Hmmm... Magnus thank you for your comments. The 5 ns figure is certainly 
helpful. My 1980 manual from HP just says "Time Interval" A -> B = 0 ns to 10E7 
seconds. Not very helpful.

May I ask where you found your 1992 reference? Perhaps you have a link?

Many thanks,

Jim...
N5SPE

On Feb 22, 2014, at 5:20 PM, Magnus Danielson <mag...@rubidium.dyndns.org> 
wrote:

> Hi Jim,
> 
> On 22/02/14 14:17, Jimmy Burrell wrote:
>> I need some help with a 'noob' question regarding some practical examples in 
>> some of the NIST literature. When attempting to compare two clocks, I'm a 
>> bit confused on the subject of exactly how to use my counter to compare a 
>> delayed clock relative to another. Or perhaps I should just say 'comparing 
>> two clocks'. Let's take some concrete examples.
>> 
>> Let's say I want to characterize my Morion MV89 ocxo using my HP5335a. 
>> Obviously, I can tune the MV89's 10MHz by +/- 1Hz and feed it to the 
>> counter's input 'A'. Obviously, I can feed in a second, external reference 
>> clock at 10MHz into input 'B'.  Suppose, however, I didn't have an external 
>> reference clock. Can I compare against the counter's internal time base by 
>> hooking a line from the rear jack time base output to channel 'B' input? Or 
>> am I making it too complicated? Do I simply plug into input 'A' and go?
>> 
>> In a somewhat related question, in this article 
>> (http://www.wriley.com/Examples%20of%201%20PPS%20Clock%20Measuring%20Systems.pdf)
>>  where two clocks, both divided to 1PPS, were compared, W.Riley makes the 
>> following statement, "The two 1 PPS outputs were connected to a Racal Dana 
>> 1992 time internal counter having 1 nanosecond resolution, and the start and 
>> stop signals were separated sufficiently in time for the counter to function 
>> properly".  I wonder what exactly is meant by "separated sufficiently in 
>> time for the counter to function properly" and how one would go about doing 
>> this? For example, is inverting one of the signals sufficient separation? If 
>> not, how is this typically done? Delay line?
> 
> The problem is that the start trigger (Channel A) will arm the measurement on 
> the stop channel (Channel B). For this process to operate properly, allow 5 
> ns. If you look into the 1992 spec-sheet it says that TI interval is from 5 
> ns.
> 
> If you have a 10 MHz or 5 MHz signal, you only have 100 ns or 200 ns periods 
> to play with. Dividing them down to say 10 Hz gives you plenty of time such 
> that you can boot-strap one of the dividers such that you don't go through 
> zero within the measurement period.
> 
> Adding a delay doesn't help, as frequency error will get the phase into 
> through zero condition pretty quick I would guess.
> 
> The SR620 counter has a coax delay-line to give trigger look-ahead.
> 
> Cheers,
> Magnus
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.

_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.