Ralph,

This might be true but that is not how we saw it way back when. The 240VA 
"Energy Hazard" was not a consideration for the protection against Fire but a 
limit value for accessible parts by the User. We still today consider 
accessible circuits, regardless of the voltage, to be "Hazardous Live" if the 
circuit exceeds 240VA. This requirement is not specifically called out in our 
working safety standard (IEC/EN 61010-1 for Laboratory Equipment) but we still 
take this condition under consideration especially with products that exposes 
the user to high currents at low voltages such as Electrode Furnaces (similar 
to a welder).

In Tempest Computers which fell under the IEC950, the hard drives had to be 
made removable so they could be easily taken with during an invasion or 
destroyed in a giant shredder machine. The opening in the front of the computer 
gave the User access to a small backplane card and the data and power 
connectors for the hard drive. The backplane had to be limited to less than 
240VA if the User could touch it.  Fire was a completely different evaluation.

My step dad was working on a car a got his metal watch band between the starter 
solenoid and the chassis. It instantly welded his watch to the car and turned 
the band into a glowing red hot heating element within a second. He was able to 
break it loose and get the watch off but not before he was badly burned. Almost 
required skin grafts. However, according to most safety standards, 12 volts at 
high current is NOT considered hazardous live and does not limit access to 
Users. Yes, it is a fire hazard but I don't think that is where the 240VA 
requirement comes from.

Like the watch band, I have heard where people have reached inside of a piece 
of electronic gear and shorted out a circuit with their wedding ring. If this 
condition is possible, I believe the circuit would have to be limited to 240VA.

This is my recollection of where 240VA came from and how it was used. I do not 
have any current documented support for it use today. But we still consider it 
for circuits accessible to the User to determine an Energy Hazard.

The Other Brian

-----Original Message-----
From: Ralph McDiarmid [mailto:ralph.mcdiar...@schneider-electric.com]
Sent: Wednesday, June 08, 2016 4:31 PM
To: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: Re: [PSES] fire safety test methods for different country standards

So, for the protection against FIRE, we have two energy rates, 100VA and 240VA, 
used across quite a number of standards, and the units are wrong.  Should be 
Watts.

Ralph McDiarmid
Product Compliance
Engineering
Solar Business
Schneider Electric




*Please consider the environment before printing this e-mail



-----Original Message-----
From: Nyffenegger, Dave [mailto:dave.nyffeneg...@bhemail.com]
Sent: Wednesday, June 08, 2016 1:27 PM
To: Ralph McDiarmid <ralph.mcdiar...@schneider-electric.com>; 
EMC-PSTC@LISTSERV.IEEE.ORG
Subject: RE: [PSES] fire safety test methods for different country standards

EN 60950-1:2006  2.5  uses 100 VA for LPS and is also referenced for fire 
enclosure requirements in section 4.7.2.1.

-Dave

-----Original Message-----
From: Ralph McDiarmid [mailto:ralph.mcdiar...@schneider-electric.com]
Sent: Wednesday, June 08, 2016 3:11 PM
To: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: Re: [PSES] fire safety test methods for different country standards

Hi Chuck,

A poor choice of words on my part.  I should have written, "in most of the 
standards I have worked in".   Those include CSA107.1, UL1741, UL1012, and 
IEC62109-1

The 240VA (I think they meant 240W) must have come from some base standard as a 
normative reference.  I don't know what is special about that number, but some 
committee somewhere may have concluded that power (rate of energy) below that 
threshold was unlike to be a source of ignition.  I've seen 30V and 8A used to 
define an energy limited, extra-low voltage circuit. (UL calls that a Class 2 I 
think).   The product of 8A and 30V gives 240VA as a third criterion.   I'm not 
sure it's that simple though.

Regards,

Ralph McDiarmid
Product Compliance
Engineering
Solar Business
Schneider Electric





*Please consider the environment before printing this e-mail



-----Original Message-----
From: Chuck August-McDowell [mailto:chu...@meyersound.com]
Sent: Wednesday, June 08, 2016 11:22 AM
To: Ralph McDiarmid <ralph.mcdiar...@schneider-electric.com>
Subject: RE: [PSES] fire safety test methods for different country standards

Hi Ralph,

I live in the IEC/EN/UL 60065 standard world.

Could you point at "most standards appear to limit rate of energy transfer 
(e.g. 240W)" standard?
IEC/EN/UL 62368-1?
IEC/EN/UL 60950-1?

Thank you,

Chuck McDowell
Compliance Specialist
Meyer Sound Laboratories Inc.


-----Original Message-----
From: Ralph McDiarmid [mailto:ralph.mcdiar...@schneider-electric.com]
Sent: Wednesday, June 08, 2016 9:27 AM
To: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: Re: [PSES] fire safety test methods for different country standards

Not following instructions is foreseeable misuse and needs a FMEA and maybe a 
Fault Tree analysis too, if a hazard is the anticipated result.

Getting back to this HB enclosure discussion earlier in this discussion thread, 
I see that most standards appear to limit rate of energy transfer (e.g. 240W) 
and may also place limit on available current.   The expectation is, I think, 
that a power-limited device cannot ignite something.  I assume there is lots of 
history that assumption.

Ralph McDiarmid
Product Compliance
Engineering
Solar Business
Schneider Electric



-----Original Message-----
From: Richard Nute [mailto:ri...@ieee.org]
Sent: Sunday, May 22, 2016 5:40 PM
To: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: Re: [PSES] fire safety test methods for different country standards

Hi John:


Thanks for your additional comments.

> Could it be that the scenarios which the standards committees envisage
> are not "the real deal"

In my opinion, this is the case.

> OR that the
> products which cause the fires just don't comply with the standards?

Of course, counterfeit and non-complying products are in the marketplace.  Some 
of these do catch fire.

My interest is the cause of fires in products which comply with the standards.  
The "In Compliance" reports do identify the counterfeit products, but these 
seem to be in the minority.

Fires occur under fault conditions.  Not following instructions is a sort-of 
fault condition, but rarely the cause of a fire.


Rich

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<emc-p...@ieee.org>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://www.ieee-pses.org/list.html (including how to 
unsubscribe) List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <sdoug...@ieee.org>
Mike Cantwell <mcantw...@ieee.org>

For policy questions, send mail to:
Jim Bacher:  <j.bac...@ieee.org>
David Heald: <dhe...@gmail.com>

______________________________________________________________________
This email has been scanned by the Symantec Email Security.cloud service.
______________________________________________________________________
________________________________
 This message was scanned by Exchange Online Protection Services.
________________________________

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<emc-p...@ieee.org>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://www.ieee-pses.org/list.html (including how to 
unsubscribe) List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <sdoug...@ieee.org>
Mike Cantwell <mcantw...@ieee.org>

For policy questions, send mail to:
Jim Bacher:  <j.bac...@ieee.org>
David Heald: <dhe...@gmail.com>

NOTICE: This email may contain confidential information.  Please see 
http://www.meyersound.com/confidential/ for our complete policy.

______________________________________________________________________
This email has been scanned by the Symantec Email Security.cloud service.
______________________________________________________________________
________________________________
 This message was scanned by Exchange Online Protection Services.
________________________________

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<emc-p...@ieee.org>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://www.ieee-pses.org/list.html (including how to 
unsubscribe) List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <sdoug...@ieee.org>
Mike Cantwell <mcantw...@ieee.org>

For policy questions, send mail to:
Jim Bacher:  <j.bac...@ieee.org>
David Heald: <dhe...@gmail.com>

______________________________________________________________________
This email has been scanned by the Symantec Email Security.cloud service.
______________________________________________________________________
________________________________
 This message was scanned by Exchange Online Protection Services.
________________________________

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<emc-p...@ieee.org>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://www.ieee-pses.org/list.html (including how to 
unsubscribe) List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <sdoug...@ieee.org>
Mike Cantwell <mcantw...@ieee.org>

For policy questions, send mail to:
Jim Bacher:  <j.bac...@ieee.org>
David Heald: <dhe...@gmail.com>
________________________________

LECO Corporation Notice: This communication may contain confidential 
information intended for the named recipient(s) only. If you received this by 
mistake, please destroy it and notify us of the error. Thank you.

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<emc-p...@ieee.org>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://www.ieee-pses.org/list.html (including how to 
unsubscribe) List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <sdoug...@ieee.org>
Mike Cantwell <mcantw...@ieee.org>

For policy questions, send mail to:
Jim Bacher:  <j.bac...@ieee.org>
David Heald: <dhe...@gmail.com>

______________________________________________________________________
This email has been scanned by the Symantec Email Security.cloud service.
______________________________________________________________________
________________________________
 This message was scanned by Exchange Online Protection Services.
________________________________

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<emc-p...@ieee.org>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://www.ieee-pses.org/list.html (including how to unsubscribe)
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <sdoug...@ieee.org>
Mike Cantwell <mcantw...@ieee.org>

For policy questions, send mail to:
Jim Bacher:  <j.bac...@ieee.org>
David Heald: <dhe...@gmail.com>

Reply via email to