There are several different part numbers listed in the sensor chart in the pdf.

A number of them are rated to 150C.

also.

It would appear that the measurement would benefit from a measurement of 
pressure inside the reactor in order to confirm the steam is super heated.

Date: Thu, 23 Jun 2011 03:46:29 -0500
Subject: Re: [Vo]:E-Cat vs. Water Heater for coffee/tea...
From: joshua.c...@gmail.com
To: vortex-l@eskimo.com



On Thu, Jun 23, 2011 at 1:50 AM, Finlay MacNab <finlaymac...@hotmail.com> wrote:






The delta ohm probe in question is rated to 150C with an accuracy of +/- 3.5% 
above 95% RH from this spec sheet  
http://www.deltaohm.com/ver2010/uk/st_airQ.php?str=HD37AB1347.



I think you're reading that spec sheet wrong. The 150C refers to the range of 
temperature measurement. But the RH "sensor operating temperature" is given as 
-20 to 80C in the section on common characteristics of RH sensors below the 
table.

So, there appears to be no claim that it can measure even RH at 100C.           
                          