Re: [PSES] Thoughts on what may be the emission source

2020-11-27 Thread Jon Griver
overclocking options to see what affects the emissions. Best Wishes, Jon Griver On Fri, 2020-11-27 at 17:49 -0600, DEREK WALTON wrote: Hi Folks, I’m working on a product containing a Raspberry PI as the main controller. I see two very strong emissions at about 240 MHz and 480

Re: [PSES] ISO 13485-2016

2017-01-18 Thread Jon Griver
comply with ISO 13485:2003, EN ISO 13485:2012 and ISO 13485:2016, together with whatever other regulatory requirements you need. Regards, Jon Griver http://www.601help.com The Medical Device Designer's Guide to IEC 60601-1 From: Stephen Whalen [mailto:scwha...@bellsouth.net] Sent

Re: [PSES] One DoCs vs multiple DoCs

2015-11-05 Thread Jon Griver
ar on the MDD Declaration, and if the RoHS Declaration was included in the same Declaration, then it may be assumed that the Notified Body was involved in the supervision of the manufacturer's compliance with RoHS, which it is not. Jon Griver www.601help.com -Original Message- From: N

Re: [PSES] Medical Device approval for U.S. market

2015-06-11 Thread Jon Griver
likely to have more stringent requirements for the proof of clinical efficacy. You will find all the requirements on the FDA’s site. I would start here: http://www.fda.gov/MedicalDevices/DeviceRegulationandGuidance/Overview/default.htm Regards, Jon Griver http://www.601help.com The

Fw: Grounding

2003-11-15 Thread Jon Griver
tinuity test of clause 18. In this case additional earthing arrangements, beyond a hinge, are not required. Jon Griver http://www.601help.com The Medical Device Designers Guide to IEC 60601-1 From: "Crabb, John" To: "Ned Devine" ; "IEEE EMC/Product Safety (IEEE, EMC/Pro

Equipment Calibration

2003-11-08 Thread Jon Griver
Corrective Action proposal to amend your current calibration procedures in order to reduce calibration costs, without in any way compromising quality or ISO 9001 compliance. This should bring the problem to the attention of Quality Assurance and bypass the interests of the your Metrology department

Equipotential Terminal - Yellow/Green Disc?

2003-11-03 Thread Jon Griver
60601-1, which just specifies the marking which is required adjacent to the terminal, but does not specify the form of the terminal itself, or when it should be used. Regards, Jon Griver http://www.601help.com The Medical Device Designer's Guide to IEC 60601-1 This message is from the IEE

Fw: centrifuges and the MDD

2003-10-15 Thread Jon Griver
it "intended by their manufacturer to be used for in vitro diagnostic examination", then it falls under the IVD. If not, and the same model is also sold for non-medical use, then its LVD. Regards, Jon Griver http://www.601help.com The Medical Device Developer's Guide to IEC 60601-1

Re: Czech medical grounding?

2003-06-29 Thread Jon Griver
onnection caused the enclosure to be earthed at a potential different from that of other equipment to which the patient may be connected. Regards, Jon Griver http://www.601help.com The Medical Device Developers Guide to IEC 60601-1 This message is from the IEEE EMC Society Product Safety Technica

Re: Medical grade PSU and Grounding techniques

2003-04-07 Thread Jon Griver
pplies are not designed to be in direct conductive connection with the patient. Additional insulation is usually required. 5. Red or yellow indicators must not be used for the ON/OFF switch. Green is usually used. Regards, Jon Griver http://www.601help.com The Medical Device Designer's Guide

Term in Electricity

2003-03-18 Thread Jon Griver
t at the transformer, except at the load end? I am interested in both US and UK terms. Regards, Jon Griver http://www.601help.com The Medical Device Developers Guide to IEC 60601-1 This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit ou

Re: D of C for medical P.S.

2003-02-26 Thread Jon Griver
the LVD, specifying IEC 60601-1 as the applied standard. You can present your comparison table to anyone who questions this, demonstrating the engineering basis for your declaration. Regards, Jon Griver http://www.601help.com On 25 Feb 2003 at 13:40, boconn...@t-yuden.com wrote: > >

RE: IEC 61010 requirements

2003-02-22 Thread Jon Griver
er day, I called a surgeon and he happened to be in the operating room with his cellphone > performing an operation. Does that make his cellular comply with IEC 601-1? > Peter > > Jon Griver wrote: > > Delphina, > > It's not only an issue of intended

RE: IEC 61010 requirements

2003-02-21 Thread Jon Griver
h other medical instruments with which he is in contact. This could be particularly nasty if there is a fault in the PC's earth wire. Regards, Jon Griver http://www.601help.com The Medical Device Developers Guide to IEC 60601-1 On 20 Feb 2003 at 12:38, Han, Delphina wrote: > > H

(Fwd) RE: IEC 61010 requirements

2003-02-20 Thread Jon Griver
would probably be acceptable if the instrument is based on a PC. Regards, Jon Griver http://www.601help.com The Medical Device Developers Guide to IEC 60601-1 Hello Greg, I have to disagree with your interpretation. There are many in vitro devices that are classified as medical equipment and

(Fwd) MRI compatible?

2003-02-20 Thread Jon Griver
Brent, Can you clarify your question? What exactly do you mean by MRI compatible? An accessory to an MRI device? Please give some more details. Regards, Jon Griver http://www.601help.com The Medical Device Developers Guide to IEC 60601-1 Subject:MRI compatible? To

Fw: Medical Devices in Japan and Australia

2002-01-27 Thread Jon Griver
.html for Japan > > Any help/direction would be appreciated. > > Thanks, > > Mike Cantwell > > Regards, Jon Griver www.601help.com The Medical Device Developer's Guide to IEC 60601-1 --- This message is from the IEEE EMC So

Fw: Websites for world-wide requirements

2002-01-27 Thread Jon Griver
Paul, I posted this a few days ago, but it only refers to regulatory requirements for medical equipment: http://www.ita.doc.gov/td/mdequip/regulations.html As it's the site of the US Department of Commerce, it may also have regulatory requirements in any fields. Regards, Jon Grive

Fw: 601-1 Factory Tests

2002-01-24 Thread Jon Griver
your production tests are not suitable. There are obviously also legal liability issues involved. You should have a written engineering justification for the tests chosen, as well as written traceable records of production testing performed. Regards, Jon Griver www.601help.com The Medical Device De

Re: Medical devices in Asia

2002-01-23 Thread Jon Griver
Peter, An excellent place to start your search is: http://www.ita.doc.gov/td/mdequip/regulations.html which is the US Department of Commerce site, giving information on the medical device requirements for many countries. Good Luck Jon Griver www.601help.com The Medical Device Developer&#

IEC 60601-1/EN 60601-1 Question

2002-01-22 Thread Jon Griver
general requirements for batteries. Regards, Jon Griver www.601help.com The Medical Device Developers Guide to IEC 60601-1 --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at

Fw: emc for FDA

2001-12-01 Thread Jon Griver
Brian, Most definitely yes, see http://www.fda.gov/cdrh/emc/ The FDA is concerned with the correct operation of medical equipment, so immunity requirements are also included. Regards, Jon Griver www.601help.com The Medical Device Developer's Guide to IEC 60601-1 > > Do FDA specif

Fw: Notified Body Question

2001-11-26 Thread Jon Griver
;s accreditation. I do not believe that a test report produced under EN 45001 accreditation becomes invalid because this standard has now been superceded. Best Regards, Jon Griver www.i-spec.com The Engineer's On-line Guide to EN 60950 > > Hi, > > I have a Notified Body test re

FW: FDA & FCC

2001-10-30 Thread Jon Griver
t question. Ultrasound equipment does not have to be certified to Part 18, but the FCC does consider this equipment to be in the 'domain' of Part 18, and can, for instance, demand to inspect the equipment (under Sec 18.113). Best Regards, Jon Griver http://www.601help.com - The Medic

FW: FDA & FCC

2001-10-29 Thread Jon Griver
sections of Part 18 that deal mainly with allowed frequencies. Best Regards, Jon Griver http://www.601help.com - The Medical Device Developers' Guide to IEC 60601-1 > > Jon, Ultrasound equipment does us use RF directly to treat or diagnose > patients. I don't believe Ultraso

FDA & FCC

2001-10-28 Thread Jon Griver
found to cause interference can be stopped from operating. Medical devices that intentionally use EM radiation are subject to FCC Part 18. This includes ultrasound equipment, diathermy equipment, microwave therapeutic devices. Regards, Jon Griver http://www.601help.com - The Medical Device

FW: standard for Epilation device?

2001-09-04 Thread Jon Griver
applicable. Nevetheless, I believe that due diligence requires you to get a written opinion from an EU MDD Notified Body, or even a Competent Authority (e.g. the Medical Devices Agency in the UK), rather than depending on free advice (like this). Regards Jon Griver http://www.601help.com The

RE: Medical application-Isolation voltage

2001-08-30 Thread Jon Griver
ich is not specifically designed for compliance with the medical standard, is very unlikely to comply with this standard. Once your product is 'front-ended' by a medical grade isolation transformer, you can then use other non-medical components in many cases. I hope this helps. Jon Griver ht

FW: Electroshock exercising devices ...

2001-08-23 Thread Jon Griver
ve would appear to be the most suitable. Regards, Jon Griver http://www.601help.com > > > > Actually, I have experienced such a unit at my former chiropractor, and it > actually felt pretty good to loosen my back before the back adjustment > session. I was really astonis

FW: Manufacturing Hipot Testing

2001-08-20 Thread Jon Griver
n I assume they would be obligatory. Nevertheless, If you have a good engineering reason why the production test in the standard is not as good as a different test you wish to perform, and you document the reasoning, and feel that you have no problem justifying it, then you should perform your preferred te

FW: Secondary Grounding

2001-05-31 Thread Jon Griver
alues, rather than by adding a second earth connection. Best Wishes, Jon Griver http://www.601help.com The Medical Device Developers' Guide to IEC 60601-1 > > Group, > > This is a general product safety question: > > In order to avoid test failures during a sing

RE: Production Line Test Requirements - Medical Devices

2001-05-22 Thread Jon Griver
UL test, which might miss them. Presumably if you are selling your medical product in the EU, you have a Notified Body (unless your product is Group 1). You should agree your production test plan with them. Be aware that they may insist on the high current earth continuity test. Good Luck, Jo

FW: Production Line Test Requirements - Medical Devices

2001-05-20 Thread Jon Griver
Dick, To the best of my knowledge there is no EN standard or guidance on production tests for medical equipment (There is a standard, EN 50116, for IT equipment). Your best bet is to discuss this with your Notified Body and to come to a mutually acceptable set of tests. Regards, Jon Griver

RE: PC power supply with PFC

2001-05-16 Thread Jon Griver
harmonics, but the EU requirement concerns harmonics, not power factor. Best Wishes, Jon Griver Medson Ltd. > > > Dear Colleagues, > > I appreciate if you could help me with my problem please. We use > PC power supply in our control machine for feeding our > motherboard. The en

FW: Medical Directive

2001-04-29 Thread Jon Griver
gent leakage current, spacings and dielectric requirements. Also, depending on the classification of the product, your company may need to have its Quality System approved by a European Notified Body. Regards, Jon Griver Medson Ltd. jo...@medson.com > -Original Message- > Fro

FW: Harmonics -- WSJ opinion.

2001-02-27 Thread Jon Griver
ishes, Jon Griver One comment of Mr Hunter's that stood out in particular was the very last . . . " . . . the only ones who benefit from the harmonic current emission standard are the European electricity distributors. They "avoid investments in bolstering their

RE: 61326-1

2001-02-18 Thread Jon Griver
nsure that you are using the most updated version. Best Wishes, Jon Griver Medson Ltd. VP Quality Assurance > > Good people of the group: > > Are requirements in IEC61326-1 (EMC for lab instruments) similiar > to/same/much different than CISPR11/16 and/or the "good stuff" i

FW: Medical devices

2001-01-17 Thread Jon Griver
r these Directives. It sounds like your product falls under the In Vitro Diagnostic Directive, but check the scope. As an accessory, your product still has to fully comply with the directive. Best Wishes, Jon Griver Medson Ltd. VP Quality Assurance email: jo...@medson.com > Hi all > &g

RE: EMC on Medical Devices?

2001-01-09 Thread Jon Griver
country to country. Hope this helps. Jon Griver Medson Ltd. VP Quality Assurance > Subject: EMC on Medical Devices? > > Looking for a list of standards that apply to EMC on medical > devices and the > markets which they apply ( EU, US, and Asia Pacific). > > Thanks > >

FW: DoC & TCF

2000-10-11 Thread Jon Griver
Richard, DoC is the same except that instead of a list of harmonised standards, you list the name and address of the Competent Body, and the serial number and date of the Certificate they issued to you. Regards, Jon Griver VP Quality Assurance Medson Ltd. > > I have seen s

RE: Verifying functionality of the equipment for Production Safety Testing

2000-09-26 Thread Jon Griver
tter than testing for weeks or months with a faulty lead. Your Quality Manager would be quite pleased if you detect a failure and have to redo a day's testing. He will be able to demonstrate to his auditors how your Quality System picks up and corrects a potentially serious problem. :-) Re

LVD voltage range.

2000-01-13 Thread Jon Griver
e his 'due diligence' with regard to the safety of the product. In this case it may be a safety test report with most paragraphs marked as 'Not Applicable', but that may be the difference between being found criminally negligent, or just liable for damages. Regards, Jon Griver i-Spec.

RE: Y3K

2000-01-05 Thread Jon Griver
As an ex-programmer, I can say that Y3K shouldn't be a problem. But we really do have to be concerned about Y10K when we move from 4 digit to 5 digit years :) Jon Griver i-Spec.com The On-Line Compliance Guide http://www.i-spec.com >Y3K preparedness seminar >For those who t

TCF & DOC

1999-12-20 Thread Jon Griver
do go the TCF route, your DoC should include the name of the Competent Body that assessed your TCF, together with the number and date of issue of the certificate the Competent Body issued. Regards, Jon Griver i-Spec.com Ltd. jgri...@i-spec.com i-Spec, The Online Guide to Compliance http://www.i-

Re: Marketing/Offering for Sale in Europe

1999-12-08 Thread Jon Griver
ious version of this document is available at http://www.rcic.com The wording and section number are different from those quoted above. Maybe someone on the list can give an address for the latest version of this document. I hope this is helpful. Jon Griver i-Spec.com Ltd. i-Spec IT - The Online Guid

SAR reducing antennas

1999-12-02 Thread Jon Griver
Ron, Be very careful with this subject. There is a feedback loop between the remote transmitter and the cell phone. A reduction in the efficiency of the cell phone antenna will cause the cell phone to transmit at higher power, defeating the intention. Regards, Jon Griver i-Spec IT - The IEC

Cell Phone Hazards?

1999-12-02 Thread Jon Griver
t of our way of life. I use a cell phone, though as little as possible, knowing that there is a possible risk, in the same way as I know I risk my life every time I get in my car. Jon Griver i-Spec IT - The IEC 60950 Compliance Guide http://www.i-spec.com >From: geor...@lexmark.com > &g

EN 61000-3-2 and medical equipment

1999-06-21 Thread Jon Griver
lusion of a product group within the scope of an EN standard does not necessarily imply that compliance with that standard is mandatory. Jon Griver ITL (Product Testing) Ltd http://www.itl.co.il - This message is coming from the emc-pstc discussion list. To cancel your subscrip

RE: Medical Device Directive, IEC 601-1-2, and IEC 1000-3-2

1999-06-19 Thread Jon Griver
s not apply the medical devices, so EN 61000-3-2 does not apply. Jon Griver ITL (Product Testing) Ltd. http://www.itl.co.il At 14:25 18/06/99 -0400, you wrote: > >Pat: >61000-3-2 is a horizontal standard and it applies to all products unless >specifically excluded by 61000-3-2.

Medical Device Directive, IEC 601-1-2, and IEC 1000-3-2

1999-06-18 Thread Jon Griver
ctive, not the EMC Directive. So if harmonics are not mentioned in EN >60601-1-2, then it is not required. > > >Jon Griver >ITL (Product Testing) Ltd. >http://www.itl.co.il > > > >> >> >>I was asked the following question, and wasn't sure about t

Israel

1998-03-17 Thread Jon Griver
equired. Application to the Ministry of Communications can only be made by local importer/distributor to ensure local support for product. For further information try: http://www.itl.co.il Regards, Jon Griver ITL (Product Testing) Ltd. Tel: +972-3-5339022 Fax: +972-3-5339019 email: stand...@itl.co.il

One Safety Standard?!

1996-09-27 Thread Jon Griver
y specific conclusions on this subject, but I would be interested to hear the opinions of other maling list members. Jon Griver ITL (Product Testing) Ltd. i...@shani.net

One safety standard?!

1996-09-25 Thread Jon Griver
he body, particularly with internal body parts, e.g. via endoscopes, intracardiac probes, etc. The isolation concepts used in IEC 950, IEC 1010 and IEC 65 are sufficiently different from those required for medical equipment to make a unified standard not feasible in this case. Jon Griver ITL (Product Testing) Ltd. i...@shani.net

Re: Power Factor Correction, Medical Equipment?

1996-05-26 Thread Jon Griver
The Medical Devices Directive does not have any specific requirements for Power Factor Correction. The EMC standard for medical equipment, EN60601-1-2, does not call out any limitations on harmonics, a related issue. Jon Griver i...@shani.net