Hi

If you grab the schematic of pretty much any of these counters, you find a 
fairly common approach. 
Somewhere inside is a VHF oscillator. The internal TCXO, OCXO, or Rb acts as a 
phase lock source
for that VHF oscillator. Typical PLL bandwidths are pretty low (10’s of Hz). 
When you put in an external
reference, it acts as the reference to the same PLL. 

This does a couple of things. You take care of spurs on the external reference. 
That is the same 
thing you do locking up a VHF oscillator to your GPSD for microwave work. You 
get the long term
accuracy (1 second and out) of the external reference. Your phase noise (and 
thus broadband mask
jitter) is improved. 

Does every counter on the planet work this way? - of course not. Somewhere 
somebody did it a 
different way. As long as they *did* do it this way, the internal reference 
does not matter once you 
switch to an external standard. Even if they did it in some other way, if they 
did it right, the same 
statement would apply. 

Bob 

> On May 8, 2017, at 9:02 PM, Tom Knox <act...@hotmail.com> wrote:
> 
> Hi All;
> 
> How important is the standard, medium, or high stability reference in 
> counters like the 53230A, or FCA3120 when locked to an ultra low phase noise 
> external reference? Particularly when making measurements and adjustments on 
> other ultra high performance references?
> 
> I know Agilent sold 53132A counters with an option (H01 I believe) that 
> bypassed the internal reference completely when and external reference is 
> applied, but I think standard configuration on most counter is to discipline 
> the internal reference.
> 
> Thanks for your thoughts.
> 
> Thomas Knox
> 
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.

_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.