 50 black ) of Pomona #6733, which is extremely flexable copper wire with Silicone double insulation. I will then twist these together and the put a shield braid over that and finally put a heat shrink layer over the shield to make my own copper twisted pair shielded wire which I will then attach to the #4892 or #5406 banana plugs to create a more perfect low thermal test lead ( I hope ). This is all when I get a round toit of course. In the meanwhile I have purchased 6 ea Pomona #1756-48 leads. Frankly I cannot measure a difference between the #1756 and the homemade leads, once I wait about 3 minutes for all Thermals to settle down. I also have 2 ea Fluke P/N 738716 ( 24 inches ) and 1 ea Fluke #738724 twin shielded test leads. These are the Fluke 5440A-7002 Low Thermal test lead set that is recommended by Fluke for use with 732A/B, 5440A/B and other high precision calibrators. They are RG-58 with Tellurium Copper conductors/ Beryllium Copper Tip Springs. Again I cannot see differences between all of these cables with the best measuring equipment I have. HP 3458A DVM and Fluke 845AB Null Detector and measuring between 732A/Bs, 5440B for nulls. I am sure that Fluke has done extensive research on this subject. It may well be that measurements can be made faster using the purest approach of pure copper and gold plated connectors. I don't know at the present time. Hope this helps you in making up what you need. Bill - Original Message - From: Randy Evans randyevans2...@gmail.com To: volt-nuts@febo.com Sent: Tuesday, July 21, 2015 9:33 AM Subject: [volt-nuts] Test Leads I recently acquired a Fluke 752A to go with my Fluke 732A and HP-3458A. Now I need to address the need for low thermal test leads. Does anyone have any suggestions for test leads that have low thermal contributions to measurements? I have looked at Pomona 1756 Low Thermal EMF Cables as a starting point but haven't found any other candidates. I have heard that old TV twin lead works well since most are stranded copper wire. I have tried it between my 732A and 3458A and it seems to work fine but I would like to use more professional looking test leads, particularly with correct copper spade lugs (Pomona 2305 Low thermal EMF spade lug, Gold-plated?). Any other suggestions? Thanks, Randy Evans AE6YG ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] HP 3456A -3 Fault
Dave: These things are NOTORIOUS for filter cap failures in the power supplies. Check both the voltages and the RIPPLE on the various power supplies. It is not sufficent to just check the voltages. You must check the ripple with an o'scope. At least this is the first step before looking for anything else. Good luck, Bill - Original Message - From: David Smith w...@msn.com To: volt-nuts@febo.com Sent: Friday, December 26, 2014 3:34 PM Subject: [volt-nuts] HP 3456A -3 Fault Hello Friends, I have two HP 3456A's that have the same -3 fault in the test mode. Both DVM's started exhibiting this same fault about the same time. I've troubleshot one of the DVM's to a failure in the Isolation Logic Transmitter. I have no waveform at the output of A3T1. Before I continue with troubleshooting I thought I'd ask a question of the group. Is there a common mode failure of these DVM? Is there something in particular or area that I should be looking at/for? It's probably coincidental that they would both fail at about the same time with the same -3 fault. Your input is greatly appreciated. Best regards, Dave - W6TE ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] 845ab Led Mod Update
Sorry I should have said uv where ever I said mv. Too early in the morning where I live in the world. Bill - Original Message - From: Bill Gold wpgold3...@att.net To: Discussion of precise voltage measurement volt-nuts@febo.com Sent: Thursday, September 25, 2014 8:10 AM Subject: Re: [volt-nuts] 845ab Led Mod Update Dallas: I check my 845ab yesterday, S/N 4470002. It still has the neon bulbs/photo resistors chopper. From what I can observe the Peak to Peak noise on the 1 mv range is about +/- 0.1 mv, or a total of around 0.2 mv. As far as temperature drift I don't know. Never tried to measure that parameter. Sure would love to know if the FLUKE modification to the 845ab made in 1993, when the new circuit was put into production, makes the noise lower. Or better, if your efforts can make the noise lower on the 1 mv range. Bill - Original Message - From: Dallas Smith dosm...@outlook.com To: Discussion of precise voltage measurement volt-nuts@febo.com Sent: Wednesday, September 24, 2014 12:06 PM Subject: [volt-nuts] 845ab Led Mod Update Well I finally got the 615nm Led's installed and the results seem so so. The zero drifted was significantly on the 1uV range. So I used two LT1634 1.25 Volt references with the current set by two 43k resistors from the ±15 volt supplies. The previous reference used came from the -0.4 volts at TP1 and +0.6 volts at TP8. Removed the 150k resistor at the positive end of the zero control and lifted the negative end from the circuit board the connected the ±1.25 ref to the zero pot. Much better stability, drifts by noise about ±.1 to ±.2uv now on 1uv range. Is this normal now? Never used a new Fluke 845ab before. The white led's seemed to filter the noise better with its persistent, may go back. The zero seems stable with ±2 F change in room temperature now. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Old HP3458A - SN: 2823A 03939
TIA: Push blue key then blue R ( the shifted TRIG key) and ENTER. Then press the downarrow V key 4 times to get to the REV? on the display. Then press ENTER and two numbers will be displayed with a comma in the middle. The first number is the multimeter's master processor firmware revision. The second number is the slave processor's firmware revision. This is if the full menu has been set. If not, see the MENU command in the Operator's Manual and set it to FULL. Mine are both 7,2 and I have no problems. The latest is 9. Bill - Original Message - From: cfo xne...@luna.dyndns.dk To: volt-nuts@febo.com Sent: Sunday, September 21, 2014 8:22 AM Subject: [volt-nuts] Old HP3458A - SN: 2823A 03939 Gents I'm looking for a 3458A , and have found one : SN: 2823A 03939 w. opt 002 Seller says it's passing selftest. But it's an old serial# , is there any way one can verify that it has an up-to date firmware , and maybe via that see if it has has all the needed upgrades or ServiceNotes applied ? TIA CFO Denmark ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Fluke 5200A extender board.
Ken: I have a 5200 extender kit including the PC Board and the cables. In looking at it the center PC board that has the traces on it is enclosed on both sides with a piece of blank PC board material (no copper or traces) so you wind up with 3 ea 0.062 PC Boards riveted together. The center board has the necessary traces to connect the edge connector fingers to the socket. The outer two boards are just plain PC material with the necessary holes machined out so that the center board can have the connector and fingers open. There has been test points place on either side of the socket so you can connect a meter or scope to the lines. The connector has the following printing on it. 345-086-520-201 below that it has what looks like a trade mark of either a C or O with letters inside that which I can't make out, then 3A 250V EADC ?? the ?? means I can't read this. When you say photo copy the PC Board I will assume something that would show how the traces are laid out on both sides. This is impossible without removing the 5 rivets that hold this sandwich together which I don't want to do. I will do a PM to you with a pdf attached of the backside of the extender. Bill - Original Message - From: Kgoodhew kgood...@iinet.net.au To: volt-nuts@febo.com Sent: Wednesday, August 27, 2014 11:44 PM Subject: [volt-nuts] Fluke 5200A extender board. Hi, The pin spacing on the card edge connectors is 0.1 (2.54 mm) and is a 43/86 dual row socket. I know what the extender looks like as I found one on ebay several months ago that had already sold for $50 (damm it!!) and I have been looking ever since, but it had a picture of it. It is a double sided pcb board that plugs into the female connector that is mounted vertically in the instrument and then rises up above the top of the instrument case where there is a female socket mounted horizontally so in effect the board now sits horizontally above the instrument whereas normally it sits vertically in the instrument. That way you have access to the board to take measurements, something you cannot do when the boards are in the instrument due to the close spacing of the various boards. As the 43/86 pin card edge connectors appear unobtainable I have sourced a 50/100 pin female connector that I can make do with, but I need either a 43/86 pin male card edge connector (which are also unobtainable) to make up an extender using cables as you have done, so the only option appears to be to make up a double sided pcb board with pads spaced at 0.1 centres on both sides and then either use it to terminate the cables to the female socket , or probably better make the pcb the same as the original extender and just use my 100 pin socket on that. That is why I was hoping someone may be able to photo copy the pcb board if they have an extender they do not want to sell. Thanks, Ken Goodhew. --- This email is free from viruses and malware because avast! Antivirus protection is active. http://www.avast.com ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] 732A and Prologix received
Randy: Something I had never tried to measure. As I have found out in the past there is a lot of overhead going on in the meter during and after a measurement. In thinking about this I turned OFF the autozero AZERO and the time for each SMPL was cut in half to around the estimated 16.66 seconds for 1000 PLC. So it becomes obvious that the meter makes an autozero measurement for 1000 PLC and then the actual measurement for 1000 PLC which explains the 33 seconds. Makes sense. This is probably why the AZERO menu gives you ON, OFF and ONCE. For short measurement sequences you just autozero ONCE at the start. Thanks for the observation, this helps me. Everyday I learn something new. Bill Thanks for the information - Original Message - From: Randy Evans randyevans2...@gmail.com To: Discussion of precise voltage measurement volt-nuts@febo.com Sent: Tuesday, August 26, 2014 2:16 PM Subject: Re: [volt-nuts] 732A and Prologix received Bill, I measured the time between SMPL symbols with NPLC set to 1000 and it is approximately 33 seconds. It takes an hour to complete 100 readings. Randy On Mon, Aug 25, 2014 at 10:01 PM, Bill Gold wpgold3...@att.net wrote: Not that I know of. Just wait and when you don't see a SMPL on the display it is done. But then with 1000 PLC that is around 16.66 seconds per reading times 100 readings is somewhere around 28 minutes and there is probably some overhead time so around 30 minutes. Not from the front panel at any rate. Bill - Original Message - From: Randy Evans randyevans2...@gmail.com To: Discussion of precise voltage measurement volt-nuts@febo.com Sent: Monday, August 25, 2014 9:23 PM Subject: Re: [volt-nuts] 732A and Prologix received Is there any way to tell when the function key routine is complete? In the case of taking multiple readings using the DEFKEY and MATH function, I don't see any indication when the routine is complete. In one particular case, I am taking a 100 readings with NLPC set for 1000 so its a long while before it's complete, but i have to guess when it's done. Thanks, Randy On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans randyevans2...@gmail.com wrote: Bill, I am trying to figure out the MATH function without much success. I input the sequence you said (I looked up the instructions to understand what you did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG; and it shows up on the display when I input BLUE F1. I hit ENTER and it takes the 40 measurements and the MATH symbol shows on the display during the measurements. After the SMPL symbol no longer blinks I hit MATH 2 and I get a MATH ERR symbol on the display. I tried it a couple of times and the same result so I am doing something wrong. Is there a better source for explaining how to do front panel masurements than the User Guide, which seems oriented at programming automatic rather than manual measurements. Randy On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold wpgold3...@att.net wrote: Randy: The MATH function is accessible from the keypad. I don't have an IEEE interface right now that works. You can also program the numeric keypad keys to have preprogrammed functions. DEFKEY I have made my own low thermal measurement leads from Pomona #4892 banana plugs and Belden #9272 wire. Why 9272, because it was handy at the time. It is tin plated copper, shielded twisted pair 20 ga. I have plans to do custom cables with 16 ga. bare copper wire that I will twist and then put a braided shield over it. I simply cannot find what I want so I will build my own cable. I have done something like this before and it worked fine. When I get a round toit. I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have used in the past to verify my homemade low thermal leads as described above. Frankly I cannot see any difference between using the 1756 cables and my homemade cables once I give them a few minutes for the thermals to go away. As far as I can tell and measure the differences, if any, are below 0.1 ppm at 10 volts. Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are all adjustable you may be seeing a misadjusted 1 volt from the 732A. As far as the instability of the readings it is hard to determine which is causing the problem. I have programed (DEFKEY) a numeric keypad key #1 with the following code. MATH 14;NRDGS 40;TRIG 4;TRIG ; So what this does is set the MATH to Statistics (store high reading/low reading/ and mean of the readings) in the registers, the number of readings to 40, the trigger to hold (which keeps the meter from triggering until I
Re: [volt-nuts] 732A and Prologix received
Randy: Sorry my fault. You have to use the RMATH command to get the various values stored in the registers. See the RMATH command in the User's Guide for a list of what registers you can read. I sure haven't found any other guides other than the 4 manuals. User's Guide, Quick Reference Guide, Calibration Manual, and Assembly Level Repair. It is just a matter of reading the guides and trying to remember what commands are available. It took me a lot of time to figure out what commands I use now. I am sure I am missing other commands that might be useful. Bill - Original Message - From: Randy Evans randyevans2...@gmail.com To: Discussion of precise voltage measurement volt-nuts@febo.com Sent: Sunday, August 24, 2014 6:04 PM Subject: Re: [volt-nuts] 732A and Prologix received Bill, I am trying to figure out the MATH function without much success. I input the sequence you said (I looked up the instructions to understand what you did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG; and it shows up on the display when I input BLUE F1. I hit ENTER and it takes the 40 measurements and the MATH symbol shows on the display during the measurements. After the SMPL symbol no longer blinks I hit MATH 2 and I get a MATH ERR symbol on the display. I tried it a couple of times and the same result so I am doing something wrong. Is there a better source for explaining how to do front panel masurements than the User Guide, which seems oriented at programming automatic rather than manual measurements. Randy On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold wpgold3...@att.net wrote: Randy: The MATH function is accessible from the keypad. I don't have an IEEE interface right now that works. You can also program the numeric keypad keys to have preprogrammed functions. DEFKEY I have made my own low thermal measurement leads from Pomona #4892 banana plugs and Belden #9272 wire. Why 9272, because it was handy at the time. It is tin plated copper, shielded twisted pair 20 ga. I have plans to do custom cables with 16 ga. bare copper wire that I will twist and then put a braided shield over it. I simply cannot find what I want so I will build my own cable. I have done something like this before and it worked fine. When I get a round toit. I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have used in the past to verify my homemade low thermal leads as described above. Frankly I cannot see any difference between using the 1756 cables and my homemade cables once I give them a few minutes for the thermals to go away. As far as I can tell and measure the differences, if any, are below 0.1 ppm at 10 volts. Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are all adjustable you may be seeing a misadjusted 1 volt from the 732A. As far as the instability of the readings it is hard to determine which is causing the problem. I have programed (DEFKEY) a numeric keypad key #1 with the following code. MATH 14;NRDGS 40;TRIG 4;TRIG ; So what this does is set the MATH to Statistics (store high reading/low reading/ and mean of the readings) in the registers, the number of readings to 40, the trigger to hold (which keeps the meter from triggering until I press ENTER and then trigger the sequence of 40 readings when I push the ENTER button. You can do all of this manually from the keypads but since I use this sequence a lot I have preprogrammed it. This is after I set digits to 8 and PLC to 100. Once those 40 readings are finished then you can access the various MATH statistic registers, using the menu, by entering MATH and then a 2 for low, a 4 for mean, and 13 for high. Of course you could do all of this through the IEEE also. The 3458A has a very rich set of measurement commands. I am still learning all of them. It depends upon what I am trying to accomplish. Since the 1.018 and 1.0 volt outputs are passive and derived from resistive dividers from the 10 volt, I don't see how they could contribute to the varying readings you are measuring. I think I would put a short on the input of the 3458A and manually set the range to 1 volt and then observe the variations that way without the 732A involved. When I do this I see a variation from low reading to high reading of 0.125 uVolts and then another 40 I get 0.155 uVolts. This is without the GUARD connected to the low side of the measurment terminals, GUARD connected doesn't seem to affect the readings. So that is the base noise of the 3458A without the 732A, somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output I got a variation of 0.159 uVolts using the same 40 reading method above. I would use this to determine where your problem might exist. Just having the meter input shorted will point you in the right direction. Meter, cables
Re: [volt-nuts] Access to Volt-Nuts Web site
Randy: It doesn't work for me either. Same error message. Must be down for some reason. Time-Nuts is down also. Bill - Original Message - From: Randy Evans randyevans2...@gmail.com To: volt-nuts@febo.com Sent: Monday, August 25, 2014 7:28 AM Subject: [volt-nuts] Access to Volt-Nuts Web site For some reason I can no longer access the Volt-Nuts web site. Has anything changed on the permissions list? Is anyone else having this problem? I get the following message: You don't have permission to access /pipermail/volt-nuts/ on this server. Thanks, Randy ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] AC calibration
There is one more thing that enters into this discussion and that is reversal errors on the DC. The complicates the transfer somewhat. AC is always going + and -. DC is in one direction so you have to then reverse the voltage to the Thermal Transfer Standard and then take the average of the two readings. That is why there is a Reversal switch on the 540B. When you are using a fixed voltage High Frequency Thermal Converter you need an external DC reversal switch in addition to other equipment. You also need an AC/DC transfer switch so that you don't have to disconnect the AC source and then hook up the DC source manually. See the 540B again. All Thermal Transfer Standards have some reversal error. This is controlled by the internal construction of the unit and exactly where the glass isolation bead is located on the heating element. The thermocouple converter used in the 540B is selected to have a very low reversal error, but always will have some error. The error is fixed so you can approximate a DC measurement once you have characterized the particular converter. I can't remember now but I think there can be up to around .05% reversal errors on some converters, while the ones selected for the 540B are under .01%. Read the FLUKE Calibration: Philosophy in Practice for further information. Bill - Original Message - From: Dave M dgmin...@mediacombb.net To: Discussion of precise voltage measurement volt-nuts@febo.com Sent: Monday, August 25, 2014 11:38 AM Subject: Re: [volt-nuts] AC calibration Well, you sort of answered your own question. The equipment is called a Thermal Transfer Standard, but instead of thermistors, it uses a thermocouple. Look at the manual for the Fluke 540B (http://bama.edebris.com/manuals/fluke/540b/) and you'll see how it's done. Basically, the AC source is input into the transfer standard, and the standard's internal reference voltage is adjusted for a null on the galvanometer. Leaving the reference voltage setting alone, a DC voltage is input into the unit, and the DC source is adjusted for a null on the galvanometer. At that point, the AC voltage source is equal to that of the DC voltage source. Ther are thermocouple-type thermal converters used for RF voltage measurements with the transfer standard. They aren't cheap, and you have to have a converter for each range of voltages that you need to measure. The thermal converters used with this type of transfer standard isn't great (50 MHz or so typical), but their accuracy far surpasses that of the thermistor type sensors. There are other brands and models of thermal transfer standards, but I have a Fluke model 540 and a few thermal converters. That's why I referred you to the manual for it. Cheers, Dave M pa4...@gmail.com wrote: Is there a way to link an AC voltage to a DC source for compare. I can check my calibrators (like a Fluke 332, 760 , 731 and a Philips) against standardcells. But for AC I can not do that. I have two AC+DC TRMS 7,5 digit meters but the last calibration was 2 years ago. My idea is in theory simple. It is based on the thermal converters used in RF powermeters. Two resistors, two high resolution temperature meters. AC on the first en DC on the second. If both are the same temperature the AC voltage is the same as the DC voltage. But I'm sure some people here have done this in the past. I would like to use it for 50 to 100 kHz (or less) and something like for 1V, 10V and 100V (and use several resistors/heaters.) Or mabey there is an other way to convert AC (for RF it can be done with lightbubs but I never tryed that) I do not mind if it is slow etc, I like this sort of experiments. You can learn a lot from it. Fred, pa4tim ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] 732A and Prologix received
Not that I know of. Just wait and when you don't see a SMPL on the display it is done. But then with 1000 PLC that is around 16.66 seconds per reading times 100 readings is somewhere around 28 minutes and there is probably some overhead time so around 30 minutes. Not from the front panel at any rate. Bill - Original Message - From: Randy Evans randyevans2...@gmail.com To: Discussion of precise voltage measurement volt-nuts@febo.com Sent: Monday, August 25, 2014 9:23 PM Subject: Re: [volt-nuts] 732A and Prologix received Is there any way to tell when the function key routine is complete? In the case of taking multiple readings using the DEFKEY and MATH function, I don't see any indication when the routine is complete. In one particular case, I am taking a 100 readings with NLPC set for 1000 so its a long while before it's complete, but i have to guess when it's done. Thanks, Randy On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans randyevans2...@gmail.com wrote: Bill, I am trying to figure out the MATH function without much success. I input the sequence you said (I looked up the instructions to understand what you did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG; and it shows up on the display when I input BLUE F1. I hit ENTER and it takes the 40 measurements and the MATH symbol shows on the display during the measurements. After the SMPL symbol no longer blinks I hit MATH 2 and I get a MATH ERR symbol on the display. I tried it a couple of times and the same result so I am doing something wrong. Is there a better source for explaining how to do front panel masurements than the User Guide, which seems oriented at programming automatic rather than manual measurements. Randy On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold wpgold3...@att.net wrote: Randy: The MATH function is accessible from the keypad. I don't have an IEEE interface right now that works. You can also program the numeric keypad keys to have preprogrammed functions. DEFKEY I have made my own low thermal measurement leads from Pomona #4892 banana plugs and Belden #9272 wire. Why 9272, because it was handy at the time. It is tin plated copper, shielded twisted pair 20 ga. I have plans to do custom cables with 16 ga. bare copper wire that I will twist and then put a braided shield over it. I simply cannot find what I want so I will build my own cable. I have done something like this before and it worked fine. When I get a round toit. I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have used in the past to verify my homemade low thermal leads as described above. Frankly I cannot see any difference between using the 1756 cables and my homemade cables once I give them a few minutes for the thermals to go away. As far as I can tell and measure the differences, if any, are below 0.1 ppm at 10 volts. Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are all adjustable you may be seeing a misadjusted 1 volt from the 732A. As far as the instability of the readings it is hard to determine which is causing the problem. I have programed (DEFKEY) a numeric keypad key #1 with the following code. MATH 14;NRDGS 40;TRIG 4;TRIG ; So what this does is set the MATH to Statistics (store high reading/low reading/ and mean of the readings) in the registers, the number of readings to 40, the trigger to hold (which keeps the meter from triggering until I press ENTER and then trigger the sequence of 40 readings when I push the ENTER button. You can do all of this manually from the keypads but since I use this sequence a lot I have preprogrammed it. This is after I set digits to 8 and PLC to 100. Once those 40 readings are finished then you can access the various MATH statistic registers, using the menu, by entering MATH and then a 2 for low, a 4 for mean, and 13 for high. Of course you could do all of this through the IEEE also. The 3458A has a very rich set of measurement commands. I am still learning all of them. It depends upon what I am trying to accomplish. Since the 1.018 and 1.0 volt outputs are passive and derived from resistive dividers from the 10 volt, I don't see how they could contribute to the varying readings you are measuring. I think I would put a short on the input of the 3458A and manually set the range to 1 volt and then observe the variations that way without the 732A involved. When I do this I see a variation from low reading to high reading of 0.125 uVolts and then another 40 I get 0.155 uVolts. This is without the GUARD connected to the low side of the measurment terminals, GUARD connected doesn't seem to affect the readings. So that is the base noise of the 3458A without the 732A, somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output I got
Re: [volt-nuts] 732A and Prologix received
Todd: I check the batteries every month per the 732A manual. Usually I keep a few used cells from previous packs around that seem to be ok after checking them for capacity just like checking the battery pack. I can usually spot cells going bad with that method. If I don't have a spare set then I just substitute a used cell until I receive a new set. I agree with you that the cheap ones are crap. I have used Power Sonic a lot with some a few EaglePicher and one set of Enersys. The EaglePicher are crap so I will never use them. EaglePicher have had a couple out of 8 that have failed very early. I have never tried Panasonic. So I stick with the Power Sonic's. I am not sure why the 12 volt ones would be better than the 6 volts because 12 volt is just 6 lead acid cells while the 6volt are 3 lead acid cells. If one of the 6 cells in the 12 volt battery go bad then the whole battery has to be changed. So you better buy good quality 12 volt batteries. I just can't seem to source batteries locally easily at a reasonable price. So I order them from Allied usually. Bill - Original Message - From: Todd Micallef tmical...@gmail.com To: Discussion of precise voltage measurement volt-nuts@febo.com Sent: Sunday, August 24, 2014 8:13 AM Subject: Re: [volt-nuts] 732A and Prologix received Bill, I use a lot of the 12v 7ah batteries for my UPS backups and 732B. But yes, I was mistaken. The 12V 5ah batteries are the ones that I am using in one of my 732A. There is more play inside the tray with the 12v batteries by several mm as compared to the 6v which only has a 2-3mm. The 12v conversion is not difficult, but it is easier if the battery tray has already been machined for the 12v batteries. The battery tabs don't line up well with the existing holes and need to be widened. Two additional holes must also be added. Sourcing the batteries locally is more of a convenience than waiting for the delivery. The el-cheapo 6v batteries are a waste of money and I have 8 batteries so far to prove it. They died the first time I had a 6hr outage and would not hold a charge after that. Todd On Sat, Aug 23, 2014 at 1:21 PM, Bill Gold wpgold3...@att.net wrote: Randy: I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and others that are in this size and package. Order from one of the usual electronics distributors like Allied, Mouser, Digikey. This is a very common battery as it is used in a lot of EXIT signs so they are lighted when the power goes out. I don't see how the 12V 7AH will fit as they are too large. I guess you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4 AH but the terminals are in the wrong place so you will have to nibble out the aluminum plate that holds them in the 732A battery pack. You have to be careful if you use the 12v 5AH as you will have 4 extra battery connection leads to deal with and connect correctly. I would stick with the 6V 4AH. New batteries will last around 12 to 14 hours before the CAL light goes out when AC power is not applied. So shipping to Cal Lab can be a problem if it is a distance away, or you have to use a shipper like UPS or FEDEX and you ship the night before and then use their Morning delivery and the Cal Lab is expecting your 732A. Same on the way back to you. Of course you could always strap another battery on the 732A and hook it up to the ext power plug to last longer. I have seen it done. The issue is to get the Cal Lab to charge the extra battery before they ship the 732A back to you. When you remove and work on the battery pack always have the AC power plugged in. The CAL led will stay on because the 18.6 v regulated supply is working. The CAL light is to indicate that power has not been lost to the Reference Amp or other associated circuits. When the raw supply (battery) voltage drops below about 21 volts the CAL light will go out. Below that voltage the heater circuits will not work correctly and the 18.6 volt regulated supply will not regulate. The requirement is that the Reference Amp be kept alive at all times to maintain the output voltage that was measured at the time of the most recent Calibration or Certification. When the semiconductor junctions are unbiased and cool off when power is lost, and then power is restored the result will be a different 10 volts than before the power failure. My experience is that after all of the years that these units have been powered up, this won't happen and when power is lost and then restored, even months later, the 732A will come back to almost exactly the same 10 volts as when they lost power, usually with in 0.2 PPM after 24 hours of warm up. What type of hookup leads are you using when measuring the 1 volt output? What is the PLC set to? I always
Re: [volt-nuts] 732A and Prologix received
Randy: The MATH function is accessible from the keypad. I don't have an IEEE interface right now that works. You can also program the numeric keypad keys to have preprogrammed functions. DEFKEY I have made my own low thermal measurement leads from Pomona #4892 banana plugs and Belden #9272 wire. Why 9272, because it was handy at the time. It is tin plated copper, shielded twisted pair 20 ga. I have plans to do custom cables with 16 ga. bare copper wire that I will twist and then put a braided shield over it. I simply cannot find what I want so I will build my own cable. I have done something like this before and it worked fine. When I get a round toit. I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have used in the past to verify my homemade low thermal leads as described above. Frankly I cannot see any difference between using the 1756 cables and my homemade cables once I give them a few minutes for the thermals to go away. As far as I can tell and measure the differences, if any, are below 0.1 ppm at 10 volts. Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are all adjustable you may be seeing a misadjusted 1 volt from the 732A. As far as the instability of the readings it is hard to determine which is causing the problem. I have programed (DEFKEY) a numeric keypad key #1 with the following code. MATH 14;NRDGS 40;TRIG 4;TRIG ; So what this does is set the MATH to Statistics (store high reading/low reading/ and mean of the readings) in the registers, the number of readings to 40, the trigger to hold (which keeps the meter from triggering until I press ENTER and then trigger the sequence of 40 readings when I push the ENTER button. You can do all of this manually from the keypads but since I use this sequence a lot I have preprogrammed it. This is after I set digits to 8 and PLC to 100. Once those 40 readings are finished then you can access the various MATH statistic registers, using the menu, by entering MATH and then a 2 for low, a 4 for mean, and 13 for high. Of course you could do all of this through the IEEE also. The 3458A has a very rich set of measurement commands. I am still learning all of them. It depends upon what I am trying to accomplish. Since the 1.018 and 1.0 volt outputs are passive and derived from resistive dividers from the 10 volt, I don't see how they could contribute to the varying readings you are measuring. I think I would put a short on the input of the 3458A and manually set the range to 1 volt and then observe the variations that way without the 732A involved. When I do this I see a variation from low reading to high reading of 0.125 uVolts and then another 40 I get 0.155 uVolts. This is without the GUARD connected to the low side of the measurment terminals, GUARD connected doesn't seem to affect the readings. So that is the base noise of the 3458A without the 732A, somewhere below .2uVolts. When hooked up to the 732A 1.0 volt output I got a variation of 0.159 uVolts using the same 40 reading method above. I would use this to determine where your problem might exist. Just having the meter input shorted will point you in the right direction. Meter, cables or 732A. Sorry for the long dissertation. Friends get mad at me for being so detailed sometimes. Bill - Original Message - From: Randy Evans randyevans2...@gmail.com To: Discussion of precise voltage measurement volt-nuts@febo.com Sent: Sunday, August 24, 2014 7:22 AM Subject: Re: [volt-nuts] 732A and Prologix received Bill, I have convinced myself that the problem I an seeing is due to thermals. If I move the cables (with gold-plated banana plugs) using a small towel rather than letting my hand touch the plugs, it is much more stable. If I then hold the banana plug with my hand after the reading has stabilized, the reading drifts rapidly upward. I am trying to check the stability of the reading but I haven't figured out the MATH function yet. I assume this is a programmed function using GPIB only? The stability I am seeing by manually recording readings (using NLPC of 100 and 1000) is much greater than what you are measuring on your system. Not sure how to ascertain if it's the 3458A or the 732. The value of the readings are very different between the two - the 3458 reads about 50 uV high on the 10 V output and about 12 uV low on the 1V output. Rather large differences (this is after an ACAL). I need to find some better cables to make sure the errors are not due to thermals again. Randy On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold wpgold3...@att.net wrote: Randy: I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and others that are in this size and package. Order from one of the usual electronics distributors like Allied, Mouser, Digikey. This is a very common battery as it is used
Re: [volt-nuts] 732A and Prologix received
Randy: I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit perfect. Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and others that are in this size and package. Order from one of the usual electronics distributors like Allied, Mouser, Digikey. This is a very common battery as it is used in a lot of EXIT signs so they are lighted when the power goes out. I don't see how the 12V 7AH will fit as they are too large. I guess you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4 AH but the terminals are in the wrong place so you will have to nibble out the aluminum plate that holds them in the 732A battery pack. You have to be careful if you use the 12v 5AH as you will have 4 extra battery connection leads to deal with and connect correctly. I would stick with the 6V 4AH. New batteries will last around 12 to 14 hours before the CAL light goes out when AC power is not applied. So shipping to Cal Lab can be a problem if it is a distance away, or you have to use a shipper like UPS or FEDEX and you ship the night before and then use their Morning delivery and the Cal Lab is expecting your 732A. Same on the way back to you. Of course you could always strap another battery on the 732A and hook it up to the ext power plug to last longer. I have seen it done. The issue is to get the Cal Lab to charge the extra battery before they ship the 732A back to you. When you remove and work on the battery pack always have the AC power plugged in. The CAL led will stay on because the 18.6 v regulated supply is working. The CAL light is to indicate that power has not been lost to the Reference Amp or other associated circuits. When the raw supply (battery) voltage drops below about 21 volts the CAL light will go out. Below that voltage the heater circuits will not work correctly and the 18.6 volt regulated supply will not regulate. The requirement is that the Reference Amp be kept alive at all times to maintain the output voltage that was measured at the time of the most recent Calibration or Certification. When the semiconductor junctions are unbiased and cool off when power is lost, and then power is restored the result will be a different 10 volts than before the power failure. My experience is that after all of the years that these units have been powered up, this won't happen and when power is lost and then restored, even months later, the 732A will come back to almost exactly the same 10 volts as when they lost power, usually with in 0.2 PPM after 24 hours of warm up. What type of hookup leads are you using when measuring the 1 volt output? What is the PLC set to? I always use 100 PLC to measure this. If you don't have low thermal connection leads you can experience uV changes for a minute or more after plugging in the leads due to the thermals generated because of the difference in temperature between the banana jacks on the 732A and the banana plugs of the leads. I have found that even just plugging in the lead will generate a thermal difference because of difference of temps and some heating due to the physical act of just inserting the plug because of friction between the jack and plug (my theory at any rate). You have to allow at least a minute or more before being able to make a measurement after plugging in the leads. I just measured the variation of the 1 volt output of my 732A and using my 3458A and I got a total difference of 0.159 uV over 40 measurements using 100 PLC on the 1 volt range of the 3458A. Using the MATH function and all of the data you can collect. That was after waiting for several minutes after plugging in the leads. I hope all of this helps. Bill - Original Message - From: Randy Evans randyevans2...@gmail.com To: Discussion of precise voltage measurement volt-nuts@febo.com Sent: Friday, August 22, 2014 9:03 PM Subject: Re: [volt-nuts] 732A and Prologix received Todd, Thanks for the info. I have several Panasonic 12V 7 AH batteries that I keep topped off and they have very low current draw (~2 to 3 mA at 13.5 VDC) when charged and at their float voltage, so I am pretty sure they are in good condition. I will look at getting those in the units after I ascertain the condition of the 732. So now I have a what appears to be a functioning 3458A and a 732A but they slightly disagree. I am like the man with two watches that disagree on the time - which is correct? For the moment, i am only concerned with stability. The need for absolute accuracy will come later. Randy On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef tmical...@gmail.com wrote: Randy, You have two possible choices. It can be configured with 4 x 6v 4Ah batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner has modified the battery pack already. A couple of mine needed a nibbler tool to remove enough of the aluminum cover that fits over the tops of the batteries. The original cover will short
Re: [volt-nuts] HP 3458A Mem test 1 hight. How do you read 32K memorychips?
When I removed the original NVRAM devices and then put in sockets, I plugged the original devices back in just to see if everything worked. Apparently some cal data was lost just desoldering the devices as I got error messages. It really didn't concern me as I intended to completely recal the meter with my local standards anyhow. It was hours before I plugged the original devices back in so they were completely at room temp. My guess was that since the date codes were 1992 and I did this in 2010 the batteries were almost dead after 18 years and just the shock of desoldering caused the end. Bill - Original Message - From: Mark Sims hol...@hotmail.com To: volt-nuts@febo.com Sent: Wednesday, August 20, 2014 7:20 PM Subject: [volt-nuts] HP 3458A Mem test 1 hight. How do you read 32K memorychips? The problem with removing the chip and copying it is the very real chance of glitching the contents in the process. Much better to make a backup copy first. The 32Kx8 NVRAMS are used for system memory and things like storage of user programs and data. Note that the two 32Kx8 devices are form a 16-bit word. The 2kx8 cal ram is only on the high byte of the data bus. MREAD returns memory contents as a signed integer value (-32768..32767) in ASCII. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] HP 3458A Mem test 1 hight. How do youread 32K memorychips?
Correct, I didn't read the NVRAMS. I didn't have a prom reader/writer that would handle these devices anyhow and no GPIB interface that worked anymore. Actually I have two 3458A. Once I was completely done with the first one including the cal I didn't even try to put the original NVRAMS removed from the second meter, just put in the new ones and proceeded to the cal procedure. Bill - Original Message - From: J. L. Trantham jlt...@att.net To: 'Discussion of precise voltage measurement' volt-nuts@febo.com Sent: Thursday, August 21, 2014 7:53 AM Subject: Re: [volt-nuts] HP 3458A Mem test 1 hight. How do youread 32K memorychips? Bill, Thanks for the insight. I always thought that I lost my CALRAM data by trying to read the chip in my programmer while the chip was still warm. What you are saying is that you did not try to read your chip, just unsoldered, installed a socket, plugged it back in and the data was corrupted. Very interesting. I have done this to three 3458A's and only lost the data on one of the units. This further supports the need to be able to read the contents before unsoldering and have that data be able to be used to program a new chip. Thanks for the insight and I would love to try some programs to do that. I only have Windows systems but I have several USB to GPIB adapter choices, PCI to HPIB adapter choices, and RS232 to GPIB adapter choices. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Bill Gold Sent: Thursday, August 21, 2014 9:33 AM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] HP 3458A Mem test 1 hight. How do you read 32K memorychips? When I removed the original NVRAM devices and then put in sockets, I plugged the original devices back in just to see if everything worked. Apparently some cal data was lost just desoldering the devices as I got error messages. It really didn't concern me as I intended to completely recal the meter with my local standards anyhow. It was hours before I plugged the original devices back in so they were completely at room temp. My guess was that since the date codes were 1992 and I did this in 2010 the batteries were almost dead after 18 years and just the shock of desoldering caused the end. Bill - Original Message - From: Mark Sims hol...@hotmail.com To: volt-nuts@febo.com Sent: Wednesday, August 20, 2014 7:20 PM Subject: [volt-nuts] HP 3458A Mem test 1 hight. How do you read 32K memorychips? The problem with removing the chip and copying it is the very real chance of glitching the contents in the process. Much better to make a backup copy first. The 32Kx8 NVRAMS are used for system memory and things like storage of user programs and data. Note that the two 32Kx8 devices are form a 16-bit word. The 2kx8 cal ram is only on the high byte of the data bus. MREAD returns memory contents as a signed integer value (-32768..32767) in ASCII. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] HP 3458A Mem test 1 hight. How do you read 32K memorychips?
John: I have replaced the 3 NVRAM devices in two different 3458A meters. I didn't do anything to the devices other than to install sockets to make future replacements easier when that issue comes up again. I did not program them or anything, just put the blank NVRAMS in and turned on the meters. I did get a lot of error messages at turn on because there were no calibration constants in the devices. But after clearing those messages the meter worked normally, but was way off on all readings which was expected. I was worried that the other limit constants that were in the old NVRAM would be reprogrammed into the new NVRAMS but it was all there just all of the calibration constants were 0 for the most part. Since I have a complete set of necessary calibration equipment that is required to completely recalibrate the 3458A I went through all of the calibration commands with the proper standards hooked up to the meter, CAL 0, CAL 10.000, CAL 10.000E3 and SCAL and whatever else was required. When I got done both meters would power on with no problems or complaints about calibration needed or whatever. My point here is that putting new blank NVRAMS into the meter does not cause a Mem test 1 high error. Of course you must calibrate the entire meter or send it to HP/Agilent/Keysight or a local lab if you don't have the necessary equipment to do the cal. So my guess is that you have a problem on your A5 Outguard Controller board with addressing memory that is NOT being caused by replacing the NVRAMS. Or you have a problem with one of the RAMs on that board or something. Unfortunately HP/Agilent/Keysight did not ever publish a full set of error messages and what they may mean and what to do about them. Good luck. Bill - Original Message - From: John Phillips john.philli...@gmail.com To: Discussion of precise voltage measurement volt-nuts@febo.com Sent: Wednesday, August 20, 2014 1:53 PM Subject: [volt-nuts] HP 3458A Mem test 1 hight. How do you read 32K memorychips? Hi, I have removed my 32k memory chips form my 3458A. Doe anyone have an easy way to connect them to a windows computer. USB would be my preferred connection to a 28 and 24 pin socket. I would like to be able to read and write to the chips to validate the data. I have replace the chips and still get Mem test 1 High on start up. I really do not want to spend close to $3000 for KeySight to fix this. -- *John Phillips* ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] RE new 3458A
Randy: Look at page #23 of the 3458A Calibration Manual. Make yourself a 4 Terminal Short as shown and put it in as shown. Before I did an ACAL I had around -000.00025 mVDC. After ACAL I now read +000.2 mVDC. I did the CAL 0 myself so I would expect that the meter should return to a low value, and it does. I do get a variation of +/- 30 nVDC using 100 PLC and just observing the variations. As I remember I have never seen a spec on the ZERO stability over temperature. If I turn on the MATH function and then do 40 measurements with 100 PLC the statistics show: Low reading-70 nVDC Mean reading -28 nVDC High reading+3.5 nVDC Total Variation73 nVDC So that correlates with my visual observation of 60 nVDC. After an hour the room had gone up around 1 degree C. Then I observed -000.00023 mVDC. After another ACAL the reading was again +000.2 mVDC. This particular meter has a negative tempco as the room temp goes up. Obviously do an ACAL before any precision measurements requiring low nanovolts. Go to the Keysight website and go to Technical Support and choose Parts. Then enter in the Part Number 03458-66517 which is the replacement 03458-66507 assembly and you will see the replacement part number on the right hand side. Click on that and you will get the information about the exchange program and so on. Looks like you can just order this part online and pay for it with a credit card, but you have to create or use an existing login account. I needed a new display a few years ago. At that time you could order just the display for around $80. Being extremely good at removing and then inserting and soldering I ordered the part. The problem was that the spacing from top to bottom of the pins had changed. It went from around 1.3 inches to around 1.5 inches. So I had to bend the pins to fit my display board and then get something like 72 pins into the holes on the PC Board. It took hours. This change is probably why HP/Agilent/Keysight doesn't let you just get the display anymore but wants you to get the whole PC Board assembly. It did work just fine once installed. Hope this helps your decision to keep or not. Bill - Original Message - From: Randy Evans randyevans2...@gmail.com To: Discussion of precise voltage measurement volt-nuts@febo.com Sent: Saturday, August 16, 2014 10:36 PM Subject: Re: [volt-nuts] RE new 3458A Interesting note. After the room cooled down from about 79F to 73F, and another ACAL, the meter now reads +000.00035 mVDC, a more reasonable value, although it does bounce around a couple of tenths of a uV. Maybe that is OK? If so, then the only issue would seem to be the display has some faint pixels, which a new display should fix. Randy On Sat, Aug 16, 2014 at 9:59 PM, Randy Evans randyevans2...@gmail.com wrote: The unit seems to be working so far except for one issue. After doing an ACAL, and making sure the Auto Zero is ON, I short the input leads with a copper wire shunt across the inputs and the reading is approximately -000.0023 mVDC. That seems rather high. I would expect the unit to short the input leads internally and force a zero reading during the ACAL. Anyone have any comments on this reading? Thanks, Randy On Sat, Aug 16, 2014 at 3:37 PM, Richard Moore richiem5...@gmail.com wrote: Hi Randy -- sounds like your unit is in cal, based on your measurements of DCV and precision 10k resistor. Using autocal all is recommended before doing precision measurements, and I do that if it's been more than a day or two since last use. The autocal uses the internal Vref and an internal 10K resistor to do cal on everything else, so that tells you what the basic cal procedure is. I just got my 3458 back from Loveland, and that's what they did for me -- warmed it up, then ran autocal, then measured everything against a Fluke 5700, aided by an HP 3325, and another 3458. It has been 5 years since I replaced the display board (no exchange deal was available then AFAIK, so I don't know what's changed) and also the NVRAM board, which was dead, with one with the Snap-cap RAM chips. I did those replacements, then sent it home for cal, which was complete, since all the RAM was new. Now after 5 years, the unit passed all incoming performance tests and was sent back to me without a cal process of any kind. This tells me that an old, well-aged Vref module is a good thing. The 10VDC test had changed by a bit under 5ppm, or roughly 1ppm/year. They have a cal deal -- use code 1.090 -- press them for it -- and that saved me 30% off the normal price. I think this deal lasts until mid-September, so my recent cal ended up at just under $400 including shipping. I'm not sure the deal is available on new or first-time cals; my unit was in their data bank. But this is a long way of saying I don't think you need to send
Re: [volt-nuts] Fluke 732A themistor reading
Randy: I have 6 ea Fluke 732A instruments. The results of being an over the edge and deeply obsessed Volt Nut, they just kept adding (reproducing?) over the years. My readings for the themistors range from 3.407K OHMS to 4.514K OHMS and all work just fine. 3.6K OHMS would seem just fine to me. It really depends upon the final inside temp of the oven assembly. The one with 3.407K OHMS measures around 47.5 degrees C while the others are around 45 C. It is the stability of this reading over time that is very important. Fluke says that it could change as much as 10 ohms per month but mine are rock solid for years and years. Bill - Original Message - From: Randy Evans randyevans2...@gmail.com To: volt-nuts@febo.com Sent: Sunday, August 17, 2014 11:08 AM Subject: [volt-nuts] Fluke 732A themistor reading Can anyone tell me what value to expect for a Fluke 732A thermistor reading? I am looking to get a used 732A and one of the things I was told to look for is to measure the thermistor reading after the unit has heated up for at least an hour. I saw in one of the posts that a typical reading was 4950 ohms. One vendor I asked said he measured about 3600 ohms after a two hour warmup. I was concerned that this is too low. What do you guys think? Thanks, Randy ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] A Fluke 732A: Return it or keep it?
Roy: I use a ( General Cement ) GC 8276 adjustment tool to make the pot adjustments in the 732A. I use the end that has the recessed metal blade. The pots are about 3.5 inches back from the front panel. It sometimes is very difficult to engage the tool slot in the pot. It helps to look through the front panel hole with a small flashlight and observe the position of the slot in the pot ( i.e. 9, 10, 11 o'clock for example ). Then when you insert the adjustment tool you can get close to the point when you can engage the adjustment pot. Sometimes it will seem impossible to engage the tool into the slot, but with patience it can be done. Once you do engage the pot do whatever adjustment you want, don't pull the tool out, just leave the tool engaged with the pot so that if you want to make another small change you won't have to go through the same problem of trying again and again to engage. I have found that turning the pots may cause a drift which will show up days later due to the pot being dirty or whatever. So what I usually do is to turn the pot back and forth over several revolutions so that I can clean the contact wiper. I usually make an initial adjustment and then turn the pot about 1/32 of a turn in the opposite direction just to relieve any stress that might be there and cause a small change in the output voltage. I usually have very good results with these methods. But I will agree with Joe that sometimes it is better just to leave things alone and just use the known offset from 10 volts when checking or doing cals. Bill - Original Message - From: Todd Micallef tmical...@gmail.com To: Discussion of precise voltage measurement volt-nuts@febo.com Sent: Thursday, June 12, 2014 7:20 AM Subject: Re: [volt-nuts] A Fluke 732A: Return it or keep it? Roy, The 732A uses the 3059Y-1-101 type cermet pot for the 10V adjustment. You can use a simple plastic tweaking tool for adjustment. I believe you will need a fairly long tool. You can shine a light into one of the other holes to see the orientation of the pot. However, many people may prefer to leave the pot as is and just record the readings. It is going to drift anyway and won't stay adjusted for long. The drift may worsen once the pot position has changed too. I have adjusted mine, but only after the broken pots were replaced. Todd On Thu, Jun 12, 2014 at 9:57 AM, R.Phillips phill...@btinternet.com wrote: Hi Joe Sorry if I am retracing steps, but I have recently had my 3458A restored/recal'd by Agilent - its good to have it back. I also have a Fluke 732A which I can now check. Currently it is 10.24 volts, I now feel confident to trim it to the 3458A. I have tried to see the trimmers in each of the three holes - they appear to be some way within, and I am wondering if you could confirm the type of pot. that is used and just how long a trimming device is required. Is this a special tool only available from Fluke ? Regards Roy Phillips. -Original Message- From: J. L. Trantham Sent: Saturday, March 08, 2014 10:09 PM To: 'Discussion of precise voltage measurement' Subject: Re: [volt-nuts] A Fluke 732A: Return it or keep it? Charles, What size batteries did you use for your external pack and where are you located? In other words, how long can your external battery pack 'survive' keeping the 732A 'hot'? Is the external pack recharged at Fluke or does it have to make a round trip on the initial charge? Once I am convinced my unit is stable (which will probably be another six months or so, after I get my 3458A's re-calibrated by Agilent), I hope to be able to send it to Fluke for calibration. Fluke made a transit case (732A-7002) and a Battery Charger and Auxiliary Battery Case (732A-7003). However, I have never seen a picture of either of these. Does anyone have any information on these? How did you make your case? Did you include a charger? Combination of battery pack and shipping container or separate battery pack and shipping container? I was thinking of building a case with built-in charger, fuse, and AC connector and cord such that when the unit arrived at the CAL facility all that would be needed would be to plug it in and allow the external battery pack to recharge while the unit being calibrated was plugged in, recharge the internal batteries, and calibrated. Or does the 732A also charge the external battery pack along with the internal batteries? The connector you need (complete with female contacts) is a Hypertronics P/N D01PB306FSTAH and is in stock at Kensington Electronics at $8.96 each (plus tax and shipping). Only problem is their $50 minimum order. The 'shell' only is P/N D01PB306NT and the female contact is P/N YSK006-010ANH (three needed). http://ecommerce.keiconn.com/hypertronics/D01PB306FSTAH I ordered two of the connectors from Fluke using a P/N that one of
Re: [volt-nuts] What made a HP3458A so expensive
Joe: I would guess that a local JJ would have to be compared with a standard JJ. Since NIST says Ours is correct and yours could be a little more or less than ours, you need to know the difference. But this all get into a really gray area in the 0.01 ppm region. So what are you going to do? Move your JJ to NIST? Not practical! My suggestion about a JJ in a 9.5 digit was just to get a more stable voltage reference. Of course at the present time given the fact that the JJ has to be cooled to below 4.2 K makes this idea more than a little impractical. Of course you could setup to reference the voltmeter to a stand alone JJ in your workshop. See this website for a turnkey unit you can get. www.supracon.comThey have two models, one which uses liquid He and the other has a Cryocooler. I am sure that you can talk the little lady into this as a birthday present? I am not sure what Fluke does to be traceable to NIST but I would imagine that Fluke regularly sends carefully characterized 734A units to NIST in a round robin so that they can accomplish this. There is an application note somewhere that talks about this. Same with the Hz, the ohm and so on, you have to somehow reference it to NIST, to claim you are traceable to NIST. Bill - Original Message - From: Joe Hobart n...@npgcable.com To: Tom Miller tmiller11...@verizon.net; Discussion of precise voltage measurement volt-nuts@febo.com Sent: Saturday, January 18, 2014 4:44 PM Subject: Re: [volt-nuts] What made a HP3458A so expensive Bill, Would a Josephson Junction standard need to be calibrated? Adjusted and maybe compared, yes, but you should not need to calibrate a primary standard? Years ago we had HP Cesium Frequency Standards at work. There were primary standards and good to 4E-12 with no additional calibration. Joe On 1/18/2014 3:37 PM, Tom Miller wrote: Can you imagine what it would cost to get that calibrated? Tom - Original Message - From: Bill Gold wpgold3...@att.net To: Discussion of precise voltage measurement volt-nuts@febo.com Sent: Saturday, January 18, 2014 5:17 PM Subject: Re: [volt-nuts] What made a HP3458A so expensive In fact if you look at a 1982 HP catalog you will see that the 3456A was selling for around $3,700 and given inflation the 3458A is still a pretty good bargain when it was introduced in 1989. In 1989 the 3456A was selling for $4,600 while the 3458A was selling for $5,900. I wonder if there will ever be a 3459A 9.5 digit meter? With a super miniture Josehpson Junction for a reference? Bill - Original Message - From: Tom Knox act...@hotmail.com To: Discussion of precise voltage measurement volt-nuts@febo.com Sent: Saturday, January 18, 2014 1:29 PM Subject: Re: [volt-nuts] What made a HP3458A so expensive Actually in spite of it's high price I feel the 3458A represents a bargain. When it was introduced in 1989 it was $5900. A price it held for decades. Although designed in then 80's it's accuracy is still unsurpassed. The 8846A is much less expensive because it has only about a quarter the parts. With few that are hand selected and/or aged. And all though both meter have well executed designs the result is the 3458A has about a magnitude greater accuracy. And anyone in Metrology can attest to the fact that the cost of accuracy is exponential. But the thought I wanted to contribute to the dialog is the real value of the 3458A is the body of knowledge built around the thousands of 358A's some running continuously for nearly three decades. The characteristics of the 3458A are perhaps the best documented of any electronic instrument ever made. That is priceless. When comparing that body of knowledge to individual units I have found every 3458A is a little different and seem to each have their own personality. In a side note, recently I owned what appeared to be the original 3458A. I cannot remember the serial number, but it had the numbered stickers identifying the boards that can be seen in the original service manual. And it was still working flawlessly. Thomas Knox Date: Sat, 18 Jan 2014 17:41:26 +0100 From: frank.stellm...@freenet.de To: volt-nuts@febo.com Subject: [volt-nuts] What made a HP3458A so expensive Well, the 3458A now is in production for about 25 years , and only around 50k units have been sold. (That's an estimation by Joe Gellers series number collection) As the 3458A is a niche product with homeopathic series volume, the development, verification/validation, special reliability engineering, selection and burn-in measures, QA costs had to be paid per unit in the beginning. To my opinion, especially the 3458A was designed mainly for military requirements (Tamb 55°C). The military was willing to pay a premium (HP = High Price) at that time. So HP was able to realize that price. All that development budget is long paid, and after end of the Cold War, the military does not order
Re: [volt-nuts] What made a HP3458A so expensive
In fact if you look at a 1982 HP catalog you will see that the 3456A was selling for around $3,700 and given inflation the 3458A is still a pretty good bargain when it was introduced in 1989. In 1989 the 3456A was selling for $4,600 while the 3458A was selling for $5,900. I wonder if there will ever be a 3459A 9.5 digit meter? With a super miniture Josehpson Junction for a reference? Bill - Original Message - From: Tom Knox act...@hotmail.com To: Discussion of precise voltage measurement volt-nuts@febo.com Sent: Saturday, January 18, 2014 1:29 PM Subject: Re: [volt-nuts] What made a HP3458A so expensive Actually in spite of it's high price I feel the 3458A represents a bargain. When it was introduced in 1989 it was $5900. A price it held for decades. Although designed in then 80's it's accuracy is still unsurpassed. The 8846A is much less expensive because it has only about a quarter the parts. With few that are hand selected and/or aged. And all though both meter have well executed designs the result is the 3458A has about a magnitude greater accuracy. And anyone in Metrology can attest to the fact that the cost of accuracy is exponential. But the thought I wanted to contribute to the dialog is the real value of the 3458A is the body of knowledge built around the thousands of 358A's some running continuously for nearly three decades. The characteristics of the 3458A are perhaps the best documented of any electronic instrument ever made. That is priceless. When comparing that body of knowledge to individual units I have found every 3458A is a little different and seem to each have their own personality. In a side note, recently I owned what appeared to be the original 3458A. I cannot remember the serial number, but it had the numbered stickers identifying the boards that can be seen in the original service manual. And it was still working flawlessly. Thomas Knox Date: Sat, 18 Jan 2014 17:41:26 +0100 From: frank.stellm...@freenet.de To: volt-nuts@febo.com Subject: [volt-nuts] What made a HP3458A so expensive Well, the 3458A now is in production for about 25 years , and only around 50k units have been sold. (That's an estimation by Joe Gellers series number collection) As the 3458A is a niche product with homeopathic series volume, the development, verification/validation, special reliability engineering, selection and burn-in measures, QA costs had to be paid per unit in the beginning. To my opinion, especially the 3458A was designed mainly for military requirements (Tamb 55°C). The military was willing to pay a premium (HP = High Price) at that time. So HP was able to realize that price. All that development budget is long paid, and after end of the Cold War, the military does not order so many devices anymore, what caused the problems of the TM business (finally = Keysight, urgh!). But as that market is tight, competitors are few, so the price is not going down, instead it's increasing over the years, from $5900 to around $8500. It would be interesting to calcualte the BOM of the 3458A. As they use many custom specific components, it should be relatively expensive. There is no parameter in the HP3458A specifications, that the device has to be powered constantly to meet the specs. Those very high 8ppm/yr. drift might apply only during continuous operation. During power down, the LTZ should not drift at all, as the ageing mechanism is driven by temperature. But there might be (there are indeed) considerable hysteresis effects. I have set the temperature of my HP3458A to ~ 60°C, I shut it down after usage, and the periodical comparison to 3 other references shows a drift of less than 1ppm/year. Frank ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Connector PN
Randy: The 3458A CLIP shows this connector as a Molex 22-17-2052. Looking on Molex this seems to be a current number. Mouser and Digikey don't have any stock. Alliedelect doesn't show anything. Good luck. Bill - Original Message - From: Randy Evans randallgrayev...@yahoo.com To: volt-nuts@febo.com Sent: Wednesday, January 15, 2014 6:54 PM Subject: [volt-nuts] Connector PN Does anyone know the manufacturer and part number of the connectors used on the HP-3458 voltage reference board? thanks, Randy Evans ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Connector PN
update The HP/AGILENT/KEYSIGHT P/N is 1251-6041 and is in stock at $5.13 ea. Bill - Original Message - From: Bill Gold wpgold3...@att.net To: Randy Evans randallgrayev...@yahoo.com; Discussion of precise voltage measurement volt-nuts@febo.com Sent: Thursday, January 16, 2014 7:17 AM Subject: Re: [volt-nuts] Connector PN Randy: The 3458A CLIP shows this connector as a Molex 22-17-2052. Looking on Molex this seems to be a current number. Mouser and Digikey don't have any stock. Alliedelect doesn't show anything. Good luck. Bill - Original Message - From: Randy Evans randallgrayev...@yahoo.com To: volt-nuts@febo.com Sent: Wednesday, January 15, 2014 6:54 PM Subject: [volt-nuts] Connector PN Does anyone know the manufacturer and part number of the connectors used on the HP-3458 voltage reference board? thanks, Randy Evans ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Fluke 732A/AN Manual
Joe: Not to steal Todd's thunder but you can find these files on Didier's ftp site and you can download them from there until he gets a chance to place them on his regular website. I have already got them and they are just like the manual I have. Great job Todd! This is the first time I have seen the latest manual on the 732A/AN. I was thinking about doing this but I don't have a great scanner and it takes time to scan a manual like this. He even has the schematics in B size. ftp.ko4bb.com username u47755086-manuals password manuals Bill - Original Message - From: J. L. Trantham jlt...@att.net To: 'Discussion of precise voltage measurement' volt-nuts@febo.com Sent: Monday, January 06, 2014 7:57 PM Subject: Re: [volt-nuts] Fluke 732A/AN Manual Todd, Looking forward to the 7332A/AN manual but where did you upload it? I looked on Didier's site but did not see anything that looked related to the 732A/AN. Thanks. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of T. Micallef Sent: Monday, January 06, 2014 12:38 PM To: volt-nuts@febo.com Subject: [volt-nuts] Fluke 732A/AN Manual I just received the last couple of missing pages from my manual and uploaded it. I deleted two uploads that are now redundant and was able to scan the schematics with a large format scanner. Everything should be a lot easier to read now. Todd ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Fluke 732A Questions
Joe: A correction on the dropout voltage when the In Cal light goes out, I found my notes and it is 21 volts instead of 22 volts. The way I did this was to remove the battery pack, AC voltage and then just feed in a power supply where the battery pack would have delivered it's voltage to the back-plane mother board. Then I just reduced the voltage until the light turned off. So that means that Fluke was following the battery manufacturers specs of 5.25 volts at the end of the discharge curve. The banana jacks on the battery pack have two jobs. One is to provide a very good chassis ground to the battery pack and the other is probably to guide the pack accurately into the mother board. One other thing I have found out the hard way is that the holes in the top plate of the battery pack where the battery terminals come through are very close to the battery terminals and can short out to the chassis. I have made the holes a LOT bigger to avoid this problem. This really depends upon the brand of batteries you are using but after having to repair the regulator board I would rather be safe than sorry. I have 6 Pomona 1756-48 low thermal leads for when I need to be careful with thermals. I also have made my own low thermal(I think) using Pomona 4892 gold plated double banana plugs and Belden 9272 twisted pair twinax, using a Pomona 1825 for the ground lead. If I need single banana plugs to connect to the 732A I use Pomona 4897 gold plated. I know that this isn't the best possible low thermal because the wire is tin plated but I have looked and I can't find a good source of bare copper twisted pair shielded wire. When I am making measurements with the home made cables I just wait a few minutes for the thermals to settle down. Why Belden 9272? It was the best choice from what was available at work. From what I have seen Pomona was acquired by Fluke and makes special low thermal cables for Fluke that are not available through the Pomona catalog. When you find out from Fluke what services they offer and prices please let all of us know. Bill - Original Message - From: J. L. Trantham jlt...@att.net To: 'Discussion of precise voltage measurement' volt-nuts@febo.com Sent: Tuesday, December 31, 2013 5:13 PM Subject: Re: [volt-nuts] Fluke 732A Questions Bill, I have yet to put my 732A though it's paces but I could connect an external power supply, disconnect the AC supply and the battery module and slowly decrease the supplied voltage and see what happens when. I'll try your measurement algorithm for the battery to see when the 'IN CAL' LED goes out and what the current draws are. If the measured voltage at J10 is 27 V, then the actual battery voltage should be about 27.6 V, accounting for the diode in the circuit (unless the current draw during measurement is so low that no voltage drop occurs across the diode). I found the NIST/NBS Tech Note easily. I think if I decide to send it out for calibration, I'll use Fluke, if they still calibrate the unit. I'll call them later this week of next week to see what services/accessories they might still have available. Yes, I wound up with the 735C. It was easy to replace the NiCd battery pack and bring it 'on scale' by adjusting the 'jumpers' and the front panel adjustment. I now only have to deal with the periodic 'NO CAL' LED indication. I need to study the 732A manual then take a close look at the 735C to see how it works on it. This also brings up the question of 'low EMF cables' for accurate measurement. I have a Fluke 720A K/V Divider on the shelf that I have yet to bring out to check to see if it is operational. Another one of those 'projects' for a later time. Where would one look for 'low EMF cables' to be used in precise measurement experiments and what errors might one expect to find using Pomona gold plated Banana Plug cables (which is as good as I have at this time)? Thanks for the info. Happy New Year to all. Joe ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Fluke 732A Questions
they are for? I always like to keep equipment as 'original' as possible. If I can't find the correct connector, I'll find something similar that will fit in the same hole and use it instead. We'll see what I hear from Fluke and Hypertronics. Also, how do you get your 732A calibrated? Do the batteries last long enough to ship overnight to a facility? Thanks for the info. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Bill Gold Sent: Monday, December 30, 2013 7:03 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 732A Questions Joe: My 732A units vary in thermistor resistance from 3.408K ohms to 4.514K ohms and have been extremely stable to with in +/- 5 ohms for years now. I did measure the oven temperature of the one with the 3.408K ohms just to see what was causing the thermistor resistance to be a little low and came up with about 47 degrees centigrade, as best as my equipment would allow me to measure it. Equipment was a Fluke 80TK thermocouple module, a Fluke 80 PK-1 thermocouple bead and a HP 3456A DVM. I put shrink wrap around the bead and then threaded the bead into the temperature oven through one of the adjustment holes in the front panel and made sure that it was far into the oven, probably around the middle. Then I checked another unit that read 4.5K ohms and it read around 43 degrees C so all of that made sense. I believe I saw somewhere that the spec was 45 degrees C +/- 2 C. I did have to change the jumpers on the Calibration PCB Assembly on one of my units so that it would match other 732A units that were In Cal and had certified values. But that unit has since drifted down at a rate of around 1.2 ppm per year which is with in specs but a little more than I have seen from other units. Recently this unit has suddenly quit drifting down and seems a lot more stable now. That was S/N 459. I have a S/N 343 which has been proven rock solid for years now at around +/- .3 ppm and just seems to have DC noise stability as it just goes up a little and then down a little and never needed any adjustment. I also have a S/N 460 which is extremely stable also. So I guess that age does seem to factor into stability as well as how long they have been powered up. But the only way to insure that you have a good volt is to have at least 4 units and then inter-compare them periodically following NIST (NBS) Technical Note #430. While 430 was written for Saturated Standard Cells, the technique seems to work just fine at 10 volts also. You can also find this measurement technique in the 1st Fluke Calibration - Philosophy in Practice book published around 1974. I gave up on the unobtainium connector on the back of the battery pack very quickly. I drilled a 1/4 holes on either side of the unobtainium connector, after removing it, and used two single miniature banana jacks to allow me to connect to the batteries for the purpose of checking their performance as described in the 732A manual page 2-9 in manual P/N 788414 May 1986. I should probably turn in my resignation as a volt nut for this action but it works. KISS. Bill - Original Message - From: J. L. Trantham jlt...@att.net To: volt-nuts@febo.com Sent: Saturday, December 28, 2013 2:51 PM Subject: [volt-nuts] Fluke 732A Questions I have reviewed the prior postings on Volt-nuts (a very rewarding review, BTW)) regarding the 732A and have two questions: 1.Has anyone determined the part number, or a source, for the mating plug to J10, the external power connector for the 732A-7005 battery pack? I noted some postings recently about this but did not find a definitive response regarding the identity of this plug. The alternative would appear to be a complete replacement of J10 and its plug, as long is it all fits in the opening in the panel. I hate to 'bore holes' in vintage equipment. 2. The oven thermistor in my 'new to me' unit measures 4229 ohms (+/- an ohm or two) after the unit has warmed up for a week or so. I note the manual refers to 3K to 4K for the value of this thermistor when the unit is 'stable' as well as other's posting values in the mid 3500's ohms for their units. Should I be concerned? The unit seems to be stable to within about 2 uV over about a week (as measured by my 3458A - see below). Should I open the unit and try to measure the oven temperature or just be satisfied that the unit seems to be working? I had to replace the four 6V 4AH SLA batteries and they charged up appropriately as judged by the front panel LED's. I had to remove the 'jumper' for the '40' option on the A7 board and connect the jumper to the '20' and '10' options (total of '30') in order to get the unit to adjust to 10.000 VDC on my 'Agilent In Cal'd' 3458A. The need
Re: [volt-nuts] Fluke 732B
Joe: The 731A and 731B were run at room temp and therefore were subject to room variations. They both use the same Reference Amplifier as the 732A B use. I have one 731B and it is amazingly stable for the room temp variations I experience at home. When I was working we had a 731B that was also very good. I had sent it for calibration over a 10 year period and the drift was equal to, or better than the 732 series. At the end of the 10 years it was absolutely stable at less than 1 ppm. This was all being done at a Fluke Service Center that was located in Milpitas California, and their 732As were being routinely sent to Fluke in Everett WA. The difference between the 731 732 is that starting with the 732A all of the temperature sensitive components were enclosed in a oven assembly that had heaters surrounding the oven enclosure which is running at 45 degrees C. This includes all of the resistors and adjustment pots. I have torn down a 732A to the oven assembly and gone inside the assembly to replace a component. It is a real trick to do this, but by being careful it can be done. The fact that the 732As have been in operation for a much longer period of time probably explains the lower drift rates than the 732Bs. The 732As have been aged much longer. I know people who prefer the 732A over the 732B because of this. The only problem with the 732A is that the heaters use more power than the 732Bs. Early manuals spec 24 hour battery operation, but later manuals have changed that to 12 hours. All of the 732As I have observed will only run for about 12 to 15 hours before the in cal light goes out meaning that the batteries have dropped below about 22 volts. The 732B is speced at 72 hours battery operation but I have no experience if this is true. I guess I am a real Volt-Nut in that I have 6 732As at home right now. I inter-compare them once a week. The last time I was able to verify what the official volt is was in 2005. Right now I need to find a good Cal Lab in the South SF Bay Area where I can have one of them certified against their traceable standards. The Fluke service center was closed years ago. Bill ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Fluke 732B
Joe: Thanks for the upgrade, but they are only 732As not Bs. I got all of these at less than the going price of one 732B that recently sold on the infamous action site. Bill ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Fluke 732B
Joe: That was a very interesting teardown of the 732B on eevblog. The one thing that this answered for me was if Fluke was using their Reference Amplifier in all of the 732B production and Fluke did. If you look at a paper by David Deaver called Predictability of Solid State References which he wrote, and you can download from the Fluke website http://us.fluke.com/. you will see on page three of the paper that he talks about the reference drifts of a number of 732Bs that were checked over a period of time. Deaver talks about Motorola References and Linear Technology References and that the Motorola's drift upwards while the Linear Tech's drift downwards over time. Also look at Fig. 10. I was assuming (wrongly of course as I see now) that Fluke had switched from the Reference Amplifier to the LTZ1000 in the production of the 732B. Now I see that it is just what company made the Reference Amplifiers for Fluke. Of course the 4 legged device is a Reference Amplifier that they have used for a least 25 years or more as the voltage reference in a lot of their instruments. I have never seen a schematic of the super secret oven assembly but you can probably bet that it looks a lot like the 732A but with a lot more black magic built into it. I don't think that Fluke would ever expect someone to fix this beyond the unit replacement level as I am sure that this would require factory precedures far beyond anything that we could muster in the field. Part of the black magic is adjusting the collector current of the reference amplifier to obtain a zero TC around a very limited range of operating temperatures. The digital adjustments also add something inside the oven assembly, probably a DAC, and it would be interesting what they are doing there, so a schematic would be interesting to see. Anyway I will quit rambling. So many things to do, so little time. Bill ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] (New to me) 3458A Trivia
Speaking of the 03458-66529 reference board added to the HP 3458A by Fluke (before Fluke offered the 8508A), has anyone ever seen a schematic of that board or how Fluke could rate that reference at 2 ppm? One person on this forum theorized that by lowering the operating temperature of the LTZ1000A to around 70 degrees C, which is what the LTC device specification note recommends, this might have what was done to achieve the better drift specification. It was noted that HP was running the LTZ1000A at around 105 degrees C which might accelerate the aging rate. I lowered my references to around 70 C and they have seemed to be a little more stable when checked against the Fluke 732A DC Reference Standard. But I never started a project to follow up on this theory. Just another project to play with when I get a round toit Bill ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.